SIRIUS® XHS Technical Specifications
- SIRIUS® XHS system specifications
- SIRIUS® XHS HybridADC specifications
- SIRIUS® XHS standard counter specifications
- SIRIUS® XHS counter with variable trigger specifications
- SIRIUS® XHS amplifiers overview
- SIRIUS® XHS UNI amplifier specifications
- SIRIUS® XHS ACC amplifier specifications
- SIRIUS® XHS CHG amplifier specifications
- SIRIUS® XHS HV amplifier specifications
- SIRIUS® XHS LV amplifier specifications
SIRIUS® XHS system specifications
| Power specs | |
|---|---|
| Power Supply | 9 - 48 V DC |
| Power consumption | Typ. 30 W |
| Environmental | |
| Operating Temperature | -10 to 50 °C |
| Storage Temperature | -40 to 85 °C |
| Humidity | 5 to 95 % RH non condensing @ 60 °C |
| IP rating | IP20 |
| Shock & Vibration | Shock (EN 60068-2-27:2009) 75 g, 6 ms, half-sine (25x pos./neg in each axis) Random Vibration (EN 60721-3-2: 1997 - Class 2M2) Sweep sinus Vibration (EN 60068-2-6:2008) |
| Dimensions | 266 mm x 149 mm x 55 mm 266 mm x 149 mm x 69 mm (extended slice) |
| Weight | 2,4 kg |
| Interfaces | |
| Ethernet | 1 GbE (XCP, OPC UA) incl. IEEE1588v2 synchronization (PTP) (RJ45) |
| USB | USB 3.0 (Type C) |
| CAN | CAN 2.0, CAN-FD (DSUB9) |
| Sync Input/Ouput | |
| Level (Input/Output) | TTL compatible |
| Max. Output Current | ±24 mA (±50 mA for 1 sec) |
| Synchronization | |
| Delay between slices | ± 1 μs ( to the reference clock ) |
| Max. Sync-cable length | 100 m (Master/Slave), 200 m (IRIG) |
SIRIUS® XHS HybridADC specifications
| ADC | |
|---|---|
| Type | Hybrid ADC |
| High bandwidth mode | 16-bit SAR, up to 15 MS/s |
| High dynamic mode | Alias-free up to 2 MS/s |
| Low pass filter | Software selectable (Off, 1 MHz, AAF) |
| Analog HW filter | Software selectable, Analog 1 MHz 6th order (Butterworth) |
| Alias free filter (AAF) | 0.5*fs cutoff, 0.45*fs pass band, 0.55*fs stop band to -100 dB, zero phase distortion, available below 2 MS/s |
SIRIUS® XHS standard counter specifications
| Counter Input | |
|---|---|
| Input Channels (See 1) | 8 |
| Modes | Event counting, waveform timing, encoder, tacho, geartooth, ... |
| Counter timebase | 150 MHz |
| Time base accuracy | Typical: 0.5 ppm, Max: 2 ppm |
| Max. Sampling Rate | 5 MS/s |
| Max. Bandwidth | 15 MHz |
| Input Filter | Off, 500 ns, 1 μs, 2 μs, 4 μs, 5 μs and 7.5 μs |
| Counter resolution | 32-bit |
| Input Compatibility, Level | 3.3 V TTL/CMOS (Low: <0.8 V, High >2 V) |
| Input termination, impedance | 100 kΩ pull-up to +3.3 V |
| Input Protection | ±25 V continuous |
| Digital Output | |
| Output Channels | 8 connectors, 1 line per connector |
| Output | 5 V digital output |
| Maximum current | 32 mA single output, 100 mA total |
| Output series resistance | 50 Ω |
| Sensor Power Supply | |
| Output Voltage +5 V | 5 V +/- 10 % 500 mA typ. Output short protected (Shared between 8 ports) |
| Output Voltage +12 V (See 2) | 12 V +/-10 % 300 mA typ. Output short protected (Shared between 8 ports) |
| 1) Digital inputs on Counter (individual A, B, Z input) can be used only for Trigger condition 2) Maximum 2x DS-TACHO4 laser probes supported. | |
| Rev: 1775113000 | |
SIRIUS® XHS counter with variable trigger specifications
| Input | |
|---|---|
| Input Channels | 8 |
| Modes | Event counting, waveform timing, encoder, tacho, geartooth, ... |
| Counter timebase | 150 MHz |
| Time base accuracy | Typical: ±0.5 ppm, Max: ±2 ppm |
| Max. Sampling Rate | 5 MS/s |
| Analog bandwidth | -3 dB @ 8 MHz min, 15 MHz typ |
| Input Filter | Off, 500 ns, 1 μs, 2 μs, 4 μs, 5 μs and 7.5 μs |
| Counter resolution | 32-bit |
| Input impedance | 1 MΩ | 8 pF |
| Digital Output | |
| Output Channels | 8 |
| Output | 5 V |
| Maximum current | 32 mA single output, 100 mA total |
| Output series resistance | 50 Ω |
| Sensor Power Supply | |
| Output Voltage +5 V | 5 V ±10 % 500 mA typ. Output short protected (Shared between 8 ports) |
| Output Voltage +12 V (See 1) | 12 V ±10 % 500 mA typ. Output short protected (Shared between 8 ports) |
| Additional features | |
| Input compatibility | Adjustable trigger levels |
| Trigger adjustment range | -2 V to 50 V |
| Trigger resolution | 13 mV |
| Trigger level accuracy | ±60 mV ±0.2 % of threshold / retrigger level |
| Max. input DC voltage @AC coupling | ±50 VDC |
| Overvoltage protection | ±100 VDC |
| 1) Maximum 2x Tacho-4 laser probes supported. | |
| Rev: 1775113000 | |
SIRIUS® XHS amplifiers overview
| XHS-HV | XHS-LV | XHS-ACC | XHS-UNI | XHS-CHG | |
|---|---|---|---|---|---|
| Connectors | BANANA | DB9, BNC, BANANA | BNC | DB9 | BNC |
| Channels per slice | 8 | 8 | 8 | 8 | 8 |
| Data rate / channel | 15 MS/s | 15 MS/s | 15 MS/s | 15 MS/s | 15 MS/s |
| Resolution | 16-bit (24-bit @ 2 MS/s) | 16-bit (24-bit @ 2 MS/s) | 16-bit (24-bit @ 2 MS/s) | 16-bit (24-bit @ 2 MS/s) | 16-bit (24-bit @ 2 MS/s) |
| Bandwidth | 5 MHz | 5 MHz | 5 MHz | 5 MHz | 5 MHz |
| Voltage ranges | ±2000 V ... ±200 V | ±100 V ... ±50 mV | ±10 V ... ±200 mV | ±100 V ... ±50 mV | ±10 V ... ±100 mV |
| Input coupling | DC | DC, AC 1 Hz | DC, AC 0.1 Hz, AC 1 Hz | DC, AC 0.1 Hz, AC 1 Hz | |
| Sensor excitation | ✗ | 2.5..30 V bipolar, 2..24 V unipolar, max. 0.2 A / 2 W | IEPE 2 mA, 4 mA, 8 mA, 12 mA, 16 mA, 20 mA | Voltage; Unipolar and Bipolar 0.15 V .. 24 V Current: 100 µA - 100 mA, IEPE | IEPE 2 mA, 4 mA, 8 mA, 12 mA, 16 mA, 20 mA |
| Bridge connection | ✗ | ✓ | ✗ | ✓ | ✗ |
| Programmable shunt | ✗ | ✗ | ✗ | ✗ | ✗ |
| IEPE input | ✗ | DSI-ACC | ✓ | ✓ | ✓ |
| Resistance | ✗ | ✗ | ✗ | ✓ | ✗ |
| Temperature (PTx) | ✗ | DSI-RTD | ✗ | ✓ | ✗ |
| Thermocouple | ✗ | DSI-TH | ✗ | DSI-TH | ✗ |
| Potentiometer | ✗ | ✗ | ✗ | ✓ | ✗ |
| LVDT | ✗ | DSI-LVDT | ✗ | DSI-LVDT | ✗ |
| Charge | ✗ | DSI-CHG | ✗ | DSI-CHG | ✓ |
| Current | ✗ | ext. shunt DSI20mA, DSI5A | ext. shunt | ext. shunt DSI20mA, DSI5A | ext. shunt |
| TEDS | ✗ | ✓ | ✓ | ✓ | ✓ |
| Isolation voltage | CATII 1000 V | 420 Vdc (300 Vacrms) | 140 Vdc (100 Vacrms) | 420 Vdc (300 Vacrms) | 100 Vdc (100 Vacrms) |
| Maximum Power consumption | 28 W | 57 W | 45 W | 49 W | 43 W |
| Advanced functions | High voltage, high bandwidth, high isolation | High sensor excitation and multi range | Sensor error detection, high speed | Sensor error detection, high speed | High-bandwidth and DC charge |
| Optional counters | ✓ | ✓ | ✓ | ✓ | ✓ |
| Rev: 1792562400 | |||||
SIRIUS® XHS UNI amplifier specifications
| Inputs | |||||
|---|---|---|---|---|---|
| Input types | Voltage Full / Half / Quarter bridge strain (120 Ω and 350 Ω) Potentiometer, RTD, Resistance Current (ext. Shunt) IEPE | ||||
| ADC Type | Hybrid ADC - alias free up to 2 MS/s, 16-bit up to 15 MS/s | ||||
| Sampling Rate | Simultaneous 15 MS/s | ||||
| Filtering | AAF 1MHz (6th order Butterworth), AAF 2MHz (6th order Butterworth) available on request | ||||
| Analog bandwidth | 5 MHz | ||||
| Voltage Mode | |||||
| Ranges | ±100, 50, 20, 10, 5, 2.5, 1, 0.5, 0.25, 0.1 and 0.05 V | ||||
| Input Accuracy | Signal frequency | Accuracy | |||
| DC | ±0.03 % of reading ±0.025 % of range ±40 μV (1 mV for Ranges ≥ 10 V) | ||||
| Up to 1 kHz | ±0.03 % of reading ±0.02 % of range | ||||
| Up to 10 kHz | ±0.1 % of reading ±0.05 % of range | ||||
| Up to 100 kHz | ±2 % of reading ±0.1 % of range | ||||
| Up to 1000 kHz | ±5 % of reading ±0.5 % of range | ||||
| Typ. Noise floor | Sample rate, filter \ Range → | 100 V | 5 V | 0.5 V | 0.05 V |
| 15 MS/s, Filter off | -81 dB | -81 dB | -75 dB | -56 dB | |
| 1 MS/s, Filter on | -94 dB | -94 dB | -84 dB | -64 dB | |
| 100 kS/s, Filter on | -104 dB | -105 dB | -100 dB | -81 dB | |
| 10 kS/s, Filter on | -113 dB | -114 dB | -107 dB | -88 dB | |
| Typical CMR | Frequency \ Range → | 100 V | 5 V | 0.5 V | 0.05 V |
| 50 Hz | 83 dB | 92 dB | 100 dB | 100 dB | |
| 400 Hz | 82 dB | 91 dB | 89 dB | 88 dB | |
| 10 kHz | 67 dB | 77 dB | 77 dB | 77 dB | |
| Gain Drift | Typical 15 ppm/K, Max. 30 ppm/K | ||||
| Offset Drift | Range < 10 V: Typ. 1 μV/K + 2 ppm of range/K, Max 4 μV/K + 8 ppm of range/K Range ≥ 10 V: Typ. 50 μV/K + 2 ppm of range/K, Max. 150 μV/K + 6 ppm of range/K | ||||
| Gain Linearity | < 0.01 % | ||||
| Inter channel phase mismatch | 0.005° * fin [kHz] + 0.0002° | ||||
| Channel Cross talk | Range ≥ 10 V: -100 dB @ 10 kHz Range < 10 V: -140 dB @ 10 kHz | ||||
| Input Coupling | DC AC 0.15 Hz for all ranges | ||||
| Input Impedance | Range < 10 V: > 1 GΩ; Range ≥ 10 V: 1 MΩ | 110 pF between INx to GND | ||||
| Overvoltage Protection | Range ≥ 10 V: 300 V continuous Range < 10 V: 100 V (200 V peak for 10 ms) | ||||
| Max. common mode voltage Isolation voltage | 420 Vdc (300 Vacrms) | ||||
| Excitation Voltage | |||||
| Polarity | Unipolar or Bipolar Software selectable (programmable 16-bit DAC) | ||||
| Excitation Level | Unipolar and Bipolar 0.15 V .. 24 V; Predefined levels: 1, 2, 5, 10, 12, 15, 20 and 24 V | ||||
| Accuracy | ±0.05 % ±2 mV | ||||
| Drift | ±10 ppm/K ±100 μV/K | ||||
| Load stability 0 % to 100 % load | < 0.01 % | ||||
| Line regulation over 20 Ω of change | < 0.005 % @ 120 Ω load | ||||
| Noise @ 10 Volt / 350 Ω | < 300 μVrms @ 10 kHz BW | ||||
| Impedance Sense to Exc Sense to GND | 100 kΩ (0.5 Ohm with internal excitation sense) > 100 MΩ | ||||
| Current limit | 100 mA (max. 800 mW) | ||||
| Protection | Continuous short to ground | ||||
| Excitation Current | |||||
| Excitation Current | Free programmable (16-bit DAC) | ||||
| Ranges | 10 µA - 1 mA | 1 mA - 20 mA | 20 mA - 100 mA | ||
| Accuracy | ±0.2 % ±0.5 μA | ±0.15 % ±2 μA | ±0.01 % ±40 μA | ||
| Drift | 15 ppm/K | 15 ppm/K | 20 ppm/K | ||
| Output Impedance | > 1 MΩ | ||||
| Bridge Connection Types | |||||
| Bridge types | Full / Half / Quarter bridge (3-wire or 4-wire) / Quarter Constant Current (2-wire or 4-wire) | ||||
| Ranges | 2 mV/V…1000 mV/V free programmable | ||||
| Internal Bridge Completion | ½ bridge and ¼ bridge 120 Ω and 350 Ω | ||||
| Bridge Completion Accuracy | 0.05 %; TCR: 2 ppm/K (others on request) | ||||
| Internal Shunt Resistor | 59.88 kΩ and 175 kΩ, software selectable (others on request) | ||||
| Typ. Shunt Resistor Accuracy | 0.05 %; TCR: 10 ppm/K (others on request) | ||||
| Input Short, Sensor Offset Adjust | Software selectable | ||||
| Lead-wire compensation | Automatic | ||||
| IEPE mode | |||||
| IEPE Current | Free programmable (16-bit DAC) | ||||
| Predefined levels | 2, 4, 8, 12, 16 and 20 mA | ||||
| Compliance voltage | 24 V | ||||
| Output Impedance | > 1 MΩ | ||||
| Sensor detection | Shortcut: < 4 V; Open: > 19 V | ||||
| Additional Specifications | |||||
| Input connector | DSUB-9 (others on request) | ||||
| TEDS support | Class 1 (IEPE) and Class 2 | ||||
| DSI® adapters support | Yes | ||||
| Rev: 1792755220 | |||||
SIRIUS® XHS ACC amplifier specifications
| Inputs | |||||||
|---|---|---|---|---|---|---|---|
| Input types | Voltage, IEPE | ||||||
| ADC Type | Hybrid ADC - alias free up to 2 MS/s, 16-bit up to 15 MS/s | ||||||
| Sampling Rate | Simultaneous 15 MS/s | ||||||
| Filtering | AAF 1 MHz (6th order) | ||||||
| Analog bandwidth (-3 dB) | 5 MHz | ||||||
| Voltage Mode | |||||||
| Ranges | ±10, 5, 2, 1, 0.4 and 0.2 V | ||||||
| Input Accuracy | DC | ±0.03 % of reading ±0.02 % of range ±50 μV | |||||
| Up to 1 kHz | ±0.04 % of reading ±0.02 % of range | ||||||
| Up to 10 kHz | ±0.06 % of reading ±0.05 % of range | ||||||
| Up to 100 kHz | ±0.1 % of reading ±0.1 % of range | ||||||
| Up to 1000 kHz | ±5 % of reading ±0.5 % of range | ||||||
| Noise floor, Typ. | Sample rate, filter \ Range → | 10 V | 5 V | 2 V | 1 V | 0.4 V | 0.2 V |
| 15 MS/s, Filter off | -85 dB | -84 dB | -83 dB | -82 dB | -77 dB | -72 dB | |
| 1 MS/s, Filter on | -98 dB | -98 dB | -97 dB | -95 dB | -92 dB | -87 dB | |
| 100 kS/s, Filter on | -109 dB | -107 dB | -108 dB | -107 dB | -103 dB | -98 dB | |
| 10 kS/s, Filter on | -119 dB | -119 dB | -119 dB | -118 dB | -114 dB | -109 dB | |
| CMR, Typ. | Frequency \ Range → | 10 V | 5 V | 2 V | 1 V | 0.4 V | 0.2 V |
| 50 Hz | 141 dB | 141 dB | 139 dB | 133 dB | 121 dB | 120 dB | |
| 400 Hz | 130 dB | 133 dB | 136 dB | 137 dB | 137 dB | 131 dB | |
| 1 kHz | 122 dB | 125 dB | 127 dB | 129 dB | 130 dB | 131 dB | |
| Gain Drift | Typical 10 ppm/K, max. 30 ppm/K | ||||||
| Offset Drift | Typical 0.5 μV/K + 1 ppm of range/K, max. 2 μV/K + 4 ppm of range/K | ||||||
| Gain Linearity | < 0.02 % | ||||||
| Inter Channel Phase-mismatch | 0.005° * fin [kHz] + 0.0002° | ||||||
| Channel Cross talk | -145 dB typ. @ 1 kHz | ||||||
| Input Coupling | DC or AC (1 Hz, 0.1 Hz) | ||||||
| Input Impedance | 1 MΩ | ||||||
| Overvoltage Protection | 50 V continuous; 200 V peak (10 msec) | ||||||
| Max. common mode voltage Isolation voltage | 140 Vdc (100 Vacrms) | ||||||
| IEPE mode | |||||||
| Excitation | 2 mA, 4 mA, 8 mA, 12 mA, 16 mA, 20 mA | ||||||
| Compliance voltage | 24 V | ||||||
| Output Impedance | > 10 kΩ | ||||||
| Sensor detection | Shortcut: < 4 V; Open: > 19 V | ||||||
| Additional Specifications | |||||||
| Input connector | BNC | ||||||
| TEDS support | IEPE mode only | ||||||
| Rev: 1623052800 | |||||||
SIRIUS® XHS CHG amplifier specifications
| Inputs | ||||||
|---|---|---|---|---|---|---|
| Input types | Voltage IEPE Charge | |||||
| ADC Type | Hybrid ADC - alias free up to 2 MS/s, 16-bit up to 15 MS/s | |||||
| Sampling Rate | Simultaneous 15 MS/s | |||||
| Filtering | AAF 1 MHz (6th order) | |||||
| Analog bandwidth (-3 dB) | 5 MHz | |||||
| Voltage Mode | ||||||
| Ranges | ±10 V, ±5 V, ±1 V, ±0.5 V, ±0.1 V | |||||
| Input Accuracy | DC | ±0.03 % of reading ±0.02 % of range ±50 μV | ||||
| Up to 1 kHz | ±0.04 % of reading ±0.02 % of range | |||||
| Up to 10 kHz | ±0.06 % of reading ±0.05 % of range | |||||
| Up to 100 kHz | ±0.1 % of reading ±0.1 % of range | |||||
| Up to 1000 kHz | ±5 % of reading ±0.5 % of range | |||||
| Noise floor, Typ. | Sample rate, filter \ Range → | 10 V | 5 V | 1 V | 0.5 V | 0.1 V |
| 15 MS/s, Filter off | -75 dB | -75 dB | -74 dB | -73 dB | -53 dB | |
| 1 MS/s, Filter on | -89 dB | -89 dB | -88 dB | -88 dB | -80 dB | |
| 100 kS/s, Filter on | -99 dB | -99 dB | -98 dB | -98 dB | -89 dB | |
| 10 kS/s, Filter on | -109 dB | -109 dB | -107 dB | -107 dB | -95 dB | |
| CMR, Typ. | Frequency \ Range → | 10 V | 5 V | 1 V | 0.5 V | 0.1 V |
| 50 Hz | 133 dB | 135 dB | 122 dB | 124 dB | 108 dB | |
| 400 Hz | 128 dB | 130 dB | 133 dB | 131 dB | 117 dB | |
| 10 kHz | 98 dB | 99 dB | 106 dB | 108 dB | 107 dB | |
| Channel Cross talk | < -160 dB typ. @ 1 kHz | |||||
| Gain Drift | Typical 10 ppm/K, max. 30 ppm/K | |||||
| Offset Drift | Typical 2 μV/K + 1 ppm of range/K, max 8 μV/K + 3 ppm of range/K | |||||
| Gain Linearity | < 0.02 % | |||||
| Inter Channel Phase-mismatch | 0.005° * fin [kHz] + 0.0002° | |||||
| Input Coupling | DC or AC (0.1 Hz, 1 Hz) | |||||
| Input Impedance | 10 MOhm | |||||
| Overvoltage Protection | 50 V continuous; 200 V peak (10 msec) | |||||
| Max. common mode voltage Isolation voltage | 100 Vdc (100 Vacrms) | |||||
| IEPE mode | ||||||
| Ranges | ±10 V, ±5 V, ±1 V, ±0.5 V, ±0.1 V | |||||
| Excitation | 2 mA, 4 mA, 8 mA, 12 mA, 16 mA, 20 mA | |||||
| Compliance voltage | 24 V | |||||
| Output Impedance | > 10 kΩ | |||||
| Sensor detection | Shortcut: < 4 V; Open: > 19 V | |||||
| Charge mode | ||||||
| Charge ranges | ±500000 pC, ±50000 pC, ±10000 pC, ±5000 pC, ±1000 pC | |||||
| Input Accuracy | ±0.5 % of reading ±0.05 % of range ±2 pC | |||||
| Input Coupling | ±500000 pC range: - DC (filter off) - 1.45 mHz (time constant > 100 s) - 14.5 mHz (time constant > 10 s) - 145 mHz (time constant > 1 s) ±50000 pC, ±10000 pC, ±5000 pC, ±1000 pC range: - DC (filter off) - 14.5 mHz (time constant > 10 s) - 145 mHz (time constant > 1 s) - 1.45 Hz (time constant > 0.1 s) ±20 % cutoff frequency accuracy | |||||
| DC drift | typ. < 1 pC/s, max. 5 pC/s | |||||
| Charge injection | 1 Vrms, 20 Hz, ±3 % accuracy | |||||
| Additional Specifications | ||||||
| Input connector | BNC (others on request) | |||||
| TEDS support | IEPE mode only | |||||
| Rev: 1775113000 | ||||||
SIRIUS® XHS HV amplifier specifications
| Inputs | |||||
|---|---|---|---|---|---|
| ADC Type | Hybrid ADC - alias free up to 2 MS/s, 16-bit up to 15 MS/s | ||||
| Sampling Rate | Simultaneous 15 MS/s | ||||
| Filtering | AAF 1 MHz (6th order) | ||||
| Analog bandwidth (-3 dB) | 5 MHz | ||||
| Voltage Mode | |||||
| Ranges | ±2000 V, ±1000 V, ±400 V, ±200 V | ||||
| Input Accuracy | Signal frequency | Accuracy | |||
| DC to 1 kHz | ±0.03 % of reading ±0.02 % of range | ||||
| Up to 10 kHz | ±0.1 % of reading ±0.05 % of range | ||||
| Up to 100 kHz | ±4 % of reading ±0.1 % of range | ||||
| Up to 1000 kHz | ±5 % of reading ±0.5 % of range | ||||
| Noise floor, Typ. | Sample rate / Range → | 2000 V | 1000 V | 400 V | 200 V |
| 15 MS/s | -84 dB | -82 dB | -79 dB | -75 dB | |
| 1 MS/s | -95 dB | -93 dB | -85 dB | -78 dB | |
| 100 kS/s | -108 dB | -106 dB | -100 dB | -95 dB | |
| 10 kS/s | -117 dB | -115 dB | -109 dB | -104 dB | |
| CMR, Typ. | 101 dB @ 50 Hz, 74 dB @ 400 Hz | ||||
| Channel Crosstalk | -116 dB typ. @ 50 Hz, -89 dB typ. @ 1 kHz | ||||
| Gain Drift | Typical 20 ppm/K, max. 40 ppm/K | ||||
| Offset Drift | Typical 1.5 mV/K + 1 ppm of range/K, max 3 mV/K + 2 ppm of range/K | ||||
| Gain Linearity | < 0.01 % | ||||
| Input Coupling | DC | ||||
| Input Impedance | 10 MΩ | 1 pF | ||||
| Overvoltage category | CAT II 1000 V | ||||
| Maximum input voltage | 2000 Vdc or 2000 Vpk (1414 Vacrms) | ||||
| Additional Specifications | |||||
| Input connector | Banana | ||||
| Rev: 1698322800 | |||||
SIRIUS® XHS LV amplifier specifications
| Inputs | Voltage | ||||
|---|---|---|---|---|---|
| ADC Type | Hybrid ADC - alias free up to 2 MS/s, 16-bit up to 15 MS/s | ||||
| Sampling Rate | Simultaneous 15 MS/s | ||||
| Filtering | AAF 1 MHz (6th order) | ||||
| Analog bandwidth (-3 dB) | 5 MHz | ||||
| Voltage Mode | |||||
| Ranges | ±100, 50, 20, 10, 5, 2.5, 1, 0.5, 0.25, 0.1 and 0.05 V | ||||
| Input Accuracy | DC | ±0.03 % of reading ±0.02 % of range ±100 μV (2 mV for Ranges ≥ 10 V) | |||
| Up to 1 kHz | ±0.03 % of reading ±0.02 % of range | ||||
| Up to 10 kHz | ±0.1 % of reading ±0.05 % of range | ||||
| Up to 100 kHz | ±2 % of reading ±0.1 % of range | ||||
| Up to 1000 kHz | ±5 % of reading ±0.5 % of range | ||||
| Noise floor, Typ. | Sample rate, filter \ Range → | 100 V | 5 V | 0.5 V | 0.05 V |
| 15 MS/s, Filter off | -82 dB | -78 dB | -65 dB | -51 dB | |
| 1 MS/s, Filter on | -94 dB | -95 dB | -88 dB | -69 dB | |
| 100 kS/s, Filter on | -104 dB | -105 dB | -100 dB | -81 dB | |
| 10 kS/s, Filter on | -113 dB | -114 dB | -107 dB | -87 dB | |
| CMR, Typ. | Frequency \ Range → | 100 V | 5 V | 0.5 V | 0.05 V |
| 50 Hz | 91 dB | 102 dB | 98 dB | 100 dB | |
| 400 Hz | 86 dB | 75 dB | 90 dB | 98 dB | |
| 10 kHz | 66 dB | 74 dB | 82 dB | 83 dB | |
| Channel Crosstalk | Range ≥ 10 V: -105 dB @ 10 kHz Range < 10 V: -120 dB @ 10 kHz | ||||
| Gain Drift | Typical 10 ppm/K, max. 30 ppm/K | ||||
| Offset Drift | Range ≥ 10 V: Typ. 100 μV/K + 1 ppm of range/K, Max 250 μV/K + 2 ppm of range/K Range < 10 V: Typ. 4 μV/K + 1 ppm of range/K, Max 8 μV/K + 2 ppm of range/K | ||||
| Gain Linearity | < 0.03 % | ||||
| Input Coupling (see 1) | DC, AC 1 Hz | ||||
| Input Impedance | Range < 10 V: 10 MΩ; Range ≥ 10 V: 1 MΩ | 110 pF between INx to GND | ||||
| Overvoltage Protection | Range ≥ 10 V: 300 V continuous Range < 10 V: 100 V (200 V peak for 10 msec) | ||||
| Max. common mode voltage Isolation voltage | 420 Vdc (300 Vacrms) | ||||
| Excitation Voltage | |||||
| Polarity | Unipolar or Bipolar Software selectable (programmable 16-bit DAC) | ||||
| Excitation Level | Unipolar 1 .. 24 V; Predefined levels: 1, 2, 2.5, 5, 10, 12, 15, 20 and 24 VDC | ||||
| Excitation Level | Bipolar 2.5 .. 30 V; Predefined levels: 2.5, 5, 10, 12, 15, 24 and 30 VDC | ||||
| Drift | ±10 ppm/K ±150 μV/K | ||||
| Stability 10% to 90% load (bipolar) | < 0.01 % | ||||
| Current limit | 200 mA (2 Watt max. per Channel, 12 Watt max. per Slice) | ||||
| Additional Specifications | |||||
| Input connector | DSUB 9, BNC, Banana (others on request) | ||||
| TEDS support | Standard + DSI® adapters | ||||
| Rev: 1701403200 | |||||
| 1) In- must be within ±10 V referred to GND (iso); for Ranges ≥ 10 V the DC value of In- is not rejected | |||||