KRYPTON® Technical Specifications
- KRYPTON® ONE general specifications
- KRYPTON® MULTI general system specs
- KRYPTON® ONE amplifier overview
- KRYPTON® MULTI amplifier overview
- KRYPTONi-1xACC specifications
- KRYPTONi-1xSTG specifications
- KRYPTONi-1xLV specifications
- KRYPTONi-1xHV specifications
- KRYPTONi-1xTH-HV specifications
- KRYPTON-1xCNT specifications
- KRYPTON-1xAO specifications
- KRYPTONi-4xDI specifications
- KRYPTONi-4xDO specifications
- KRYPTONi-1xCAN technical specifications
- KRYPTONi-1xCAN-FD technical specifications
- KRYPTON-8xACC specifications
- KRYPTON-4xACC specifications
- KRYPTON-3xSTG specifications
- KRYPTON-6xSTG specifications
- KRYPTONi-4xLV specifications
- KRYPTONi-8xLV specifications
- KRYPTONi-8xLA specifications
- KRYPTONi-8xTH specifications
- KRYPTONi-16xTH specifications
- KRYPTON TH accuracy
- KRYPTONi-8xTH-HS specifications
- KRYPTONi-16xTH-HS specifications
- KRYPTON TH-HS accuracy
- KRYPTONi-8xRTD specifications
- KRYPTONi 8xRTD-HS specifications
- KRYPTON-4xCNT specifications
- KRYPTONi-8xDI-8xDO specifications
- KRYPTONi-16xDO specifications
- KRYPTONi-16xDI specifications
KRYPTON® ONE general specifications
| Interface | |
|---|---|
| Data interface | EtherCAT® |
| Data rate | 100Mbps Full Duplex bus speed |
| Sync Accuracy | below 1 sample to Sirius® |
| EtherCAT® | Max. Cable Length 100 m |
| Max. Throughput per Chain | From 6 MB/s to 10 MB/s |
| Max. Number of Units | 64 or more (power injectors may be required) |
| Data Interface Connection | LEMO 1T EtherCAT® hybrid cable Single cable for data, power and sync, daisy chainable |
| Power | |
| Power Supply | 9 - 50 V DC |
| Power consumption | 1 - 3 W |
| Environmental | |
| Operating Temperature | -40 to 85 °C |
| Storage Temperature | -40 to 85 °C |
| Humidity | 5 to 100 % RH |
| IP rating | IP65 / IP67 (see individual specifications for details) |
| Shock | Shock: SIST EN 60068-2-27:2009 100 g, 6 ms, half-sine 100x pos./neg in each axis 600x total |
| Vibration | Random vibration: MIL-STD-202G Method 214A, test condition II-D Profile break points: 50 Hz @ 0.025 g²/Hz 100 Hz @ 0.1 g²/Hz 2000 Hz @ 0.1 g²/Hz Profile RMS / Peak = 13.9 g / 41.7 g 8 hrs in each axis, 24 hrs total |
| Rev: 1557765000 | |
KRYPTON® MULTI general system specs
| Interface | |
|---|---|
| Data interface | EtherCAT® |
| Data Rate | 100 Mbps Full Duplex bus speed |
| Max. Throughput per Chain | From 6 MB/s to 10 MB/s (see 1) |
| Sync Accuracy | below 1 sample to Sirius® |
| EtherCAT® | Max. Cable Length 100 m |
| Max. Number of Units | 64 or more (power injectors may be required) |
| Data Interface Connection | LEMO 1T EtherCAT® hybrid cable Single cable for data, power and sync, daisy chainable |
| Additional Specifications | |
| Power supply ground to mechanical ground (housing) resistance | > 100 kΩ |
| Power | |
| Power Supply | 9 - 48 V DC |
| Environmental | |
| Operating Temperature | -40 to 85°C |
| Storage Temperature | -40 to 85°C |
| Humidity | 5 to 100 % RH |
| IP rating | IP65 / IP67 |
| Shock | Shock: SIST EN 60068-2-27:2009 100 g, 6 ms, half-sine 100x pos./neg in each axis 600x total |
| Vibration | Random vibration: MIL-STD-202G Method 214A, test condition II-D Profile break points: 50 Hz @ 0.025 g²/Hz 100 Hz @ 0.1 g²/Hz 2000 Hz @ 0.1 g²/Hz Profile RMS / Peak = 13.9 g / 41.7 g 8 hrs in each axis, 24 hrs total |
| 1) It depends on the devices connected, numbers of channels, configuration, length of cables and other parameters. | |
| Rev: 1673526968 | |
KRYPTON® ONE amplifier overview
| AO | DI | DO | ACC | STG | LV | HV | TH-HV | CNT | CAN | CAN-FD | |
|---|---|---|---|---|---|---|---|---|---|---|---|
| Connector | BNC | DSUB15HD Male | DSUB15HD Female | BNC | DB9 | BNC | Banana jack | K-type Thermocouple LEMO REDEL | LEMO T-type 7pin female | D-SUB9 | D-SUB9 |
| Number of channels per module | 1 | 4 | 4 | 1 | 1 | 1 | 1 | 1 | 1 | 4x | 4x |
| Data rate per channel | ✗ | 40 kS/s | ✗ | 40 kS/s | 40 kS/s | 40 kS/s | 40 kS/s | 100 S/s | 20 kS/s | up to 1 Mbit/s | Up to 1 Mbit/s (CAN2.0B) Up to 5 Mbit/s (CAN-FD) |
| Resolution | 18 bit | Digital | Digital | 24-bit | 24-bit | 24-bit | 24-bit | 24-bit | 100 MHz timebase 5 ppm, 20 ppm max | Digital | Digital |
| Bandwidth | ✗ | ✗ | ✗ | 0.49 fs | 0.49 fs | 0.49 fs | 0.49 fs | 0.49 fs | 10 MHz | ✗ | ✗ |
| Voltage ranges | ✗ | Digital (Low: < 1 V, High: > 2 V) | ✗ | ±10 V ±5 V ±1 V ±200 mV | ±50 V ±10 V ±1 V ± 100 mV | ±50 V ±10 V ±1 V ±100 mV | ±1000 V | Thermocouple | TTL (Low: <0.8 V, High > 2 V) | ✗ | ✗ |
| Input coupling | DC, AC 0.1 Hz, 1 Hz | DC, AC 1 Hz | DC, AC 1 Hz | DC | ✗ | ✗ | |||||
| Sensor excitation | ±10 V | ✗ | ✗ | IEPE 4 mA, 8 mA | Unipolar 0 - 24 V Bipolar 0 - 12 V | ✗ | ✗ | ✗ | 5 V (100 mA max.), 12 V (90 mA max.) | 5 V / max. 250 mA (1.25 W) 12 V / max. 90 mA (1.1 W) | 5 V / max. 250 mA (1.25 W) 12 V / max. 90 mA (1.1 W) |
| Bridge connection | ✗ | ✗ | ✗ | ✗ | Full, ½, ¼ 350 Ω, ¼ 120 Ω 3 wire | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ |
| Programmable shunt | ✗ | ✗ | ✗ | ✗ | 100 kΩ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ |
| IEPE input | ✗ | ✗ | ✗ | ✓ | DSI-ACC | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ |
| Resistance | ✗ | ✗ | ✗ | ✗ | ✓ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ |
| Temperature (PTx) | ✗ | ✗ | ✗ | ✗ | DSI-RTD | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ |
| Thermocouple | ✗ | ✗ | ✗ | ✗ | DSI-TH | ✗ | ✗ | K-type | ✗ | ✗ | ✗ |
| Current | ✗ | ✗ | ✗ | ✗ | 20mA (internal shunt), DSI-5A, ext. shunt | ext. shunt | ✗ | ✗ | ✗ | ✗ | ✗ |
| Potentiometer | ✗ | ✗ | ✗ | ✗ | ✓ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ |
| LVDT | ✗ | ✗ | ✗ | ✗ | DSI-LVDT | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ |
| Charge | ✗ | ✗ | ✗ | ✗ | DSI-CHG | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ |
| TEDS | ✗ | ✗ | ✗ | ✓ | ✓ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ |
| Isolation | None | Galvanic isolation CH, GND | Galvanic isolation CH, GND | 125 Vrms Isolation CH, GND | 125 Vrms Isolation CH, GND | 125 Vrms Isolation CH, GND | CAT III 600 V CAT II 1000 V | CAT III 600 V CAT II 1000 V | None | CATII 150 V | CATII 150 V |
| Power consumption | 2 W | 1.5 W | 2 W | 2 W | 3 W | 1.5 W | 1 W | 1.3 W | 1.2 W | 1 W | 2 W |
| OBSIDIAN compatibility | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✓ | ✓ (only CAN) |
| Rev: 1736167759 | |||||||||||
| *Some features might not be supported. | |||||||||||
KRYPTON® MULTI amplifier overview
| STG | TH | TH-HS | RTD | RTD-HS | ACC | LV | LA | DIO | CNT | |
|---|---|---|---|---|---|---|---|---|---|---|
| Connectors | DB9, L2B10f | Thermocouple | Thermocouple | L0B6f | L0B6f | BNC | BNC | BNC | DB25 | L1T7f |
| #ch per module | 3x, 6x | 8x, 16x | 8x, 16x | 8x | 8x | 4x, 8x | 4x, 8x | 8x | 16xDI, 16xDO 8xDI, 8xDO | 4xCNT |
| Data rate / channel | 20 kS/s | 100 S/s | 100 S/s, upgradable to 10 kS/s | 100 S/s | 100 S/s, upgradable to 10 kS/s | 20 kS/s | 20 kS/s | 20 kS/s | 40 kS/s | 20 kS/s |
| Resolution | 24-bit | 24-bit | 24-bit | 24-bit | 24-bit | 24-bit | 24-bit | 24-bit | digital | 100 MHz timebase 5 ppm, 20 ppm max |
| Bandwidth | 0.49*fs | ✗ | 0.49*fs (max. 1 kHz) | ✗ | ✗ | 0.49*fs | 0.49*fs | 0.49*fs | ✗ | 10 MHz |
| Voltage ranges | ±10 V, ±1 V, ±100 mV, ±10 mV | ±1 V, ±100 mV | ±1 V, ±100 mV | ±1 V, ±100 mV | ±1 V, ±100 mV | ±10 V, ±5 V, ±1 V, ±200 mV | ±50 V | ±20 mA | CMOS compatible | TTL (Low: <0.8 V, High > 2 V) |
| Input coupling | DC | DC | DC | DC | DC | DC, AC 0.1 Hz, 1 Hz | DC | DC | DC | |
| Sensor excitation | 0...15 V max. 0.4 W/ch (45 mA limit) | ✗ | ✗ | ✗ | ✗ | IEPE 4 mA, 8 mA | - | - | 5 V / 300 mA 12 V / 120 mA Vsupply / 200 mA | 5 V /100 mA / ch 12 V / 50 mA / ch |
| Bridge connection | Full, ½, ¼ 350 Ω, ¼ 120 Ω 3 wire | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ |
| Programmable shunt | 100 kΩ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ |
| IEPE input | DSI-ACC | ✗ | ✗ | ✗ | ✗ | ✓ | ✗ | ✗ | ✗ | ✗ |
| Resistance | ✗ | ✗ | ✗ | 10 kΩ | 10 kΩ | ✗ | ✗ | ✗ | ✗ | ✗ |
| Temperature (PTx) | DSI-RTD | ✗ | ✗ | PT100, 200, 500, 1000, 2000 | PT100, 200, 500, 1000, 2000 | ✗ | ✗ | ✗ | ✗ | ✗ |
| Thermocouple | DSI-TH | K, J, T, R, S, N, E, C, B | K, J, T, R, S, N, E, C, B | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ |
| Current | ext. shunt DSI | ✗ | ✗ | ✗ | ✗ | ext. shunt | ✗ | ✓ | ✗ | ✗ |
| Potentiometer | ✓ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ |
| LVDT | DSI-LVDT | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ |
| Charge | DSI-CHG | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ |
| TEDS | ✓ | ✗ | ✗ | ✗ | ✗ | ✓ | ✗ | ✗ | ✗ | ✗ |
| Isolation voltage | Differential | 1000 V | 1000 V | 1000 V | 1000 V | Differential | 1000 V | 1000 V | 250 V | ✗ |
| Isolation arrangement | None | Ch-Ch & Ch-GND | Ch-Ch & Ch-GND | Ch-Ch & Ch-GND | Ch-Ch & Ch-GND | None | Ch-Ch & Ch-GND | Ch-Ch & Ch-GND | Isolated common GND | Isolated common GND |
| Power consumption per module | 2.4 W (4 W 120 Ω @ 5 V load) 3xSTG; 5.9 W (8.9 W 120 Ω @ 5 V load) 6xSTG | 2.5 W (8xTH) 4 W (16xTH) | 2.6 W (8xTH-HS) 4.3 W (16xTH-HS) | 2.5 W | 2.5 W | 4.6 W | 3.5 W (4xLV) 6.7 W (8xLV) | 6.8 W | 2 W | 2.3W |
| Advanced functions | Supports all strain types, TEDS support | High isolation, support of main TC types | High isolation, support of main TC types | ✗ | ✗ | TEDS support | High isolation, high input range | High isolation | ✗ | SuperCounter® |
| OBSIDIAN compatibility | Yes* up to 5kHz | Yes | Yes up to 5kHz | ✗ | ✗ | Yes* up to 5kHz | Yes* up to 5kHz | ✗ | ✗ | Yes |
| Rev: 1784267400 | ||||||||||
| *Some features might not be supported. | ||||||||||
KRYPTONi-1xACC specifications
| Analog inputs | ||||
|---|---|---|---|---|
| Input types | Voltage, IEPE | |||
| Number of channels | 1 | |||
| ADC Type | 24-bit SAR with anti-aliasing filter | |||
| Sampling Rate | Simultaneous 40 kS/s (50 kS/s while using it standalone) software-selectable | |||
| Voltage Mode | ||||
| Voltage Ranges | ±10 V | ±5 V | ±1 V | ±200 mV |
| Typ. Input Accuracy | ±0.03 % of reading ±0.02 % of range ±0.1 mV | |||
| Typ. SNR (10 kS/s, -1 dBFS sine wave @ 1 kHz) | 90 dB | 90 dB | 89 dB | 88 dB |
| Type. SFDR (10 kS/s, -1 dBFS sine wave @ 1 kHz) | 100 dB | 99 dB | 99 dB | 98 dB |
| Typ. Noise floor @ 10 kS/s | -100 dB | -99 dB | -99 dB | -98 dB |
| Typical CMR @ 400 Hz / 1 kHz | 130 dB / 120 dB | 122 dB / 119 dB | 104 dB / 97 dB | 90 dB / 84 dB |
| Gain Drift | typ. 10 ppm/K (max. 40 ppm/K) | |||
| Offset Drift | typ. 0.3 μV/K + 5 ppm of range/K (max. 2 µV/K + 10 ppm of range/K) | |||
| Gain Linearity | < 0.02 % | |||
| Input Coupling | DC, AC 0.1 Hz, 1 Hz | |||
| Input Impedance | 1 MΩ | |||
| Overvoltage Protection | In+ to In-: 50 V continuous, 200 V peak (10 msec) | |||
| Analog input performance | ||||
| Bandwidth (-3 dB) | 0.49 fs | |||
| Alias-free Bandwidth | DC to 0.453 fs | |||
| Alias Rejection | -100 dB (all sample rates) | |||
| Delay Through ADC | 37 / fs | |||
| Oversampling | 32 | |||
| IEPE input | ||||
| Excitation | 4 mA, 8 mA | |||
| Compliance voltage | 24 V | |||
| Output Impedance | ||||
| Sensor detection | Shortcut: <4 V; Open: >19 V | |||
| TEDS support | IEPE mode only | |||
| Additional specifications | ||||
| Input connector | BNC | |||
| Isolation | 125 Vrms channel to ground isolation | |||
| Power | ||||
| Power supply | 9 - 48 V DC | |||
| Power consumption | 2 W (typ.) | |||
| Environmental | ||||
| IP rating | IP65 / IP67 (see 1) | |||
| Physical | ||||
| Dimensions | 62 x 56 x 29 mm | |||
| Weight | 250 g | |||
| 1) Depending on the mating connector types | ||||
| Rev: 1702375200 | ||||
KRYPTONi-1xSTG specifications
| Analog input | ||||
|---|---|---|---|---|
| Input types | Voltage, current, full bridge strain, half bridge strain, quarter bridge strain | |||
| Number of channels | 1 | |||
| ADC Type | 24-bit SAR with anti-aliasing filter | |||
| Sampling Rate | Simultaneous 40 kS/s per channel (software-selectable) | |||
| Voltage Mode | ||||
| Ranges | ±50 V | ±10 V | ±1 V | ±100 mV |
| Input Accuracy | ±0.03 % of reading, ±0.02 % of range, ±0.1 mV | |||
| Typical SNR @ 10 kS/s | 89 dB | 88 dB | 88 dB | 87 dB |
| Typical SFDR @ 10 kS/s | 100 dB | 99 dB | 100 dB | 96 dB |
| Typical Noise floor @ 10 kS/s | -103 dB | -99 dB | -103 dB | -96 dB |
| Typical CMR @ 400 Hz / 1 kHz | 71 dB / 66 dB | 72 dB / 66 dB | 96 dB / 88 dB | 96 dB / 87 dB |
| Gain Drift | Typical 10 ppm/K (max. 40 ppm/K) | |||
| Offset Drift | Typical 0.3 μV/K + 5 ppm of range/K, max 2 μV/K + 10 ppm of range/K | |||
| Gain Linearity | < 0.02 % | |||
| Input Coupling | DC, AC 1 Hz | |||
| Input Impedance | 1 MΩ | 20 MΩ | ||
| Overvoltage Protection | 200 V | 50 V | ||
| Current Mode | ||||
| Ranges | 20 mA | 2 mA | ||
| Input Accuracy | ±0.03 % of reading, ±0.02 % of range, ±2.1 μA | |||
| Internal Shunt Resistor | 50 Ω | |||
| Analog input performance | ||||
| Bandwidth (-3 dB) | 0.49 fs | |||
| Alias-free Bandwidth | DC to 0.453 fs | |||
| Alias Rejection | -100 dB (all sample rates) | |||
| Delay Through ADC | 37 / fs | |||
| Oversampling | 32 | |||
| Excitation Voltage | ||||
| Excitation voltage | Free programmable (16-bit DAC) | |||
| Predefined Levels | Unipolar 0 - 24 V, Bipolar 0 - 12 V | |||
| Accuracy | ±0.05 % ±2 mV | |||
| Drift | ±50 ppm/K ±100 μV/K | |||
| Stability 10 % to 90 % Load | < 0.01 % | |||
| Current Limit | 42 mA (400 mW max. Power) | |||
| Protection | Continuous short to ground | |||
| Bridge mode | ||||
| Bridge Connection Types | full bridge strain, ½ bridge strain, ¼ bridge strain (3 wire) | |||
| Bridge ranges | 2…1000 mV/V free programmable | |||
| Internal Bridge Completion | ½ bridge and ¼ bridge 120 and 350 Ω | |||
| Bridge Completion Accuracy | 0.05 %; TCR: 2 ppm/K (others on request) | |||
| Internal Shunt Resistor | 100 kΩ (others on request) | |||
| Shunt Resistor Accuracy | 0.1 %; TCR: 10 ppm/K (others on request) | |||
| Input Short, Sensor Offset Adjust | Software-selectable | |||
| Additional specifications | ||||
| Input connector | DB9 | |||
| Isolation | 125 Vrms channel to ground isolation | |||
| TEDS support | Standard + DSI adapters | |||
| Power | ||||
| Power Supply | 9 - 48 V DC | |||
| Power consumption | 2.8 W (typ.) | |||
| Physical | ||||
| Dimensions | 62 x 56 x 29 mm | |||
| Weight | 250 g | |||
| Rev: 1710761400 | ||||
KRYPTONi-1xLV specifications
| Analog input | ||||
|---|---|---|---|---|
| Input types | Voltage | |||
| Number of channels | 1 | |||
| ADC Type | 24-bit SAR with anti-aliasing filter | |||
| Sampling Rate | Simultaneous 40 kS/s per channel (software-selectable) | |||
| Voltage Mode | ||||
| Ranges | ±50 V | ±10 V | ±1 V | ±100 mV |
| Input Accuracy | ±0.03 % of reading, ±0.02 % of range, ±0.1 mV | |||
| Typical SNR @ 10 kS/s | 96 dB | 96 dB | 94 dB | 88 dB |
| Typical Dynamic Range @ 10 kS | -130 dB | -132 dB | -137 dB | -138 dB |
| Typical Noise floor @ 10 kS | -108 dB | -102 dB | -107 dB | -100 dB |
| Gain Drift | Typical 10 ppm/K (max. 40 ppm/K) | |||
| Offset Drift | Typical 0.3 μV/K + 5 ppm of range/K, max 2 μV/K + 10 ppm of range/K | |||
| Gain Linearity | < 0.02 % | |||
| Input Coupling | DC, AC 1 Hz | |||
| Input Impedance | 1 MΩ | 20 MΩ | ||
| Overvoltage Protection | 200 V | 50 V | ||
| Analog input performance | ||||
| Bandwidth (-3 dB) | 0.49 fs | |||
| Alias-free Bandwidth | DC to 0.453 fs | |||
| Alias Rejection | -100 dB (all sample rates) | |||
| Delay Through ADC | 37 / fs | |||
| Oversampling | 32 | |||
| Additional specifications | ||||
| Input connector | BNC | |||
| Isolation | 125 Vrms channel to ground isolation | |||
| Power | ||||
| Power Supply | 9 - 48 V DC | |||
| Power consumption | 1.5 W (typ.) | |||
| Physical | ||||
| Dimensions | 62 x 56 x 29 mm | |||
| Weight | 250 g | |||
| Rev: 1578868200 | ||||
KRYPTONi-1xHV specifications
| Analog Input | |
|---|---|
| Input types | Voltage |
| Number of channels | 1 |
| ADC Type | 24-bit SAR with anti-aliasing filter |
| Sampling Rate | 40 kS/s (software-selectable) |
| Input range | ±1000 V |
| Input accuracy | ±0.03 % of reading ±0.2 V |
| Typ. Noise floor @ 10 kS/s | -96 dB |
| Typ. Noise floor @ 40 kS/s | -90 dB |
| Typ. CMR @ 400 Hz / 1 kHz | - |
| Gain Drift | Typ. 50 ppm (Max. 150ppm) |
| Offset Drift | Typ. 1.5 mV/K (Max. 4 mV/K) |
| Gain Linearity | - |
| Input Coupling | DC |
| Input Impedance | 10 MΩ |
| Protection class | CAT IV 600 V, CAT III 1000 V |
| Overvoltage Protection | +-2 kVpeak |
| Analog input performance | |
| Bandwidth (-3 dB) | 0.49 fs |
| Alias-free Bandwidth | DC to 0.453 fs |
| Alias Rejection | -100 dB (all sample rates) |
| Delay Through ADC | 37 / fs |
| Oversampling | 32 |
| Additional specifications | |
| Input connector | 4 mm safety banana jack |
| Power | |
| Power Supply | 9 - 48 V DC |
| Power consumption | 1 W (typ.) |
| Environmental | |
| IP rating | IP65 / IP67 (see 1) |
| Physical | |
| Dimensions | 62 x 56 x 36 mm |
| Weight | 300 g |
| 1) Depending on the mating connector types | |
| Rev: 1578868200 | |
KRYPTONi-1xTH-HV specifications
| Inputs | |
|---|---|
| Input types | Thermocouple |
| Number of channels | 1 |
| ADC Type | 24-bit delta-sigma |
| Sampling Rate | Simultaneous 100 S/s |
| Input Impedance | 100 MΩ |
| Thermocouple | |
| TC type | Type K |
| Accuracy | ±0.02 % of reading ±0.5 °C ±10 μV |
| Resolution | < 0.001 °C |
| Sampling rates | 100, 80, 40, 20, 10 S/s |
| Typical Noise | 0.007°C RMS @ Type K @ 10 S/sec |
| Additional specifications | |
| Input connector | K-type Thermocouple LEMO REDEL 2P series |
| Isolation voltage | CAT III 600 V, CAT II 1000 V |
| Power | |
| Power Supply | 9 - 48 V DC |
| Power consumption | 1.3 W (typ.) |
| Physical | |
| Dimensions | 62 x 56 x 36 mm |
| Weight | 300 g |
| Rev: 1578868200 | |
KRYPTON-1xCNT specifications
| Counter | |
|---|---|
| Input type | Digital |
| Number of channels | 1 |
| Timebase | 100 MHz |
| Time base accuracy | Typical: 5 ppm, Max: 20 ppm |
| Max. bandwidth | 10 MHz |
| Input Filter | 500 ns, 1 μs, 2 μs, 4 μs, 5 μs and 7.5 μs |
| Input Level Compatibility | TTL (Low: < 0.8 V, High > 2 V) |
| Input impedance | 100 kΩ pull-up to +3.3 V |
| Input Protection | ±25 V continuous |
| Isolated Sensor Power Supply | |
| Output Voltage +5 V | 5 V +/- 10 % 100 mA max. |
| Output Voltage +12 V | 12 V +/-10 % 90 mA max. |
| Additional specifications | |
| Input connector | LEMO 1T series 7-pin |
| Isolation | - |
| Power | |
| Power Supply | 9 - 48 V DC |
| Power consumption | 1.2 W (typ.) |
| Physical | |
| Dimensions | 62 x 56 x 29 mm |
| Weight | 250 g |
| Rev: 1578868200 | |
KRYPTON-1xAO specifications
| Analog Output | |
|---|---|
| Output types | Voltage |
| Number of channels | 1 |
| DAC Resolution | 18 bit |
| Function | File replay, conditioned AI output, FGEN (software option), channel output |
| Update Rate | 20 kS/s (5 kS/s with 3rd party master) |
| Analogue out bandwidth | 4 kHz |
| Full Scale Output Range | ±10 V |
| Accuracy (typ.) | 0.1 % ±1 mV (0 to 20 mA load) |
| Temperature gain drift (typ.) | ±10 ppm/K |
| Temperature offset drift (typ.) | ±5 uV/K |
| SNR @ 1000 S/s output rate | 92 dB (sine wave, 250 Hz, ±9 V) |
| THD | -90 dB |
| Output impedance | 0.45 Ω DC |
| Maximum output current | 20 mA |
| Settling time (0.05 % of set value) | 30 μs @ 1000 Ω load, -10 V to 10 V step |
| Max slew rate | 2.5 V/μs @ 1000 Ω load |
| Max capacitive load (2 % overshoot) | 10 nF |
| Maximum load | 500 Ω |
| Output protection | Short circuit protected |
| Additional specifications | |
| Output connector | BNC |
| Power | |
| Power supply | 9 - 48 V DC |
| Power consumption | 2 W (typ.) |
| Environmental | |
| IP rating | IP65 / IP67 (see 1) |
| Physical | |
| Dimensions | 62 x 56 x 29 mm |
| Weight | 250 g |
| 1) Depending on the mating connector types | |
| Rev: 1714120668 | |
KRYPTONi-4xDI specifications
| Digital Inputs | |
|---|---|
| Input type | Digital |
| Number of channels | 4 |
| Compatibility | TTL / CMOS Voltage Levels |
| Input low level | UIN < 1 V |
| Input high level | UIN > 2 V |
| Input high current @ 5 V UIN | 3 mA typ. |
| Input high current @ 30 V UIN | 6 mA typ. |
| Propagation delay | < 1 μs |
| Max. sampling rate | 40 kS/s |
| Non-Isolated Sensor Power Supply | |
| Output Voltage +5 V | 5 V +/- 10 % 300 mA max. (see 1) |
| Output Voltage +12 V | 12 V +/-10 % 100 mA max. (see 1) |
| Output Voltage +Vecat | EtherCAT Bus Supply Voltage 200 mA max. (see 1) |
| Additional specifications | |
| Connector | DSUB15HD Male |
| Overvoltage protection | 30 V Continuous, 65 V peak |
| Isolation | Galvanic isolation CH, GND |
| Power | |
| Power supply | 9 - 48 V DC |
| Power consumption | 1.5 W (typ.) |
| Physical | |
| Dimensions | 62 x 56 x 29 mm |
| Weight | 250 g |
| 1) Max. combined load on +5 V and +12 V pins: 1.5 W reduced by the load already present on +Vin pin | |
| Rev: 1578868200 | |
KRYPTONi-4xDO specifications
| Digital Outputs | |
|---|---|
| Output type | Digital |
| Number of channels | 4 |
| Maximum update rate | depends on EtherCAT master cycle time |
| Output type | open collector with active internal pullup to 5 V |
| Sink | max 50 V, 2 A (resistive load, unprotected) |
| Source | 5 V, max. 40 mA (internally limited) |
| Max. switching frequency for internal PWM | 100 kHz |
| Non-Isolated Sensor Power Supply | |
| Output Voltage +5 V | 5 V +/- 10 % 300 mA max. (see 1) |
| Output Voltage +12 V | 12 V +/-10 % 100 mA max. (see 1) |
| Output Voltage +Vecat | EtherCAT Bus Supply Voltage 200 mA max. (see 1) |
| Additional specifications | |
| Connector | DSUB15HD Female |
| Isolation | Galvanic isolation CH, GND |
| Power | |
| Power supply | 9 - 48 V DC |
| Power consumption | 2 W (typ.) |
| Physical | |
| Dimensions | 62 x 56 x 29 mm |
| Weight | 250 g |
| 1) Max. combined load on +5 V and +12 V pins: 1.5 W reduced by the load already present on +Vin pin | |
| Rev: 1578868200 | |
KRYPTONi-1xCAN technical specifications
| KRYPTONi-1xCAN | |
|---|---|
| CAN ports specification | |
| Number of ports | 1 |
| Interface type | CAN 2.0B |
| Interface data rate | up to 1 Mbit/sec |
| Special protocols | J1939 (see 1) |
| Galvanic isolation | Isolated |
| Bus pin fault protection | |
| Power specification | |
| Power supply | 9 - 50 V DC |
| Maximum sensor power consumption | 5 V / max. 250 mA (1.25 W) 12 V / max. 90 mA (1.1 W) |
| Maximum power consumption | |
| Environmental specification | |
| Interface | EtherCAT, 100 Mbit/s Lemo 1T (1 cable for data, power and sync, daisy chainable) |
| Connector type | DSUB-9 |
| Physical dimensions | 62 x 56 x 29 mm |
| Weight | 200 g 0.44 lbs |
| Operating temperature | -40 °C to +85 °C |
| Storage temperature | -40 °C to +85 °C |
| Humidity | - |
| IP protection | IP65 / IP67 |
| 1) Transmit mode not supported yet. OBDII and XCP plugins not supported yet. | |
| Rev.: 1671516000 | |
KRYPTONi-1xCAN-FD technical specifications
| KRYPTONi-1xCAN-FD | |
|---|---|
| CAN ports specification | |
| Number of ports | 1 |
| Interface type | CAN-FD, CAN 2.0B (See 1) |
| Interface data rate | up to 5 Mbit/sec |
| Special protocols | J1939 |
| Galvanic isolation (Sensor supply not isolated) | Isolated |
| Bus pin fault protection | ±28 V |
| Power specification | |
| Power supply | 9 - 50 V DC |
| Maximum sensor power consumption | max. 1.4 W (5 V / 12 V) (max. 1.2 W on 12 V) |
| Maximum power consumption | max. 2 W (no load) |
| Environmental specification | |
| Interface | EtherCAT, 100 Mbit/s Lemo 1T (1 cable for data, power and sync, daisy chainable) |
| Connector type | DSUB-9 |
| Physical dimensions | 62 x 56 x 29 mm |
| Weight | 200 g 0.44 lbs |
| Operating temperature | -40 °C to +85 °C |
| Storage temperature | -40 °C to +85 °C |
| IP protection | IP65 / IP67 |
| 1) Transmit mode not supported. OBDII and XCP plugins not supported yet. | |
KRYPTON-8xACC specifications
| Analog inputs | KRYPTON-8xACC | |||
|---|---|---|---|---|
| Input type | Voltage, IEPE | |||
| Number of channels | 8 | |||
| ADC Type | 24-bit delta-sigma with anti-aliasing filter | |||
| Sampling Rate | Simultaneous 20 kS/s per channel (software-selectable) | |||
| Voltage Mode | ||||
| Ranges | ±10 V | ±5 V | ±1 V | ±200 mV |
| Input Accuracy | ±0.03 % of reading ±0.02 % of range ±0.2 mV | |||
| Typical Dynamic Range @ 10 kS/s | -140 dB | -137 dB | 138 dB | -134 dB |
| Typical noise floor @ 10 kS/s | -106 dB | -106 dB | 106 dB | -102 dB |
| Type. THD (20 kS/s, -1 dBFS sine wave @ 1 kHz) | -94 dB | -94 dB | -93 dB | -95 dB |
| Type. SFDR (10 kS/s, -1 dBFS sine wave @ 1 kHz) | 96 dB | 96 dB | 96 dB | 93 dB |
| Gain Drift | Typical 10 ppm/K, max. 20 ppm/K | |||
| Offset Drift | Typical 0.5 μV/K + 1 ppm of range/K, max 2 μV/K + 3 ppm of range/K | |||
| Input Coupling | DC, AC 0.1 Hz, 1 Hz | |||
| Input Impedance | 1 MΩ | |||
| Overvoltage Protection | In+ to In-: 50 V continuous, 200 V peak (10 msec) | |||
| Analog input performance | ||||
| Bandwidth (-3 dB) | 0.49 fs | |||
| Alias-free Bandwidth | DC to 0.453 fs | |||
| Alias Rejection | -100 dB (all sample rates) | |||
| Delay Through ADC | 37 / fs | |||
| Oversampling | 32 | |||
| IEPE input | ||||
| Excitation | 4 mA, 8 mA | |||
| Compliance voltage | 25 V | |||
| Output Impedance | >100 kΩ | |||
| Sensor detection | Shortcut: <4 V; Open: >19 V | |||
| Additional Specifications | ||||
| Input connector | BNC | |||
| TEDs support | IEPE mode only | |||
| Power | ||||
| Power Supply | 9 - 48 V DC | |||
| Power Consumption | 4.6 W | |||
| Environemental | ||||
| IP rating (see 1) | IP67 | |||
| Physical | ||||
| Weight | 1000 g | |||
| Slice Dimensions | 213 x 54 x 56 mm | |||
| Dimensions with Connectors | 213 x 54 x 66.8 mm | |||
| 1) Depending on the mating connector types | ||||
| Rev: 1703235600 | ||||
KRYPTON-4xACC specifications
| Analog inputs | KRYPTON-4xACC | |||
|---|---|---|---|---|
| Input type | Voltage, IEPE | |||
| Number of channels | 4 | |||
| ADC Type | 24-bit delta-sigma with anti-aliasing filter | |||
| Sampling Rate | Simultaneous 20 kS/s per channel (software-selectable) | |||
| Voltage Mode | ||||
| Ranges | ±10 V | ±5 V | ±1 V | ±200 mV |
| Input Accuracy | ±0.03 % of reading ±0.02 % of range ±0.2 mV | |||
| Typical Dynamic Range @ 10 kS/s | -140 dB | -137 dB | 138 dB | -134 dB |
| Typical noise floor @ 10 kS/s | -106 dB | -106 dB | 106 dB | -102 dB |
| Type. THD (20 kS/s, -1 dBFS sine wave @ 1 kHz) | -94 dB | -94 dB | -93 dB | -95 dB |
| Type. SFDR (10 kS/s, -1 dBFS sine wave @ 1 kHz) | 96 dB | 96 dB | 96 dB | 93 dB |
| Gain Drift | Typical 10 ppm/K, max. 20 ppm/K | |||
| Offset Drift | Typical 0.5 μV/K + 1 ppm of range/K, max 2 μV/K + 3 ppm of range/K | |||
| Input Coupling | DC, AC 0.1 Hz, 1 Hz | |||
| Input Impedance | 1 MΩ | |||
| Overvoltage Protection | In+ to In-: 50 V continuous, 200 V peak (10 msec) | |||
| Analog input performance | ||||
| Bandwidth (-3 dB) | 0.49 fs | |||
| Alias-free Bandwidth | DC to 0.453 fs | |||
| Alias Rejection | -100 dB (all sample rates) | |||
| Delay Through ADC | 37 / fs | |||
| Oversampling | 32 | |||
| IEPE input | ||||
| Excitation | 4 mA, 8 mA | |||
| Compliance voltage | 25 V | |||
| Output Impedance | >100 kΩ | |||
| Sensor detection | Shortcut: <4 V; Open: >19 V | |||
| Additional Specifications | ||||
| Input connector | BNC | |||
| TEDs support | IEPE mode only | |||
| Power | ||||
| Power Supply | 9 - 48 V DC | |||
| Power Consumption | 3 W | |||
| Environemental | ||||
| IP rating (see 1) | IP67 | |||
| Physical | ||||
| Weight | 700 g | |||
| Slice Dimensions | 213 x 39 x 56 mm | |||
| Dimensions with Connectors | 213 x 39 x 66.8 mm | |||
| 1) Depending on the mating connector types | ||||
| Rev: 1703235600 | ||||
KRYPTON-3xSTG specifications
| Analog inputs | KRYPTON-3xSTG | |||
|---|---|---|---|---|
| Input type | Voltage Full / half / quarter strain | |||
| Number of channels | 3 | |||
| ADC Type | 24-bit oversampled SAR with 41 kHz anti-aliasing filter | |||
| Sampling Rate | Simultaneous 20k, 10k, 5k, 2k, 1k, 500, 200, 100 S/s (software-selectable) | |||
| Voltage Mode | ||||
| Ranges | ±10 V | ±1 V | ±100 mV | ±10 mV |
| Input Accuracy | ±0.03 % of reading, ±0.02 % of range, ±0.1 mV | |||
| Typical Dynamic Range @ 10 kS/s | -139 dB | -139 dB | -129 dB | -110 dB |
| Typical Noise Floor @ 10 kS/s | -106 dB | -105dB | -97 dB | -78 dB |
| Typical CMR @ 400 Hz / 1 kHz | 86 dB / 84 dB | 96 dB / 95 dB | 112 dB / 102 dB | 112 dB / 102 dB |
| Gain Drift | Typical 10 ppm/K (max. 40 ppm/K) | |||
| Offset Drift | Typical 0.3 μV/K + 5 ppm of range/K, max 2 μV/K + 10 ppm of range/K | |||
| Gain Linearity | < 0.02 % | |||
| Input Coupling | DC | |||
| Input Impedance | 10 MΩ | |||
| Overvoltage Protection | In+ to In-: 50 V continuous, 200 V peak (10 ms) | |||
| Analog input performance | ||||
| Bandwidth (-3 dB) | 0.49 fs | |||
| Alias-free Bandwidth | DC to 0.453 fs | |||
| Alias Rejection | -100 dB (all sample rates) | |||
| Delay Through ADC | 37 / fs | |||
| Oversampling | 32 | |||
| Excitation Voltage | ||||
| Excitation voltage | Free programmable (16-bit DAC) | |||
| Predefined Levels | 0, 1, 2.5, 5, 10 and 15 VDC | |||
| Accuracy | ±0.05 % ±2 mV | |||
| Drift | ±50 ppm/K ±100 μV/K | |||
| Stability 10 % to 90 % Load | < 0.01 % | |||
| Current Limit | 45 mA (400 mW max. Power) | |||
| Protection | Continuous short to ground | |||
| Bridge mode | ||||
| Bridge Connection Types | full bridge strain, ½ bridge strain, ¼ bridge strain (3 wire) | |||
| Ranges | 2…1000 mV/V free programmable | |||
| Internal Bridge Completion | ½ bridge and ¼ bridge 120 and 350 Ω | |||
| Bridge Completion Accuracy | 0.05 %; TCR: 2 ppm/K (others on request) | |||
| Internal Shunt Resistor | 100 kΩ (others on request) | |||
| Shunt Resistor Accuracy | 0.1 %; TCR: 10 ppm/K (others on request) | |||
| Input Short, Sensor Offset Adjust | Software-selectable | |||
| Additional Specifications | ||||
| Input connector | DB9; L1T7f; L2B10f | |||
| TEDs support | Standard + DSI adapters | |||
| Power | ||||
| Power Consumption | 3.5 W (4.8 W 120 Ω @ 5 V load) | |||
| Power Supply | 9 - 48 V DC | |||
| Environmental | ||||
| IP rating | IP67 | |||
| Physical | ||||
| Weight | 740 g | |||
| Slice Dimensions | 213 x 39 x 56 mm | |||
| Dimensions with Connectors | 213 x 39 x 58.2 mm | |||
KRYPTON-6xSTG specifications
| Analog inputs | KRYPTON-6xSTG | |||
|---|---|---|---|---|
| Input type | Voltage Full / half / quarter strain | |||
| Number of channels | 6 | |||
| ADC Type | 24-bit oversampled SAR with 41 kHz anti-aliasing filter | |||
| Sampling Rate | Simultaneous 20k, 10k, 5k, 2k, 1k, 500, 200, 100 S/s (software-selectable) | |||
| Voltage Mode | ||||
| Ranges | ±10 V | ±1 V | ±100 mV | ±10 mV |
| Input Accuracy | ±0.03 % of reading, ±0.02 % of range, ±0.1 mV | |||
| Typical Dynamic Range @ 10 kS/s | -139 dB | -139 dB | -129 dB | -110 dB |
| Typical Noise Floor @ 10 kS/s | -106 dB | -105dB | -97 dB | -78 dB |
| Typical CMR @ 400 Hz / 1 kHz | 86 dB / 84 dB | 96 dB / 95 dB | 112 dB / 102 dB | 112 dB / 102 dB |
| Gain Drift | Typical 10 ppm/K (max. 40 ppm/K) | |||
| Offset Drift | Typical 0.3 μV/K + 5 ppm of range/K, max 2 μV/K + 10 ppm of range/K | |||
| Gain Linearity | < 0.02 % | |||
| Input Coupling | DC | |||
| Input Impedance | 10 MΩ | |||
| Overvoltage Protection | In+ to In-: 50 V continuous, 200 V peak (10 ms) | |||
| Analog input performance | ||||
| Bandwidth (-3 dB) | 0.49 fs | |||
| Alias-free Bandwidth | DC to 0.453 fs | |||
| Alias Rejection | -100 dB (all sample rates) | |||
| Delay Through ADC | 37 / fs | |||
| Oversampling | 32 | |||
| Excitation Voltage | ||||
| Excitation voltage | Free programmable (16-bit DAC) | |||
| Predefined Levels | 0, 1, 2.5, 5, 10 and 15 VDC | |||
| Accuracy | ±0.05 % ±2 mV | |||
| Drift | ±50 ppm/K ±100 μV/K | |||
| Stability 10 % to 90 % Load | < 0.01 % | |||
| Current Limit | 45 mA (400 mW max. Power) | |||
| Protection | Continuous short to ground | |||
| Bridge mode | ||||
| Bridge Connection Types | full bridge strain, ½ bridge strain, ¼ bridge strain (3 wire) | |||
| Ranges | 2…1000 mV/V free programmable | |||
| Internal Bridge Completion | ½ bridge and ¼ bridge 120 and 350 Ω | |||
| Bridge Completion Accuracy | 0.05 %; TCR: 2 ppm/K (others on request) | |||
| Internal Shunt Resistor | 100 kΩ (others on request) | |||
| Shunt Resistor Accuracy | 0.1 %; TCR: 10 ppm/K (others on request) | |||
| Input Short, Sensor Offset Adjust | Software-selectable | |||
| Additional Specifications | ||||
| Input connector | DB9; L1T7f; L2B10f | |||
| TEDs support | Standard + DSI adapters | |||
| Power | ||||
| Power Consumption | 5.9 W (8.9 W load 120 Ω @ 5 V) | |||
| Power Supply | 9 - 48 V DC | |||
| Environmental | ||||
| IP rating | IP67 | |||
| Physical | ||||
| Weight | 1100 g | |||
| Slice Dimensions | 213 x 54 x 56 mm | |||
| Dimensions with Connectors | 213 x 54 x 58.2 mm | |||
KRYPTONi-4xLV specifications
| Analog input | KRYPTONi-4xLV |
|---|---|
| Input type | Isolated voltage |
| Number of channels | 4 |
| ADC Type | 24-bit oversampled SAR with 41 kHz anti-aliasing filter |
| Sampling Rates | Simultaneous 20k, 10k, 5k, 2k, 1k, 500, 200, 100 S/s (software-selectable) |
| Voltage Mode | |
| Range | ±50 V |
| Accuracy | ±0.03 % of reading ±5 mV |
| Typ. SNR (10 kS/s, -1 dBFS sine wave @ 1 kHz) | 104 dB |
| Typical noise floor @ 10 kS/s | -106 dB |
| Type. THD (10 kS/s, -1 dBFS sine wave @ 1 kHz) | -104 dB |
| Type. SFDR (10 kS/s, -1 dBFS sine wave @ 1 kHz) | 105 dB |
| Typical CMR @ 400 Hz / 1 kHz | 105 dB / 94 dB |
| Gain Drift | Typical 10 ppm/K, max. 20 ppm/K |
| Offset Drift | Typical 5 μV/K, max. 10 μV/K |
| Gain Linearity | < 0.01 % |
| Input Coupling | DC |
| Input Impedance | 1 MΩ |
| Noise | 0.7 mVRMS |
| Overvoltage Protection | In+ to In-: 200 V continuous, 350 Vpeak (100 msec) |
| Analog input performance | |
| Bandwidth (-3dB) | 0.49 fs |
| Alias-free Bandwidth | DC to 0.453 fs |
| Alias Rejection | -100 dB (all sample rates) |
| Delay Through ADC | 37 / fs |
| Oversampling | 32 |
| Additional Specifications | |
| Input Connector | BNC |
| Isolation Voltage | 1000 Vpeak channel/ground & channel/channel |
| Power | |
| Power Supply | 9 - 48 V DC |
| Power Consumption | 3.8 W |
| Environmental | |
| IP Rating (see 1) | IP67 |
| Physical | |
| Weight | 700 g |
| Slice Dimensions | 213 x 39 x 56 mm |
| Dimensions with Connectors | 213 x 39 x 66.8 mm |
| 1) Depending on the mating connector types | |
KRYPTONi-8xLV specifications
| Analog input | KRYPTONi-8xLV |
|---|---|
| Input type | Isolated voltage |
| Number of channels | 8 |
| ADC Type | 24-bit oversampled SAR with 41 kHz anti-aliasing filter |
| Sampling Rates | Simultaneous 20k, 10k, 5k, 2k, 1k, 500, 200, 100 S/s (software-selectable) |
| Voltage Mode | |
| Range | ±50 V, ±100 V (see 2) |
| Accuracy | ±0.03 % of reading ±5 mV |
| Typ. SNR (10 kS/s, -1 dBFS sine wave @ 1 kHz) | 104 dB |
| Typical noise floor @ 10 kS/s | -106 dB |
| Type. THD (10 kS/s, -1 dBFS sine wave @ 1 kHz) | -104 dB |
| Type. SFDR (10 kS/s, -1 dBFS sine wave @ 1 kHz) | 105 dB |
| Typical CMR @ 400 Hz / 1 kHz | 105 dB / 94 dB |
| Gain Drift | Typical 10 ppm/K, max. 20 ppm/K |
| Offset Drift | Typical 5 μV/K, max. 10 μV/K |
| Gain Linearity | < 0.01 % |
| Input Coupling | DC |
| Input Impedance | 1 MΩ |
| Noise | 0.7 mVRMS |
| Overvoltage Protection | In+ to In-: 200 V continuous, 350 Vpeak (100 msec) |
| Analog input performance | |
| Bandwidth (-3dB) | 0.49 fs |
| Alias-free Bandwidth | DC to 0.453 fs |
| Alias Rejection | -100 dB (all sample rates) |
| Delay Through ADC | 37 / fs |
| Oversampling | 32 |
| Additional Specifications | |
| Input Connector | BNC, Lemo 2 pin (see 2) |
| Isolation Voltage | 1000 Vpeak channel/ground & channel/channel |
| Power | |
| Power Supply | 9 - 48 V DC |
| Power Consumption | 6.8 W |
| Environmental | |
| IP Rating (see 1) | IP67 |
| Physical | |
| Weight | 1000 g |
| Slice Dimensions | 213 x 54 x 56 mm |
| Dimensions with Connectors | 213 x 54 x 66.8 mm |
| 1) Depending on the mating connector types | |
| 2) Upon request | |
KRYPTONi-8xLA specifications
| Analog input | KRYPTONi-8xLA |
|---|---|
| Input type | Isolated current |
| Number of channels | 8 |
| ADC Type | 24-bit oversampled SAR with 41 kHz anti-aliasing filter |
| Sampling Rates | Simultaneous 20k, 10k, 5k, 2k, 1k, 500, 200, 100 S/s (software-selectable) |
| Current Mode | |
| Range | ±20 mA |
| Accuracy | ±0.02 % of reading ±8 µA |
| Typical noise floor @ 10 kS/s | -95 dB |
| Crosstalk (20 kS/s, -1 dBFS @ 1 kHz) | -124 dB |
| Gain Drift | Typical 10 ppm/K, max. 20 ppm/K |
| Offset Drift | Typical 100 nA/K, max. 200 nA/K |
| Gain Linearity | < 0.01 % |
| Input Coupling | DC |
| Input Impedance | 50 Ω |
| Noise | 0.3 µARMS |
| Overvoltage Protection | 50 V |
| Analog input performance | |
| Bandwidth (-3 dB) | 0.49 fs |
| Alias-free Bandwidth | DC to 0.453 fs |
| Alias Rejection | -100 dB (all sample rates) |
| Delay Through ADC | 37 / fs |
| Oversampling | 32 |
| Additional Specifications | |
| Input Connector | BNC |
| Isolation Voltage | 1000 Vpeak channel/ground & channel/channel |
| Power | |
| Power Supply | 9 - 48 V DC |
| Power Consumption | 6.8 W |
| Environmental | |
| IP Rating (see 1) | IP67 |
| Physical | |
| Weight | 1000 g |
| Slice Dimensions | 213 x 54 x 56 mm |
| Dimensions with Connectors | 213 x 54 x 66.8 mm |
| 1) Depending on the mating connector types | |
KRYPTONi-8xTH specifications
| Analog inputs | KRYPTONi-8xTH | |
|---|---|---|
| Input type | Isolated universal thermocouple and voltage | |
| Number of channels | 8 | |
| ADC Type | 24-bit delta-sigma | |
| Sampling rates | Simultaneous 100, 80, 40, 20, 10 S/s (software selectable) | |
| Voltage Ranges | ||
| Ranges | ±1 V | ±100 mV |
| Input Accuracy | ±0.02 % of reading ±100 μV | ±0.02 % of reading ±10 μV |
| Typical Noise Floor @ 10 / 100 S/s | -114 dB / -105 dB | -109 dB / -100 dB |
| Gain Drift | Typical 3 ppm/K (max. 10 ppm/K) | |
| Offset Drift | 0.03 μV/K | |
| Gain Linearity | < 0.01 % | |
| Input Coupling | DC | |
| Input Impedance | 100 MΩ | |
| Thermocouple | ||
| Input type | TC Types: K, J, T, R, S, N , E, C, B | |
| Accuracy | ±0.02 % of reading ± 0.5 °C ± 10 μV | |
| Typical Noise | 0.007 °C RMS @ Type K @ 10 S/s | |
| Additional Specifications | ||
| Input connector | Mini Thermocouple connector (copper) | |
| Isolation voltage | 1000 Vpeak channel/ground & channel/channel | |
| Power | ||
| Power supply | 9 - 48 V DC | |
| Power consumption | 2.5 W | |
| Environmental | ||
| IP rating (see 1) | IP67 | |
| Physical | ||
| Weight | 720 g | |
| Dimensions | 213 x 39 x 56 mm | |
| 1) Depending on the mating connector types, for non corrosive liquids only, IP65 standard | ||
KRYPTONi-16xTH specifications
| Analog inputs | KRYPTONi-16xTH | |
|---|---|---|
| Input type | Isolated universal thermocouple and voltage | |
| Number of channels | 16 | |
| ADC Type | 24-bit delta-sigma | |
| Sampling rates | Simultaneous 100, 80, 40, 20, 10 S/s (software selectable) | |
| Voltage Mode | ||
| Ranges | ±1 V | ±100 mV |
| Input Accuracy | ±0.02 % of reading ±100 μV | ±0.02 % of reading ±10 μV |
| Typical Noise Floor @ 10 / 100 S/s | -114 dB / -105 dB | -109 dB / -100 dB |
| Gain Drift | Typical 3 ppm/K (max. 10 ppm/K) | |
| Offset Drift | 0.03 μV/K | |
| Gain Linearity | < 0.01 % | |
| Input Coupling | DC | |
| Input Impedance | 100 MΩ | |
| Thermocouple | ||
| Input type | TC Types: K, J, T, R, S, N , E, C, B | |
| Accuracy | ±0.02 % of reading ± 0.5 °C ± 10 μV | |
| Typical Noise | 0.02 °C RMS @ Type K @ 100 S/s | |
| Additional Specifications | ||
| Input connector | Mini Thermocouple connector (copper) | |
| Isolation voltage | 1000 Vpeak channel/ground & channel/channel | |
| Power | ||
| Power supply | 9 - 48 V DC | |
| Power consumption | 4 W | |
| Environmental | ||
| IP rating (see 1) | IP67 | |
| Physical | ||
| Weight | 1000 g | |
| Dimensions | 213 x 54 x 56 mm | |
| 1) Depending on the mating connector types, for non corrosive liquids only, IP65 standard | ||
KRYPTON TH accuracy
| Temperature | Type K | Type J | Type T | Type R | Type S | Type N | Type E | Type B | Type C |
|---|---|---|---|---|---|---|---|---|---|
| -200 °C | 1.19 °C | 1.00 °C | 1.17 °C | ✗ | ✗ | 1.54 °C | 0.83 °C | ✗ | ✗ |
| -100 °C | 0.85 °C | 0.76 °C | 0.87 °C | ✗ | ✗ | 1.00 °C | 0.74 °C | ✗ | ✗ |
| 0 °C | 0.75 °C | 0.70 °C | 0.76 °C | 2.38 °C | 2.34 °C | 0.89 °C | 0.67 °C | ✗ | 1.27 °C |
| 100 °C | 0.76 °C | 0.70 °C | 0.73 °C | 1.85 °C | 1.88 °C | 0.86 °C | 0.67 °C | 10.96 °C | 1.15 °C |
| 400 °C | 0.82 °C | 0.76 °C | 0.74 °C | 1.54 °C | 1.63 °C | 0.85 °C | 0.71 °C | 3.03 °C | 1.11 °C |
| 700 °C | 0.88 °C | 0.80 °C | ✗ | 1.49 °C | 1.59 °C | 0.89 °C | 0.77 °C | 2.11 °C | 1.17 °C |
| 1000 °C | 0.96 °C | 0.87 °C | ✗ | 1.46 °C | 1.57 °C | 0.96 °C | 0.85 °C | 1.80 °C | 1.23 °C |
| 1300 °C | 1.05 °C | ✗ | ✗ | 1.47 °C | 1.59 °C | 1.04 °C | ✗ | 1.68 °C | 1.35 °C |
| 1700 °C | ✗ | ✗ | ✗ | 1.58 °C | 1.71 °C | ✗ | ✗ | 1.70 °C | 1.65 °C |
KRYPTONi-8xTH-HS specifications
| Analog inputs | KRYPTONi-8xTH-HS | |
|---|---|---|
| Input type | Isolated universal thermocouple and voltage | |
| Number of channels | 8 | |
| ADC Type | 24-bit delta-sigma | |
| Sampling rates | Simultaneous 10, 20, 40, 50, 80, 100, 200, 500, 1k, 2k, 5k, 10k S/s (optional) | |
| Voltage Mode | ||
| Ranges | ±1 V | ±100 mV |
| Input Accuracy | ±0.02 % of reading ±50 μV | ±0.02 % of reading ±7 μV |
| Typical Noise Floor @ 10 / 100 / 1k / 10k S/s | -130 dB / -120 dB / -110 dB / -104 dB | -120 dB / -110 dB / -100 dB / -90 dB |
| Offset Drift | Typ. 0.05 μV/K (Max. 0.2 μV/K) | Typ. 0.01 μV/K (Max. 0.1 μV/K) |
| Gain Drift | Typ. 2 ppm/K (Max. 10 ppm/K) | |
| Gain Linearity | < 0.007 % | |
| Input Coupling | DC | |
| Input Impedance | 100 MΩ | |
| Thermocouple | ||
| Input type | TC Types: K, J, T, R, S, N , E, C, B | |
| Accuracy | ±0.02 % of reading ±0.5 °C ±7 μV | |
| Typical Noise | 0.002 °C RMS @ Type K @ 10 S/s 0.006 °C RMS @ Type K @ 100 S/s 0.02 °C RMS @ Type K @ 1 kS/s 0.05 °C RMS @ Type K @ 10 kS/s | |
| Additional Specifications | ||
| Input connector | Mini Thermocouple connector (copper) | |
| Isolation voltage | 1000 Vpeak channel/ground & channel/channel | |
| Power | ||
| Power Supply | 9 - 48 V DC | |
| Power consumption | Typ. 2.6 W (Max. 3 W) | |
| Environmental | ||
| IP rating | IP67 | |
| Physical | ||
| Weight | 750 g | |
| Dimensions | 213 x 39 x 56 mm | |
| Rev: 1660298400 | ||
KRYPTONi-16xTH-HS specifications
| Analog inputs | KRYPTONi-16xTH-HS | |
|---|---|---|
| Input type | ||
| Number of channels | 16 | |
| ADC Type | 24-bit delta-sigma | |
| Sampling rates | Simultaneous 10, 20, 40, 50, 80, 100, 200, 500, 1k, 2k, 5k, 10k S/s (optional) | |
| Voltage Mode | ||
| Ranges | ±1 V | ±100 mV |
| Input Accuracy | ±0.02 % of reading ±50 μV | ±0.02 % of reading ±7 μV |
| Typical Noise Floor @ 10 / 100 / 1k / 10k S/s | -130 dB / -120 dB / -110 dB / -104 dB | -120 dB / -110 dB / -100 dB / -90 dB |
| Offset Drift | Typ. 0.05 μV/K (Max. 0.2 μV/K) | Typ. 0.01 μV/K (Max. 0.1 μV/K) |
| Gain Drift | Typ. 2 ppm/K (Max. 10 ppm/K) | |
| Gain Linearity | < 0.007 % | |
| Input Coupling | DC | |
| Input Impedance | 100 MΩ | |
| Thermocouple | ||
| Input type | TC Types: K, J, T, R, S, N , E, C, B | |
| Accuracy | ±0.02 % of reading ±0.5 °C ±7 μV | |
| Typical Noise | 0.002 °C RMS @ Type K @ 10 S/s 0.006 °C RMS @ Type K @ 100 S/s 0.02 °C RMS @ Type K @ 1 kS/s 0.05 °C RMS @ Type K @ 10 kS/s | |
| Additional Specifications | ||
| Input connector | Mini Thermocouple connector (copper) | |
| Isolation voltage | 1000 Vpeak channel/ground & channel/channel | |
| Power | ||
| Power Supply | 9 - 48 V DC | |
| Power consumption | Typ. 4.3 W (Max. 5 W) | |
| Environmental | ||
| IP rating | IP67 | |
| Physical | ||
| Weight | 1060 g | |
| Dimensions | 213 x 54 x 56 mm | |
| Rev: 1660298400 | ||
KRYPTON TH-HS accuracy
| Temperature | Type K | Type J | Type T | Type R | Type S | Type N | Type E | Type B | Type C |
|---|---|---|---|---|---|---|---|---|---|
| -200 °C | 1.00 °C | 0.86 °C | 0.98 °C | ✗ | ✗ | 1.24 °C | 0.71 °C | ✗ | ✗ |
| -100 °C | 0.75 °C | 0.69 °C | 0.77 °C | ✗ | ✗ | 0.85 °C | 0.67 °C | ✗ | ✗ |
| 0 °C | 0.68 °C | 0.64 °C | 0.68 °C | 1.82 °C | 1.79 °C | 0.77 °C | 0.62 °C | ✗ | 1.07 °C |
| 100 °C | 0.69 °C | 0.65 °C | 0.67 °C | 1.46 °C | 1.47 °C | 0.76 °C | 0.62 °C | 7.96 °C | 0.96 °C |
| 400 °C | 0.75 °C | 0.71 °C | 0.72 °C | 1.26 °C | 1.31 °C | 0.77 °C | 0.67 °C | 2.30 °C | 0.95 °C |
| 700 °C | 0.81 °C | 0.75 °C | ✗ | 1.23 °C | 1.30 °C | 0.82 °C | 0.73 °C | 1.67 °C | 1.01 °C |
| 1000 °C | 0.88 °C | 0.82 °C | ✗ | 1.23 °C | 1.31 °C | 0.88 °C | 0.83 °C | 1.47 °C | 1.07 °C |
| 1300 °C | 0.96 °C | ✗ | ✗ | 1.26 °C | 1.34 °C | 0.98 °C | ✗ | 1.40 °C | 1.17 °C |
| 1700 °C | ✗ | ✗ | ✗ | 1.36 °C | 1.45 °C | ✗ | ✗ | 1.45 °C | 1.34 °C |
KRYPTONi-8xRTD specifications
| Analog inputs | KRYPTONi-8xRTD | |
|---|---|---|
| Input types | voltage, resistor or universal Ptxxx inputs | |
| Number of channels | 8 | |
| ADC Type | 24-bit delta-sigma | |
| Sampling Rate | Simultaneous 100 S/s | |
| Voltage Mode | ||
| Ranges | ±1 V | ±100 mV |
| Input Accuracy | ±0.02 % of reading ±0.01 % of range ±10 μV | ±0.02% of reading ±0.01% of range ±10 μV |
| Typical Noise floor @ 10/100 S/s | -108 dB / -98 dB | -108 dB / -98 dB |
| Gain Drift | Typical 3 ppm/K (max. 10 ppm/K) | |
| Offset Drift | Typical 0.05 μV/K (max. 0.2 μV/K) | |
| Gain Linearity | < 0.01 % | |
| Input Coupling | DC | |
| Input Impedance | 1 MΩ | |
| Temperature Mode | ||
| Ranges | PT500, PT1000, PT2000 | PT100, PT200 |
| Measurement range | -200 °C to +850 °C | |
| Accuracy | ±0.05 % of reading ±0.2 °C | |
| Temperature Drift | typ. ±5 ppm/K (max. ±12 ppm/K ±0.003 °C/K) | |
| Input Connection | 3 wire or 4 wire | |
| Resistance mode | ||
| Ranges | 0…10 kΩ | 0…1 kΩ |
| Accuracy | ±0.02 % of reading ±0.01 % of range | |
| Input Connection | 3 wire or 4 wire | |
| Additional Specifications | ||
| Input connector | Lemo 0B 6pin EEA.0B.306.CLN | |
| Isolation voltage | 1000 Vpeak channel/ground & channel/channel | |
| Power | ||
| Power Supply | 9 - 48 V DC | |
| Power consumption | 2.5 W | |
| Environmental | ||
| IP rating (see 1) | IP65 | |
| Physical | ||
| Weight | 720 g | |
| Dimensions | 213 x 39 x 58 mm | |
| 1) Depending on the mating connector types | ||
| Rev: 1564584000 | ||
KRYPTONi 8xRTD-HS specifications
| Analog inputs | KRYPTONi-8xRTD-HS | |
|---|---|---|
| Input types | Voltage, Resistance or Universal PTxxx inputs | |
| Number of channels | 8 | |
| ADC Type | 24-bit delta-sigma | |
| Sampling Rate IOLITE | Simultaneous 10, 20, 40, 50, 80, 100, 200, 500, 1k, 2k, 5k, 10k S/s (optional) | |
| Sampling Rate OBSIDIAN | 100 S/s | |
| Voltage Mode | ||
| Ranges | ±1 V | ±100 mV |
| Input Accuracy | ±0.02 % of reading ±0.01 % of range | ±0.02 % of reading ±0.02 % of range |
| Typical Noise Floor @ 10 / 100 / 1k / 10k S/s | -130 dB / -120 dB / -110 dB / -104 dB | -120 dB / -108 dB / -98 dB / -90 dB |
| Offset Drift | Typ. 0.05 μV/K (Max. 0.2 μV/K) | Typ. 0.01 μV/K (Max. 0.1 μV/K) |
| Gain Drift | Typ. 2 ppm/K (Max. 10 ppm/K) | |
| Gain Linearity | < 0.007 % | |
| Input Coupling | DC | |
| Input Impedance | 10 MΩ | |
| Temperature | ||
| Ranges | PT500, PT1000, PT2000 | PT100, PT200 |
| Excitation current | 150 uA | 400 uA |
| Measurement range | -200 °C to +850 °C | |
| Accuracy | ±0.02 % of reading ±0.2 °C | |
| Temperature Drift | Typ. ±2 ppm/K (Max. ±10 ppm/K) | |
| Input Connection | 3-wire or 4-wire | |
| Resistance Mode | ||
| Ranges | 0…10 kΩ | 0…1 kΩ |
| Excitation current | 150 uA | 400 uA |
| Accuracy | ±0.02 % of reading ±0.008 % of range | |
| Input Connection | 3-wire or 4-wire | |
| Additional Specifications | ||
| Input connector | Lemo 0B 6pin EEA.0B.306.CLN | |
| Isolation voltage | 1000 Vpeak channel/ground & channel/channel | |
| Power | ||
| Power Supply | 9 - 48 V DC | |
| Power consumption | Typ. 3.5 W (Max. 4.7 W) | |
| Environmental | ||
| IP rating (see 1) | IP65 | |
| Physical | ||
| Weight | 755 g | |
| Dimensions | 213 x 39 x 58 mm | |
| 1) Depending on the mating connector types | ||
| Rev: 1765454400 | ||
KRYPTON-4xCNT specifications
| Digital inputs | |
|---|---|
| Input type | Digital |
| Number of channels | 4 |
| Input connector | LEMO 1T series 7-pin |
| Timebase | 100 MHz |
| Time base accuracy | Typical: 5 ppm, Max: 20 ppm |
| Max. bandwidth | 10 MHz |
| Input Filter | 500 ns, 1 μs, 2 μs, 4 μs, 5 μs and 7.5 μs |
| Input Level Compatibility | TTL (Low: < 0.8 V, High > 2 V) |
| Input impedance | 100 kΩ pull-up to +3.3 V |
| Input Protection | ±25 V continuous |
| Isolation | Ch to GND isolation |
| Isolated Sensor Power Supply | |
| Output Voltage +5 V | 5.1 V +/- 10 % 100 mA max. (400 mA to supply 4 channels) |
| Output Voltage +12 V | 12.1 V +/-5 % 50 mA max. (200 mA to supply 4 channels) |
| Isolation | Ch to GND isolation |
| Power | |
| Power Supply | 9 - 48 V DC |
| Power consumption | Typ 2.3 W (Max 2.7 W) |
| Environmental | |
| IP rating | IP67 |
| Physical | |
| Dimensions | 213 x 39 x 56 mm |
| Weight | 700 g |
| Rev: 1765454400 | |
KRYPTONi-8xDI-8xDO specifications
| KRYPTONi-8xDI-8xDO | |
|---|---|
| Digital Input | |
| Isolated Input Channels | 8 |
| Compatibility | CMOS Configuration |
| Input low level | UIN < 0.8 V |
| Input high level | UIN > 2.4 V |
| Input high current @ 5 V | IIN < 1.8 mA |
| Input high current @ 30 V | IIN < 6 mA |
| Propagation delay | < 1 μs |
| Sampling rate | Simultaneous 40 kS/s |
| Overvoltage protection | 40 V continuous (65 V peak) |
| Isolation voltage peak | 250 V channel/ground & channel/channel |
| Digital Output | |
| Isolated Output Channels | 8 |
| Compatibility | Open collector with 10 kΩ pull-up to +5 V |
| Maximum sink current | 150 mA (not protected) |
| Maximum switching voltage | 50 V |
| Propagation delay | < 20 μs |
| Maximum update rate | depending on EtherCAT master |
| Isolation voltage peak | 250 V channel to ground, no channel to channel isolation |
| Power | |
| Power Supply | 9 - 48 V DC |
| Power consumption | 2 W |
| Environmental | |
| IP rating | IP67 |
| Physical | |
| Weight | 700 g |
| Slice Dimensions | 213 x 39 x 56 mm |
| Dimensions with Connectors | 213 x 39 x 58.2 mm |
| Rev: 1602590400 | |
KRYPTONi-16xDO specifications
| KRYPTONi-16xDO | |
|---|---|
| Digital Output | |
| Isolated Output Channels | 16 |
| Compatibility | Open collector with 10 kΩ pull-up to +5 V |
| Maximum sink current | 150 mA (not protected) |
| Maximum switching voltage | 50 V |
| Propagation delay | < 20 μs |
| Maximum update rate | depending on EtherCAT master |
| Isolation voltage peak | 250 V channel to ground, no channel to channel isolation |
| Power | |
| Power Supply | 9 - 48 V DC |
| Power consumption | 2 W |
| Environmental | |
| IP rating | IP67 |
| Physical | |
| Weight | 700 g |
| Slice Dimensions | 213 x 39 x 56 mm |
| Dimensions with Connectors | 213 x 39 x 58.2 mm |
| Rev: 1602590400 | |
KRYPTONi-16xDI specifications
| KRYPTONi-16xDI | |
|---|---|
| Digital Input | |
| Isolated Input Channels | 16 |
| Compatibility | CMOS Configuration |
| Input low level | UIN < 0.8 V |
| Input high level | UIN > 2.4 V |
| Input high current @ 5 V | IIN < 1.8 mA |
| Input high current @ 30 V | IIN < 6 mA |
| Propagation delay | < 1 μs |
| Sampling rate | Simultaneous 40 kS/s |
| Overvoltage protection | 40 V continuous (65 V peak) |
| Isolation voltage peak | 250 V channel/ground & channel/channel |
| Power | |
| Power Supply | 9 - 48 V DC |
| Power consumption | 2 W |
| Environmental | |
| IP rating | IP67 |
| Physical | |
| Weight | 700 g |
| Slice Dimensions | 213 x 39 x 56 mm |
| Dimensions with Connectors | 213 x 39 x 58.2 mm |
| Rev: 1602590400 | |