OBSIDIAN® technical specifications
- OBSIDIAN® systems overview
- OBSIDIAN® R8 system specifications
- OBSIDIAN® R8w system specifications
- OBSIDIAN® R12 system specifications
- IOLITE® multi-channel amplifier modules overview
- IOLITEr-8xUNI specifications
- IOLITEr-6xUNI specifications
- IOLITEir-4xUNI specifications
- IOLITEhir-4xUNI specifications
- IOLITEr-8xACC specifications
- IOLITEr-6xSTG specifications
- IOLITEr-8xSTGS specifications
- IOLITEr-8xSTGS (v2) specifications
- IOLITEir-4xLV-BAN specifications
- IOLITEir-8xLV specifications
- IOLITEr-8xLVe specifications
- IOLITEr-16xLV specifications
- IOLITEr-16xLV-10V specifications
- IOLITEir-4xHV specifications
- IOLITEir-8xLA specifications
- IOLITEr-16xLA specifications
- IOLITEhir-4xTH specifications
- IOLITEir-8xTH specifications
- IOLITEir-8xTH-HS specifications
- IOLITEir-8xRTD-HS specifications
- IOLITEr-8xDI-4xDO specifications
- IOLITEr-32xDI specifications
- IOLITEr-32xDO specifications
- IOLITEr-4xCNT specifications
- IOLITEr-16xAO technical specifications
- IOLITEir-4xCAN-FD technical specifications
OBSIDIAN® systems overview
| System | OBSIDIAN-R8 | OBSIDIAN-R8w | OBSIDIAN-R12 |
|---|---|---|---|
| Number of slots | 8 | 8 | 12 |
| Synchronization | 2x SYNC on L00B4f (IRIG-B) GPS PPS IEEE 1588v2 synchronization (RJ45) PTP over UDP & PTP IEEE 802.3 (Layer 2) | 2x SYNC on L00B4f (IP67) (IRIG-B) GPS PPS IEEE 1588v2 synchronization (RJ45) PTP over UDP & PTP IEEE 802.3 (Layer 2) | 2x SYNC on L00B4f (IRIG-B) GPS PPS IEEE 1588v2 synchronization (RJ45) PTP over UDP & PTP IEEE 802.3 (Layer 2) |
| Sync Accuracy | < 200 ns within same EtherCAT chain < 10 µs using sync for multiple EtherCAT chains below 1 sample to Sirius® | < 200 ns within same EtherCAT chain < 10 µs using sync for multiple EtherCAT chains below 1 sample to Sirius® | < 200 ns within same EtherCAT chain < 10 µs using sync for multiple EtherCAT chains below 1 sample to Sirius® |
| Ethernet Interface | |||
| Interface type | 1x Gigabit Ethernet (RJ45) | 1x Gigabit Ethernet RJ45, 1x WLAN | 1x Gigabit Ethernet RJ45 |
| Dual EtherCAT® interface | |||
| Number of buses | Two (both with buffered DAQ or real time) | Two (both with buffered DAQ or real time) | Two (both with buffered DAQ or real time) |
| Data Rate | Dual 100 Mbit bus speed | Dual 100 Mbit bus speed | Dual 100 Mbit bus speed |
| Max. Throughput per Chain | From 6 MB/s to 10 MB/s | From 6 MB/s to 10 MB/s | From 6 MB/s to 10 MB/s |
| Bus 1 connectors | 2x Lemo 1B | 2x Lemo 1T | OBSIDIANr: 2× Ethernet RJ45 OBSIDIANr-L: 2× Lemo 1T |
| Bus 2 connectors | 2x Ethernet RJ45 | 2x Ethernet RJ45 | 2x Ethernet RJ45 |
| Minimum delay (analog input to EtherCAT® bus) | 70 µs | 70 µs | 70 µs |
| Minimum EtherCAT® cycle time | 100 µs | 100 µs | 100 µs |
| Other Interfaces | |||
| Device protocols | OPC-UA XCP over Ethernet openDAQ (see 2) | OPC-UA XCP over Ethernet openDAQ (see 2) | OPC-UA XCP over Ethernet openDAQ (see 2) |
| CAN bus | |||
| Number of ports | 2 | 2 | 2 |
| Interface type | CAN 2.0B, up to 1 MBit/sec | CAN 2.0B, up to 1 MBit/sec | CAN 2.0B, up to 1 MBit/sec |
| Special applications | CCP, OBDII, J1939, CAN output (see 3) | CCP, OBDII, J1939, CAN output (see 3) | CCP, OBDII, J1939, CAN output (see 3) |
| Galvanic isolation | Isolated (only on isolated motherboard) | Functional isolation | Functional isolation |
| Bus pin fault protection | ±36 V | From –27 V to 40 V | From –27 V to 40 V |
| ESD protection | 8 kV | ±4 kV | ±4 kV |
| Sensor supply | 5 V (max. 1.2 W) | 5 V (max. 2.2 W) / 12 V (max. 2 W) (see 5) | 5 V (max. 2.2 W) / 12 V (max. 2 W) (see 5) |
| Options | |||
| GPS | External GNSS via USB-C interface | Option 1: Integrated GPS module with 10Hz update rate GPS L1, SBAS L1, QZSS L1, GLONASS L1OF, BeiDou B1l, Galileo E1 Option 2: Integrated GPS module with 100Hz update rate and RTK (see 4) GPS L1C/A, L1PY, L2C, L2P, L5 GLONASS L1CA, L2CA, L2P, L3 CDMA Beidou B1I, B1C, B2a, B2b, B2I, B3 Galileo E1, E5a, E5b, E5 AltBoc E6, QZSS L1C/A, L1 C/B, L2C, L5 Navic L5 External GNSS via USB-C interface | Option 1: Integrated GPS module with 10Hz update rate GPS L1, SBAS L1, QZSS L1, GLONASS L1OF, BeiDou B1l, Galileo E1 Option 2: Integrated GPS module with 100Hz update rate and RTK (see 4) GPS L1C/A, L1PY, L2C, L2P, L5 GLONASS L1CA, L2CA, L2P, L3 CDMA Beidou B1I, B1C, B2a, B2b, B2I, B3 Galileo E1, E5a, E5b, E5 AltBoc E6, QZSS L1C/A, L1 C/B, L2C, L5 Navic L5 External GNSS via USB-C interface |
| Internals | |||
| CPU type | ARM | ARM | ARM |
| Operating system | Linux | Linux | Linux |
| Software | Dewesoft RT | Dewesoft RT | Dewesoft RT |
| Internal storage | 4 GB eMMC | 4 GB eMMC | 4 GB eMMC (max. 32 GB eMMC) |
| Expandable storage | SD card up to 2 TB, USB-C 3.0 Type external storage | SD card up to 2 TB, USB-C 3.0 Type external storage | SD card up to 2 TB, USB-C 3.0 Type external storage |
| Power | |||
| Power supply | Dual redundant 9 - 48 V DC | Dual redundant 9 - 48 V DC | Dual redundant 9 - 48 V DC |
| Power consumption | 5.8 W @ 24 V (without modules) | Typ. 7.3 W (Max. 12 W) | OBSIDIAN-R12: Typ 7.7 W, Max. 12.7 W IOLITE-RACK (Incl. OBSIDIANr): Typ 16.2 W, Max. 21.5 W |
| EtherCAT Bus 1 maximum output | 10 A | 10 A | 10 A |
| Environmental | |||
| Operating Temperature | -10 to 50 °C (-40 to 85 °C optional) (see 1) | -40 to 60 °C (-40 to 85 °C optional) (see 1) | -10 to 50 °C (-40 to 85 °C optional ) (see 1) |
| Storage Temperature | -40 to 85 °C | -40 to 85 °C | -40 to 85 °C |
| Humidity | 5 to 95 % RH non-condensing at 50 °C | 5 to 100 % RH | 5 to 95 % RH non-condensing at 50 °C |
| IP rating | IP40 | IP67 | IP30 |
| Physical | |||
| Dimensions | 282 x 145 x 144 mm | 322 x 162 x 155 mm | 483 x 148 x 133 mm |
| Weight | 2.4 kg | 3.6 kg | IOLITE-RACK: 2.85 kg OBSIDIANr: 500g OBSIDIANr-L: 520g |
| 1) Extended operating temperature range depends of module configuration | |||
| 2) OBSIDIAN devices are openDAQ compliant from 25.3 version on. | |||
| 3) Some functionalities may not be available in standalone mode | |||
| 4) GPS module with 100Hz update rate and RTK is available on chassis with HW versions starting from 1.4.0.0 | |||
| 5) CAN power on chassis with HW versions up to 1.3.0.0 is limited to: 5 V (max. 1.2 W) / 12 V (max. 1 W) | |||
OBSIDIAN® R8 system specifications
| System | OBSIDIAN-R8 |
|---|---|
| Number of slots | 8 |
| Ethernet Interface | |
| Interface type | Gigabit Ethernet |
| Connector | RJ-45 |
| Dual EtherCAT® interface | |
| Number of buses | Two (both with buffered DAQ or real time) |
| Data Rate | Dual 100 Mbit bus speed |
| Max. Throughput per Chain | From 6 MB/s to 10 MB/s |
| Bus 1 connectors | 2x Lemo 1B |
| Bus 2 connectors | 2x Ethernet RJ45 |
| Minimum delay (analog input to EtherCAT® bus) | 70 µs |
| Minimum EtherCAT® cycle time | 100 µs |
| CAN bus | |
| Number of ports | 2 |
| Interface type | CAN 2.0B, up to 1 MBit/sec |
| Special applications | CCP, OBDII, J1939, CAN output |
| Galvanic isolation | Isolated (only on isolated motherboard) |
| Bus pin fault protection | ±36 V |
| ESD protection | 8 kV |
| Internals | |
| CPU type | ARM |
| Operating system | Linux |
| Software | Dewesoft RT |
| Internal storage | 4 GB eMMC |
| Expandable storage | SD card up to 2 TB, USB-C 3.0 Type external storage |
| Power | |
| Power supply | Dual redundant 9 - 48 V DC |
| Power consumption | 5.8 W @ 24 V (without modules) |
| EtherCAT Bus 1 maximum output | 10 A |
| Environmental | |
| Operating Temperature | -10 to 50 °C (-40 to 85 °C optional) (see 1) |
| Storage Temperature | -40 to 85 °C |
| Humidity | 5 to 95 % RH non-condensing at 50 °C |
| IP rating | IP40 |
| Physical | |
| Dimensions | 282 x 145 x 144 mm |
| Weight | 2.4 kg |
| 1) Extended operating temperature range depends of module configuration | |
OBSIDIAN® R8w system specifications
| System | OBSIDIAN-R8w | |
|---|---|---|
| Number of slots | 8 | |
| Ethernet Interface | ||
| Interface type | Gigabit Ethernet, WLAN | |
| Connector | RJ45 | |
| Dual EtherCAT® interface | ||
| Number of buses | Two (both with buffered DAQ or real time) | |
| Data Rate | Dual 100 Mbit bus speed | |
| Max. Throughput per Chain | From 6 MB/s to 10 MB/s | |
| Bus 1 connectors | 2x Lemo 1T | |
| Bus 2 connectors | 2x Ethernet RJ45 | |
| Minimum delay (analog input to EtherCAT® bus) | 70 µs | |
| Minimum EtherCAT® cycle time | 100 µs | |
| WiFi | ||
| Supported standards | 802.11ac/a/b/g/n (2T2R), Industrial-Grade | |
| Operating frequency | Dual band 2.4 GHz or 5 GHz | |
| Operation modes | STA (client), SoftAP (access point) | |
| Security | WPA2 / WPA3 | |
| Power consumption | TX mode: 610 mA RX mode: 285 mA | |
| Compliance | CE, FCC, TELEC | |
| CAN bus | ||
| Number of ports | 2 | |
| Interface type | CAN 2.0B, up to 1 MBit/sec | |
| Special applications | CCP, OBDII, J1939, CAN output | |
| Isolation | Functional isolation | |
| Bus pin fault protection | From –27 V to 40 V | |
| ESD protection | ±4 kV | |
| Sensor supply | 5 V (max. 2.2 W) / 12 V (max. 2 W) (see 1) | |
| GPS | Option 1: Integrated GPS module with 10Hz update rate | Option 2: Integrated GPS module with 100Hz update rate and RTK (see 2) |
| Supported navigation systems | GPS L1, SBAS L1, QZSS L1, GLONASS L1OF, BeiDou B1l, Galileo E1 | GPS L1C/A, L1PY, L2C, L2P, L5 GLONASS L1CA, L2CA, L2P, L3 CDMA Beidou B1I, B1C, B2a, B2b, B2I, B3 Galileo E1, E5a, E5b, E5 AltBoc E6, QZSS L1C/A, L1 C/B, L2C, L5 Navic L5 |
| Position accuracy | 2.5 m | 1.2 m |
| Position accuracy SBAS | 2 m | 0.6 m |
| Velocity accuracy | 0.05 m/s | 0.03 m/s |
| Heading accuracy | 0.3 degres | 0.1 degres |
| Output data rate | 10 Hz | 100 Hz |
| Hot start | < 2s | < 20s |
| Internals | ||
| CPU type | ARM | |
| Operating system | Linux | |
| Software | Dewesoft RT | |
| Internal storage | 4 GB eMMC | |
| Expandable storage | SD card up to 2 TB, USB-C 3.0 Type external storage | |
| USB | USB-C (SuperSpeed, up to 5 Gb/s) Host mode, 5 V (max. 1.3 A) | |
| WiFi, Bluetooth | Wifi supported in Acces mode and client mode. Bluetooth software support missing. | |
| Power | ||
| Power supply | Dual redundant 9 - 48 V DC | |
| Power consumption | Typ. 7.3 W (Max. 12 W) | |
| EtherCAT Bus 1 maximum output | 10 A | |
| Environmental | ||
| Operating Temperature | -40 to 60 °C (-40 to 85 °C optional) (see 3) | |
| Storage Temperature | -40 to 85 °C | |
| Humidity | 5 to 100 % RH | |
| IP rating | IP67 | |
| Shock | Shock: SIST EN 60068-2-27:2009 50 g, 6 ms, half-sine 75 g, 6 ms, half-sine 25x pos./neg in each axis 300x total | |
| Vibration | Random: Profile breakpoints: 24 Hz - 0.04 g²/Hz 60 Hz - 0.5 g²/Hz 100 Hz - 0.5 g²/Hz 240 Hz - 0.1 g²/Hz 2 kHz - 0.1 g²/Hz Profile RMS / Peak = 15 g / 45 g 4 hrs in each axis, 12 hrs total | |
| Physical | ||
| Dimensions | 322 x 162 x 155 mm | |
| Weight | 3.6 kg (only chassis with front blends) | |
| 1) CAN power on chassis with HW version from 1.1.0.0 to 1.3.0.0 is limited to: 5 V (max. 1.2 W) / 12 V (max. 1 W) | ||
| 2) GPS module with 100Hz update rate and RTK is available on chassis with HW version starting from 1.4.0.0 and above | ||
| 3) Extended operating temperature range depends of module configuration | ||
| Rev: 1772535600 | ||
OBSIDIAN® R12 system specifications
| System | OBSIDIAN-R12 |
|---|---|
| Number of slots | 12 |
| Ethernet Interface | |
| Interface type | Gigabit Ethernet |
| Connector | RJ45 |
| Dual EtherCAT® interface | |
| Number of buses | Two (both with buffered DAQ or real time) |
| Data Rate | Dual 100 Mbit bus speed |
| Max. Throughput per Chain | From 6 MB/s to 10 MB/s |
| Bus 1 connectors | OBSIDIAN-R12: 2× Ethernet RJ45 OBSIDIAN-R12-L: 2× Lemo 1T |
| Bus 2 connectors | 2x Ethernet RJ45 |
| Minimum delay (analog input to EtherCAT® bus) | 70 µs |
| Minimum EtherCAT® cycle time | 100 µs |
| CAN bus | |
| Number of ports | 2 |
| Interface type | CAN 2.0B, up to 1 MBit/s |
| Special applications | CCP, OBDII, J1939, CAN output |
| Isolation | Functional isolation |
| Bus pin fault protection | From –27 V to 40 V |
| ESD protection | ±4 kV |
| Sensor supply | 5 V (max. 1.2 W) / 12 V (max. 1 W) |
| GPS | |
| Supported navigation systems | GPS L1, SBAS L1, QZSS L1, GLONASS L1OF, BeiDou B1l, Galileo E1 |
| Position accuracy | 2.5 m |
| Position accuracy SBAS | 2 m |
| Velocity accuracy | 0.05 m/s |
| Heading accuracy | 0.3 degres |
| Output data rate | 10 Hz |
| Hot start | 1s |
| Internals | |
| CPU type | ARM |
| Operating system | Linux |
| Software | Dewesoft RT |
| Internal storage | 4 GB eMMC (max. 32 GB eMMC) |
| Expandable storage | SD card up to 2 TB, USB-C 3.0 Type external storage |
| WiFi, Bluetooth | Software support missing |
| Power | |
| Power supply | Dual redundant 9 - 48 V DC |
| Power consumption | OBSIDIAN-R12: Typ 7 W, Max. 11 W IOLITE-RACK (Incl. OBSIDIAN-R12): Typ 15.5 W, Max. 19 W |
| EtherCAT Bus 1 maximum output | 10 A |
| Environmental | |
| Operating Temperature | -10 to 50 °C (-40 to 85 °C optional ) (see 1) |
| Storage Temperature | -40 to 85 °C |
| Humidity | 5 to 95 % RH non-condensing at 50 °C |
| IP rating | IP30 |
| Shock | Shock (EN 60068-2-27:2009) |
| Vibration | Vibration random (MIL-STD-810D) Vibration random (EN 60721-3-2: 1997 - Class 2M2) |
| Physical | |
| Dimensions | 483 x 148 x 133 mm |
| Weight | IOLITE-RACK: 2.85 kg OBSIDIAN-R12 (4× RJ45): 500 g OBSIDIAN-R12-L (2× LEMO, 2× RJ45): 470g |
| 1) Extended operating temperature range depends of module configuration | |
| Rev: | |
IOLITE® multi-channel amplifier modules overview
| 4xUNI | 6xUNI | 8xUNI | 6xSTG | 8xSTGS | 8xLV | 8xLVe | 16xLV | 16xLV-10V | 4xLV | 4xHV | 8xLA | 16xLA | 8xACC | 8xTH | 8xTH-HS | 4xTH | 8xRTD | 8xRTD-HS | 8xRTDp | 8xRTD v3¹ | 32xDI | 32xDO | 8xDI-4xDO | 16xAO | 4xCNT | 4xCAN-FD | |
|---|---|---|---|---|---|---|---|---|---|---|---|---|---|---|---|---|---|---|---|---|---|---|---|---|---|---|---|
| Connectors | D-SUB9, LEMO 2T 10-pin, LEMO REDEL 2P 8-pin, LEMO L2B10f | D-SUB9 | DSUB-37 | DSUB-9 | Terminal block, DSUB-37 | BNC, Terminal block | Terminal block, DSUB-37 | Terminal block | Terminal block | BANANA | Banana plug, LEMO CNB.H02.SLKG | BNC, Terminal block | Terminal block | BNC | MINI TC | MINI TC | LEMO REDEL 2P series | L0B6f, Terminal block | L0B6f, Terminal block | L0B6f | Terminal block 2-way 24-pole | Terminal block | Terminal block | Terminal block | Terminal block | L1B7f | D-SUB9 |
| Channels per module | 4x | 6x | 8x | 6x | 8x | 8x | 8x | 16x | 16x | 4x | 4x | 8x | 16x | 8x | 8x | 8x | 4x | 8x | 8x | 8x | 8x | 32x | 32x | 8x digital in, 4x digital out | 16x | 4x | 4x |
| Data rate per channel IOLITE | 40 kS/s | 40 kS/s | 40 kS/s | 20 kS/s | 20 kS/s | 20 kS/s | 20 kS/s | 20 kS/s | 20 kS/s | 40 kS/s | 40 kS/s | 20 kS/s | 20 kS/s | 40 kS/s | 100 S/s | 100 S/s, upgradable to 10 kS/s | 10 kS/s | 100 S/s | 100 S/s, upgradable to 10 kS/s | 100 S/s | 10 kS/s | 20 kS/s | ✗ | ✗ | ✗ | 20 kS/s | up to 5.9 Mbit/s |
| Data rate per channel OBSIDIAN | 20 kS/s | 20 kS/s | 20 kS/s | 20 kS/s | 20 kS/s | 20 kS/s | 20 kS/s | 5 kS/s | 5 kS/s | 20 kS/s | 20 kS/s | 20 kS/s | not supported | 20 kS/s | 100 S/s | 100 S/s, upgradable to 5 kS/s | not supported | 100 S/s | 100 S/s, upgradable to 5 kS/s | 100 S/s | 5 kS/s | 10 kS/s | not supported | not supported | 1 kS/s | 20 kS/s | up to 5.9 Mbit/s |
| Resolution | 24-bit | 24-bit | 24-bit | 24-bit | 24-bit | 24-bit | 24-bit | 24-bit | 24-bit | 24-bit | 24-bit | 24-bit | 24-bit | 24-bit | 24-bit | 24-bit | 24-bit | 24-bit | 24-bit | 24-bit | 24-bit | digital | digital | digital | 16-bit | 100 MHz timebase 5 ppm, 20 ppm max | Digital |
| Bandwidth | 0.437*fs | 0.433*fs | 0.433*fs | 0.49*fs | 0.433*fs | 0.49*fs | 0.433*fs | 0.433*fs | 0.433*fs | 0.437*fs | 0.433*fs | 0.49*fs | 0.433*fs | 0.433*fs | ✗ | 0.49*fs (max. 1 kHz) | 0.49*fs (max. 1 kHz) | ✗ | 0.49*fs (max. 1 kHz) | ✗ | 0.49*fs (max. 1 kHz) | ✗ | ✗ | ✗ | ✗ | 10 MHz | ✗ |
| Voltage mode ranges | ±100 V, ±10 V, ±1 V, ±100 mV | ±100 V, ±10 V, ±1 V, ±100 mV | ±100 V, ±10 V, ±1 V, ±100 mV | ±50 V, ±10 V, ±1 V, ±100 mV | ✗ | ±100 V, ±10 V (±10 V, ±1 V on request) | ±100 V, ±5 V | ±200 V, ±10 V | ±10 V | ±100 V, ±10 V, ±1 V, ±100 mV | ±2000 V, ±1000 V, ±400 V, ±200 V | ✗ | ✗ | ±10 V, ±5 V, ±1 V, ±200 mV | ±1 V, ±100 mV | ±1 V, ±100 mV | ±100 mV | ±1 V, ±100 mV | ±1 V, ±100 mV | ±1 V, ±100 mV | ±1 V, ±100 mV | ✗ | open collector | ✗ | ±10 V | TTL (Low: <0.8 V, High > 2 V) | ✗ |
| Input coupling | DC, AC 0.15 Hz | DC, AC 0.15 Hz | DC, AC 0.15 Hz | DC, AC 1 Hz | DC | DC | DC | DC | DC | DC, AC 0.15 Hz | DC | DC | DC | DC, AC 1 Hz, AC 0.1 Hz | DC | DC | DC | DC | DC | DC | DC | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ |
| Sensor excitation | 0..24 V (unipolar) 0..12 V (bipolar) 0..100 mA (current) Max. 0.8 W/ch 2 mA, 4 mA, 6 mA (IEPE) | 0..24 V (unipolar), 0..12 V (bipolar), 0..100 mA (current), Max. 0.6 W/chn (IOLITEr), Max. 0.8 W/chn (IOLITE) 2 mA ... 20 mA (IEPE) | 0..24 V (unipolar), 0..12 V (bipolar), 0..100 mA (current), Max. 0.6 W/chn (IOLITEr), Max. 0.8 W/chn (IOLITE) 2 mA ... 20 mA (IEPE) | 0..24 V (unipolar), 0..12 V (bipolar), 0..44 mA (current), Max. 0.55 W/chn | 1 V, 2 V, 5 V | ✗ | 0...24V (unipolar) | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | 2 mA, 4 mA, 6 mA | ✗ | ✗ | ✗ | 244 uA / 440 uA | 150 uA / 400 uA | 0.4 mA / 2 mA | 225 nA / 2 uA / 17.2 uA / 145 uA / 420 uA / 1.7 mA | ✗ | ✗ | ✗ | ✗ | 5 V, 12 V | 5 V (max. 240 mA), 12 V (max. 100 mA) |
| Bridge connection | Full, ½, ¼ 350 Ω or ¼ 120 Ω (3-wire or 4-wire), ¼ CC (2-wire or 4-wire) | Full, ½, ¼ 350 Ω or ¼ 120 Ω (3-wire or 4-wire), ¼ CC (2-wire or 4-wire) | Full, ½, ¼ 350 Ω or ¼ 120 Ω (3-wire or 4-wire), ¼ CC (2-wire or 4-wire) | Full, ½, ¼ 350 Ω, ¼ 120 Ω 3-wire | Full, ½, ¼ 350 Ω, ¼ 120 Ω 3-wire, 4-wire (optional) | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ |
| Programmable shunt | 100 kΩ | 100 kΩ | 100 kΩ | 100 kΩ | 100 kΩ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ |
| IEPE input | ✓ | ✓ | ✓ | DSI-ACC⁴ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✓ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ |
| Resistance | ✓ | ✓ | ✓ | ✓ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | 1 kΩ, 10 kΩ | 1 kΩ, 10 kΩ | 100 Ω, 1 kΩ, 10 kΩ | 100 Ω, 1 kΩ, 10 kΩ, 100 kΩ, 1 MΩ, 10 MΩ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ |
| Temperature (PTx) | ✓ | ✓ | ✓ | ✓ (possible FW upgrade needed) | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | PT100, 200, 500, 1000, 2000 | PT100, 200, 500, 1000, 2000 | PT100, 200, 500, 1000, 2000 | PT100, 200, 500, 1000, 2000 | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ |
| Thermocouple | DSI-TH⁴ | DSI-TH⁴ | DSI-TH⁴ | DSI-TH⁴ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | K, J, T, R, S, N, E, C, B | K, J, T, R, S, N, E, C, B | K | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ |
| Current | 20 mA (internal shunt), external shunt, DSI-5A, DSIi-10A, DSIi-20A | 20 mA (internal shunt), external shunt, DSI-5A, DSIi-10A, DSIi-20A | 20 mA (internal shunt), external shunt, DSI-5A, DSIi-10A, DSIi-20A | 20 mA (internal shunt), DSI-5A, DSIi-10A, DSIi-20A | ✗ | ✗ | ✗ | ✗ | ✗ | 20 mA (internal shunt), external shunt | ✗ | ±20 mA, ±2 mA | ±20 mA | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ |
| Potentiometer | ✓ | ✓ | ✓ | ✓ | ✓ | ✗ | ✓ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ |
| LVDT | DSI-LVDT | DSI-LVDT | DSI-LVDT | DSI-LVDT | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ |
| Charge | DSI-CHG⁴ | DSI-CHG⁴ | DSI-CHG⁴ | DSI-CHG⁴ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | |
| TEDS | ✓ | ✓ | ✓ | ✓ | ✓ (except DB37) | ✗ | ✓ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✓ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ |
| Isolation voltage | CATII 150 V (IOLITEhi 4xUNI: CATII 1000V) | 250V | 250V | Differential | Differential | 450 V | 250 V | 250 V | 250 V | CATII 1000V | CATII 1000 V | 450 V | 250 V | 250V | 1000 V | 1000 V | CATII 1000 V | 1000 V | 1000 V | 150 V | 150 V | 1000 V | 1000 V | 1000 V | 150 V | ✗ | CATII 150 V |
| Isolation arrangement | Ch-Ch & Ch-GND | Ch-GND | Ch-GND | None | Ch-GND | Ch-Ch & Ch-GND | Ch-GND | Ch-GND | Ch-GND | Ch-Ch & Ch-GND | Ch-Ch & Ch-GND | Ch-Ch & Ch-GND | Ch-GND | Ch-GND | Ch-Ch & Ch-GND | Ch-Ch & Ch-GND | Ch-Ch & Ch-GND | Ch-Ch & Ch-GND | Ch-Ch & Ch-GND | Ch-Ch & Ch-GND | Ch-Ch & Ch-GND | Ch-GND, groups of 8 ch. | Ch-GND, groups of 8 ch. | Ch-GND, group of 4/8 ch | Ch-GND | Ch-GND | Ch-Ch & Ch-GND |
| Power consumption per module | Typ. 4.8 W Max. 5.5 W | Typ. 6.1 W Max. 7.3 W | Typ. 6.9 W, Max. 8.2 W | Typ. 5.4 W, Max. 11.1 W | Typ. 2.7 W, Max. 5.1 W | Typ. 2.4 W, Max. 3.5 W | Typ. 3.3 W Max. 4.2 W | Typ. 3.4 W, Max. 4.2 W | Typ. 3.2 W, Max. 3.7 W | Typ. 4.4 W Max. 5 W | Typ. 2.6 W Max. 3 W | Typ. 2.4 W, Max. 3.5 W | Typ. 2.9 W (Max. 3.3 W) | Typ. 4.5 W, Max. 6 W | 3.2 W | Typ. 2.6 W, Max. 3.4 W | Typ. 2.6 W, Max. 3.7 W | Typ. 2.1 W, Max. 2.7 W | Typ. 3.5 W Max. 4.3 W | Typ. 2.1 W, Max. 2.7 W | Typ. 4 W Max. 5.5 W | Typ. 1.2 W, Max. 1.9 W | Typ. 1.2 W, Max. 2.0 W | Typ. 1.1 W, Max. 1.8 W | Typ. 4.3 W, Max. 7.2 W | Typ. 1.9 W | Typ. 2.5 W Max. 10.5 W |
| Waterproof version | ✓ (IP67) | ✓ (IP67) | ✓ (IP67) | ✓ (IP67) | ✓ (IP67) | ✗ | ✓ (IP67) | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✓ (IP67) | ✗ | ✓ (IP65) | ✗ | ✗ | ✓ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✓ (IP67) | ✓ (IP67) |
| Passive PoE functionality | ✓³ | ✓³ | ✓³ | ✓³ | ✓ | ✓³ | ✓ | ✓ | ✓ | ✓³ | ✓ | ✓³ | ✓ | ✓ | ✓³ | ✓ | ✗ (not available in modular) | ✓³ | ✓ | ✓³ | ✓ | ✓³ | ✓³ | ✓³ | ✓³ | ✓³ | ✓ |
| Advanced functions | Fully isolated, high input range, connect almost any sensor, supports TEDS | High input range, connect almost any sensor, supports TEDS, high channel density | High input range, connect almost any sensor, supports TEDS, high channel density | Supports all strain types, high input range | Supports all strain types, low power consumption | High isolation, high input range | High input range, powerful sensor supply | High density low voltage | High density low voltage | Fully isolated, high input range | High isolation, high input range | High isolation, high input range | High isolation, high input range | IEPE, supports TEDS, high channel density | High isolation, support of main TC types | High isolation, support of main TC types | High isolation | High isolation | High isolation | Pulsed excitation, Cryogenic temperature range, High isolation | Fully isolated, support of Pt sensors and thermistors, high resistance range | ✗ | Watchdog | High sink current, watchdog | Signal conditioning | Supercounter technology | CAN bus, Isolated |
| Standalone version support in OBSIDIAN | ✗ | ✗ | ✗ | ✗ | ✓ | ✓ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✓ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✓ | ✓ | ✓ | ✗ | ✗ | ✓ |
| ¹ Not supported with OBSIDIAN ² Third party ECAT master Data rate = 1 kS/s ³ Available for devices produced after September 2024. ⁴ Not supported with OBSIDIAN in Standalone mode | |||||||||||||||||||||||||||
| Rev: 1788177400 | |||||||||||||||||||||||||||
IOLITEr-8xUNI specifications
| Inputs | ||||
|---|---|---|---|---|
| Input types | Voltage Full/Half/Quarter bridge strain (120 Ω and 350 Ω) IEPE Potentiometer, RTD, Resistance Current (ext. and internal Shunt) | |||
| ADC Type | 24-bit delta-sigma with anti-aliasing filter | |||
| Sampling Rate IOLITE | Simultaneous 40 kS/s per channel (software-selectable) | |||
| Sampling Rate OBSIDIAN | Simultaneous 20 kS/s per channel (software-selectable) | |||
| Bandwith | SR ≤ 2 kS/s 0.49*SR / SR > 2kS/s 0.433*SR | |||
| Voltage Mode | ||||
| Voltage Ranges | ±100 V | ±10 V | ±1 V | ±100 mV |
| Gain accuracy | ±0.03 % of reading | |||
| Offset accuracy | ±0.02 % of range | |||
| Typ. Dynamic Range @ 10 kS/s | -150 dB | -150 dB | -150 dB | -137 dB |
| Typ. Noise floor @ 10 kS/s | 115 dB | 115 dB | 114 dB | 99 dB |
| Typ. CMR @ DC...50 Hz / 400 Hz / 1 kHz | 85 dB / 69 dB / 60 dB | 85 dB / 85 dB / 85 dB | 105 dB / 105 dB / 102 dB | 113 dB / 110 dB / 104 dB |
| Typ. THD 1kHz @ 10kS/s | -91 dB | -101 dB | -100 dB | -107 dB |
| Gain Drift | Typical 10 PPM/K, max. 40 PPM/K | |||
| Offset Drift | Typical 0.3 μV/K + 2 ppm of range/K, max 0.8 μV/K + 10 ppm of range/K | |||
| Gain Linearity | <0.02 % | |||
| Channel Crosstalk @ 1kHz/10 kHz | -102 dB / -86 dB | -141 dB / -123 dB | -141 dB / -120 dB | -123 dB / -109 dB |
| Input Coupling | DC | |||
| AC 0.15 Hz for all ranges; | ||||
| Input Impedance | 2 MΩ between IN+ and IN- for 100 V Range; all other Ranges > 1 GΩ | |||
| Max. Common Mode Voltage | 100 V Range: ±120 V; all other Ranges: ±12 V | |||
| Overvoltage Protection | 100 V Range: 200 V continuously ; all other Ranges: 100 V continuously or 200 V peak for 10 ms | |||
| Analog input performance | ||||
| Sample rate | SR ≤ 2 kS/s | SR > 2 kS/s | ||
| Bandwidth (-3 dB) | 0.489*fs | 0.433*fs | ||
| Alias-free Bandwidth | DC to 0.489*fs | DC to 0.433*fs | ||
| Alias Rejection | -105 dB (all sample rates) | |||
| Delay Through ADC | 34 / fs | |||
| Oversampling | 128 | |||
| Current mode (internal shunt) | ||||
| Ranges | 20 mA | 2 mA | ||
| Input Accuracy | ±0.03 % of reading, ±0.02 % of range, ±2.1 μA | |||
| Internal Shunt Resistor | 47 Ω | |||
| Protection | TYP Trip Im>31mA - auto retry 800ms / MAX volt 33V | |||
| Excitation Voltage | ||||
| Excitation Voltage | Free programmable (16-bit DAC) | |||
| Predefined levels | 0, 1, 2.5, 5, 10, 12, 15, 24 V DC | |||
| Accuracy | ±0.05 % ±2 mV | |||
| Drift | ±10 ppm/K ±100 μV/K (offset could be 10uV/k) | |||
| Load stability | 0 % to 100 % load <0.01 % | |||
| Noise @ 10 V / 350 Ω | <150 μVrms @ 10 kS/s | |||
| Line regulation over 20 Ω of change | <0.005 % @ 120 Ω load | |||
| Sense Impedance to Exc / to GND | 100 kΩ / >100 MΩ | |||
| Current limit | 100 mA (max standalone 800 mW / max Rack 600mW) | |||
| Protection | Continuous short to ground | |||
| Excitation Current | ||||
| Excitation Current | Free programmable (16-bit DAC) | |||
| Predefined levels | 0.1, 1, 2, 5, 10, 20,60 and 100 mA DC | |||
| Ranges | 0.1 mA - 1 mA | >1 mA - 20mA | >20-100mA | |
| Accuracy | ±0.2 % ±0.5 μA | ±0.15 % ±2 μA | ±0.01 % ±40 μA | |
| Drift Typical | 5 ppm/K | 10 ppm/K | 25 ppm/K | |
| Compliance voltage | 24 Volt, max. 600 mW | |||
| Output Impedance | >100 MΩ | |||
| Bridge Connection Types | ||||
| Bridge types | Full / Half / Quarter bridge (3-wire or 4-wire) / Quarter Constant Current (2-wire or 4-wire) | |||
| Ranges | 2 mV/V…1000 mV/V free programmable | |||
| Internal Bridge Completion | Half and Quarter bridge 120 Ω and 350 Ω | |||
| Typ. Bridge Completion Accuracy | 0.05 %; TCR: 2 ppm/K (others on request) | |||
| Internal Shunt Resistor | 100 kΩ, software selectable (others on request 175k / 59.88k ) | |||
| Typ. Shunt Resistor Accuracy | 0.05 %; TCR: 10 ppm/K (others on request) | |||
| Input Short, Sensor Offset Adjust | Software selectable | |||
| IEPE mode | ||||
| IEPE Current | Free programmable (16-bit DAC) | |||
| Predefined levels | 2, 4, 8, 12, 16 or 20 mA | |||
| Compliance voltage | 25 Volt | |||
| Output Impedance | >100 kΩ | |||
| Sensor detection | Shortcut: <4 Volt; Open: >19 Volt | |||
| Additional Specifications | ||||
| Rack module type | IOLITEr-8xUNI (Water proof option available) | |||
| Power supply | 9 - 48 Vdc | |||
| Power consumption Typ. 24 V / MAX. 48 V | Typ: 6.9 W MAX: 8.15 W (40 kS/s -only Ain EXC 0 V) | |||
| Typ: 18.95 W MAX: 20.7 W (40 kS/s EXC: 0.6 W per channel) | ||||
| Typ: 12.6 W MAX: 14 W (40 kS/s - EXC: 5 V / 120 ohms per channel) | ||||
| Weight | 420 g (WP: 435 g ) | |||
| Device Dimensions | 128.4 x 115.4 x 30.1 mm | |||
| Shake | 7.7 G 10-2 kHz 1 h each axis (MIL-STD-810F, Category 24) | |||
| Shock | 50 G - 6 ms 25 times on each axis, 150 total | |||
| Input connector | D-SUB37 | |||
| Standalone | IOLITE 8xUNI | |||
| Power supply | 9 - 48 Vdc | |||
| Power consumption Typ. 24 V / MAX. 48 V | Typ: 7.25 W MAX: 8.9 W (40 kS/s -only Ain EXC 0 V) | |||
| Typ: 21.7 W MAX: 23.5 W (40 kS/s EXC: 0.8 W per channel) | ||||
| Typ: 12.9 W MAX:15.7 W (40 kS/s - EXC: 5 V / 120 ohms per channel) | ||||
| Weight | 780 g | |||
| Device Dimensions | 82 x 62 x 52 mm | |||
| Shake | 1.5 G 5-100 Hz 30 min each axis | |||
| Shock | 10 G - 6 ms 25 times on each axis, 150 total | |||
| Input connector | D-SUB37 | |||
| Additional Specifications | ||||
| TEDS support | Standard + DSI® adapters | |||
| Rev: 1789552800 | ||||
IOLITEr-6xUNI specifications
| Inputs | ||||
|---|---|---|---|---|
| Input types | Voltage Full/Half/Quarter bridge strain (120 Ω and 350 Ω) IEPE Potentiometer, RTD, Resistance Current (ext. and internal Shunt) | |||
| ADC Type | 24-bit delta-sigma with anti-aliasing filter | |||
| Sampling Rate IOLITE | Simultaneous 40 kS/s per channel (software-selectable) | |||
| Sampling Rate OBSIDIAN | Simultaneous 20 kS/s per channel (software-selectable) | |||
| Bandwith | SR ≤ 2 kS/s 0.49*SR / SR > 2 kS/s 0.433*SR | |||
| Voltage Mode | ||||
| Voltage Ranges | ±100 V | ±10 V | ±1 V | ±100 mV |
| Gain accuracy | ±0.03 % of reading | |||
| Offset accuracy | ±0.02 % of range | |||
| Typ. Dynamic Range @ 10 kS/s | -150 dB | -150 dB | -150 dB | -137 dB |
| Typ. Noise floor @ 10 kS/s | 115 dB | 115 dB | 114 dB | 99 dB |
| Typ. CMR @ DC...50 Hz / 400 Hz / 1 kHz | 85 dB / 69 dB / 60 dB | 85 dB / 85 dB / 85 dB | 105 dB / 105 dB / 102 dB | 113 dB / 110 dB / 104 dB |
| Typ. THD 1kHz @ 10kS/s | -91 dB | -101 dB | -100 dB | -107 dB |
| Gain Drift | Typical 10 PPM/K, max. 40 PPM/K | |||
| Offset Drift | Typical 0.3 μV/K + 2 ppm of range/K, max 0.8 μV/K + 10 ppm of range/K | |||
| Gain Linearity | <0.02 % | |||
| Channel Crosstalk @ 1kHz/10 kHz | -102 dB / -86 dB | -141 dB / -123 dB | -141 dB / -120 dB | -123 dB / -109 dB |
| Input Coupling | DC AC 0.15 Hz for all ranges | |||
| Input Impedance | 2 MΩ between IN+ and IN- for 100 V Range; all other Ranges > 1 GΩ | |||
| Max. Common Mode Voltage | 100 V Range: ±120 V; all other Ranges: ±12 V | |||
| Overvoltage Protection | 100 V Range: 200 V continuously ; all other Ranges: 100 V continuously or 200 V peak for 10 ms | |||
| Analog input performance | ||||
| Sample rate | SR ≤ 2 kS/s | SR > 2 kS/s | ||
| Bandwidth (-3 dB) | 0.489*fs | 0.433*fs | ||
| Alias-free Bandwidth | DC to 0.489*fs | DC to 0.433*fs | ||
| Alias Rejection | -105 dB (all sample rates) | |||
| Delay Through ADC | 34 / fs | |||
| Current mode (internal shunt) | ||||
| Ranges | 20 mA | 2 mA | ||
| Input Accuracy | ±0.03 % of reading, ±0.02 % of range, ±2.1 μA | |||
| Internal Shunt Resistor | 47 Ω | |||
| Protection | TYP Trip Im>31mA - auto retry 800ms / MAX volt 33V | |||
| Excitation Voltage | ||||
| Excitation Voltage | Free programmable (16-bit DAC) | |||
| Predefined levels | 0, 1, 2.5, 5, 10, 12, 15, 24 V DC | |||
| Accuracy | ±0.05 % ±2 mV | |||
| Drift | ±10 ppm/K ±100 μV/K (offset could be 10uV/k) | |||
| Load stability | 0 % to 100 % load <0.01 % | |||
| Noise @ 10 V / 350 Ω | <150 μVrms @ 10 kS/s | |||
| Line regulation over 20 Ω of change | <0.005 % @ 120 Ω load | |||
| Sense Impedance to Exc / to GND | 100 kΩ / >100 MΩ | |||
| Current limit | 100 mA (MAX 600 mW ) | |||
| Protection | Continuous short to ground | |||
| Excitation Current | ||||
| Excitation Current | Free programmable (16-bit DAC) | |||
| Predefined levels | 0.1, 1, 2, 5, 10, 20,60 and 100 mA DC | |||
| Ranges | 0.1 mA - 1 mA | >1 mA - 20mA | >20-100mA | |
| Accuracy | ±0.2 % ±0.5 μA | ±0.15 % ±2 μA | ±0.01 % ±40 μA | |
| Drift Typical | 5 ppm/K | 10 ppm/K | 25 ppm/K | |
| Compliance voltage | 24 Volt, max. | |||
| Output Impedance | >100 MΩ | |||
| Bridge Connection Types | ||||
| Bridge types | Full / Half / Quarter bridge (3-wire or 4-wire) / Quarter Constant Current (2-wire or 4-wire) | |||
| Ranges | 2 mV/V…1000 mV/V free programmable | |||
| Internal Bridge Completion | Half and Quarter bridge 120 Ω and 350 Ω | |||
| Typ. Bridge Completion Accuracy | 0.05 %; TCR: 2 ppm/K (others on request) | |||
| Internal Shunt Resistor | 100 kΩ, software selectable (others on request 175k / 59.88k ) | |||
| Typ. Shunt Resistor Accuracy | 0.05 %; TCR: 10 ppm/K (others on request) | |||
| Input Short, Sensor Offset Adjust | Software selectable | |||
| IEPE mode | ||||
| IEPE Current | Free programmable (16-bit DAC) | |||
| Predefined levels | 2, 4, 8, 12, 16 or 20 mA | |||
| Compliance voltage | 25 Volt | |||
| Output Impedance | >100 kΩ | |||
| Sensor detection | Shortcut: <4 Volt; Open: >19 Volt | |||
| Additional Specifications | ||||
| Rack | IOLITErw 6xUNI | |||
| Power supply | 9 - 48 Vdc | |||
| Power consumption Typ. 24 V / MAX. 48 V | Typ: 6.05 W MAX:7.25 W (40 kS/s -only Ain EXC 0 V) | |||
| Typ: 16 W MAX: 14.7 W (40 kS/s EXC: 0.6 W per channel | ||||
| Typ: 10.6 W MAX: 12 W (40 kS/s - EXC: 5 V / 120 ohms per channel) | ||||
| Weight | 430 g | |||
| Device Dimensions | 128.4 x 115.4 x 30.1 mm | |||
| Shake | 7.7 G 10-2 kHz 1 h each axis (MIL-STD-810F, Category 24) | |||
| shock | 50 G - 6 ms 25 times on each axis, 150 total | |||
| Input connector | D-SUB9 | |||
| Additional Specifications | ||||
| TEDS support | Standard + DSI® adapters | |||
| Rev: 1789552800 | ||||
IOLITEir-4xUNI specifications
| Analog inputs | ||||||
|---|---|---|---|---|---|---|
| Input type | Voltage Full/Half/Quarter bridge strain (120 Ω and 350 Ω) IEPE Potentiometer, RTD, Resistance Current (int. and ext. Shunt) | |||||
| Number of channels | 4 | |||||
| ADC Type | 24-bit delta-sigma with anti-aliasing filter | |||||
| Sampling Rate IOLITE | Simultaneous 40 kS/s per channel (software-selectable) | |||||
| Sampling Rate OBSIDIAN | Simultaneous 20 kS/s per channel (software-selectable) | |||||
| Voltage Mode | ||||||
| Ranges | ±100 V | ±10 V | ±1 V | ±0.1 V | ||
| Input Accuracy | DC | ±0.03 % of reading ±0.02 % of range | ||||
| Noise floor, Typ. | 10 kS/s | -117 dB | -117 dB | -116 dB | -100 dB | |
| Dynamic Range, Typ. | 10 kS/s | -148 dB | -149 dB | -147 dB | -132 dB | |
| CMR, Typ. @ 40 kS/s | 0 dBFS sine wave @ 50 Hz | 82 dB | 87 dB | 105 dB | 113 dB | |
| 0 dBFS sine wave @ 400 Hz | 67 dB | 87 dB | 105 dB | 110 dB | ||
| 0 dBFS sine wave @ 1 kHz | 59 dB | 87 dB | 102 dB | 104 dB | ||
| THD, Typ. | 40 kS/s, -1 dBFS sine wave @ 1 kHz | -91 dB | -101 dB | -100 dB | -107 dB | |
| Gain Linearity | < 0.01 % | |||||
| Gain Drift | Typ. 10 ppm/K, max. 40 ppm/K | |||||
| Offset Drift | Typ. 0.3 μV/K + 2 ppm of range/K, max. 0.8 μV/K + 10 ppm of range/K | |||||
| Channel Cross talk, Typ. @ 1 kHz | -116 dB | -143 dB | -141 dB | -126 dB | ||
| Input Coupling | DC AC 0.15 Hz | |||||
| Input Impedance | 2 MΩ between IN+ and IN- | >1 GΩ between IN+ and IN- | ||||
| Max. Common Mode Voltage | ±120 V | ±12 V | ||||
| Overvoltage Protection | ±200 V continuously | ±100 V continuously / 200 Vpeak for 10 ms | ||||
| Current mode (internal shunt) | ||||||
| Ranges | 20 mA | 2 mA | ||||
| Input Accuracy | ±0.03 % of reading, ±0.02 % of range, ±2.1 μA | |||||
| Internal Shunt Resistor | 50 Ω | |||||
| Protection | Typ. trip current: Imax ≥ 31 mA; auto retry on 800 ms (clamp voltage 33 V) | |||||
| Excitation Voltage | ||||||
| Excitation Voltage | Free programmable (16-bit DAC) | |||||
| Predefined levels | 24 V DC | |||||
| Accuracy | ±0.05 % ±2 mV | |||||
| Drift | ±10 ppm/K ±100 μV/K | |||||
| Load stability | <0.01 % | |||||
| Noise @ 10 V / 350 Ω | < 300 μVrms @ 10 kHz BW | |||||
| Line regulation over 20 Ω of change | <0.005 % @ 120 Ω load | |||||
| Sense Impedance to Exc / to GND | 100 kΩ | |||||
| Current limit | 100 mA (max. 800 mW) | |||||
| Protection | Continuous short to ground | |||||
| Excitation Current | ||||||
| Excitation Current | Free programmable (16-bit DAC) | |||||
| Predefined levels | 0.1, 1, 2, 5, 10, 20,60 and 100 mA DC | |||||
| Ranges | 0.1 - 1 mA | >1 - 20 mA | >20 - 100 mA | |||
| Accuracy | ±0.2 % ±0.5 μA | ±0.15 % ±2 μA | ±0.01 % ±40 μA | |||
| Drift Typical | 5 ppm/K | 10 ppm/K | 40 ppm/K | |||
| Compliance voltage | 24 V | |||||
| Output Impedance | >100 MΩ | |||||
| Bridge Connection Types | ||||||
| Bridge types | Full / Half / Quarter bridge (3-wire or 4-wire) / Quarter Constant Current (2-wire or 4-wire) | |||||
| Ranges | 2 - 1000 mV/V free programmable | |||||
| Internal Bridge Completion | Half and Quarter bridge 120 Ω and 350 Ω | |||||
| Typ. Bridge Completion Accuracy | 0.05 %; TCR: 2 ppm/K | |||||
| Internal Shunt Resistor | 100 kΩ, software selectable | |||||
| Typ. Shunt Resistor Accuracy | 0.05 %; TCR: 10 ppm/K | |||||
| Input Short, Sensor Offset Adjust | Software selectable | |||||
| IEPE input | ||||||
| IEPE Current | Predefined levels | |||||
| Predefined levels | 2, 4, 6 mA | |||||
| Compliance voltage | 24 V | |||||
| Output Impedance | >100 kΩ | |||||
| Sensor detection | Shortcut: <4 V; Open: >19 V | |||||
| Additional Specifications | ||||||
| Rack | IOLITEir-4xUNI | IOLITEirw-4xUNI-L2T10f | IOLITEhir-4xUNI | |||
| Input connector | D-SUB9 | LEMO 2T Series 10-pin female | LEMO Redel 2P Series 8-pin | |||
| TEDS support | TEDS (Class 1 & Class 2) + DSI® adapters | TEDS (Class 1 & Class 2) | TEDS (Class 1 & Class 2) | |||
| Power supply | 9 - 48 V DC | 9 - 48 V DC | 9 - 48 V DC | |||
| Power Consumption | Voltage mode: Typ. 4.8 W / Max. 5.5 W Voltage mode with load: Typ. 12.6 W / Max. 13.4 W (Load: 0.8W per channel) | Voltage mode: Typ. 4.8 W / Max. 5.5 W Voltage mode with load: Typ. 12.6 W / Max. 13.4 W (Load: 0.8W per channel) | Voltage mode: Typ. 4.8 W / Max. 5.5 W Voltage mode with load: Typ. 12.6 W / Max. 13.4 W (Load: 0.8W per channel) | |||
| Isolation | CH - GND and CH - CH isolated Secondary circuits CAT II 150V | CH - GND and CH - CH isolated Secondary circuits CAT II 150V | Secondary circuits 1000V CATII 300 V | |||
| Weight | 370 g | 400g | 390g 440g (with anti-vibration cover) | |||
| Slice Dimensions | 128.4 x 127.6 x 30.1 mm | 128.4 x 124.1 x 30.1 mm | 128.4 x 125.5 x 30.1 mm 128.4 x 152.4 x 30.1 mm (with anti-vibration cover) | |||
| Rev: 1775217600 | ||||||
IOLITEhir-4xUNI specifications
| Analog inputs | ||||||
|---|---|---|---|---|---|---|
| Input type | Voltage Full/Half/Quarter bridge strain (120 Ω and 350 Ω) IEPE Potentiometer, RTD, Resistance Current (int. and ext. Shunt) | |||||
| Number of channels | 4 | |||||
| ADC Type | 24-bit delta-sigma with anti-aliasing filter | |||||
| Sampling Rate IOLITE | Simultaneous 40 kS/s per channel (software-selectable) | |||||
| Sampling Rate OBSIDIAN | Simultaneous 20 kS/s per channel (software-selectable) | |||||
| Voltage Mode | ||||||
| Ranges | ±100 V | ±10 V | ±1 V | ±0.1 V | ||
| Input Accuracy | DC | ±0.03 % of reading ±0.02 % of range | ||||
| Noise floor, Typ. | 10 kS/s | -117 dB | -117 dB | -116 dB | -100 dB | |
| Dynamic Range, Typ. | 10 kS/s | -148 dB | -149 dB | -147 dB | -132 dB | |
| CMR, Typ. @ 40 kS/s | 0 dBFS sine wave @ 50 Hz | 82 dB | 87 dB | 105 dB | 113 dB | |
| 0 dBFS sine wave @ 400 Hz | 67 dB | 87 dB | 105 dB | 110 dB | ||
| 0 dBFS sine wave @ 1 kHz | 59 dB | 87 dB | 102 dB | 104 dB | ||
| THD, Typ. | 40 kS/s, -1 dBFS sine wave @ 1 kHz | -91 dB | -101 dB | -100 dB | -107 dB | |
| Gain Linearity | < 0.01 % | |||||
| Gain Drift | Typ. 10 ppm/K, max. 40 ppm/K | |||||
| Offset Drift | Typ. 0.3 μV/K + 2 ppm of range/K, max. 0.8 μV/K + 10 ppm of range/K | |||||
| Channel Cross talk, Typ. @ 1 kHz | -116 dB | -143 dB | -141 dB | -126 dB | ||
| Input Coupling | DC AC 0.15 Hz | |||||
| Input Impedance | 2 MΩ between IN+ and IN- | >1 GΩ between IN+ and IN- | ||||
| Max. Common Mode Voltage | ±120 V | ±12 V | ||||
| Overvoltage Protection | ±200 V continuously | ±100 V continuously / 200 Vpeak for 10 ms | ||||
| Current mode (internal shunt) | ||||||
| Ranges | 20 mA | 2 mA | ||||
| Input Accuracy | ±0.03 % of reading, ±0.02 % of range, ±2.1 μA | |||||
| Internal Shunt Resistor | 50 Ω | |||||
| Protection | Typ. trip current: Imax ≥ 31 mA; auto retry on 800 ms (clamp voltage 33 V) | |||||
| Excitation Voltage | ||||||
| Excitation Voltage | Free programmable (16-bit DAC) | |||||
| Predefined levels | 24 V DC | |||||
| Accuracy | ±0.05 % ±2 mV | |||||
| Drift | ±10 ppm/K ±100 μV/K | |||||
| Load stability | <0.01 % | |||||
| Noise @ 10 V / 350 Ω | < 300 μVrms @ 10 kHz BW | |||||
| Line regulation over 20 Ω of change | <0.005 % @ 120 Ω load | |||||
| Sense Impedance to Exc / to GND | 100 kΩ | |||||
| Current limit | 100 mA (max. 800 mW) | |||||
| Protection | Continuous short to ground | |||||
| Excitation Current | ||||||
| Excitation Current | Free programmable (16-bit DAC) | |||||
| Predefined levels | 0.1, 1, 2, 5, 10, 20,60 and 100 mA DC | |||||
| Ranges | 0.1 - 1 mA | >1 - 20 mA | >20 - 100 mA | |||
| Accuracy | ±0.2 % ±0.5 μA | ±0.15 % ±2 μA | ±0.01 % ±40 μA | |||
| Drift Typical | 5 ppm/K | 10 ppm/K | 40 ppm/K | |||
| Compliance voltage | 24 V | |||||
| Output Impedance | >100 MΩ | |||||
| Bridge Connection Types | ||||||
| Bridge types | Full / Half / Quarter bridge (3-wire or 4-wire) / Quarter Constant Current (2-wire or 4-wire) | |||||
| Ranges | 2 - 1000 mV/V free programmable | |||||
| Internal Bridge Completion | Half and Quarter bridge 120 Ω and 350 Ω | |||||
| Typ. Bridge Completion Accuracy | 0.05 %; TCR: 2 ppm/K | |||||
| Internal Shunt Resistor | 100 kΩ, software selectable | |||||
| Typ. Shunt Resistor Accuracy | 0.05 %; TCR: 10 ppm/K | |||||
| Input Short, Sensor Offset Adjust | Software selectable | |||||
| IEPE input | ||||||
| IEPE Current | Predefined levels | |||||
| Predefined levels | 2, 4, 6 mA | |||||
| Compliance voltage | 24 V | |||||
| Output Impedance | >100 kΩ | |||||
| Sensor detection | Shortcut: <4 V; Open: >19 V | |||||
| Additional Specifications | ||||||
| Rack | IOLITEir-4xUNI | IOLITEirw-4xUNI-L2T10f | IOLITEhir-4xUNI | |||
| Input connector | D-SUB9 | LEMO 2T Series 10-pin female | LEMO Redel 2P Series 8-pin | |||
| TEDS support | TEDS (Class 1 & Class 2) + DSI® adapters | TEDS (Class 1 & Class 2) | TEDS (Class 1 & Class 2) | |||
| Power supply | 9 - 48 V DC | 9 - 48 V DC | 9 - 48 V DC | |||
| Power Consumption | Voltage mode: Typ. 4.8 W / Max. 5.5 W Voltage mode with load: Typ. 12.6 W / Max. 13.4 W (Load: 0.8W per channel) | Voltage mode: Typ. 4.8 W / Max. 5.5 W Voltage mode with load: Typ. 12.6 W / Max. 13.4 W (Load: 0.8W per channel) | Voltage mode: Typ. 4.8 W / Max. 5.5 W Voltage mode with load: Typ. 12.6 W / Max. 13.4 W (Load: 0.8W per channel) | |||
| Isolation | CH - GND and CH - CH isolated Secondary circuits CAT II 150V | CH - GND and CH - CH isolated Secondary circuits CAT II 150V | Secondary circuits 1000V CATII 300 V | |||
| Weight | 370 g | 400g | 390g 440g (with anti-vibration cover) | |||
| Slice Dimensions | 128.4 x 127.6 x 30.1 mm | 128.4 x 124.1 x 30.1 mm | 128.4 x 125.5 x 30.1 mm 128.4 x 152.4 x 30.1 mm (with anti-vibration cover) | |||
| Rev: 1775217600 | ||||||
IOLITEr-8xACC specifications
| Analog inputs | |||||
|---|---|---|---|---|---|
| Input type | Voltage, IEPE | ||||
| Number of channels | 8 | ||||
| ADC Type | 24-bit delta-sigma with anti-aliasing filter | ||||
| Sampling Rate IOLITE | Simultaneous 40 kS/s per channel (software-selectable) | ||||
| Sampling Rate OBSIDIAN | Simultaneous 20 kS/s per channel (software-selectable) | ||||
| Voltage Mode | |||||
| Ranges | ±10 V | ±5 V | ±1 V | ±0.2 V | |
| Input Accuracy | DC | ±0.02 % of reading ±0.02 % of range | |||
| Noise floor, Typ. | 10 kS/s | -116 dB | -116 dB | -114 dB | -105 dB |
| Dynamic Range, Typ. | 10 kS/s | -151 dB | -151 dB | -150 dB | -140 dB |
| CMR, Typ. @ 40 kS/s | 0 dBFS sine wave @ 50 Hz | 113 dB | 113 dB | 111 dB | 102 dB |
| 0 dBFS sine wave @ 400 Hz | 112 dB | 112 dB | 110 dB | 102 dB | |
| 0 dBFS sine wave @ 1 kHz | 109 dB | 108 dB | 108 dB | 102 dB | |
| THD, Typ. | 40 kS/s, -1 dBFS sine wave @ 1 kHz | -104 dB | -104 dB | -104 dB | -102 dB |
| SFDR, Typ. | 40 kS/s, -1 dBFS sine wave @ 1 kHz | 106 dB | 106 dB | 99 dB | 96 dB |
| Gain Linearity | < 0.02 % | ||||
| Gain Drift | Typ. 5 ppm/K (Max. 15 ppm/K) | ||||
| Offset Drift | Typ. 0.5 μV/K + 0.5 ppm of range/K (Max. 2 μV/K + 2 ppm of range/K) | ||||
| Channel Cross talk, Typ. @ 1 kHz | -140 dB | -140 dB | -140 dB | -130 dB | |
| Input Coupling | DC, AC 0.1 Hz, 1 Hz | ||||
| Input Impedance | 1 MΩ | ||||
| Overvoltage Protection | In+ to In-: 50 V continuous, 200 V peak (10 msec) | ||||
| Isolation voltage | 250 Vpeak channel to ground | ||||
| Analog input performance | |||||
| Sampling rate | SR < 5 kS/s | SR ≥ 5 kS/s | |||
| Bandwidth (-3 dB) | 0.49*fs | 0.433*fs | |||
| Alias-free Bandwidth | DC to 0.44*fs | DC to 0.499*fs | |||
| Alias Rejection | -105 dB | ||||
| Delay Through ADC | 34 / fs | ||||
| Oversampling | 128 | ||||
| IEPE input | |||||
| Excitation | 2 mA, 4 mA, 6 mA | ||||
| Compliance voltage | 28 V | ||||
| Output Impedance | >50 kΩ | ||||
| Sensor detection | Shortcut: <4 V; Open: >19 V | ||||
| Additional Specifications | |||||
| Rack | IOLITEr-8xACC | ||||
| Input connector | BNC | ||||
| TEDs support | IEPE mode only (Class 1) | ||||
| Power supply | 9 - 48 V DC | ||||
| Power Consumption | Analog input only: Typ. 3.6 W (Max. 4 W) IEPE: Typ. 4.8 W @ 4 mA (Max. 6 W @ 6 mA) | ||||
| Weight | 320 g | ||||
| Slice Dimensions | 128.4 x 127.6 x 30.1 mm | ||||
| Additional Specifications | |||||
| Modular | IOLITE-8xACC | ||||
| Input connector | BNC | ||||
| TEDs support | IEPE mode only (Class 1) | ||||
| Power supply | 9 - 48 V DC | ||||
| Power Consumption | Analog input only: Typ. 3.7 W (Max. 4.5 W) IEPE: Typ. 4.9 W @ 4 mA (Max. 6.4 W @ 6 mA) | ||||
| Weight | 575 g | ||||
| Slice Dimensions | 136.75 x 127.2 x 35.0 mm | ||||
| Rev: 1681991028 | |||||
IOLITEr-6xSTG specifications
| Analog inputs | ||||
|---|---|---|---|---|
| Input type | Voltage Full / half / quarter bridge strain Current Potentiometer Resistance Temperature | |||
| Number of channels | 6 | |||
| ADC Type | 24-bit OSAR with anti-aliasing filter | |||
| Sampling Rate | Simultaneous 20 kS/s per channel (software-selectable) | |||
| Voltage Mode | ||||
| Ranges | ±50 V | ±10 V | ±1 V | ±100 mV |
| Input Accuracy | ±0.03 % of reading, ±0.02 % of range, ±0.1 mV | |||
| Typical Dynamic Range @ 10 kS | -100 dB | -110 dB | -130 dB | -145 dB |
| Typical Noise floor @ 10 kS | -103 dB | -97 dB | -103 dB | -94 dB |
| Typical CMR @ 400 Hz / 1 kHz | 71 dB / 66 dB | 72 dB / 66 dB | 96 dB / 88 dB | 96 dB / 87 dB |
| Gain Drift | Typical 10 ppm/K (max. 40 ppm/K) | |||
| Offset Drift | Typical 0.3 μV/K + 5 ppm of range/K, max 2 μV/K + 10 ppm of range/K | |||
| Gain Linearity | < 0.02% | |||
| Input Coupling | DC, AC 1Hz | |||
| Input Impedance | Range ≥ 10 V; 1 MΩ between IN+ and IN- Range < 10 V: 20 MΩ | |||
| Overvoltage Protection | In+ to In-: 50 V continuous, 200 V peak (10 msec) | |||
| Max. Common Mode Voltage | ±50 V (signal + common mode) | ±5 V (signal + common mode) | ||
| Current Mode | ||||
| Ranges | 20 mA | 2 mA | ||
| Input Accuracy | ±0.03 % of reading, ±0.02 % of range, ±2.1 μA | |||
| Internal Shunt Resistor | 50 Ω | |||
| Analog input performance | ||||
| Bandwidth (-3 dB) | 0.49*fs | |||
| Alias-free Bandwidth | DC to 0.453*fs | |||
| Alias Rejection | -100 dB (all sample rates) | |||
| Delay Through ADC | 37 / fs | |||
| Oversampling | 32 | |||
| Excitation Voltage | ||||
| Excitation Voltage | Free programmable (16-bit DAC) | |||
| Predefined Levels | Bipolar: 0, 1, 2, 5, 10 and 12 VDC Unipolar: 0, 2, 5, 10, 15, 24 VDC | |||
| Accuracy | ±0.05 % ±2 mV | |||
| Drift | ±50 ppm/K ±100 μV/K | |||
| Stability 10 % to 90 % Load | < 0.01 % | |||
| Current Limit | 42 mA (550 mW max. power) | |||
| Protection | Continuous short to ground | |||
| Excitation Current | ||||
| Excitation Current | Free programmable (16-bit DAC) | |||
| Predefined Levels | 0, 2, 4, 8, 16, 44 mA | |||
| Accuracy | ±0.1 % ±2 μA (<10 mA), ±0.5 % ±5 μA (>10 mA) | |||
| Bridge measurement | ||||
| Bridge Connection Types | full bridge strain, ½ bridge strain, ¼ bridge strain (3-wire) | |||
| Ranges | 2…1000 mV/V free programmable | |||
| Internal Bridge Completion | ½ bridge and ¼ bridge 120 and 350 Ω | |||
| Bridge Completion Accuracy | 0.05 %; TCR: 5 ppm/K (others on request) | |||
| Internal Shunt Resistor | 100 kΩ (others on request) | |||
| Shunt Resistor Accuracy | 0.05 %; TCR: 10 ppm/K (others on request) | |||
| Input Short, Sensor Offset Adjust | Software-selectable | |||
| Resistance measurement | ||||
| Supported ranges | 50000 Ω, 10000 Ω, 3000 Ω, 1000 Ω, 100 Ω (software support) | |||
| Temperature Mode ¹ | ||||
| RTD Types | PT100, PT200, PT500, PT1000, PT2000 | |||
| Measurement range | -200 °C to +850 °C | |||
| Accuracy | ±0.05 % of reading ±0.2 °C | |||
| Sampling Rate | Simultaneous 100 S/s | |||
| Temperature Drift | typ. ±5 ppm/K (max. ±12 ppm/K ±0.003 °C/K) | |||
| Input Connection | 3-wire or 4-wire | |||
| Additional Specifications | ||||
| Input connectors | DSUB-9 | |||
| TEDS support | Standard + DSI adapters | |||
| Power supply | 9 - 48 V DC | |||
| Power Consumption | 5.4 W, Max. 11.1 W (7.9 W 120 Ω @ 5 V load, 8.8 W 350 Ω @ 10 V load) | |||
| Weight | 340 g | |||
| Slice Dimensions | 128.4 x 115.4 x 30.1 mm | |||
| ¹Calibration and firmware update needed | ||||
| Rev: 1784278300 | ||||
IOLITEr-8xSTGS specifications
| Analog inputs | IOLITE 8xSTGS | |||
|---|---|---|---|---|
| Input type | Full / half / quarter bridge strain (3-wire), Potentiometer | |||
| Number of channels | 8 | |||
| ADC Type | 24-bit oversampled sigma-delta with anti-aliasing filter | |||
| Sampling Rate | Simultaneous 20 kS/s per channel (software-selectable) | |||
| Analog input performance (see 1) | ||||
| Ranges | 1 V | 0.1 V | 0.02 V | |
| Gain accuracy | 0.02 % of reading | 0.02 % of reading | 0.02 % of reading | |
| Offset accuracy | 210 uV | 30 uV | 14 uV | |
| Offset Accuracy after Balance Amplifier | 21 uV | 3 uV | 1.4 uV | |
| Typical Noise floor @ 10 kS/s / 1 kS/s / 100 S/s | -107 / -114 / -122 dB | -99 / -107 / -117 dB | -87 / -96 / -108 dB | |
| Typical Dynamic Range @ 10 kS/s / 1 kS/s / 100 S/s | -132 / -139 / -145 dB | -122 / -132 / -139 dB | -112 / -120 / -130 dB | |
| Typ. Crosstalk @ 1 kHz | -120 dB | -126 dB | -114 dB | |
| Typ. CMR @ 50 Hz / 400 Hz / 1 kHz | 76 / 74 / 73 dB | 97 / 90 / 84 dB | 108 / 91 / 83 dB | |
| Gain Drift | Typical 10 ppm/K (max. 20 ppm/K) | |||
| Offset Drift | Typical 0.03 μV/K + 2.5 ppm of range/K, max 0.1 μV/K + 9 ppm of range/K | |||
| Gain Linearity | < 0.01 % | |||
| Input Coupling | DC | |||
| Input Impedance | >10 MΩ | |||
| Overvoltage Protection | In+ to In-: 6 V continuous | |||
| Bandwidth (-3 dB) | 0.433*fs (0.49*fs < 5 kS/s) | |||
| Alias-free Bandwidth | DC to 0.499*fs (DC to 0.44*fs @ < 5 kS/s) | |||
| Alias Rejection | -105 dB (all sample rates) | |||
| Delay Through ADC | 34 / fs | |||
| Oversampling | 128 | |||
| Excitation Voltage | ||||
| Excitation Voltage | Unipolar: 1 V, 2 V, 5 V (see 2) | |||
| Excitation Accuracy | ±3 % of set ±50 mV | |||
| Sense Accuracy | ±0.02 % of reading ±0.25 mV | |||
| Sense Drift | Typical 10 μV/K + 10 ppm of reading/K, max 20 μV/K + 20 ppm of reading/K | |||
| Current Limit | 42 mA | |||
| Protection | Continuous short to ground | |||
| Bridge measurement | ||||
| Bridge Connection Types | full bridge strain, ½ bridge strain, ¼ bridge strain (3-wire) | |||
| Ranges | 2 mV/V ... 1000 mV/V free programmable (software ranges) | |||
| Accuracy | Range \ Excitation | 1 V | 2 V | 5 V |
| 200 mV/V | ±0.035 % of reading ±210 μV/V | ±0.028 % of reading ±105 μV/V | ±0.025 % of reading ±42 μV/V | |
| 20 mV/V | ±0.027 % of reading ±14 μV/V | ±0.027 % of reading ±15 μV/V | ±0.024 % of reading ±6 μV/V | |
| 2 mV/V | ±0.027 % of reading ±14 μV/V | ±0.024 % of reading ±7 μV/V | ±0.022 % of reading ±2.8 μV/V | |
| Internal Bridge Completion | ¼ bridge 120 Ω and 350 Ω | |||
| Bridge Completion Accuracy | ±125 µV/V, TCR: 5 ppm/K (others on request) | |||
| Internal Shunt Resistor | 100 kΩ software selectable to SNS+ or SNS- and In+ or In- | |||
| Shunt Resistor Accuracy | 0.05 %; TCR: 10 ppm/K (others on request) | |||
| Input Short, Sensor Offset Adjust | Software-selectable | |||
| Additional Specifications | ||||
| Rack | IOLITEr-8xSTGS | |||
| Input connectors | Terminal Block Weidmueller SL 2.50-04 | DSUB-37 | ||
| TEDS support | Available | Not available | ||
| Power supply | 9 - 48 V DC | |||
| Power Consumption | Typ. 2.7 W Max. 5.1 W (120 Ω @ 5 V load) | |||
| Weight | 340 g | 330 g | ||
| Slice Dimensions | 128.4 x 118.4 x 30.14 mm | |||
| Additional Specifications | ||||
| Modular | IOLITE-8xSTGS | |||
| Input connectors | Terminal Block Weidmueller SL 2.50-04 | DSUB-37 | ||
| TEDS support | Available | Not available | ||
| Power supply | 9 - 48 V DC | |||
| Power Consumption | Typ. 2.6 W; Max. 4.7 W (120 Ω @ 5 V load) | |||
| Weight | 570 g | 560 g | ||
| Slice Dimensions | 136.75 x 115.0 x 35.0 mm | |||
| Rev: 1697547600 | ||||
| 1) Voltage ranges are describing the input for the bridge mode and are not supported as voltage mode. | ||||
| 2) Ratiometric method | ||||
IOLITEr-8xSTGS (v2) specifications
| Analog inputs | IOLITE STGSv2 | |||
|---|---|---|---|---|
| Input type | Full / half / quarter bridge strain (3-wire, 4-wire), Potentiometer | |||
| Number of channels | 8 | |||
| ADC Type | 24-bit oversampled sigma-delta with anti-aliasing filter | |||
| Sampling Rate | Simultaneous 20 kS/s per channel (software-selectable) | |||
| Voltage Mode | ||||
| Ranges | ±1 V | ±100 mV | ±20 mV | |
| Gain accuracy | 0.02 % of reading | 0.02 % of reading | 0.02 % of reading | |
| Offset accuracy | 210 uV | 30 uV | 14 uV | |
| Typical Noise floor @ 10 kS/s / 1 kS/s / 100 S/s | -107 / -114 / -122 dB | -99 / -107 / -117 dB | -87 / -96 / -108 dB | |
| Typical Dynamic Range @ 10 kS/s / 1 kS/s / 100 S/s | -131/ -139/ -144 dB | -123/ -133/ -141 dB | -111 / -123 / -131 dB | |
| Typ. Crosstalk @ 1 kHz | -137dB | -131dB | -121dB | |
| Typ. CMR @ 50 Hz / 400 Hz / 1 kHz | 81 / 80 / 77 dB | 100 / 92 / 84 dB | 110 / 91 / 84 dB | |
| Gain Drift | Typical 10 ppm/K (max. 20 ppm/K) | |||
| Offset Drift | Typical 0.03 μV/K + 2.5 ppm of range/K, max 0.1 μV/K + 9 ppm of range/K | |||
| Gain Linearity | < 0.01 % | |||
| Input Coupling | DC | |||
| Overvoltage Protection | In+ to In-: 6 V continuous | |||
| Analog input performance | ||||
| Bandwidth (-3 dB) | 0.433*fs (0.49*fs < 5 kS/s) | |||
| Alias-free Bandwidth | DC to 0.499*fs (DC to 0.44*fs @ < 5 kS/s) | |||
| Alias Rejection | -105 dB (all sample rates) | |||
| Delay Through ADC | 34 / fs | |||
| Oversampling | 128 | |||
| Excitation Voltage | ||||
| Excitation Voltage | Unipolar: 1 V, 2 V, 5 V (see 1) | |||
| Excitation Accuracy | ±3 % of set ±50 mV | |||
| Sense Accuracy | ±0.02 % of reading ±0.25 mV | |||
| Sense Drift | Typical 10 μV/K + 10 ppm of reading/K, max 20 μV/K + 20 ppm of reading/K | |||
| Current Limit | 42 mA | |||
| Protection | Continuous short to ground | |||
| Bridge measurement | ||||
| Bridge Connection Types | full bridge strain, ½ bridge strain, ¼ bridge strain (4-wire), ¼ bridge strain (3-wire) | |||
| Ranges | 2 mV/V ... 1000 mV/V free programmable (software ranges) | |||
| Accuracy | Range \ Excitation | 1 V | 2 V | 5 V |
| 200 mV/V | ±0.035 % of reading ±210 μV/V | ±0.028 % of reading ±105 μV/V | ±0.025 % of reading ±42 μV/V | |
| 20 mV/V | ±0.027 % of reading ±14 μV/V | ±0.027 % of reading ±15 μV/V | ±0.024 % of reading ±6 μV/V | |
| 2 mV/V | ±0.027 % of reading ±14 μV/V | ±0.024 % of reading ±7 μV/V | ±0.022 % of reading ±2.8 μV/V | |
| Internal Bridge Completion | ¼ bridge 120 Ω and 350 Ω | |||
| Bridge Completion Accuracy | ±125 µV/V, TCR: 5 ppm/K (others on request) | |||
| Internal Shunt Resistor | 100 kΩ software selectable to SNS+ or SNS- and In+ or In- | |||
| Shunt Resistor Accuracy | 0.05 %; TCR: 10 ppm/K (others on request) | |||
| Input Short, Sensor Offset Adjust | Software-selectable | |||
| Internal switch configurations | Bridge type | Selectable connections | Internal SW on: | Additonal compensation |
| Full | 6 or 4 Wires | 4 wires SNS + and EXC + & SNS- and EXC - | ||
| Half | 5 or 3 Wires | 3 wires SNS + and EXC + & SNS- and EXC - | ||
| QB 3W | SNS + and EXC + & SNS- and EXC - | SNS + and EXC + & SNS- and EXC - | Lead wire compensation available | |
| QB 4W | SNS + and EXC + | SNS + and EXC + | ||
| Sensor current measurement | ||||
| Bridge configuration | Full / Half | QB | ||
| Current measuring line | EXC+ line | R Line | ||
| Current limit | 50 mAdc | |||
| Resolution | 0.1 mAdc | |||
| Additional math calculus | Resistance, Power | |||
| Sample rate | 1 Hz | |||
| Additional Specifications | ||||
| Rack | IOLITEr-8xSTGSv2 | |||
| Input connectors | Terminal block, DSUB-37 | |||
| TEDS support | Available for Terminal block | |||
| Power supply | 9 - 48 V DC | |||
| Power Consumption | Typ. 3.1 W Max. 6 W (120 Ω @ 5 V load) | |||
| Weight | 340 g | |||
| Slice Dimensions | 128.4 x 118.4 x 30.14 mm | |||
| Additional Specifications | ||||
| Modular | IOLITE-8xSTGSv2 | |||
| Input connectors | Terminal block, DSUB-37 | |||
| TEDS support | Available for Terminal block | |||
| Power supply | 9 - 48 V DC | |||
| Power Consumption | Typ. 3.1 W Max. 6 W (120 Ω @ 5 V load) | |||
| Weight | 570 g | |||
| Slice Dimensions | 136.75 x 115.0 x 35.0 mm | |||
| Rev: 1671786000 | ||||
| 1) Ratiometric method | ||||
IOLITEir-4xLV-BAN specifications
| Analog input | ||||
|---|---|---|---|---|
| Input type | Voltage Current (ext. Shunt) | |||
| Number of channels | 4 | |||
| ADC Type | 24-bit, delta-sigma ADC | |||
| Sampling Rate IOLITE | Simultaneous 40 kS/s | |||
| Sampling Rate OBSIDIAN | Simultaneous 20 kS/s | |||
| Voltage Mode | ||||
| Ranges | ±100 V | ±10 V | ±1 V | ±0.1 V |
| Accuracy | ±0.03 % of reading ±0.025 % of range ±40 μV (1 mV for 100 V Range) | |||
| Typ. Noise floor @ 10 kS | -116 dB | -117 dB | -116 dB | -99 dB |
| Typ. THD (40 kS/s, -1 dBFS sine wave @ 1 kHz) | -93 db | -105 dB | -111 dB | -106 dB |
| Typical IMRR @ 50 Hz / 400 Hz / 1 kHz | -99 dB / -80 dB / -71 dB | -112 dB / -112 dB / -111 dB | -129 dB / -130 dB / -118 dB | -110 dB / -128 dB / -114 dB |
| Gain Drift | Typical 10 ppm/K, Max. 40 ppm/K | |||
| Offset Drift | Typical 0.3 μV/K + 2 ppm of range/K, max 0.8 μV/K + 10 ppm of range/K | |||
| Channel Crosstalk | -105 dB | -140 dB | -145 dB | -129 dB |
| Linearity | < 0.01 % | |||
| Input Coupling | DC AC 0.015 Hz | |||
| Input Impedance | 10 MΩ | >1 GΩ | ||
| Analog input performance | ||||
| SR ≤ 1 kS/s | SR > 1 kS/s | |||
| Bandwidth (-3 dB) | 0.49*fs | 0.437*fs | ||
| Alias-free Bandwidth | DC to 0.49*fs | DC to 0.437*fs | ||
| Alias Rejection | -105 dB (all sample rates) | |||
| Delay Through ADC | 34 / fs | |||
| Additional Specifications | ||||
| Rack | IOLITEir-4xLV-BAN | |||
| Input Connectors | Banana | |||
| Power supply | 9 - 48 V DC | |||
| Isolation Voltage | CAT II 1000 V | |||
| Power Consumption | Typ. 4.4 W (Max. 5 W) | |||
| Weight | 280 g | |||
| Slice Dimensions | 128.4 x 127.4 x 30.1 mm | |||
| Operating Temperature | -40 to 85 °C | |||
| Rev: 1775217600 | ||||
IOLITEir-8xLV specifications
| Analog input | ||
|---|---|---|
| Input type | Isolated voltage | |
| Number of channels | 8 | |
| ADC Type | 24-bit oversampled SAR | |
| Sampling Rates | Simultaneous 20k, 10k, 5k, 2k, 1k, 500, 200, 100 S/s (software-selectable) | |
| Voltage Mode | ||
| Ranges | ±100 V | ±10 V |
| Accuracy | ±0.03 % of reading ±50 mV | ±0.03 % of reading ±5 mV |
| Typ. SNR (10 kS/s, -1 dBFS sine wave @ 1 kHz) | 91 dB | 90 dB |
| Typical noise floor @ 10 kS | -105 dB | -98 dB |
| Typ. THD (10 kS/s, -1 dBFS sine wave @ 1 kHz) | -91 dB | -91 dB |
| Typ. SFDR (10 kS/s, -1 dBFS sine wave @ 1 kHz) | 93 dB | 92 dB |
| Typical IMR @ 400 Hz / 1 kHz | 98 dB / 90 dB | 83 dB / 75 dB |
| Gain Drift | Typical 10 ppm/K, Max. 40 ppm/K | |
| Offset Drift | Typical 2 ppm of range/K, Max. 10 ppm of range/K | |
| Gain Linearity | < 0.02 % | |
| Input Coupling | DC | |
| Input Impedance | 1 MΩ | |
| Overvoltage Protection | In+ to In-: 200 V continuous, 350 Vpeak (100 ms) | |
| Analog input performance | ||
| Bandwidth (-3 dB) | 0.49*fs | |
| Alias-free Bandwidth | DC to 0.453*fs | |
| Alias Rejection | -100 dB (all sample rates) | |
| Delay Through ADC | 37 / fs | |
| Oversampling | 32 | |
| Additional Specifications | ||
| Rack | IOLITEir-8xLV | |
| Input Connectors | BNC | Terminal block 2 pole |
| Power supply | 9 - 48 V DC | |
| Isolation Voltage | 450 Vpeak channel to ground & channel to channel | |
| Power Consumption | Typ. 2.4 W, Max. 3.5 W | |
| Weight | 310 g | 270 g |
| Slice Dimensions | 128.4 x 127.6 x 30.1 mm | |
| Additional Specifications | ||
| Modular | IOLITEi-8xLV | |
| Input Connectors | BNC | Terminal block 2 pole |
| Power supply | 9 - 48 V DC | |
| Isolation Voltage | 450 Vpeak channel to ground & channel to channel | |
| Power Consumption | Typ. 2.7 W, Max. 3.3 W | |
| Weight | 545 g | 505 g |
| Slice Dimensions | 137 x 115.0 x 35.0 mm | |
| Rev: 1681991028 | ||
IOLITEr-8xLVe specifications
| Inputs | |||
|---|---|---|---|
| Input types | Voltage, Potentiometer, Current (Ext. shunt) | ||
| ADC Type | 24-bit delta-sigma with anti-aliasing filter (-3 dB @ 160 kHz) Butterworth 2nd order | ||
| Sampling Rate | Simultaneous 20 kS/s | ||
| Voltage Mode | |||
| Ranges | ±100 V | ±5 V | |
| Accuracy | ±0.03 % of reading ±0.02 % of range | ||
| Typ. Noise floor @ 10 kS/sec | -114 dB | -99 dB | |
| Typ. THD (10 kS/s, -1 dBFS sine wave @ 1 kHz) | - | -103 dB | |
| Typ. SFDR (10 kS/s, -1 dBFS sine wave @ 1 kHz) | - | 114 dB | |
| Typ. CMR @ 50 Hz / 400 Hz / 1 kHz | 100 dB / 85 dB / 85 dB | 96 dB / 88 dB / 88 dB | |
| Gain Linearity | <0.02 % | <0.01 % | |
| Gain Drift | Typical 5 ppm/K, Max. 20 ppm/K | ||
| Offset Drift | Typical 5 ppm of range/K, Max. 10 ppm of range/K | ||
| Channel Crosstalk | -130 dB | ||
| Input Coupling | DC | ||
| Input Impedance | 800 kΩ (between In+ and In-) | ||
| Max. common mode voltage | ±200 V | ||
| Input Overvoltage Protection | In+ to In-: 200 V continuous, 350 Vpeak (100 ms) | ||
| Isolation | 250 V functional isolation (channel to power supply ground) | ||
| Analog input performance | |||
| Bandwidth (-3 dB) | 0.433*fs | ||
| Alias-free Bandwidth | DC to 0.499*fs (DC to 0.44*fs @ < 5 kS/s) | ||
| Alias Rejection | -105 dB (all sample rates) | ||
| Delay Through ADC | 34 / fs | ||
| Oversampling | 128 | ||
| Excitation Voltage | |||
| Excitation voltage | Unipolar 2 ... 24V DC | ||
| Current limit | 2.8 W max. power per channel , 22.4 W per slice (MAX: 500mA ) | ||
| Excitation accuracy | 0.05 % ± 2 mV | ||
| Drift | ±20 ppm/K ± 60 μV/K | ||
| Stability 10 % to 90 % load | < 0.06 % @ 5 V exc (< 0.01 % @ 24 V exc) | ||
| Noise @ 10 V / 350 Ω | <150 μVrms @ 10 kS | ||
| Line regulation over 10 Ω of change | < 0.04 % @ 120 Ω load | ||
| Protection | Continuous short to ground | ||
| Additional Specifications | |||
| Rack | IOLITEr-8xLVe | ||
| Input connector | Terminal block | DSUB-37 | |
| TEDS support | Yes | ||
| Power supply | 9 - 48 V DC | ||
| Power consumption | Analog input only: Typ. 3.3 W (Max. 4.2 W) With Excitation: 6.9 W 120 Ω load @ 5 V, 38 W 120 Ω load @ 20 V | ||
| Weight | 310 g | 300 g | |
| Slice Dimensions | 128.4 x 118.4 x 30.14 mm | ||
| Additional Specifications | |||
| Modular | IOLITE-8xLVe | ||
| Input connector | Terminal block | DSUB-37 | |
| TEDS support | Yes | ||
| Power supply | 9 - 48 V DC | ||
| Power consumption | Analog input only: Typ. 3.3 W (Max. 4.2 W) With Excitation: 6.9 W 120 Ω load @ 5 V, 38 W 120 Ω load @ 20 V | ||
| Weight | 550 g | 540 g | |
| Slice Dimensions | 136.75 x 115.0 x 35.0 mm | ||
| Rev.: 1721714400 | |||
IOLITEr-16xLV specifications
| Analog input | ||
|---|---|---|
| Input type | Voltage | |
| Number of channels | 16 | |
| ADC Type | 24-bit oversampled sigma-delta | |
| Sampling Rates IOLITE | Simultaneous 20k, 10k, 5k, 2k, 1k, 500, 200, 100 S/s (software-selectable) | |
| Sampling Rates OBSIDIAN | Simultaneous 5k, 2k, 1k, 500, 200, 100 S/s (software-selectable) | |
| Voltage Mode | ||
| Ranges | ±200 V | ±10 V |
| Accuracy | ±0.03 % of reading ±40 mV | ±0.03 % of reading ±2 mV |
| Typ. SNR (10 kS/s, -1 dBFS sine wave @ 1 kHz) | 103 dB | 101 dB |
| Typical noise floor @ 10 kS | -115 dB | -105 dB |
| Typ. THD (10 kS/s, -1 dBFS sine wave @ 1 kHz) | -93 dB | -108 dB |
| Typ. SFDR (10 kS/s, -1 dBFS sine wave @ 1 kHz) | 95 dB | 111 dB |
| Typical CMR @ 400 Hz / 1 kHz | 88 dB / 87 dB | 82 dB / 80 dB |
| Crosstalk @ 1 kHz | -118 dB | -120 dB |
| Gain Drift | Typical 10 ppm/K, Max. 30 ppm/K | |
| Offset Drift | Typical 2 ppm of range/K, Max. 5 ppm of range/K | |
| Gain Linearity | < 0.01 % | |
| Input Coupling | DC | |
| Input Impedance | 800 kΩ | |
| Max. common mode voltage | ±200 V | |
| Overvoltage Protection | In+ to In-: 200 V continuous, 350 Vpeak (100 ms) | |
| Analog input performance | ||
| Bandwidth (-3 dB) | 0.433*fs | |
| Alias-free Bandwidth | DC to 0.45*fs (DC to 0.499*fs @ 20 kS/s) | |
| Alias Rejection | -105 dB (all sample rates) | |
| Delay Through ADC | 34 / fs | |
| Oversampling | 128 | |
| Additional Specifications | ||
| Rack | IOLITEr-16xLV | |
| Input Connectors | Terminal block 9 pole OMNIMATE SL 2.50 / BLF 2.50/180 | |
| Power supply | 9 - 48 V DC | |
| Isolation Voltage | 250 V functional isolation (channel to power supply ground) | |
| Power Consumption | Typ. 3.4 W (Max. 3.9 W) | |
| Weight | 230 g | |
| Slice Dimensions | 128.4 x 115.4 x 30.1 mm | |
| Modular | IOLITE-16xLV | |
| Input Connectors | Terminal block 9 pole OMNIMATE SL 2.50 / BLF 2.50/180 | |
| Power supply | 9 - 48 V DC | |
| Isolation Voltage | 250 V functional isolation (channel to power supply ground) | |
| Power Consumption | Typ. 3.6 W (Max. 4.5 W) | |
| Weight | 530g | |
| Slice Dimensions | 136.75 x 115.0 x 35.0 mm | |
| Rev: 1685344484 | ||
IOLITEr-16xLV-10V specifications
| Analog input | |
|---|---|
| Input type | Voltage |
| Number of channels | 16 |
| ADC Type | 24-bit oversampled sigma-delta |
| Sampling Rates IOLITE | Simultaneous 20k, 10k, 5k, 2k, 1k, 500, 200, 100 S/s (software-selectable) |
| Sampling Rates OBSIDIAN | Simultaneous 5k, 2k, 1k, 500, 200, 100 S/s (software-selectable) |
| Voltage Mode | |
| Ranges | ±10 V |
| Accuracy | ±0.03 % of reading ±2 mV |
| Typ. SNR (10 kS/s, -1 dBFS sine wave @ 1 kHz) | 99 dB |
| Typical noise floor @ 0.10/1/10/20 kS/s | 116dB/106dB/97dB/94dB |
| Typ. THD (10 kS/s, -1 dBFS sine wave @ 1 kHz) | -108 dB |
| Typ. SFDR (10 kS/s, -1 dBFS sine wave @ 1 kHz) | 110 dB |
| Typical CMR @ 50Hz / 400 Hz / 1 kHz | 92 dB / 92 dB / 88dB |
| Crosstalk @ 1 kHz | -120 dB |
| Gain Drift | Typical 10 ppm/K, Max. 30 ppm/K |
| Offset Drift | Typical 2 ppm of range/K, Max. 5 ppm of range/K |
| Gain Linearity | < 0.01 % |
| Input Coupling | DC |
| Input Impedance | 2MEG between In+ and In- |
| Max. common mode voltage | ±200 V |
| Overvoltage Protection | In+ to In-: 200 V continuous, 350 Vpeak (100 ms) |
| Analog input performance | |
| Bandwidth (-3 dB) | 0.433*fs |
| Alias-free Bandwidth | DC to 0.45*fs (DC to 0.499*fs @ 20 kS/s) |
| Alias Rejection | -105 dB (all sample rates) |
| Delay Through ADC | 34 / fs |
| Oversampling | 128 |
| Additional Specifications | |
| Rack | IOLITEr-16xLV-10 |
| Input Connectors | Terminal block 9 pole OMNIMATE SL 2.50 / BLF 2.50/180 |
| Power supply | 9 - 48 V DC |
| Isolation Voltage | 250 V functional isolation (channel to power supply ground) |
| Power Consumption | Typ. 3.15 W (Max. 3.65 W) |
| Weight | 230 g |
| Slice Dimensions | 128.4 x 115.4 x 30.1 mm |
| Modular | IOLITE-16xLV-10 |
| Input Connectors | Terminal block 9 pole OMNIMATE SL 2.50 / BLF 2.50/180 |
| Power supply | 9 - 48 V DC |
| Isolation Voltage | 250 V functional isolation (channel to power supply ground) |
| Power Consumption | Typ. 3.45 W (Max. 4.35 W) |
| Weight | 530g |
| Slice Dimensions | 136.75 x 115.0 x 35.0 mm |
| Rev: | |
IOLITEir-4xHV specifications
| Analog inputs | IOLITEi 4xHV | |||
|---|---|---|---|---|
| Input type | Voltage | |||
| Number of channels | 4 | |||
| ADC Type | 24-bit delta-sigma with anti-aliasing filter | |||
| Sampling Rate IOLITE | Simultaneous 40 kS/s per channel (software-selectable) | |||
| Sampling Rate OBSIDIAN | Simultaneous 20 kS/s per channel (software-selectable) | |||
| Anti-aliasing filter | -3 dB @ 150 kHz (Butterworth 3rd order) | |||
| Voltage mode | ||||
| Ranges | ±2000 V | ±1000 V | ±400 V | ±200 V |
| Input Accuracy | ±0.03 % of reading ±0.02 % of range | |||
| Typ. Dynamic Range @ 10 kS | -137 dB | -137 dB | -137 dB | - 132 dB |
| Typ. Noise floor @ 10 kS | -113 dB | -112 dB | -109 dB | -104 dB |
| Typ. THD (40 kS/s, -6 dBFS sine wave @ 1 kHz) | N/A | N/A | N/A | -103 dB |
| Typ. CMR @ 20 kS/s | 80 dB @ 50 Hz, 64 dB @ 400 Hz | |||
| Gain Drift | Typical 10 ppm/K, max. 25 ppm/K | |||
| Offset Drift | Typical 100 μV/K + 2 ppm of range/K, max 200 μV/K + 4 ppm of range/K | |||
| Gain Linearity | < 0.02% (2000 V range), < 0.01% (other ranges) | |||
| Input Coupling | DC | |||
| Input Impedance | 10 MΩ | |||
| Isolation voltage | CAT II 1000 V | |||
| Analog input performance | ||||
| Bandwidth (-3 dB) | 0.433*fs (0.49*fs < 5 kS/s) | |||
| Alias-free Bandwidth | DC to 0.499*fs (DC to 0.45*fs @ < 5 kS/s) | |||
| Alias Rejection | -105 dB (all sample rates) | |||
| Delay Through ADC | 34 / fs | |||
| Oversampling | 128 | |||
| Additional Specifications | ||||
| Rack | IOLITEir-4xHV | |||
| Input connector | 4mm safety banana jack | |||
| Power supply | 9 - 48 V DC | |||
| Power Consumption | Typ. 2.6 W (max. 3 W) | |||
| Weight | 310 g | |||
| Slice Dimensions | 128.4 x 115.4 x 30.1 mm | |||
| Additional Specifications | ||||
| Modular | IOLITEi-4xHV | |||
| Input connector | 4mm safety banana jack | |||
| Power supply | 9 - 48 V DC | |||
| Power Consumption | Typ. 2.6 W (max. 3 W) | |||
| Weight | 575 g | |||
| Slice Dimensions | 136.75 x 127.2 x 35.0 mm | |||
| Rev: 1701151200 | ||||
IOLITEir-8xLA specifications
| Analog input | ||
|---|---|---|
| Input type | Isolated current | |
| Number of channels | 8 | |
| ADC Type | 24-bit oversampled SAR | |
| Sampling Rates | Simultaneous 20k, 10k, 5k, 2k, 1k, 500, 200, 100 S/s (software-selectable) | |
| Current Mode | ||
| Ranges | ±20 mA | ±2 mA |
| Accuracy | ±0.05 % of reading ±10 µA | ±0.05 % of reading ±2 µA |
| Typical noise floor @ 10 kS/sec | -100 dB | -100 dB |
| Crosstalk (20 kS/sec, -1 dBFS @ 1 kHz) | -121 dB | -128 dB |
| Gain Drift | Typical 10 ppm/K, Max. 40 ppm/K | |
| Offset Drift | Typical 5 ppm of range/K, Max. 15 ppm of range/K | |
| Gain Linearity | < 0.05 % | |
| Input Coupling | DC | |
| Input Impedance | 50 Ω | |
| Overvoltage Protection | In+ to In-: 30 V continuous | |
| Overcurrent Protection | In+ to In-: 70 mA continuous | |
| Analog input performance | ||
| Bandwidth (-3 dB) | 0.49 fs | |
| Alias-free Bandwidth | DC to 0.453 fs | |
| Alias Rejection | -100 dB (all sample rates) | |
| Delay Through ADC | 37 / fs | |
| Oversampling | 32 | |
| Additional Specifications | ||
| Power supply | 9 - 48 V DC | |
| Isolation Voltage | 450 Vpeak channel to ground & channel to channel | |
| Power Consumption | Typ. 2.8 W, Max. 3.5 W | |
| Input Connectors | BNC | Terminal block 2 pole |
| Weight | 310 g | 270 g |
| Slice Dimensions | 128.4 x 127.6 x 30.1 mm | 128.4 x 115.4 x 30.1 mm |
| Rev: 1632117100 | ||
IOLITEr-16xLA specifications
| Analog input | |
|---|---|
| Input type | Current External loop-powered 4-20 mA |
| Number of channels | 16 |
| ADC Type | 24-bit oversampled sigma-delta |
| Sampling Rates IOLITE | Simultaneous 20k, 10k, 5k, 2k, 1k, 500, 200, 100 S/s (software-selectable) |
| Sampling Rates OBSIDIAN | Simultaneous 5k, 2k, 1k, 500, 200, 100 S/s (software-selectable) |
| Current Input Mode | |
| Ranges | ±20 mA |
| Accuracy | ±0.05 % of reading ±10 µA |
| Typ. SNR (10 kS/s, -1 dBFS sine wave @ 1 kHz) | 112 dB |
| Typical noise floor @ 10 kS | -91 dB |
| Crosstalk @ 1 kHz | -120 dB |
| Gain Drift | Typical 3 ppm/K, Max. 9 ppm/K |
| Offset Drift | Typical 4 ppm of range/K, Max. 12 ppm of range/K |
| Gain Linearity | < 0.02 % |
| Input Coupling | DC |
| Input Impedance | 100 Ω |
| Overvoltage Protection | In+ to In-: 30 V continuous |
| Overcurrent Protection | In+ to In-: 50 mA continuous |
| Analog input performance | |
| Bandwidth (-3 dB) | 0.433*fs |
| Alias-free Bandwidth | DC to 0.45*fs (DC to 0.499*fs @ 20 kS/s) |
| Alias Rejection | -105 dB (all sample rates) |
| Delay Through ADC | 34 / fs |
| Oversampling | 128 |
| Additional Specifications | |
| Rack | IOLITEr-16xLA |
| Input Connectors | Terminal block 9 pole (OMNIMATE SL 2.50 / BLF 2.50/180), D-SUB37 female |
| Power supply | 9 - 48 V DC |
| Isolation Voltage | 250 Vpeak channel to ground (functional isolation) |
| Power Consumption | Typ. 2.9 W (Max. 3.3 W) |
| Weight | 310 g |
| Slice Dimensions | 128.4 x 115.4 x 30.1 mm |
| Modular | IOLITE-16xLA |
| Input Connectors | Terminal block 9 pole (OMNIMATE S2L 3.50/18/180F), D-SUB37 female |
| Power supply | 9 - 48 V DC |
| Isolation Voltage | 250 Vpeak channel to ground (functional isolation) |
| Power Consumption | Typ. 2.9 W (Max. 3.3 W) |
| Weight | 560 g |
| Slice Dimensions | 136.75 x 115.0 x 35.0 mm |
| Rev: 1705294800 | |
IOLITEhir-4xTH specifications
| Analog inputs | IOLITEhi-4xTH | |
|---|---|---|
| Input type | Thermocouple | |
| Number of channels | 4 | |
| ADC Type | 24-bit delta-sigma | |
| Sampling Rate IOLITE | Simultaneous 10, 20, 40, 50, 80, 100, 200, 500, 1k, 2k, 5k, 10k S/s | |
| Sampling Rate OBSIDIAN | 100 S/s | |
| Voltage mode | ||
| Voltage Ranges | ±100 mV | |
| Input Accuracy | ±0.02 % of reading ±7 μV | |
| Typical Noise Floor @ 10 / 100 / 1k / 10k S/s | -120 dB / -110 dB / -100 dB / -90 dB | |
| Offset Drift | Typ. 0.01 μV/K (Max. 0.1 μV/K) | |
| Gain Drift | Typ. 2 ppm/K (Max. 10 ppm/K) | |
| Gain Linearity | < 0.007% | |
| Input Coupling | DC | |
| Input Impedance | 100 MΩ | |
| Thermocouple | ||
| Input type | TC Type: K | |
| Accuracy | ±0.02 % of reading ± 0.5 °C ± 7 μV | |
| Typical Noise | 0.002 °C RMS@Type K @ 10 S/sec 0.006 °C RMS@Type K @ 100 S/sec 0.02 °C RMS@Type K @ 1k S/sec 0.05 °C RMS@Type K @ 10k S/sec | |
| Additional Specifications | ||
| Rack | IOLITEhir-4xTH-HV | |
| Input connector | K-type Thermocouple LEMO REDEL 2P series | |
| Isolation voltage | CAT III 600 V, CAT II 1000 V | |
| Power supply | 9 - 48 V DC | |
| Power consumption | Typ. 2.6 W (Max. 3.7 W) | |
| Weight | 270 g | |
| Dimensions | 128.4 x 115.4 x 30.1 mm | |
| Modular | IOLITEhi-4xTH-HV | |
| Input connector | K-type Thermocouple LEMO REDEL 2P series | |
| Isolation voltage | CAT III 600 V, CAT II 1000 V | |
| Power supply | 9 - 48 V DC | |
| Power consumption | Typ. 2.6 W (Max. 3.7 W) | |
| Weight | 270 g | |
| Dimensions | 128.4 x 115.4 x 30.1 mm | |
| Rev: 1676980800 | ||
IOLITEir-8xTH specifications
| Inputs | ||
|---|---|---|
| Input type | Isolated universal thermocouple and voltage | |
| Number of channels | 8 | |
| ADC Type | 24-bit delta-sigma | |
| Sampling Rate | Simultaneous 100 S/s | |
| Voltage Mode | ||
| Ranges | ±1 V | ±100 mV |
| Input Accuracy | ±0.02 % of reading ±100 μV | ±0.02 % of reading ±10 μV |
| Typical Noise floor @ 10/100 S/sec | -114 dB / -105 dB | -109 dB / -100 dB |
| Gain Drift | Typical 4 ppm/K (max. 10 ppm/K) | Typical 4 ppm/K (max. 10 ppm/K) |
| Offset Drift | Typical 0.05 μV/K (max. 0.2 μV/K) | Typical 0.05 μV/K (max. 0.2 μV/K) |
| Gain Linearity | <0.01% | <0.01% |
| Input Coupling | DC | DC |
| Input Impedance | 100 MΩ | 100 MΩ |
| Temperature Mode | ||
| Thermocouple Types | K, J, T, R, S, N, E, C, B | |
| Accuracy | ±0.02 % of reading ±0.5 °C ±10 μV | |
| Sampling rates | 10, 20, 40, 80, 100 S/sec | |
| Typical Noise | 0.007 °C RMS@Type K @ 10 S/sec 0.02 °C RMS@Type K @ 100 S/sec | |
| Additional Specifications | ||
| Input connectors | Mini Thermocouple connector (copper) | |
| Isolation voltage | 1000 Vpeak channel to ground & channel to channel | |
| Power supply | 9 - 48 V DC | |
| Power consumption | 3.2 W | |
| Weight | 230 g | |
| Slice Dimensions | 128.4 x 115.4 x 30.1 mm | |
| Rev: 1578677400 | ||
IOLITEir-8xTH-HS specifications
| Inputs | ||
|---|---|---|
| Input type | Isolated universal thermocouple and voltage | |
| Number of channels | 8 | |
| ADC Type | 24-bit delta-sigma | |
| Sampling Rate IOLITE | Simultaneous 10, 20, 40, 50, 80, 100, 200, 500, 1k, 2k, 5k, 10k S/s (license for high rates) | |
| Sampling Rate OBSIDIAN | Simultaneous 10, 20, 40, 50, 80, 100, 200, 500, 1k, 2k, 5k (license for high rates) | |
| Voltage Mode | ||
| Ranges | ±1 V | ±100 mV |
| Input Accuracy | ±0.02 % of reading ±50 μV | ±0.02 % of reading ±7 μV |
| Typical Noise Floor @ 10 / 100 / 1k / 10k S/s | -130 dB / -120 dB / -110 dB / -104 dB | -120 dB / -110 dB / -100 dB / -90 dB |
| Offset Drift | Typ. 0.05 μV/K (Max. 0.2 μV/K) | Typ. 0.01 μV/K (Max. 0.1 μV/K) |
| Gain Drift | Typ. 2 ppm/K (Max. 10 ppm/K) | |
| Gain Linearity | < 0.007% | |
| Input Coupling | DC | |
| Input Impedance | 100 MΩ | |
| Temperature Mode | ||
| Thermocouple Types | K, J, T, R, S, N , E, C, B | |
| CJC reference | PT1000 | |
| Accuracy | ±0.02 % of reading ±0.5 °C ±7 μV | |
| Typical Noise | 0.002 °C RMS @ Type K @ 10 S/s 0.006 °C RMS @ Type K @ 100 S/s 0.02 °C RMS @ Type K @ 1 kS/s 0.05 °C RMS @ Type K @ 10 kS/s | |
| Additional Specifications | ||
| Rack | IOLITEir-8xTH-HS | |
| Input connectors | Mini Thermocouple connector (copper) | |
| Isolation voltage | 1000 Vpeak channel to ground & channel to channel | |
| Power supply | 9 - 48 V DC | |
| Power consumption | Typ. 2.6 W (Max. 3.4 W) | |
| Weight | 290 g | |
| Slice Dimensions | 128.4 x 115.4 x 30.1 mm | |
| Additional Specifications | ||
| Modular | IOLITEi-8xTH-HS | |
| Input connectors | Mini Thermocouple connector (copper) | |
| Isolation voltage | 1000 Vpeak channel to ground & channel to channel | |
| Power supply | 9 - 48 V DC | |
| Power consumption | Typ. 2.6 W (Max. 3.8 W) | |
| Weight | 555 g | |
| Slice Dimensions | 137 x 115.0 x 35.0 mm | |
| Rev: 1701151200 | ||
IOLITEir-8xRTD-HS specifications
| Analog inputs | ||
|---|---|---|
| Input types | Voltage, Resistance or Universal PTxxx inputs | |
| Number of channels | 8 | |
| ADC Type | 24-bit delta-sigma | |
| Sampling Rate IOLITE | Simultaneous 10, 20, 40, 50, 80, 100, 200, 500, 1k, 2k, 5k, 10k S/s (license for high rates) | |
| Sampling Rate OBSIDIAN | Simultaneous 10, 20, 40, 50, 80, 100, 200, 500, 1k, 2k, 5k (license for high rates) | |
| Voltage mode | ||
| Ranges | ±1 V | ±100 mV |
| Input Accuracy | ±0.02 % of reading ±0.01 % of range | ±0.02 % of reading ±0.02 % of range |
| Typical Noise Floor @ 10 / 100 / 1k / 10k S/s | -130 dB / -120 dB / -110 dB / -104 dB | -120 dB / -108 dB / -98 dB / -90 dB |
| Offset Drift | Typ. 0.05 μV/K (Max. 0.2 μV/K) | Typ. 0.01 μV/K (Max. 0.1 μV/K) |
| Gain Drift | Typ. 2 ppm/K (Max. 10 ppm/K) | |
| Gain Linearity | < 0.007 % | |
| Input Coupling | DC | |
| Input Impedance | 10 MΩ | |
| Temperature mode | ||
| Ranges | PT500, PT1000, PT2000 | PT100, PT200 |
| Excitation current | 150 uA | 400 uA |
| Measurement range | -200 °C to +850 °C | |
| Accuracy | ±0.02 % of reading ±0.2 °C | |
| Temperature Drift | Typ. ±2 ppm/K (Max. ±10 ppm/K) | |
| Input Connection | 3-wire or 4-wire | |
| Resistance mode | ||
| Ranges | 0…10 kΩ | 0…1 kΩ |
| Excitation current | 150 uA | 400 uA |
| Accuracy | ±0.02 % of reading ±0.008 % of range | |
| Input Connection | 3-wire or 4-wire | |
| Additional Specifications | ||
| Rack | IOLITEir-8xRTD-HS | |
| Input connectors | Lemo 0B 6pin EEA.0B.306.CLN Waterproof version: Lemo 0T 6pin EEA.0T.306.CLN | Terminal Block Weidmueller SL 2.50-04 |
| Isolation voltage | 1000 Vpeak channel to ground & channel to channel | |
| Power supply | 9 - 48 V DC | |
| Power consumption | Typ. 3.5 W (Max. 4.3 W) | |
| Weight | 325 g | 225 g |
| Dimensions | 128.4 x 115.4 x 30.1 mm | |
| Additional Specifications | ||
| Modular | IOLITEi-8xRTD-HS | |
| Input connectors | Lemo 0B 6pin EEA.0B.306.CLN | Terminal Block Weidmueller SL 2.50-04 |
| Isolation voltage | 1000 Vpeak channel to ground & channel to channel | |
| Power supply | 9 - 48 V DC | |
| Power consumption | Typ. 3.5 W (Max. 4.7 W) | |
| Weight | 585 g | 485 g |
| Dimensions | 137 x 115.0 x 35.0 mm | |
| Rev: 1765454400 | ||
IOLITEr-8xDI-4xDO specifications
| Digital Input | |
|---|---|
| Isolated Input Channels | 8 |
| Input low level | -1 V ~ +1 V |
| Input high level | -48 V ~ -3 V, +3 V ~ +48 V |
| Input high current @ 5 V | UIN < 1 mA |
| Input high current @ 30 V | UIN < 1 mA |
| Sampling rate | Simultaneous 40 kS/s |
| Overvoltage protection | 100 V continuous (250 Vpeak) |
| Isolation voltage peak | 1000 V Isolated common GND |
| Digital Output | |
| Isolated Output Channels | 4 |
| Compatibility | Solid state relay |
| Maximum sink current | 0.5 A |
| Maximum switching voltage | 50 V |
| Maximum update rate | depending on EtherCAT master |
| Isolation voltage peak | 1000 V Isolated common GND |
| Additional Specifications | |
| Input connectors | Terminal block (3 x 2 pole, 2 x 12 pole, 1 x 8 pole) OMNIMATE SL 2.50 / BLF 2.50/180 |
| Power supply | 9 - 48 V DC |
| Power out | max. 2 A (unprotected) |
| Power consumption | 1.8 |
| Weight | 184 g |
| Slice Dimensions | 128.4 x 115.4 x 30.1 mm |
| Rev: 1781866800 | |
IOLITEr-32xDI specifications
| Digital Input | |
|---|---|
| Isolated Input Channels | 32 |
| Input low level | -1 V ~ +1 V |
| Input high level | -48 V ~ -3 V, +3 V ~ +48 V |
| Input high current @ 5 V | UIN < 1 mA |
| Input high current @ 30 V | UIN < 1 mA |
| Sampling rate | Simultaneous 20 kS/s |
| Overvoltage protection | 100 V continuous (250 Vpeak) |
| Isolation voltage peak | 1000 V Isolated common GND, groups of 8 ch |
| Additional Specifications | |
| Input connectors | Terminal block, 2 pole, 4 x 9 pole OMNIMATE SL 2.50 / BLF 2.50/180 |
| Power supply | 9 - 48 V DC |
| Power out | max. 2 A (unprotected) |
| Power consumption | Typ. 1.2 W, Max. 1.9 W |
| Weight | 220 g |
| Slice Dimensions | 128.4 x 115.4 x 30.1 mm |
| Rev: 1668186000 | |
IOLITEr-32xDO specifications
| Digital Output | |
|---|---|
| Isolated Output Channels | 32 |
| Compatibility | Solid state relay |
| Maximum sink current | 0.5 A |
| Maximum switching voltage | 50 V |
| Maximum update rate | depending on EtherCAT master |
| Isolation voltage peak | 1000 V Isolated common GND, groups of 8 ch |
| Additional Specifications | |
| Input connectors | Terminal block, 2 pole, 4 x 9 pole OMNIMATE SL 2.50 / BLF 2.50/180 |
| Power supply | 9 - 48 V DC |
| Power out | max. 2 A (unprotected) |
| Power consumption | Typ. 1.2 W, Max. 2.0 W |
| Weight | 230 g |
| Slice Dimensions | 128.4 x 115.4 x 30.1 mm |
| Rev: 1781866800 | |
IOLITEr-4xCNT specifications
| Counter | |
|---|---|
| Number of channels | 4 |
| Timebase | 100 MHz |
| Time base accuracy | Typical: 5 ppm, Max: 20 ppm |
| Max. bandwidth | 10 MHz |
| Input Filter | 500 ns, 1 μs, 2 μs, 4 μs, 5 μs and 7.5 μs |
| Input Level Compatibility | TTL (Low: <0.8 V, High >2 V) |
| Input impedance | 100 kΩ pull-up to +3.3 V |
| Input Protection | ±25 V continuous |
| Counter modes | counting, waveform timing, encoder, gear-tooth sensor |
| Isolated Sensor Power Supply | |
| Output Voltage +5 V | 5.1 V +/- 10 % 100 mA max. (400 mA to supply 4 channels) |
| Output Voltage +12 V | 12.1 V +/-5 % 50 mA max. (100 mA to supply 4 channels) |
| Isolation | Ch to GND isolation |
| Additional Specifications | |
| Rack | IOLITEr-4xCNT |
| Input connector | 7-pin LEMO 1B series EEG.1B.307.CLNY Waterproof version: 7-pin Lemo 1T series EEG.1T.307.CLN |
| Power Consumption @ 24 V | Typ. 1.9 W |
| Power supply | 9 - 48 V DC |
| Slice Dimensions | 128.4 x 115.4 x 30.1 mm |
| Modular | IOLITE-4xCNT |
| Input connector | 7-pin LEMO 1B series EEG.1B.307.CLNY |
| Power Consumption @ 24 V | Typ. 1.9 W |
| Power supply | 9 - 48 V DC |
| Slice Dimensions | 137 x 115.0 x 35.0 mm |
IOLITEr-16xAO technical specifications
| Analog Output | |
|---|---|
| Output types | Voltage |
| Number of channels | 16 |
| Output connector | Terminal block |
| DAC Resolution | 16-bit |
| Update Rate | 20000 S/s (see 1) |
| Full Scale Output Range | ±10 V (see 2) |
| Accuracy (typ.) | 0.05 % ±1.5 mV (0 to 20 mA load) |
| Temperature gain drift (typ.) | ±10 ppm/K |
| Temperature offset drift (typ.) | ±5 uV/K |
| Output impedance | < 1 Ω DC |
| Minimal load resistance | 500 Ω load @ 10 V output |
| Settling time (0.05 % of set value) | 140 µs |
| Max slew rate | 0.20 V/µs |
| Max capacitive load (2 % overshoot) | 10 nF |
| Current limit | 20 mA per channel (160 mA per module) |
| Output protection | Short circuit protected |
| Overvoltage protection | AO to iGND: 11 V continuous, 100 Vpeak (10 msec) |
| Isolation | 200 Vdc or 150 Vacrms channel to ground |
| Additional Specifications | |
| Output connectors | Terminal block, 2 pole, 4 x 9 pole OMNIMATE SL 2.50 / BLF 2.50/180 |
| Power supply | 9 - 48 V DC |
| Power out | max. 2 A |
| Power consumption | Typ. 4.5 W, Max. 8 W |
| Weight | 210 g |
| Slice Dimensions | 128.4 x 115.4 x 30.1 mm |
| Rev: 1770714000 | |
| 1) Only asynchronous mode is supported in Dewesoft, function generator is not supported | |
| 2) Only DC output voltage, function generator not supported | |
IOLITEir-4xCAN-FD technical specifications
| CAN ports specification | |
|---|---|
| Number of ports | 4 |
| Interface type | CAN-FD, CAN 2.0B |
| CAN 2.0B data rate | Up to 1 Mbit/s (see 1) |
| CAN-FD data rate | Up to 5 Mbit/s (see 1) |
| Special protocols | J1939, XCP, OBDII |
| Galvanic isolation | Isolated |
| Bus pin fault protection | ±36 V |
| Oscillator frequency | 40 MHz |
| Power specification | |
| Power supply | 9 - 48 V DC |
| Maximum sensor power consumption | 5 V / max. 250 mA (1.25 W) 12 V / max. 90 mA (1.1 W) |
| Maximum power consumption | max. 2 W (no load) |
| Sensor supply isolation | Channel to ground |
| Additional Specifications | |
| Rack | IOLITEir-4xCAN-FD |
| Input connector | DSUB9 |
| Power supply | 9 - 48 V DC |
| Weight | 310 g |
| Slice Dimensions | 128.4 x 118.4 x 30.14 mm |
| Additional Specifications | |
| Modular | IOLITEi-4xCAN-FD |
| Input connector | DSUB9 |
| Power supply | 9 - 48 V DC |
| Weight | 550 g |
| Slice Dimensions | 136.75 x 115.0 x 35.0 mm |
| 1) For more information please see the manual. | |