SIRIUS® R1DB/R2DB Technical Specifications
- SIRIUS® R1DB system specifications
- SIRIUS® R2DB system specifications
- SIRIUS® EtherCAT specifications
- SIRIUS® DualCoreADC® filtering specifications
- SIRIUS® HS (High Speed) filtering specifications
- SIRIUS® CNT (counter) specifications
- SIRIUS® analog out specifications
- SIRIUS® CAN bus specifications
- SIRIUS® DualCoreADC® amplifiers overview
- SIRIUS® DualCoreADC® UNI amplifier specifications
- SIRIUS® DualCoreADC® STG amplifier specifications
- SIRIUS® DualCoreADC® STGM amplifier specifications
- SIRIUS® DualCoreADC® ACC amplifier specifications
- SIRIUS® DualCoreADC® CHG amplifier specifications
- SIRIUS® DualCoreADC® MIC200 amplifier specifications
- SIRIUS® DualCoreADC® HV amplifier specifications
- SIRIUS® DualCoreADC® LV amplifier specifications
- SIRIUS® DualCoreADC® MULTI amplifier specifications
- SIRIUS® HD (High Density) amplifiers overview
- SIRIUS® HD-STGS amplifier specifications
- SIRIUS® HD-ACC amplifier specifications
- SIRIUS® HD-LV amplifier specifications
- SIRIUS® HS (High Speed) amplifiers overview
- SIRIUS® HS-ACC amplifier specifications
- SIRIUS® HS-CHG amplifier specifications
- SIRIUS® HS-HV (v2) amplifier specifications
- SIRIUS® HS-HV (v3) amplifier specifications
- SIRIUS® HS-LV amplifier specifications
- SIRIUS® HS-STG amplifier specifications
SIRIUS® R1DB system specifications
| Computer SBOXse v2 (See 1) | |
|---|---|
| Processor | Intel® Core™ i7 (I7-1185G7E) - Kontron 36030-0000-18-7 4x 1.8 GHz base, 4.4 GHz max 8 threads |
| Memory | 8 GB (optional up to 32 GB) |
| Storage | Non-removable NVMe M2 1 TB (2 TB as option) |
| Interfaces and options | |
| USB Front | 4x USB 3.0 |
| Ethernet | 2x 2.5 Gbit/s (RJ45, option: Lemo 1T), 1x WLAN (RP-SMA Female Jack) |
| EtherCAT® | 1x EtherCAT® 100 Mbps Full Duplex, 8-pin LEMO female |
| Synchronisation | 2x SIRIUS® SYNC |
| Video | 1x HDMI |
| GPS (option) | 10 Hz or 100 Hz or 100 Hz + RTK |
| GPS display (option) | External on DSUB-9 female connector + remote power on |
| Display | |
| Type | TFT 12.1'' |
| Resolution | WXGA, 1280 x 800 pixels @ 60Hz |
| Active display Area | 261.1 x 163.2 mm |
| Luminance | 1200 cd/m² |
| Contrast Ratio | 1000:1 (typ.) |
| Viewing Angle (CR>=10) | -89 ~ +89 (H); -89 ~ +89 (V) |
| Synchronization Range Horizontal / Vertical | 30 ~ 83 kHz / 45 ~ 75Hz |
| Batteries | |
| Battery type | RRC2040-2 |
| Number of batteries | 2, hot swappable |
| Voltage | 11.25 V (at fully charged state 13.05 V) |
| Total capacity of single battery | 6.4 Ah |
| Discharge current of single battery | 6.5 A |
| Wrong polarity protection | YES |
| Weight of single battery | 0.34 kg |
| Power | |
| Power Supply | 12 - 36 V, tolerated 9 V voltage dips |
| Power Consumption | Display: 10 W Computer: 20 W |
| Power out | |
| Type | Switched input supply on 2-pin LEMO female |
| Maximum power | 40 W |
| Output Voltage | 11 - 13 V without external supply, 24 V with external supply |
| EtherCAT® Power out | |
| Type | Switched input supply on EtherCAT® connector, 8-pin LEMO female |
| Maximum power | 40 W |
| Output Voltage | 11 - 13 V without external supply, 24 V with external supply |
| Environmental | |
| Operating Temperature | 0 to 40 °C |
| Storage Temperature | -20 to 60 °C |
| Humidity | 5 to 95 % RH non condensing @ 40 °C |
| IP rating | IP20 |
| Shock & Vibration | Vibration sweep sinus (EN 60068-2-6:2008) Vibration random (EN 60721-3-2: 1997 - Class 2M2) Shock (EN 60068-2-27:2009) MIL-STD-810D |
| Physical | |
| Dimensions | 331 x 224 x 145 mm |
| Weight | 6.4 kg (excl. SIRIUS® slices) 8.0 kg (std. system, with batteries) 0.76 kg (SBOXse v2) |
| Rev: 1772442000 | |
| 1) Available after February 2026. From July 2025 to January 2026 Intel® Core™ i5 (I5-1145G7E) - 36030-0000-15-5, 4x 1.5 GHz base, 4.1 GHz max. 8 threads | |
SIRIUS® R2DB system specifications
| Computer SBOXse v2 (See 1) | |
|---|---|
| Processor | Intel® Core™ i7 (I7-1185G7E) - Kontron 36030-0000-18-7 4x 1.8 GHz base, 4.4 GHz max 8 threads |
| Memory | 8 GB (optional up to 32 GB) |
| Storage | Non-removable NVMe M2 1 TB (2 TB as option) |
| Interfaces and options | |
| USB Front | 4x USB 3.0 |
| Ethernet | 2x 2.5 Gbit/s (RJ45, option: Lemo 1T), 1x WLAN (RP-SMA Female Jack) |
| EtherCAT® | 1x EtherCAT® 100 Mbps Full Duplex, 8-pin LEMO female |
| Synchronisation | 2x SIRIUS® SYNC |
| Video | 1x HDMI |
| GPS (option) | 10 Hz or 100 Hz or 100 Hz + RTK |
| GPS display (option) | External on DSUB-9 female connector + remote power on |
| Display | |
| Type | TFT 12.1'' |
| Resolution | WXGA, 1280 x 800 pixels @ 60Hz |
| Active display Area | 261.1 x 163.2 mm |
| Luminance | 1200 cd/m² |
| Contrast Ratio | 1000:1 (typ.) |
| Viewing Angle (CR>=10) | -89 ~ +89 (H); -89 ~ +89 (V) |
| Synchronization Range Horizontal / Vertical | 30 ~ 83 kHz / 45 ~ 75Hz |
| Batteries | |
| Battery type | NL2024HD |
| Number of batteries | 2, hot swappable |
| Voltage | 14.4 V (at fully charged state 16.8 V) |
| Total capacity of single battery | 6.25 Ah |
| Discharge current of single battery | 8 A |
| Wrong polarity protection | YES |
| Weight of single battery | 0.65 kg |
| Power | |
| Power Supply | 12 - 36 V, tolerated 9 V voltage dips |
| Charging power | 55 - 65 W total |
| Power Consumption | Display: 10 W Computer: 20 W |
| Power out | |
| Type | Switched input supply on 2-pin LEMO female |
| Maximum power | 60 W |
| Output Voltage | 13 - 16 V without external supply, 24 V with external supply |
| EtherCAT® Power out | |
| Type | Switched input supply on EtherCAT® connector, 8-pin LEMO female |
| Maximum power | 40 W |
| Output Voltage | 13 - 16 V without external supply, 24 V with external supply |
| Environmental | |
| Operating Temperature | 0 to 40 °C |
| Storage Temperature | -20 to 60 °C |
| Humidity | 5 to 95 % RH non-condensing at 40 °C |
| IP rating | IP20 |
| Shock & Vibration | Vibration sweep sinus (EN 60068-2-6:2008) Vibration random (EN 60721-3-2: 1997 - Class 2M2) Shock (EN 60068-2-27:2009) MIL-STD-810D |
| Physical | |
| Dimensions | 332 x 225 x 194 mm |
| Weight | 7.8 kg (excl. SIRIUS® slices) 11 kg (std. system, with batteries) 0.76 kg (SBOXse v2) |
| Rev: 1772442000 | |
| 1) Available after February 2026. From July 2025 to January 2026 Intel® Core™ i5 (I5-1145G7E) - 36030-0000-15-5, 4x 1.5 GHz base, 4.1 GHz max. 8 threads | |
SIRIUS® EtherCAT specifications
| Interface | |
|---|---|
| Data interface | EtherCAT® |
| Data Rate | 100 Mbps Full Duplex bus speed |
| Max. Sample Rate | DualCore: 20 kS/s, HD 10 kS/s, HS 40 kS/s |
| Max. Throughput per Chain | From 6 MB/s to 10 MB/s |
| Sync Accuracy | below 1 sample to Sirius® |
| EtherCAT® | Max. Cable Length 100 m |
| Max. Number of Units | 64 or more (power injectors may be required) |
| Data Interface Connection | LEMO 1T EtherCAT® hybrid cable Single cable for data, power and sync, daisy chainable |
| Power Supply via EtherCAT | 9 - 36 V DC |
| Rev: 1593507600 | |
SIRIUS® DualCoreADC® filtering specifications
| Digital Filter (vs. Sample Rate) | 0.1 ... 51.2 kS/s | 51.2 ... 102.4 kS/s | 102.4 ... 204.8 kS/s |
|---|---|---|---|
| Bandwidth (-3 dB) | DC to 0.494 fs | DC to 0.49 fs | DC to 0.38 fs |
| Passband (-0.1 dB) | DC to 0.47 fs | DC to 0.45 fs | DC to 0.24 fs |
| Passband Ripple | -0.035 to +0.01 dB | -0.035 to +0.01 dB | -0.035 to +0.01 dB |
| Alias-free Bandwidth | DC to 0.42 fs | DC to 0.32 fs | DC to 0.22 fs |
| Alias Rejection | -95 dB | -92 dB | -97 dB |
| Delay through ADC | 12 / fs | 9 / fs | 5 / fs |
| Oversampling | 256 | 128 | 64 |
SIRIUS® HS (High Speed) filtering specifications
| ADC | |
|---|---|
| Type | 16-bit SAR |
| Sampling Rate | Simultaneous 1 MS/sec |
| Analog anti alias filter | Bypass or analog 100 kHz 5th order Bessel |
| Digital low pass filter (software selectable) | |
| Digital filter Characteristic | Bypass, Butterworth or Bessel |
| Filter Order | 2nd , 4th, 6th or 8th |
| Ratio Sample rate to Filter Freq. | From 2 to 100 |
| Topology | Cascaded IIR Filter (up to 4 sections) |
SIRIUS® CNT (counter) specifications
| Counter | |
|---|---|
| Modes | Counting, waveform timing, encoder, tacho, geartooth sensor |
| Inputs | 3 Digital inputs (individual A, B, Z input) |
| Outputs | 1 Digital output (Alarm; Open collector, max. 100 mA / 30 V) |
| Counter timebase | 102.4 MHz |
| Time base accuracy | Typical: 5 ppm, Max: 20 ppm |
| Max. Bandwidth | 10 MHz |
| Input Filter | 500 ns, 1 μs, 2 μs, 4 μs, 5 μs and 7.5 μs |
| Counter resolution | 32-bit |
| Input Compatibility, Level | TTL/CMOS (Low: <0.8 V, High >2 V) |
| Input termination, impedance | 100 kΩ pull-up to +3.3 V |
| Input Protection | ±25 V continuous |
| Sensor supply | 5 V / 100 mA; 12 V / 50 mA |
| Rev: 1738156469 | |
SIRIUS® analog out specifications
| Voltage Output | |||
|---|---|---|---|
| DAC Type | 24-bit delta-sigma | ||
| Sampling Rate | Simultaneous 200 kS/s | ||
| Number Of Channels | 8 | ||
| Function | File replay, conditioned AI output, FGEN (software option), channel output | ||
| Specifications | |||
| Full Scale | ±10 V | ||
| Analogue out bandwidth | 40 kHz | ||
| Accuracy | ±0.1 % of reading ±0.02 V | ||
| Temperature Drift | ±50 ppm/K of reading ±200 μV/K | ||
| Output configuration | Single Ended | ||
| Typ. Noise floor @ 50 (200) kS/s output rate | -95 dB (-86 dB) @ 100 kHz Bandwidth | ||
| Inter channel phase mismatch | 0.1° * fout [kHz] + 0.1° | ||
| THD | < -90 dB @ 1 kHz, 10 Vpeak-peak | ||
| Output Impedance | < 1 Ω | ||
| Maximum Output Current | 20 mA | ||
| Maximum load | > 1000 Ω | ||
| Output Protection | Continuous short to ground | ||
| Signal Delay | |||
| Sampling Rate | 100 S/s…50 kS/s | 50 kS/s…100 kS/s | 100 kS/s…200 kS/s |
| Signal conditioning mode (see 1) | 14 Samples + 50 μs | 19 Samples + 2 μs | 12 Samples + 35 μs |
| Additional Specifications | |||
| Output connector | BNC | ||
| 1) SIRIUS-HD series supports signal conditioning only for channels 1-8 | |||
SIRIUS® CAN bus specifications
| CAN bus | |
|---|---|
| Number of ports | 1 |
| Interface type | CAN 2.0B, up to 1 MBit/s CAN-FD, up to 8 Mbit/s (See 1) |
| Special applications | CCP, OBDII, J1939, CAN output |
| Galvanic isolation | Isolated (only on isolated motherboard) |
| Bus pin fault protection | ±36 V |
| ESD protection | 8 kV |
| 1) CAN-FD is available for SIRIUS DualCore, HighSpeed and SIRIUS HD modular slices, also available for SIRIUS DualCore and SIRIUS HS rack slices. | |
SIRIUS® DualCoreADC® amplifiers overview
| UNI | STG | STGM | MULTI | ACC | MIC200 | CHG | HV | LV | |
|---|---|---|---|---|---|---|---|---|---|
| Connectors | DSUB-9, L2B10f | DSUB-9, L2B7f, L2B10f | DSUB-9, L2B7f, L2B10f | DSUB-15, L2B16f | BNC, TNC | L1B7f | BNC, TNC | BANANA | DSUB-9, BNC, BANANA |
| Counter version | ✓ | ✓ | ✓ | ✓ | ✓ | ✗ | ✓ | ✗ | ✓ |
| Isolated version | ✓ | ✓ | ✓ | ✓ | ✓ | ✓ | ✓ | ✓ | ✓ |
| Differential version | ✓ | ✓ | ✓ | ✓ | ✓ | ✗ | ✗ | ✗ | ✓ |
| EtherCAT version | ✓ | ✓ | ✓ | ✓ | ✓ | ✓ | ✓ | ✓ | ✓ |
| Rack version | ✓ | ✓ | ✓ | ✓ | ✓ | ✓ | ✓ | ✓ | ✓ |
| Fanless version | ✗ | ✗ | ✓ | ✗ | ✓ | ✓ | ✗ | ✓ | ✓(see 1) |
| Channels per slice | 8 | 8 | 8 | 8 | 8 | 8 | 8 | 8 | 8 |
| Data rate / channel | 200 kS/s USB, 20 kS/s EtherCAT | 200 kS/s USB, 20 kS/s EtherCAT | 200 kS/s USB, 20 kS/s EtherCAT | 200 kS/s USB, 20 kS/s EtherCAT | 200 kS/s USB, 20 kS/s EtherCAT | 200 kS/s USB, 20 kS/s EtherCAT | 200 kS/s USB, 20 kS/s EtherCAT | 200 kS/s USB, 20 kS/s EtherCAT | 200 kS/s USB, 20 kS/s EtherCAT |
| Resolution | 24-bit dual core | 24-bit dual core | 24-bit dual core | 24-bit dual core | 24-bit dual core | 24-bit dual core | 24-bit dual core | 24-bit dual core | 24-bit dual core |
| Bandwidth | 70 kHz | 70 kHz | 70 kHz | 70 kHz | 70 kHz | 70 kHz | 70 kHz | 70 kHz | 70 kHz |
| Voltage ranges | ±100 V, ±10 V, ±1 V, ±100 mV, ±10 mV | ±50 V, ±10 V, ±1 V, ±100 mV | ±10 V, ±1 V, ±100 mV, ±10 mV | ±10 V, ±1 V, ±100 mV, ±50 mV | ±10 V, ±500 mV | ±10 V, ±500 mV | ±10 V, ±500 mV | ±1200 V, ±50 V | ±200 V, ±10 V, ±1 V, ±100 mV |
| Input coupling | DC, AC 0.3 Hz / 0.16 Hz (1, 3, 5, 10, 20 Hz SW) | DC, AC 1 Hz (3, 10 Hz SW) | DC | DC | DC, AC 0.1 Hz, 1 Hz | AC 0.35 Hz (3, 10 Hz SW) | DC, AC 0.1 Hz, 1 Hz, 10 Hz, 100 Hz | DC | DC, AC 1 Hz (3, 10 Hz SW) |
| Sensor excitation | 0..20 V max. 0.8 W, 0..60 mA max 0.5 W, IEPE 2..20 mA | 0..20 V max. 0.8 W, 0..60 mA max 0.5 W | 0..15 V max. 44 mA | 0..12 V max. 44 mA | IEPE 2, 4, 8, 12, 16, 20 mA | 14 V bipolar, 200 V polarization | IEPE 4, 8, 12 mA | ✗ | 2..30 V bipolar, 0..24 V unipolar, max. 0.2 A / 2 W |
| Bridge connection | Full, ½, ¼ 350 Ω, ¼ 120 Ω 3, 4 wire | Full, ½, ¼ 350 Ω, ¼ 120 Ω 3, 4 wire | Full, ½, ¼ 350 Ω, ¼ 120 Ω 3 wire | Full, ½, ¼ 350 Ω, ¼ 120 Ω 3 wire | ✗ | ✗ | ✗ | ✗ | Full |
| Programmable shunt | 59.88 kΩ, 175 kΩ bipolar | 59.88 kΩ, 175 kΩ bipolar | 100 kΩ bipolar | 59.88 kΩ | ✗ | ✗ | ✗ | ✗ | ✗ |
| IEPE input | ✓ | DSI-ACC | DSI-ACC | DSI-ACC | ✓ | ✗ | ✓ | ✗ | DSI-ACC |
| Resistance | ✓ | ✓ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ | ✗ |
| Temperature (PTx) | ✓ | ✓ | DSI-RTD | DSI-RTD | ✗ | ✗ | ✗ | ✗ | DSI-RTD |
| Thermocouple | DSI-TH | DSI-TH | DSI-TH | DSI-TH | ✗ | ✗ | ✗ | ✗ | DSI-TH |
| Current | ext. shunt, DSI | ext. shunt, DSI | ext. shunt, DSI | ext. shunt, DSI | ext. shunt | ext. shunt | ext. shunt | ✗ | ext. shunt, DSI |
| Potentiometer | ✓ | ✓ | ✓ | ✓ | ✗ | ✗ | ✗ | ✗ | ✗ |
| LVDT | DSI-LVDT | DSI-LVDT | DSI-LVDT | DSI-LVDT | ✗ | ✗ | ✗ | ✗ | DSI-LVDT |
| Charge | DSI-CHG | DSI-CHG | DSI-CHG | DSI-CHG | ✗ | ✗ | 100 k, 10 k pC | ✗ | DSI-CHG |
| TEDS | ✓ | ✓ | ✓ | ✓ | ✓ | ✓ | ✓ | ✗ | ✓ |
| Isolation voltage | 1000 V | 1000 V | 1000 V | 1000 V | 1000 V | 1000 V | 1000 V | CAT II 1000 V | 1000 V |
| Power consumption per channel | 2 W/ch | 2 W/ch | 1.3 W/ch | 1.3 W/ch | 1 W/ch | 1.1 W/ch | 1.2 W/ch | 1 W/ch | 1.2 W/ch |
| Advanced functions | Supports all strain types, IEPE sensors and high input range | Supports all strain types and high input range | Low power, sensor and amplifier balance, bipolar shunt | Analog Output, Low power, sensor and amplifier balance | Sensor error detection, high dynamic range | Supports polarized condenser microphones | Sensor error detection in IEPE and CHG mode (injection) | High voltage, high isolation | High sensor power and multi range |
| (1) Fanless operation only for BNC or Banana version (without excitation) | |||||||||
| Rev: 1646204700 | |||||||||
SIRIUS® DualCoreADC® UNI amplifier specifications
| Inputs | |||||
|---|---|---|---|---|---|
| Input types | Voltage Full/Half/Quarter bridge strain (120 Ω and 350 Ω) IEPE Potentiometer, RTD, Resistance Current (ext. Shunt) | ||||
| ADC Type | 24-bit delta-sigma dual core with anti-aliasing filter | ||||
| Sampling Rate | Simultaneous 200 kS/s | ||||
| Voltage Mode | |||||
| Dual Core Ranges (Low) | ±100 V (±5 V) | ±10 V (±500 mV) | ±1 V (±50 mV) | ±100 mV (±5 mV) | ±10 mV (±5 mV) |
| Gain accuracy | ±0.05 % of reading | ||||
| Offset accuracy (Dual Core) | ±20 (10) mV | ±2 (1) mV | ±0.2 (0.2) mV | ±0.1 (0.1) mV | ±0.1 (0.1) mV |
| Offset Accuracy after Balance Amplifier | ±1 mV | ±0.1 mV | ±0.02 mV | ±0.01 mV | ±0.01 mV |
| Typ. Dynamic Range @ 10 kS/s (Dual Core) | -147 dB (-160 dB) | -147 dB (-165 dB) | -144 dB (-153 dB) | -137 dB (-137 dB) | -117 dB (-117 dB) |
| Typ. Noise floor @ 10 kS/s (Dual Core) | -109 dB (-122 dB) | -109 dB (-126 dB) | -108 dB (-114 dB) | -100 dB (-100 dB) | -80 dB (-80 dB) |
| Typ. CMR @ DC...50 Hz / 400 Hz / 1 kHz | 84 / 74 / 67 dB | 104 / 103 / 100 dB | 112 / 105 / 99 dB | 121 / 107 / 98 dB | 121 / 107 / 98 dB |
| Gain Drift | Typical 10 PPM/K, max. 40 PPM/K | ||||
| Offset Drift | range ≥ 100 mV: Typical 0.3 μV/K + 2 ppm of range/K, max 0.8 μV/K + 10 ppm of range/K range 10 mV: Typical 0.5 μV/K, max 1.8 μV/K | ||||
| Gain Linearity | range ≥ 100 mV: <0.02 % range 10 mV: <0.03 % | ||||
| Inter Channel Phase-mismatch | 0.02° * fin [kHz] + 0.1° (@ 200 kS/s) | ||||
| Channel Crosstalk | -120 dB @ 10 kHz | ||||
| Input Coupling | DC AC 0.3 Hz for 100 V range; AC 0.16 Hz for all other ranges; (1 Hz, 3 Hz, 5 Hz, 10 Hz, 20 Hz per SW) | ||||
| Input Impedance | 2 MΩ between IN+ and IN- for 100 V Range; all other Ranges > 1 GΩ | ||||
| Max. Common Mode Voltage | Isolated version: ±500 V | ||||
| Differential version: 100 V Range: ±120 V; all other Ranges: ±12 V | |||||
| Overvoltage Protection | 100 V Range: 200 V; all other Ranges: 100 V (200 V peak for 10 ms) | ||||
| Excitation Voltage | |||||
| Excitation Voltage | Free programmable (16-bit DAC) | ||||
| Predefined levels | 0, 1, 2.5, 5, 10, 15 and 20 VDC | ||||
| Accuracy | ±0.05 % ±2 mV | ||||
| Drift | ±10 ppm/K ±100 μV/K | ||||
| Load stability | 0 % to 100 % load <0.01 % | ||||
| Noise @ 10 V / 350 Ω | <150 μVrms @ 10 kS/s | ||||
| Line regulation over 20 Ω of change | <0.005 % @ 120 Ω load | ||||
| Sense Impedance to Exc / to GND | 100 kΩ / >100 MΩ | ||||
| Current limit | 100 mA (max. 800 mW) | ||||
| Protection | Continuous short to ground | ||||
| Excitation Current | |||||
| Excitation Current | Free programmable (16-bit DAC) | ||||
| Predefined levels | 0.1, 1, 2, 5, 10, 20 and 60 mA DC | ||||
| Ranges | 0.1 mA - 10 mA | >10 mA - 60 mA | |||
| Accuracy | 0.05 % ±2 μA | 0.1 % ±20 μA | |||
| Drift Typical | 15 ppm/K | 45 ppm/K | |||
| Compliance voltage | 20 Volt, max. 500 mW | ||||
| Output Impedance | >100 MΩ | ||||
| Bridge Connection Types | |||||
| Bridge types | Full / Half / Quarter bridge (3-wire or 4-wire) / Quarter Constant Current (2-wire or 4-wire) | ||||
| Ranges | 2 mV/V…1000 mV/V free programmable with Dual Core | ||||
| Internal Bridge Completion | Half and Quarter bridge 120 Ω and 350 Ω | ||||
| Typ. Bridge Completion Accuracy | 0.05 %; TCR: 2 ppm/K (others on request) | ||||
| Internal Shunt Resistor | 59.88 kΩ, 175 kΩ, software selectable (others on request) | ||||
| Typ. Shunt Resistor Accuracy | 0.05 %; TCR: 10 ppm/K (others on request) | ||||
| Input Short, Sensor Offset Adjust | Software selectable | ||||
| IEPE mode | |||||
| IEPE Current | Free programmable (16-bit DAC) | ||||
| Predefined levels | 2, 4, 8, 12, 16 or 20 mA | ||||
| Compliance voltage | 25 Volt | ||||
| Output Impedance | >100 kΩ | ||||
| Sensor detection | Shortcut: <4 Volt; Open: >19 Volt | ||||
| Counters (only on UNI+ type) | |||||
| Inputs | 1 digital counter input 3 digital inputs Fully synchronised with analog data Alarm output | ||||
| Counter Modes | counting, waveform timing, encoder, tacho, gear-tooth sensor | ||||
| General Counter Specifications | See "SIRIUS counter specifications" | ||||
| Additional Specifications | |||||
| Input connector | DSUB-9, 8-pin LEMO-2B, 10-pin LEMO-2B, 16-pin LEMO-2B (others on request) | ||||
| TEDS support | Standard + DSI® adapters | ||||
| Rev: 1657094400 | |||||
SIRIUS® DualCoreADC® STG amplifier specifications
| Inputs | ||||
|---|---|---|---|---|
| Input types | Voltage Full/Half/Quarter bridge strain (120 Ω and 350 Ω) Potentiometer, RTD, Resistance Current (ext. Shunt) | |||
| ADC Type | 24-bit delta-sigma dual core with anti-aliasing filter | |||
| Sampling Rate | Simultaneous 200 kS/s | |||
| Voltage Mode | ||||
| Dual Core Ranges (Low) | ±50 V (±2.5 V) | ±10 V (±500 mV) | ±1 V (±50 mV) | ±100 mV (±5 mV) |
| Gain accuracy | ±0.05 % of reading | |||
| Offset accuracy (Dual Core) | ±20 (10) mV | ±2 (1) mV | ±0.2 (0.2) mV | ±0.1 (0.1) mV |
| Offset Accuracy after Balance Amplifier | ±1 mV | ±0.1 mV | ±0.02 mV | ±0.01 mV |
| Typ. Dynamic Range @ 10 kS/s (Dual Core) | -137 dB (-147 dB) | -137 dB (-152 dB) | -137 dB (-147 dB) | -135 dB (-137 dB) |
| Typ. Noise floor @ 10 kS/s (Dual Core) | -108 dB (-118 dB) | -107 dB (-125 dB) | -107 dB (-113 dB) | -100 dB (-100 dB) |
| Typ. CMR @ DC...50 Hz / 400 Hz / 1 kHz | 56 / 56 / 56 dB | 88 / 86 / 84 dB | 97 / 96 / 95 dB | 115 / 112 / 102 dB |
| Gain Drift | Typical 10 PPM/K, max. 40 PPM/K | |||
| Offset Drift | Typical 0.3 μV/K + 2 ppm of range/K, max 0.8 μV/K + 10 ppm of range/K | |||
| Gain Linearity | <0.02 % | |||
| Inter Channel Phase-mismatch | 0.02° * fin [kHz] + 0.1° (@ 200 kS/s) | |||
| Channel Crosstalk | -120 dB @ 10 kHz | |||
| Input Coupling | DC, AC 1 Hz (3 Hz, 10 Hz per SW) | |||
| Input Impedance | 1 MΩ between IN+ and IN- for 50 V Range; all other Ranges > 1 GΩ | |||
| Max. Common Mode Voltage | Isolated version: ±500 V | |||
| Differential version: 50 V Range: ±60 V; all other Ranges: ±12 V | ||||
| Overvoltage Protection | 50 V Range: 300 V; all other Ranges: 50 V (200 V peak for 10 ms) | |||
| Excitation Voltage | ||||
| Excitation Voltage | Free programmable (16-bit DAC) | |||
| Predefined levels | 0, 1, 2.5, 5, 10, 15 and 20 VDC | |||
| Accuracy | ±0.05 % ±2 mV | |||
| Drift | ±10 ppm/K ±100 μV/K | |||
| Load stability | 0 % to 100 % load <0.01 % | |||
| Noise @ 10 V / 350 Ω | <150 μVrms @ 10 kS/s | |||
| Line regulation over 20 Ω of change | <0.005 % @ 120 Ω load | |||
| Sense Impedance to Exc / to GND | 100 kΩ / >100 MΩ | |||
| Current limit | 100 mA (max. 800 mW) | |||
| Protection | Continuous short to ground | |||
| Excitation Current | ||||
| Amplifier version | STGv2 | STGv3 | ||
| Excitation Current | Free programmable (16-bit DAC) | |||
| Predefined levels | 0.1, 1, 2, 5, 10, 20 and 60 mA DC | |||
| Ranges | 0.1 - 10 mA | 10 - 60 mA | 0.01 - 1 mA | 1 - 60 mA |
| Accuracy | 0.1 % ±2 μA | 0.5 % ±50 μA | preview 0.02 % ±2 μA | preview 0.25 % ±20 μA |
| Drift | 15 ppm/K | 100 ppm/K | preview 15 ppm/K | preview 100 ppm/K |
| Compliance voltage | 20 Volt, max. 500 mW | |||
| Output Impedance | >1 MΩ | |||
| Bridge Connection Types | ||||
| Bridge types | Full / Half / Quarter bridge (3-wire or 4-wire) | |||
| Ranges | 2 mV/V…1000 mV/V free programmable with Dual Core | |||
| Internal Bridge Completion | Half and Quarter bridge 120 Ω and 350 Ω | |||
| Maximum sensor resistance | 5 kΩ for Full and Half bridge measurements | |||
| Typ. Bridge Completion Accuracy | 0.05 %; TCR: 2 ppm/K (others on request) | |||
| Internal Shunt Resistor | 59.88 kΩ, 175 kΩ, software selectable (others on request) | |||
| Typ. Shunt Resistor Accuracy | 0.05 %; TCR: 10 ppm/K (others on request) | |||
| Input Short, Sensor Offset Adjust | Software selectable | |||
| Counters (only on STG+ type) | ||||
| Inputs | 1 digital counter input 3 digital inputs Fully synchronised with analog data Alarm output | |||
| Counter Modes | counting, waveform timing, encoder, tacho, gear-tooth sensor | |||
| General Counter Specifications | See "SIRIUS counter specifications" | |||
| Additional Specifications | ||||
| Input connector | DSUB-9, 8-pin LEMO-2B, 16-pin LEMO-2B (others on request) | |||
| TEDS support | Standard + DSI® adapters | |||
| Rev: 1583245200 | ||||
SIRIUS® DualCoreADC® STGM amplifier specifications
| Inputs | ||||
|---|---|---|---|---|
| Input types | Voltage Full / Half / Quarter bridge strain (120 Ω and 350 Ω) | |||
| ADC Type | 24-bit delta-sigma dual core with anti-aliasing filter | |||
| Sampling Rate | Simultaneous 200 kS/s | |||
| Voltage Mode | ||||
| Dual Core Ranges (Low range) | ±10 V (±500 mV) | ±1 V (50 mV) | ±100 mV (±5 mV) | ±10 mV (±0.5 mV) |
| Gain accuracy | ±0.05 % of reading | |||
| Offset accuracy (Dual Core) | ±5 (2) mV | ±0.5 (0.2) mV | ±0.1 (0.1) mV | ±0.1 (0.1) mV |
| Offset accuracy after Balance Amplifier | 0.2 mV | 0.02 mV | 0.02 mV | 0.02 mV |
| Typ. Dynamic Range @ 10 kS/s (Dual Core) | -137 dB (-152 dB) | -137 dB (-147 dB) | -130 dB (-132 dB) | -112 dB (-112 dB) |
| Typ. Noise floor @ 10 kS/s (Dual Core) | -105 dB (-121 dB) | -104 dB (-111 dB) | -95 dB (-95 dB) | -75 dB (-75 dB) |
| Typ. CMR @ DC…50 Hz / 400 Hz / 1 kHz | 88 / 86 / 84 dB | 97 / 96 / 95 dB | 111 / 110 / 102 dB | 111 / 110 / 102 dB |
| Gain Drift | Typical 10 ppm/K, max. 40 ppm/K | |||
| Offset Drift | Typical 0.3 μV/K + 5 ppm of range/K, max 2 μV/K + 10 ppm of range/K | |||
| Gain Linearity | <0.02 % | |||
| Inter Channel Phase-mismatch | 0.02° * fin [kHz] + 0.1° (@ 200 kS/s) | |||
| Channel Crosstalk | -120 dB @ 10 kHz | |||
| Input Coupling | DC | |||
| Input Impedance | 10 MΩ | |||
| Max. Common Mode Voltage | Isolated version: ±500 V | |||
| Differential version: ±12 V | ||||
| Overvoltage Protection | In+ to In-: 50 V continuous; 200 V peak (10 ms) | |||
| Excitation Voltage | ||||
| Excitation Voltage | Free programmable (16-bit DAC) | |||
| Predefined levels | 0, 1, 2.5, 5, 10 and 15 VDC | |||
| Accuracy | ±0.05 % ±2 mV | |||
| Drift | ±50 ppm/K ±100 μV/K | |||
| Load stability | 0 % to 100 % load <0.01 % | |||
| Noise @ 10 V / 350 Ω | <150 μVrms @ 10 kS/s | |||
| Line regulation over 20 Ω of change | <0.005 % @ 120 Ω load | |||
| Sense Impedance | 100 kΩ to Exc / >100 MΩ to GND | |||
| Current limit | 45 mA | |||
| Protection | Continuous short to ground | |||
| Bridge Connection Types | ||||
| Bridge Types | Full / Half / Quarter bridge (3-wire) | |||
| Ranges | 2 mV/V…1000 mV/V free programmable with Dual Core | |||
| Internal Bridge Completion | ½ bridge 1 kΩ and ¼ bridge 120 Ω and 350 Ω | |||
| Typ. Bridge Completion Accuracy | 0.05 %; TCR: 5 ppm/K (others on request) | |||
| Maximum sensor resistance | 5 kΩ for Full and Half bridge measurements | |||
| Internal Shunt Resistor | 100 kΩ, software selectable (others on request) | |||
| Typ. Shunt Resistor Accuracy | 0.05 %; TCR: 10 ppm/K (others on request) | |||
| Input Short, Sensor Offset Adjust | Software selectable | |||
| Counters (only on STGM+ type) | ||||
| Inputs | 1 digital counter input 3 digital inputs Fully synchronised with analog data Alarm output | |||
| Counter Modes | counting, waveform timing, encoder, tacho, gear-tooth sensor | |||
| General Counter Specifications | counting, waveform timing, encoder, tacho, gear-tooth sensor | |||
| Additional Specifications | ||||
| Input connector | DSUB-9, 8-pin LEMO-2B, 16-pin LEMO-2B (others on request) | |||
| TEDS support | Standard + DSI® adapters | |||
SIRIUS® DualCoreADC® ACC amplifier specifications
| Inputs | ||
|---|---|---|
| Input types | Voltage IEPE | |
| ADC Type | 24-bit delta-sigma dual core with anti-aliasing filter | |
| Sampling Rate | Simultaneous 200 kS/s | |
| Voltage Mode | ||
| Ranges (Dual Core Low Range) | ±10 V (±500 mV) | ±500 mV (NA) |
| Input Accuracy (Dual Core) | ±0.1 % of reading ±10 (1) mV | ±0.1 % of reading ±1 (NA) mV |
| Dynamic Range @ 10 kS/s (Dual Core) | -140 dB (-160 dB) | -135 dB (NA) |
| Typ. Noise floor @ 50 kS/s (Dual Core) | -107 dB (-125 dB) | -100 dB (NA) |
| Typ. CMR @ 50 Hz / 1 kHz | 140 / 120 dB | 140 / 120 dB |
| Gain Drift | Typical 10 ppm/K, max. 30 ppm/K | |
| Offset Drift | Typical 0.5 μV/K + 2 ppm of range/K, max 2 μV/K + 10 ppm of range/K | |
| Gain Linearity | <0.02 % | |
| Inter Channel Phase-mismatch | 0.02° * fin [kHz] + 0.1° (@ 200 kS/s) | |
| Channel Crosstalk | <-160 dB @ 1 kHz | |
| Input Coupling | DC, AC (0.1 Hz, 1 Hz, 3 Hz or 10 Hz) | |
| Input Impedance | 3.4 MΩ (340 kΩ for AC coupling ≥1 Hz) in parallel with 270 pF | |
| Overvoltage Protection | In+ to In-: 50 V continuous; 200 V peak (10 ms) | |
| IEPE Mode | ||
| Excitation | 2, 4, 8, 12, 16 or 20 mA | |
| Compliance voltage | 25 Volt | |
| Output Impedance | >100 kΩ | |
| Sensor detection | Shortcut: <4 Volt; Open: >19 Volt | |
| Input Coupling | 0.1 Hz, 1 Hz, 3 Hz or 10 Hz | |
| Counters (ACC+ type only) | ||
| Inputs | 1 digital counter input 3 digital inputs Fully synchronised with analog data | |
| Counter Modes | counting, waveform timing, encoder, tacho, gear-tooth sensor | |
| General Counter Specifications | See "SIRIUS counter specifications" | |
| Additional Specifications | ||
| Input connector BNC | BNC or TNC (others on request) | |
| TEDS support | IEPE mode only | |
| Rev: 1767110400 | ||
SIRIUS® DualCoreADC® CHG amplifier specifications
| Inputs | ||
|---|---|---|
| Input types | Voltage IEPE Charge | |
| ADC Type | 24-bit delta-sigma dual core with anti-aliasing filter | |
| Sampling Rate | Simultaneous 200 kS/s | |
| Voltage Mode | ||
| Ranges (Dual Core Low Range) | ±10 V (500 mV) | ±500 mV (NA) |
| Input Accuracy (Dual Core) | ±0.05 % of reading ±5 (0.5) mV | ±0.05 % of reading ±0.5 mV |
| Typ. Dynamic Range @ 10kS/s (Dual Core) | -140 dB (-155 dB) | -130 dB (NA) |
| Typ. Noise floor @ 50 kS/s (Dual Core) | -107 dB (-124 dB) | -98 dB (NA) |
| Typ. CMR @ 50 Hz / 1 kHz | 140 dB / 120 dB | 140 dB / 120 dB |
| Gain Drift | Typical 10 ppm/K, max. 30 ppm/K | |
| Offset Drift | Typical 1 μV/K + 5 ppm of range/K, max 5 μV/K + 10 ppm of range/K | |
| Gain Linearity | <0.02 % | |
| Inter Channel Phase-mismatch | 0.02° * fin [kHz] + 0.1° (@ 200 kS/s) | |
| Channel Crosstalk | -180 dB @ 50 Hz; -160 dB @ 1 kHz | |
| Input Coupling | DC, AC (0.1 Hz, 1 Hz, 10 Hz or 100 Hz) | |
| Input Impedance | 3.3 MΩ in parallel with 100 pF | |
| Overvoltage Protection | In+ to In-: 50 V continuous; 200 V peak (10 ms) | |
| IEPE Mode | ||
| Excitation | 4, 8 or 12 mA | |
| Compliance voltage | 25 Volt | |
| Output Impedance | >100 kΩ | |
| Sensor detection | Shortcut: <4 Volt; Open: >19 Volt | |
| Input Coupling | 0.1 Hz, 1 Hz, 10 Hz, 100 Hz | |
| Charge Mode | ||
| Charge ranges (Low Range) | ±100 000 pC (5000 pC) | ±10 000 pC (500 pC) |
| Input accuracy (HPF 0.1 Hz) | ±0.5 % of reading ±20 pC | ±0.5 % of reading ±5 pC |
| Typ. Dynamic Range @ 10 kS/s (Dual Core) | -130 dB (-150 dB) | -120 dB (-140 dB) |
| Typ. Noise floor @ 50 kS/s (Dual Core) | 107 dB (120 dB) | 105 dB (118 dB) |
| Input coupling | 0.01 Hz, 0.03 Hz, 0.1 Hz, 0.5 Hz, 1 Hz, 10 Hz or 100 Hz | |
| Charge injection | 1 Vrms, 20 Hz, 0.5 % accuracy | |
| Counters (only in CHG+ type) | ||
| Inputs | 1 digital counter input 3 digital inputs Fully synchronised with analog data | |
| Counter modes | counting, waveform timing, encoder, tacho, gear-tooth sensor | |
| General counter specifications | See "SIRIUS counter specifications" | |
| Additional specifications | ||
| Input connector | BNC or TNC (others on request) | |
| TEDS support | IEPE mode only | |
| Rev: 1767110400 | ||
SIRIUS® DualCoreADC® MIC200 amplifier specifications
| Inputs | ||
|---|---|---|
| Input types | Voltage (Externally polarized microphones) | |
| ADC Type | 24-bit delta-sigma dual core with anti-aliasing filter | |
| Sampling Rate | Simultaneous 200 kS/s | |
| Voltage Mode | ||
| Dual Core Ranges (Dual Core) | ±10 V (±500 mV) | ±500 mV (NA) |
| Input Accuracy | ±0.1 % of reading ±1 mV | ±0.1 % of reading ±50 μV |
| Input Accuracy (Signal: -1 dBFS @ 1 kHz) | ±0.02 dB | |
| Frequency response (10 Hz - 10 kHz) | ±0.02 dB | |
| Frequency response (1 Hz - 25.6 kHz) | ±0.1 dB | |
| Dynamic Range @ 10 kS/s (Dual Core) | -139 dB (-160 dB) | -134 dB (NA) |
| Typ. Noise floor @ 50 kS/s (Dual Core) | -109 dB (-125 dB) | -100 dB (NA) |
| Typ. THD @ 200 kS/s (Signal: -1 dBFS @ 1 kHz) | -94 dB | -92 dB |
| Typ. SFDR @ 200 kS/s (Signal: -1 dBFS @ 1 kHz) | 97 dB | 93 dB |
| Typ. CMR @ 50 Hz / 400 Hz / 1 kHz | 140 / 133 / 126 dB | 140 / 133 / 126 dB |
| Gain Drift | typ. 10 ppm/K, max. 40 ppm/K | |
| Gain Linearity | <0.03 % | |
| Input Coupling | AC 0.35 Hz (3 Hz, 10 Hz per SW) | |
| Excitation | ||
| Preamplifier Excitation Voltage | ±14 V | |
| Polarization Voltage | 200 V | |
| Polarization Voltage Accuracy | ±0.5 % | |
| Protection | Continuous short to ground | |
| Additional Specifications | ||
| Input connector | 7-pin LEMO 1B | |
| TEDS support | Standard (Class 2) | |
| Rev: 1683968400 | ||
SIRIUS® DualCoreADC® HV amplifier specifications
| Inputs Voltage | ||
|---|---|---|
| ADC Type | 24-bit delta-sigma dual core with anti-aliasing filter | |
| Sampling Rate | Simultaneous 200 kS/s | |
| Voltage Mode | ||
| Ranges (Dual Core Low Range) | ±1200 V (±50 V) | ±50 V (NA) |
| Gain accuracy | ±0.05 % of reading | |
| Offset accuracy (Dual Core) | ±100 (50) mV | ±50 mV |
| Typ. Dynamic Range @ 10 kS/s (Dual Core) | -142 dB (-158 dB) | -132 dB (NA) |
| Typ. Noise floor @ 50 kS/s (Dual Core) | -107 dB (-120 dB) | -95 dB (NA) |
| Typ. CMR @ 50 Hz / 1 kHz | 85 / 60 dB | 85 / 60 dB |
| Gain Drift | Typical 5 ppm/K, max. 30 ppm/K | |
| Offset Drift | Typical 1 mV/K + 1 ppm of range/K, max 2 mV/K + 5 ppm of range/K | |
| Gain Linearity | <0.02 % | |
| Inter Channel Phase-mismatch | 0.04° * fin [kHz] + 0.2° (@ 200 kS/s) | |
| Channel Crosstalk | -115 dB @ 50 Hz; -90 dB @ 1 kHz | |
| Input Coupling | DC | |
| Input Impedance | 10 MΩ | 2 pF | |
| Overvoltage category | CAT II 1000 V | |
| Maximum input voltage | 2000 Vdc or 2000 Vpk (1414 Vacrms) | |
| Additional Specifications | ||
| Input connector | Banana | |
| TEDS support | Not available | |
| Rev: 1698322800 | ||
SIRIUS® DualCoreADC® LV amplifier specifications
| Inputs | ||||
|---|---|---|---|---|
| Input types | Voltage Full bridge strain Current (ext. Shunt) | |||
| ADC Type | 24-bit delta-sigma dual core with anti-aliasing filter | |||
| Sampling Rate | Simultaneous 200 kS/s | |||
| Voltage Mode | ||||
| Dual Core Ranges (Low Range) | ±200 V (10 V) | ±10 V (500 mV) | ±1 V (50 mV) | ±100 mV (5 mV) |
| Gain accuracy | ±0.05 % of reading | |||
| Offset accuracy (Dual Core) | ±40 (20) mV | ±2 (1) mV | ±0.2 (0.2) mV | ±0.1 (0.1) mV |
| Offset accuracy after Balance Amplifier | 2 mV | 0.1 mV | 0.02 mV | 0.01 mV |
| Typ. Dynamic Range @ 10 kS/s (Dual Core) | -136 dB (-146 dB) | -137 dB (-152 dB) | -137 dB (-147 dB) | -130 dB (-132 dB) |
| Typ. Noise floor @ 10 kS/s (Dual Core) | -109 dB (-118 dB) | -109 dB (-126 dB) | -109 dB (-116 dB) | -97 dB (-97 dB) |
| Typ. CMR @ 50 Hz / 400 Hz / 1 kHz / 10 kHz | 70 / 70 / 60 / 55 dB | 95 / 95 / 89 / 84 dB | 105 / 105 / 100 / 95 dB | 115 / 112 / 107 / 102 dB |
| Gain Drift | Typical 10 ppm/K, max. 30 ppm/K | |||
| Offset Drift | Typical 0.3 μV/K + 5 ppm of range/K, max: 2 μV/k + 10 ppm of range/K | |||
| Gain Linearity | <0.02 % | |||
| Inter Channel Phase-mismatch | 0.02° * fin [kHz] + 0.1° (@ 200 kS/s and 10 V range) | |||
| Channel Crosstalk | -120 dB @ 10 kHz (range ≤10 V); -95 dB @ 10 kHz (range = 100 V) | |||
| Input Coupling (see 1) | DC, AC 1 Hz (3 Hz, 10 Hz per SW) | |||
| Input Impedance | 200 V Range: 1 MΩ; all other ranges 10 MΩ between IN+ or IN- against GND | |||
| Max. common mode voltage | Isolated version ±500 V | |||
| Differential version: 200 V range: ±200 V; all other Ranges: ±12 V | ||||
| Overvoltage Protection | 200 V Range: 300 V; all other ranges: 100 V (250 V peak for 10 ms) | |||
| Excitation Voltage | ||||
| Excitation voltage | Unipolar or Bipolar Software selectable (programmable with 16-bit DAC) | |||
| Excitation Level unipolar | 0…24 Volt; Predefined levels: 1, 2.5, 5, 10, 12, 15 and 24 VDC | |||
| Excitation Level bipolar | 2…30 Volt; Predefined levels: 2.5, 5, 10, 12, 15, 24 and 30 VDC | |||
| Accuracy | ±0.1 % ±5 mV | |||
| Drift | ±50 ppm/K ±100 μV/K | |||
| Stability | 10 % to 90 % load (bipolar) <0.01 % | |||
| Current limit | 200 mA (2 W max. per channel, 12 W max. per Slice) | |||
| Protection | Continuous short to ground | |||
| Bridge connection types | ||||
| Bridge type | Full bridge | |||
| Ranges @ 10 Vexc(low range) | 2 mV/V…1000 mV/V free programmable with Dual Core | |||
| Input short, Sensor offset adjust | Software selectable | |||
| Counters (only on LV+ type) | ||||
| Inputs | 1 digital counter input 3 digital inputs Fully synchronised with analog data Alarm output | |||
| Counter modes | Counting, waveform timing, encoder, tacho, gear-tooth sensor | |||
| General Counter Specifications | See "SIRIUS counter specifications" | |||
| Additional Specifications | ||||
| Misc function | Excitation control monitoring, Amplifier Short, Single Ended/Differential | |||
| Input connector | DSUB-9, BNC, Banana (others on request) | |||
| TEDS support | Standard + DSI® adapters | |||
| 1) In- must be within ±10 V referred to GND (iso); for Ranges >10 V the DC value of In- is not rejected | ||||
SIRIUS® DualCoreADC® MULTI amplifier specifications
| Inputs | ||||
|---|---|---|---|---|
| Input types | Voltage Full / Half / Quarter bridge strain (120 Ω and 350 Ω) | |||
| ADC Type | 24-bit delta-sigma dual core with anti-aliasing filter | |||
| Sampling Rate | Simultaneous 200 kS/s | |||
| Voltage Mode | ||||
| Dual Core Ranges (Low range) | ±10 V (±500 mV) | ±1 V (50 mV) | ±100 mV (±5 mV) | ±50 mV (±2.5 mV) |
| Gain accuracy | ±0.05 % of reading | |||
| Offset accuracy (Dual Core) | ±10 (5) mV | ±1 (0.5) mV | ±0.2 (0.2) mV | ±0.2 (0.2) mV |
| Offset accuracy after Balance Amplifier | 0.2 mV | 0.02 mV | 0.02 mV | 0.02 mV |
| Typ. Dynamic Range @ 10 kS/s (Dual Core) | -137 dB (-152 dB) | -137 dB (-147 dB) | -135 dB (-137 dB) | -133 dB (-133 dB) |
| Typ. Noise floor @ 10 kS/s (Dual Core) | -105 dB (-121 dB) | -104 dB (-111 dB) | -100 dB (-101 dB) | -95 dB (-95 dB) |
| Typ. CMR @ 400 Hz / 1 kHz | 86 / 84 dB | 96 / 95 dB | 112 / 102 dB | 112 / 102 dB |
| Gain Drift | Typical 10 ppm/K, max. 40 ppm/K | |||
| Offset Drift | Typical 0.3 μV/K + 5 ppm of range/K, max 2 μV/K + 10 ppm of range/K | |||
| Gain Linearity | <0.02 % | |||
| Inter Channel Phase-mismatch | 0.02° * fin [kHz] + 0.1° (@ 200 kS/s) | |||
| Channel Crosstalk | -120 dB @ 10 kHz | |||
| Input Coupling | DC | |||
| Input Impedance | 10 MΩ | |||
| Max. Common Mode Voltage | Isolated version: ±500 V Differential version: ±12 V | |||
| Overvoltage Protection | In+ to In-: 50 V continuous; 200 V peak (10 ms) | |||
| Excitation Voltage | ||||
| Excitation Voltage | Free programmable (16-bit DAC) | |||
| Predefined levels | 0, 1, 2.5, 5, 10 and 12 VDC | |||
| Accuracy | ±0.1 % ±10 mV | |||
| Drift | ±50 ppm/K ±100 μV/K | |||
| Load stability | 0 % to 100 % load <0.01 % | |||
| Current limit | 45 mA (max. 400 mW) | |||
| Protection | Continuous short to ground | |||
| Additional Fixed Excitations | DSUB-15: 12 V (max. 50 mA); 5 V (max. 100 mA) Accuracy 5 % 16-pin LEMO-2B: +/- 14.5 V (max. 50 mA); 5 V (max. 100 mA) Accuracy 5 % | |||
| Bridge Connection Types | ||||
| Bridge Types | Full / Half / Quarter bridge (3-wire) | |||
| Ranges | 2 mV/V…1000 mV/V free programmable with Dual Core | |||
| Internal Bridge Completion | ½ bridge 1 kΩ and ¼ bridge 120 Ω and 350 Ω | |||
| Typ. Bridge Completion Accuracy | 0.05 %; TCR: 5 ppm/K (others on request) | |||
| Internal Shunt Resistor | 59.88 kΩ, software selectable (others on request) | |||
| Typ. Shunt Resistor Accuracy | 0.05 %; TCR: 10 ppm/K (others on request) | |||
| Input Short, Sensor Offset Adjust | Software selectable | |||
| Counters | ||||
| Inputs | 1 digital counter input 3 digital inputs Fully synchronised with analog data | |||
| Counter Modes | counting, waveform timing, encoder, tacho, gear-tooth sensor | |||
| Input Level Compatibility | CMOS, LVTTL (protected up to ±25 Volt continuous) | |||
| Analog Out | ||||
| Overall information | 1 channel, 24-bit sigma delta 200 kHz, ±10 V | |||
| Accuracy | ±0.1 % of reading ±0.02 V | |||
| Analogue out bandwidth | 40 kHz | |||
| Temperature Drift | ±50 ppm/K of reading ± 200 µV/K | |||
| Output Impedance | <10 Ω | |||
| Maximum Output Current / Load | 20 mA / >1000 Ω | |||
| Output Protection | Continuous short to ground | |||
| Additional Specifications | ||||
| Input connector | DSUB-15, 16-pin LEMO-2B (others on request) | |||
| TEDS support | Standard + DSI® adapters | |||
SIRIUS® HD (High Density) amplifiers overview
| HD-STGS | HD-LV | HD-ACC | |
|---|---|---|---|
| Connectors | DSUB-9, L1B10f | DSUB-9, BNC | BNC |
| Counter version | ✗ | ✗ | ✗ |
| Isolated version | ✓ | ✓ | ✓ |
| Differential version | ✓ | ✓ | ✓ |
| EtherCAT version | ✓ | ✓ | ✓ |
| Rack version | ✓ | ✓ | ✓ |
| Fanless version | ✗ | ✗ | ✗ |
| Channels per slice | 16 | 16 | 16 |
| Data rate / channel | 200 kS/s USB, 10 kS/s EtherCAT | 200 kS/s USB, 10 kS/s EtherCAT | 200 kS/s USB, 10 kS/s EtherCAT |
| Resolution | 24-bit | 24-bit | 24-bit |
| Bandwidth | 70 kHz | 71 kHz | 72 kHz |
| Voltage ranges | ±10 V, ±1 V, ±100 mV, ±10 mV | ±100 V, ±10 V, ±1 V, ±100 mV | ±10 V, ±5 V, ±1 V, ±200 mV |
| Input coupling | DC | DC | DC, AC 0.1 Hz, 1 Hz |
| Sensor excitation | 0..12 V max. 44 mA | 2…30 V bipolar, 0…24 unipolar, max. 0.2 A / 2 W | IEPE 4,8,12 mA |
| Bridge connection | Full, ½, ¼ 350 Ω, ¼ 120 Ω 3-wire | Full | ✗ |
| Programmable shunt | 100 kΩ | ✗ | ✗ |
| IEPE input | DSI-ACC | DSI-ACC | ✓ |
| Resistance | ✗ | ✗ | ✗ |
| Temperature (PTx) | DSI-RTD | DSI-RTD | ✗ |
| Thermocouple | DSI-TH | DSI-TH | ✗ |
| Potentiometer | ✓ | ✗ | ✗ |
| LVDT | DSI-LVDT | DSI-LVDT | ✗ |
| Charge | DSI-CHG | DSI-CHG | ✗ |
| Current | ext. shunt, DSI | ext. shunt, DSI | ext. shunt |
| TEDS | ✓ | ✓ | ✓ |
| Isolation voltage | 500 V in pairs | 500 V in pairs | 500 V in pairs |
| Power consumption per channel | 1.3 W/pair | 1.3 W/pair | 1.3 W/pair |
| Advanced functions | Low power, sensor and amplifier balance | High sensor power and multi range | Sensor error detection |
SIRIUS® HD-STGS amplifier specifications
| Inputs | ||||
|---|---|---|---|---|
| Input types | Voltage Full/Half/Quarter bridge strain (120 Ω and 350 Ω) Potentiometer Current (ext. Shunt) | |||
| ADC Type | 24-bit delta-sigma with anti-aliasing filter | |||
| Sampling Rate Simultaneous | 200 kS/sec | |||
| Voltage Mode | ||||
| Ranges | ±10 V | ±1 V | ±100 mV | ±10 mV |
| Gain Accuracy | ±0.05 % of reading | |||
| Offset Accuracy | ±2 mV | ±0.2 mV | ±0.1 mV | ±0.1 mV |
| Offset Accuracy after Balance Amplifier | ±0.2 mV | ±0.02 mV | ±0.01 mV | ±0.01 mV |
| Dynamic Range @ 10 kS | -137 dB | -137 dB | -130 dB | -112 dB |
| Typ. Noise floor @ 10 kS | -105 dB | -104 dB | -95 dB | -75 dB |
| Typ. CMR @ 400 Hz / 1 kHz | 86 / 84 dB | 96 / 95 dB | 112 / 102 dB | 112 / 102 dB |
| Gain Drift | Typical 10 PPM/K, max. 40 PPM/K | |||
| Offset Drift | Typical 0.3 μV/K + 5 ppm of range/K, max 2 μV/K + 10 ppm of range/K | |||
| Gain Linearity | <0.02 % | |||
| Inter Channel Phase-mismatch | 0.02° * fin [kHz] + 0.1° (@ 200 kS/sec) | |||
| Channel Cross talk | -120 dB @ 10 kHz | |||
| Input Coupling | DC | |||
| Input Impedance | 10 MΩ | |||
| Max. common mode voltage | ±12 V | |||
| Overvoltage Protection | In + to In-: 50 V continuous; 200 V peak (10 msec) | |||
| Excitation Voltage | ||||
| Excitation Voltage | Free programmable (16-bit DAC) | |||
| Predefined levels | 0, 1, 2.5, 5, 10 and 12 VDC | |||
| Accuracy | ±0.05 % ±2 mV | |||
| Drift | ±50 ppm/K ±100 μV/K | |||
| Stability 10 % to 90 % load | <0.01 % | |||
| Current limit | 45 mA (350 mW max. power) | |||
| Protection | Continuous short to ground | |||
| Bridge Connection Types | ||||
| Bridge types | Full / Half / Quarter bridge (3-wire) | |||
| Ranges | 2 mV/V…1000 mV/V free programmable | |||
| Internal Bridge Completion | ½ bridge and ¼ bridge 120 Ω and 350 Ω | |||
| Bridge Completion Accuracy | 0.05 %; TCR: 5 ppm/K (others on request) | |||
| Internal Shunt Resistor | 100 kΩ (others on request) | |||
| Maximum sensor resistance | 5 kΩ for Full and Half bridge measurements | |||
| Typ. Shunt Resistor Accuracy | 0.05 %; TCR: 10 ppm/K (others on request) | |||
| Input Short, Sensor Offset Adjust | Software selectable | |||
| Additional Specifications | ||||
| Input connector | DSUB-9 | |||
| TEDS support | Standard + DSI® adapters | |||
| Rev: 1593385200 | ||||
SIRIUS® HD-ACC amplifier specifications
| Inputs | ||||
|---|---|---|---|---|
| Input types | Voltage IEPE | |||
| ADC Type | Type 24-bit delta-sigma with anti-aliasing filter | |||
| Sampling Rate | Simultaneous 200 kS/sec | |||
| Voltage Mode | ||||
| Range | ±10 V | ±5 V | ±1 V | ±200 mV |
| Gain accuracy | ±0.05 % of reading | |||
| Offset accuracy | ±2 mV | ±1 mV | ±0.2 mV | ±0.1 mV |
| Typ. Dynamic Range @ 10 kS | -137 dB | -137 dB | -137 dB | -131 dB |
| Typ. Noise floor @ 10 kS | -109 dB | -109 dB | -108 dB | -102 dB |
| Typ. CMR @ 400 Hz / 1 kHz | 140 dB / 120 dB | |||
| Gain Drift | Typical 10 ppm/K, max. 30 ppm/K | |||
| Offset Drift | Typical 0.3 μV/K + 5 ppm of range/K, max 2 μV/K + 10 ppm of range/K | |||
| Gain Linearity | <0.02 % | |||
| Inter Channel Phase-mismatch | 0.02° * fin [kHz] + 0.1° (@ 200 kS/sec and 10 V Range) | |||
| Channel Cross talk | -150 dB @ 50 Hz; -140 dB @ 1 kHz | |||
| Input Coupling | DC, AC (0.1 Hz, 1 Hz) | |||
| Input Impedance | 3.4 MΩ (340 kΩ for AC coupling ≥1 Hz) in parallel with 270 pF | |||
| Over-voltage Protection | In+ to In-: 50 V continuous; 200 V peak (10 msec) In- to GND (differential version): 3 Volt | |||
| IEPE mode | ||||
| Excitation | 4, 8 or 12 mA | |||
| Compliance voltage | 22 Volt | |||
| Output Impedance | >100 kΩ | |||
| Sensor detection | Shortcut: <4 Volt; Open: >19 Volt | |||
| Input Coupling | AC (0.1 Hz, 1 Hz) | |||
| Additional Specifications | ||||
| Input connector | BNC | |||
| TEDS support | IEPE mode only | |||
| Rev: 1767110400 | ||||
SIRIUS® HD-LV amplifier specifications
| Inputs | ||||
|---|---|---|---|---|
| Input types | Voltage Full bridge strain Current (ext. Shunt) | |||
| ADC Type | Type 24-bit delta-sigma with anti-aliasing filter | |||
| Sampling Rate | Simultaneous 200 kS/sec | |||
| Voltage Mode | ||||
| Ranges | ±100 V | ±10 V | ±1 V | ±100 mV |
| Gain accuracy | ±0.05 % of reading | |||
| Offset accuracy | ± 20 mV | ± 2 mV | ± 0.2 mV | ± 0.1 mV |
| Dynamic Range @ 10 kS | -134 dB | -137 dB | -137 dB | -125 dB |
| Typ. Noise floor @ 10 kS | -104 dB | -104 dB | -104 dB | -95 dB |
| Typ. CMR @ 400 Hz / 1 kHz | 74 / 70 dB | 86 / 84 dB | 96 / 95 dB | 112 / 102 dB |
| Gain Drift | Typical 10 PPM/K, max. 40 PPM/K | |||
| Offset Drift | Typical 0.3 μV/K + 5 ppm of range/K, max 2 μV/K + 10 ppm of range/K | |||
| Gain Linearity | <0.02 % | |||
| Inter Channel Phase-mismatch | 0.02° * fin [kHz] + 0.1° (@ 200 kS/sec and 10 V Range) | |||
| Channel Cross talk | -120 dB @ 10 kHz (Range ≤10 V); -76 dB @ 1 kHz (Range = 100 V) | |||
| Input Coupling | DC | |||
| Input Impedance | 1 MΩ for 100 V Range; all other Ranges 10 MΩ | |||
| Max. common mode voltage | 100 V Range: ±100 V; all other Ranges: ±12 V | |||
| Overvoltage Protection | In+ to In-: 100 V Range: 200 V; all other Ranges: 50 V (200 V for 10 msec) | |||
| Excitation Voltage | ||||
| Excitation Voltage | Unipolar or Bipolar Software selectable Free programmable (16-bit DAC) | |||
| Excitation Level unipolar | 0…24 Volt; Predefined levels: 1, 2.5, 5, 10, 12, 15, 20 and 24 VDC | |||
| Excitation Level bipolar | 2…30 Volt; Predefined levels: 2.5, 5, 10, 12, 15, 24 and 30 VDC | |||
| Accuracy | ±0.1 % ±5 mV | |||
| Drift | ±50 ppm/K ±100 μV/K | |||
| Stability 10 % to 90 % load | <0.01 % | |||
| Current limit | 100 mA (1 Watt max. per Channel, 12 Watt max. per Slice) | |||
| Protection | Continuous short to ground | |||
| Bridge Connection Types | Full bridge | |||
| Ranges @ 10 VExc | 2 mV/V…1000 mV/V free programmable | |||
| Sensor Offset Adjust | Software selectable | |||
| Additional Specifications | ||||
| Input connector | DSUB-9, BNC (others on request) | |||
| TEDS support | Standard + DSI® adapters | |||
| Rev: 1593385200 | ||||
SIRIUS® HS (High Speed) amplifiers overview
| HS-HV | HS-LV | HS-ACC | HS-CHG | HS-STG | |
|---|---|---|---|---|---|
| Connectors | BANANA | DSUB-9, BNC, BANANA | BNC | BNC, TNC | DSUB-9 |
| Counter version | ✗ | ✓ | ✓ | ✓ | ✓ |
| Isolated version | ✓ | ✓ | ✓ | ✓ | ✓ |
| Differential version | ✗ | ✓ | ✓ | ✓ | ✓ |
| EtherCAT version | ✓ | ✓ | ✓ | ✗ | ✓ |
| Rack version | ✓ | ✓ | ✓ | ✓ | ✓ |
| Fanless version | ✓ | ✓ (see 1) | ✗ | ✗ | ✗ |
| Channels per slice | 8 | 8 | 8 | 8 | 8 |
| Data rate / channel | 1 MS/s | 1 MS/s | 1 MS/s | 1 MS/s | 1 MS/s |
| Resolution | 16-bit | 16-bit | 16-bit | 16-bit | 16-bit |
| Bandwidth | 2 MHz | 1 MHz | 500 kHz | 500 kHz | 1 MHz |
| Voltage ranges | ±2000 V … ±20 V | ±100 V … ±50 mV | ±10 V … ±200 mV | ±10 V … ±100 mV | ±50 V … ±20 mV |
| Input coupling | DC | DC, AC 1 Hz (3, 10 Hz SW) | DC, AC 1 Hz (3, 10 Hz SW) | DC, AC 0.1 Hz, 1 Hz, 10 Hz, 100 Hz | DC, AC 1 Hz (3, 10 Hz SW) |
| Sensor excitation | ✗ | 2…30 V bipolar, 0…24 V unipolar, max. 0.2 A / 2 W | IEPE 4 mA, 8 mA | IEPE 4, 8, 12 mA | 0…20 V max. 0.8 W, 0..60 mA max 0.5 W |
| Bridge connection | ✗ | Full | ✗ | ✗ | Full, ½, ¼ 350 Ω, ¼ 120 Ω, 3 or 4 wire |
| Programmable shunt | ✗ | ✗ | ✗ | ✗ | 59.88 kΩ, 175 kΩ bipolar |
| IEPE input | ✗ | DSI-ACC | ✓ | ✓ | DSI-ACC |
| Resistance | ✗ | ✗ | ✗ | ✗ | ✓ |
| Temperature (PTx) | ✗ | DSI-RTD | ✗ | ✗ | ✓ |
| Thermocouple | ✗ | DSI-TH | ✗ | ✗ | DSI-TH |
| Potentiometer | ✗ | ✗ | ✗ | ✗ | ✓ |
| LVDT | ✗ | DSI-LVDT | ✗ | ✗ | DSI-LVDT |
| Charge | ✗ | DSI-CHG | ✗ | 1 k …100 k pC | DSI-CHG |
| Current | ✗ | ext. shunt, DSI | ext. shunt | ext. shunt | ext. shunt, DSI |
| TEDS | ✗ | ✓ | ✓ | ✓ | ✓ |
| Isolation voltage | CAT II 1000 V | 1000 V | 1000 V | 1000 V | 1000 V |
| Power consumption per channel | 1 W/ch | 1.2 W/ch | 2 W/ch | 1.2 W/ch | 2 W/ch |
| Advanced functions | High voltage, high bandwidth, high isolation | High sensor power and multi range | Sensor error detection, high speed | Sensor error detection in IEPE and CHG mode (injection) | High speed, supports all strain types and high input range |
| (1) Fanless operation only for BNC or Banana version (without excitation) | |||||
SIRIUS® HS-ACC amplifier specifications
| Inputs | ||||
|---|---|---|---|---|
| Input types | Voltage IEPE | |||
| ADC Type | 16-bit SAR | |||
| Sampling Rate | Simultaneous 1 MS/s | |||
| Filtering | See "HS general specifications" | |||
| Analog bandwidth | 500 kHz | |||
| Voltage Mode | ||||
| Ranges | ±10 Volt | ±5 Volt | ±1 Volt | ±0.2 Volt |
| Input Accuracy | ±0.05 % of reading | |||
| Offset Accuracy | ±2 mV | ±1 mV | ±0.2 mV | ±0.1 mV |
| Typ. Noise floor @ 100 kS/s | -89 dB | -89 dB | -86 dB | -83 dB |
| Typ. CMR @ 50 Hz / 400 Hz / 1 kHz | 120 / 96 / 88 dB | 126 / 100 / 92 dB | 140 / 110 / 102 dB | 140 / 118 / 110 dB |
| Gain Drift | Typical 10 ppm/K, max. 30 ppm/K | |||
| Offset Drift | Typical 0.5 μV/K + 10 ppm of range/K, max 3 μV/K + 20 ppm of range/K | |||
| Gain Linearity | <0.02 % | |||
| Inter Channel Phase-mismatch | 0.02° * fin [kHz] + 0.1° | |||
| Channel Cross talk | -83 dB | |||
| Input Coupling | DC or AC (1 Hz, 3 Hz or 10 Hz) | |||
| Input Impedance | 1 MΩ in parallel with 100 pF | |||
| Overvoltage Protection | 50 V continuous; 200 V peak (10 msec) | |||
| IEPE mode | ||||
| Excitation | 4 or 8 mA | |||
| Compliance voltage | 25 Volt | |||
| Output Impedance | >100 kΩ | |||
| Sensor detection | Shortcut: <4 Volt; Open: >19 Volt | |||
| Input Coupling | AC (1 Hz, 3 Hz or 10 Hz) | |||
| Counters (HS-ACC+ type only) | ||||
| Inputs | 1 digital counter input 3 digital inputs Fully synchronised with analog data | |||
| Counter Modes | counting, waveform timing, encoder, tacho, gear-tooth sensor | |||
| General Counter Specifications | See "SIRIUS counter specification" | |||
| Additional Specifications | ||||
| Input connector | BNC | |||
| TEDS support | IEPE mode only | |||
| Rev: 1767110400 | ||||
SIRIUS® HS-CHG amplifier specifications
| Inputs | |||||
|---|---|---|---|---|---|
| Input types | Voltage IEPE Charge | ||||
| ADC Type | 16-bit SAR | ||||
| Sampling Rate | Simultaneous 1 MS/s | ||||
| Filtering | See "HS general specifications" | ||||
| Analog bandwidth (see 1) | 500 kHz | ||||
| Voltage Mode | |||||
| Ranges | ±10, ±5, ±2, ±1, ±0.5, ±0.2, ±0.1 Volt | ||||
| Input Accuracy | Signal frequency | Accuracy | |||
| DC | ±0.05 % of reading ±0.02 % of range ±50 μV | ||||
| up to 10kHz | ±0.1 % of reading ±0.05 % of range | ||||
| up to 100kHz | ±2 % of reading ±0.1 % of range | ||||
| Typ. Noise floor | Sample rate/Range → | 10 V | 5 V | 1 V | 0.1 V |
| 1 MS/s | -88 dB | -87 dB | -85 dB | -68 dB | |
| 200 kS/s | -89 dB | -89 dB | -88 dB | -83 dB | |
| 20 kS/s | <-100 dB | <-100 dB | <-100 dB | -92 dB | |
| Typical CMR | 50 Hz / 1 kHz | >140 / 120 dB | >140 / 124 dB | >140 / 126 dB | >140 / 126 dB |
| Gain Drift | Typical 10 ppm/K, max. 30 ppm/K | ||||
| Offset Drift | Typical 0.5 μV/K + 10 ppm of range/K, max 3 μV/K + 20 ppm of range/K | ||||
| Gain Linearity | <0.02 % | ||||
| Inter Channel Phase-mismatch | 0.02° * fin [kHz] + 0.1° | ||||
| Channel Cross talk | -110 dB @ 1 kHz; -90 dB @ 10 kHz | ||||
| Input Coupling | DC or AC (0.1 Hz, 1 Hz, 10 Hz or 100 Hz) | ||||
| Input Impedance | 3.3 MΩ in parallel with 100 pF | ||||
| Overvoltage Protection | 50 V continuous; 200 V peak (10 msec) | ||||
| IEPE mode | |||||
| Excitation | 4 mA, 8 mA or 12 mA | ||||
| Compliance voltage | 25 Volt | ||||
| Output Impedance | >100 kΩ | ||||
| Sensor detection | Shortcut: <4 Volt; Open: >19 Volt | ||||
| Input Coupling | AC (0.1 Hz, 1 Hz, 10 Hz or 100 Hz) | ||||
| Charge mode | |||||
| Charge ranges | ±100 000, 50 000, 20 000, 10 000, 5000, 2000, 1000 pC | ||||
| Input Accuracy | ±0.5 % of reading ±0.05 % of range ±2 pC | ||||
| Offset Accuracy (See 2) | ±2 pC | ||||
| Input Coupling | 0.01 Hz, 0.03 Hz, 0.1 Hz, 0.5 Hz, 1 Hz, 10 Hz or 100 Hz | ||||
| Counters (only on HS-CHG+ type) | |||||
| Inputs | 1 digital counter input 3 digital inputs Fully synchronised with analog data | ||||
| Counter Modes | counting, waveform timing, encoder, tacho, gear-tooth sensor | ||||
| General Counter Specifications | See "SIRIUS counter specification" | ||||
| Additional Specifications | |||||
| Input connector | BNC (others on request) | ||||
| TEDS support | IEPE mode only | ||||
| 1) 200 kHz for Charge ranges 2) Offset Accuracy taken into account with Input Coupling <= 0.1 Hz, for others it is negligible. | |||||
| Rev: 1767110400 | |||||
SIRIUS® HS-HV (v2) amplifier specifications
| Inputs | ||||
|---|---|---|---|---|
| Input types | Voltage | |||
| ADC Type | 16-bit SAR | |||
| Sampling Rate | Simultaneous 1 MS/s | |||
| Filtering | See "HS general specifications" | |||
| Analog bandwidth | 2 MHz | |||
| Voltage Mode | ||||
| Ranges | ±1600 V, ±800 V, ±400 V, ±200 V, ±100 V, ±50 V, ±20 V | |||
| Input Accuracy (see 1) | Signal frequency | Accuracy | ||
| DC | ±0.03 % of reading ±0.02 % of range ±0.04 V | |||
| Up to 1 kHz | ±0.03 % of reading ±0.02 % of range | |||
| Up to 10 kHz | ±0.1 % of reading ±0.05 % of range | |||
| Up to 100 kHz | ±2 % of reading ±0.1 % of range | |||
| Up to 1000 kHz | ±5 % of reading ±0.5 % of range | |||
| Typ. Noise floor | Sample rate/Range → | 1600 V | 400 V | 100 V |
| 1 MS/s | -83 dB | -79 dB | -71 dB | |
| 200 kS/s | -85 dB | -85 dB | -82 dB | |
| 20 kS/s | -100 dB | -97 dB | -90 dB | |
| Typ. CMR @ 50 Hz / 1 kHz / 10 kHz | 85 dB / 75 dB / 50 dB | |||
| Gain Drift | Typical 25 ppm/K, max. 50 ppm/K | |||
| Offset Drift | Typical 1 mV/K + 10 ppm of range/K, max 2 mV/K + 15 ppm of range/K | |||
| Gain Linearity | <0.02 % | |||
| Channel Cross talk | -115 dB @ 50 Hz; -90 dB @ 1 kHz | |||
| Input Coupling | DC | |||
| Input Impedance | 10 MΩ | 2 pF | |||
| Overvoltage category | CAT II 1000 V | |||
| Maximum input voltage | 2000 Vdc or 2000 Vpk (1414 Vacrms) | |||
| Additional Specifications | ||||
| Input connector | Banana | |||
| TEDS support | NA | |||
| 1) 50 V Range (1.2 MHz) and 20 V range (700 kHz) have limited bandwidth | ||||
| Rev: 1671188100 | ||||
SIRIUS® HS-HV (v3) amplifier specifications
| Inputs | ||||
|---|---|---|---|---|
| Input types | Voltage | |||
| ADC Type | 16-bit SAR | |||
| Sampling Rate | Simultaneous 1 MS/s | |||
| Filtering | See "HS general specifications" | |||
| Analog bandwidth | 2 MHz (1.5 MHz for 50 V range, 800 kHz for 20 V range) | |||
| Voltage Mode | ||||
| Ranges | ±2000 V, ±1000 V, ±400 V, ±200 V, ±100 V, ±50 V, ±20 V | |||
| Input Accuracy | Signal frequency | Accuracy | ||
| DC | ±0.03 % of reading ±0.02 % of range ±0.04 V | |||
| Up to 1 kHz | ±0.03 % of reading ±0.02 % of range | |||
| Up to 10 kHz | ±0.1 % of reading ±0.05 % of range | |||
| Up to 100 kHz | ±2 % of reading ±0.1 % of range | |||
| Up to 1000 kHz | ±5 % of reading ±0.5 % of range | |||
| Typ. Noise floor | Sample rate/Range → | 2000 V | 400 V | 100 V |
| 1 MS/s | -84 dB | -80 dB | -71 dB | |
| 200 kS/s | -86 dB | -85 dB | -83 dB | |
| 20 kS/s | -100 dB | -96 dB | -91 dB | |
| Typ. CMR @ 50 Hz / 1 kHz / 10 kHz | 95 dB / 75 dB / 55 dB | |||
| Gain Drift | Typical 25 ppm/K, max. 50 ppm/K | |||
| Offset Drift | Typical 1 mV/K + 10 ppm of range/K, max 2 mV/K + 15 ppm of range/K | |||
| Gain Linearity | <0.02 % | |||
| Channel Cross talk | -115 dB @ 50 Hz; -90 dB @ 1 kHz | |||
| Input Coupling | DC | |||
| Input Impedance | 10 MΩ | 2 pF | |||
| Overvoltage category | CAT II 1000 V | |||
| Maximum input voltage | 2000 Vdc or 2000 Vpk (1414 Vacrms) | |||
| Additional Specifications | ||||
| Input connector | Banana | |||
| TEDS support | NA | |||
| Rev: 1698322800 | ||||
SIRIUS® HS-LV amplifier specifications
| Inputs | |||||
|---|---|---|---|---|---|
| Input types | Voltage Full bridge strain Current (ext. Shunt) | ||||
| ADC Type | 16-bit SAR | ||||
| Sampling Rate | Simultaneous 1 MS/s | ||||
| Filtering | See "HS general specifications" | ||||
| Analog bandwidth | 1 MHz (800 kHz for ranges ≤0.2 V) | ||||
| Voltage Mode | |||||
| Ranges | ±100, 50, 20, 10, 5, 2, 1, 0.5, 0.2, 0.1 and 0.05 Volt | ||||
| Input Accuracy | Signal frequency | Accuracy | |||
| DC | ±0.03 % of reading ±0.02 % of range ±100 μV (2 mV for Ranges ≥10 Volt) | ||||
| Up to 1 kHz | ±0.03 % of reading ±0.02 % of range | ||||
| Up to 10 kHz | ±0.1 % of reading ±0.05 % of range | ||||
| Up to 100 kHz | ±2 % of reading ±0.1 % of range | ||||
| Up to 1000 kHz | ±5 % of reading ±0.5 % of range | ||||
| Offset accuracy after Balance Amplifier | Range <10 V: ±0.002 % of range ±3 μV; Range ≥10 V: ±0.005 % of range | ||||
| Typ. Noise floor | Sample rate/Range → | 100 V | 5 V | 0.5 V | 0.05 V |
| 1 MS/s | -85 dB | -86 dB | -78 dB | -59 dB | |
| 200 kS/s | -88 dB | -88 dB | -87 dB | -76 dB | |
| 20 kS/s | <-100 dB | <-100 dB | -95 dB | -86 dB | |
| Typical CMR | Bandwidth/Range → | 100 V | 5 V | 0.5 V | 0.05 V |
| 50 Hz | 70 dB | 88 dB | 102 dB | 102 dB | |
| 400 Hz | 70 dB | 86 dB | 100 dB | 100 dB | |
| 10 kHz | 55 dB | 70 dB | 80 dB | 80 dB | |
| Gain Drift | Typical 10 ppm/K, max. 30 ppm/K | ||||
| Offset Drift | Typical 0.3 μV/K + 5 ppm of range/K, max 2 μV/K + 10 ppm of range/K | ||||
| Gain Linearity | <0.02 % | ||||
| Inter channel phase mismatch | 0.02° * fin [kHz] + 0.1° (5 V Range) | ||||
| Channel Cross talk Range | <10 V: -120 dB @ 10 kHz; Range ≥10 V: -95 dB @ 10 kHz | ||||
| Input Coupling (see 1) | DC, AC 1 Hz (3 Hz, 10 Hz per SW) | ||||
| Input Impedance | Range <10 V: 10 MΩ; Range ≥10 V: 1 MΩ | 110 pF between INx to GND | ||||
| Max. Common Mode Voltage | Isolated version: ±500 V | ||||
| Differential version: Ranges ≥10 Volt: ±100 V; all other Ranges: ±12 V | |||||
| Overvoltage Protection | <10 V: 100 V (200 V peak for 10 msec); Range ≥10 V: 300 V cont. | ||||
| Excitation Voltage | |||||
| Excitation Voltage | Unipolar or Bipolar Software selectable Free programmable (16-bit DAC) | ||||
| Excitation Level | unipolar 0…24 Volt; Predefined levels: 1, 2.5, 5, 10, 12, 15, 20 and 24 VDC | ||||
| Excitation Level | bipolar 2…30 Volt; Predefined levels: 2.5, 5, 10, 12, 15, 24 and 30 VDC | ||||
| Accuracy | ±0.1 % ±5 mV | ||||
| Drift | ±50 ppm/K ±100 μV/K | ||||
| Stability 10 % to 90 % load (bipolar) | <0.01 % | ||||
| Current limit | 200 mA (2 Watt max. per Channel, 12 Watt max. per Slice) | ||||
| Protection | Continuous short to ground | ||||
| Input short, Sensor offset adjust | Software selectable | ||||
| Counters (only on HS-LV+ type) | |||||
| Inputs | 1 digital counter input 3 digital inputs Fully synchronised with analog data | ||||
| Counter modes | counting, waveform timing, encoder, tacho, gear tooth sensor | ||||
| General Counter Specifications | See "SIRIUS counter specification" | ||||
| Additional Specifications | |||||
| Input connector | DSUB-9, Banana, BNC (others on request) | ||||
| TEDS support | Standard + DSI® adapters | ||||
| Rev: 1593504000 | |||||
| 1) In- must be within ±10 V referred to GND (iso); for Ranges ≥10 V the DC value of In- is not rejected | |||||
SIRIUS® HS-STG amplifier specifications
| Inputs | |||||
|---|---|---|---|---|---|
| Input types | Voltage Full/Half/Quarter bridge strain (120 Ω and 350 Ω) Potentiometer, RTD, Resistance Current (ext. Shunt) | ||||
| ADC Type | 16-bit SAR | ||||
| Sampling Rate | Simultaneous 1 MS/s | ||||
| Filtering | See "HS general specifications" | ||||
| Analog bandwidth | 1 MHz (800 kHz for ranges ≤0.2 V) | ||||
| Voltage Mode | |||||
| Ranges | ±50, 20, 10, 5, 2, 1, 0.4, 0.2, 0.1, 0.04 and 0.02 Volt | ||||
| Input Accuracy | Signal frequency | Accuracy | |||
| DC | Range <10 V: ±0.03 % of reading ±0.04 % of range ±10 μV | ||||
| Range ≥10 V: ±0.05 % of reading ±0.01 % of range ±5 mV | |||||
| Up to 1 kHz | ±0.03 % of reading ±0.02 % of range | ||||
| Up to 10 kHz | ±0.1 % of reading ±0.05 % of range | ||||
| Up to 100 kHz | ±2 % of reading ±0.1 % of range | ||||
| Up to 500 kHz | ±10 % of reading ±0.5 % of range | ||||
| Offset accuracy after Balance Amplifier | Range <10 V: ±0.002 % of range ±3 μV; Range ≥10 V: Not performed | ||||
| Typ. Noise floor | Sample rate/Range → | 50 V | 5 V | 0.2 V | 0.02 V |
| 1 MS/s | -87 dB | -88 dB | -81 dB | -63 dB | |
| 200 kS/s | -88 dB | -88 dB | -85 dB | -74 dB | |
| 20 kS/s | <-100 dB | <-100 dB | -94 dB | -84 dB | |
| Typical CMR | Bandwidth/Range → | 50 V | 5 V | 0.2 V | 0.02 V |
| 50 Hz | 70 dB | 88 dB | 102 dB | 102 dB | |
| 400 Hz | 70 dB | 86 dB | 100 dB | 100 dB | |
| 10 kHz | 55 dB | 70 dB | 80 dB | 80 dB | |
| Gain Drift | Typical 10 ppm/K, max. 30 ppm/K | ||||
| Offset Drift | typical 0.3 μV/K + 15 ppm of range/K, max 2 μV/K + 40 ppm of range/K | ||||
| Gain Linearity | <0.02 % | ||||
| Inter channel phase mismatch | 0.02° * fin [kHz] + 0.1° (5 V Range) | ||||
| Channel Cross talk | Range <10 V: -120 dB @ 10 kHz; Range ≥10 V: -95 dB @ 10 kHz | ||||
| Input Coupling (see 1) | DC, AC 1 Hz (3 Hz, 10 Hz per SW) | ||||
| Input Impedance | Range <10 V: 10 MΩ; Range ≥10 V: 1 MΩ between INx to GND | ||||
| Max. common mode voltage | Isolated version: ±500 V | ||||
| Differential version: Ranges ≥10 Volt: ±60 V; all other Ranges: ±12 V | |||||
| Overvoltage Protection | Range <10 V: 100 V (200 V peak for 10 msec); Range ≥10 V: 300 V cont. | ||||
| Excitation Voltage | |||||
| Excitation Voltage | Free programmable (16-bit DAC) | ||||
| Predefined levels | 0, 1, 2.5, 5, 10, 15 and 20 VDC | ||||
| Accuracy | ±0.05 % ±2 mV | ||||
| Drift | ±10 ppm/K ±100 μV/K | ||||
| Load stability 0 % to 100 % load | <0.01 % | ||||
| Line regulation over 20 Ω of change | <0.005 % @ 120 Ω load | ||||
| Noise @ 10 Volt / 350 Ω | < 200 μVrms @ 10 kHz BW | ||||
| Sense Impedance to Exc / to GND | 100 kΩ / 100 MΩ | ||||
| Current limit | 100 mA (max. 800 mW) | ||||
| Protection | Continuous short to ground | ||||
| Input short, Sensor offset adjust | Software selectable | ||||
| Excitation Current | |||||
| Excitation Current | Free programmable (16-bit DAC) | ||||
| Predefined levels | 0.1, 1, 2, 5, 10, 20 and 60 mADC | ||||
| Accuracy (>10 mA) | 0.1 % ±2 μA [0.5 % ±50 μA] | ||||
| Drift (>10 mA) | 15 ppm/K [100 ppm/K] | ||||
| Compliance voltage | 20 Volt, max. 300 mW | ||||
| Output Impedance | >1 MΩ | ||||
| Bridge Connection Types | |||||
| Bridge types | Full / Half / Quarter bridge (3-wire or 4-wire) | ||||
| Ranges | 2 mV/V…1000 mV/V free programmable | ||||
| Internal Bridge Completion | ½ bridge and ¼ bridge 120 Ω and 350 Ω | ||||
| Bridge Completion Accuracy | 0.05 %; TCR: 5 ppm/K (others on request) | ||||
| Internal Shunt Resistor | 59.88 kΩ and 175 kΩ, bipolar (to +Exc and -Exc) | ||||
| Maximum sensor resistance | 5 kΩ for Full and Half bridge measurements | ||||
| Typ. Shunt Resistor Accuracy | 0.05 %; TCR: 10 ppm/K (others on request) | ||||
| Sensor Balance Range | 220 % of Range (70 % for input Range ≥5 Volt) | ||||
| Input Short, Sensor Offset Adjust | Software selectable | ||||
| Counters (only on HS STG+) | |||||
| Inputs | 1 digital counter input 3 digital inputs Fully synchronised with analog data | ||||
| Counter modes | counting, waveform timing, encoder, tacho, gear-tooth sensor | ||||
| General Counter Specifications | See "SIRIUS counter specification" | ||||
| Additional Specifications | |||||
| Input connector | DSUB-9 (others on request) | ||||
| TEDS support | Standard + DSI® adapters | ||||
| 1) In- must be within ±10 V referred to GND (iso); for Ranges ≥10 V the DC value of In- is not rejected | |||||
| Rev: 1593504000 | |||||