Tekniska specifikationer för SIRIUS® X-systemet
| SIRIUS X | |
|---|---|
| GLAN | |
| Data interface | Ethernet 1 GbE incl. IEEE1588v2 synchronization (PTP IEEE 802.3) |
| Sync Accuracy | 100 ns |
| Bus connectors | Ethernet RJ45, single cable for data, power and sync, daisy chainable |
| Power over Ethernet | 48-57 V DC (max. 90W) |
| Max. cable length between devices | 100 m |
| CAN bus | |
| Number of ports | 2 |
| Interface type | CAN 2.0B, up to 1 MBit/s CAN-FD, up to 8 Mbit/s |
| Special applications | CCP, OBDII, J1939, CAN RT output |
| Galvanic isolation | Isolated |
| Bus pin fault protection | ±36 V |
| ESD protection | 8 kV |
| IRIG Synchronisation | |
| Sync Connectors | 4-pin LEMO 00 Series female |
| Sync Accuracy | 100 ns |
| Max. Sync-cable length | 200 m (IRIG) |
| Level (Input/Output) | TTL compatible |
| Max. Output Current | ±24 mA (±50 mA for 1 sec) |
| Other interfaces | |
| USB | USB-C 3.0 Type (See 1) |
| EtherCAT | Secondary bus/RT out connection to 3rd party EtherCAT master (See 2) |
| Internals | |
| CPU type | ARM |
| Operating system | Linux |
| Software | Dewesoft RT |
| Device protocols | openDAQ, OPC-UA |
| Power | |
| Power Supply | 9 - 57 V DC |
| Power consumption | *Configuration dependent |
| Power supply connector | 2-pin LEMO 1B Series male |
| Power Out | Input supply on 2-pin LEMO 1B Series female |
| Environmental | |
| Operating Temperature | -10 to 50 °C (optional -40 to 85 °C) |
| Storage Temperature | -40 to 85 °C |
| Humidity | 5 to 95 % RH non condensing @ 60 °C |
| IP rating | IP40 |
| Shock & Vibration | Vibration sweep sinus (EN 60068-2-6:2008) Vibration random (EN 60721-3-2: 1997 - Class 2M2) Shock (EN 60068-2-27:2009) MIL-STD-810D |
| Physical | |
| Cooling | Fan-based, fanless ("f" designation) (See 3) |
| Dimensions | 266 mm x 149 mm x 55 mm |
| Rev: 1712304000 | |
| 1) Currently, this functionality is available with firmware 2025.2 or newer. 2) EtherCAT functionality is currently not supported. 3) Fanless options of actively cooled units must not be stacked. | |
Översikt över SIRIUS® X-förstärkare
| 8xUNI | hi 8xUNI | 16xUNI | 16xLVe | 32xLV | 16xACC | 16xSTGS | 8xCAN-FD | |
|---|---|---|---|---|---|---|---|---|
| Connectors | DSUB-9 | Redel 2P 8-pin | DSUB-9 | D-SUB9 | Terminal block | BNC | D-SUB9 | DSUB-9 |
| Max. Channels per slice | 8x | 8x | 16x | 16x | 32x | 16x | 16x | 4x |
| Data rate per channel | 500 kS/s | 500 kS/s | 200 kS/s | 200 kS/s | 100 kS/s | 200 kS/s | 50 kS/s | up to XY Mbit/s |
| Resolution | 24-bit | 24-bit | 24-bit | 24-bit | 24-bit | 24-bit | 24-bit | Digital |
| Bandwidth | 0.437*fs | 0.437*fs | 0.433*fs | 0.433*fs | 0.433*fs | 0.433*fs | 0.433*fs | ✗ |
| Voltage mode ranges | ±100 V, ±10 V, ±1 V, ±100 mV | ±100 V, ±10 V, ±1 V, ±100 mV | ±100 V, ±10 V, ±1 V, ±100 mV | ±100 V, ±5 V | ±200 V, ±10 V | ±10 V, ±5 V, ±1 V, ±200 mV | ✗ | ✗ |
| Input coupling | DC, AC 0.15 Hz | DC, AC 0.15 Hz | DC, AC 0.15 Hz | DC | DC | DC, AC 1 Hz, AC 0.1 Hz | DC | ✗ |
| Sensor excitation | 0..24 V (unipolar) 0..12 V (bipolar) 0..100 mA (current) Max. 0.8 W/ch 2 mA, 4 mA, 6 mA (IEPE) | 0..24 V (unipolar) 0..12 V (bipolar) 0..100 mA (current) Max. 0.8 W/ch 2 mA, 4 mA, 6 mA (IEPE) | 0..24 V (unipolar) 0..12 V (bipolar) 0..100 mA (current) Max. 0.8 W/ch 2 mA ... 20 mA (IEPE) | 0...24V (unipolar) And *D-SUB9: Fixed 10V DC bipolar excitation | ✗ | 2 mA, 4 mA, 6 mA | 1 V, 2 V, 5 V | 5 V (max. 240 mA), 12 V (max. 100 mA) |
| Bridge connection | Full, ½, ¼ 350 Ω or ¼ 120 Ω (3-wire or 4-wire), ¼ CC (2-wire or 4-wire) | Full, ½, ¼ 350 Ω or ¼ 120 Ω (3-wire or 4-wire), ¼ CC (2-wire or 4-wire) | Full, ½, ¼ 350 Ω or ¼ 120 Ω (3-wire or 4-wire), ¼ CC (2-wire or 4-wire) | ✗ | ✗ | ✗ | Full, ½, ¼ 350 Ω, ¼ 120 Ω 3-wire, 4-wire (optional) | ✗ |
| Programmable shunt | 100 kΩ | 100 kΩ | 100 kΩ | ✗ | ✗ | ✗ | 100 kΩ | ✗ |
| IEPE input | ✓, DSI-ACC | ✓, DSI-ACC | ✓, DSI-ACC | DSI-ACC | ✗ | ✓ | ✗ | ✗ |
| Resistance | ✓ | ✓ | ✓ | ✗ | ✗ | ✗ | ✗ | ✗ |
| Temperature (PTx) | ✓ | ✓ | ✓ | DSI-RTD | ✗ | ✗ | ✗ | ✗ |
| Thermocouple | DSI-TH | DSI-TH | DSI-TH | DSI-TH | ✗ | ✗ | ✗ | ✗ |
| Current | 100 mA (int. shunt), ext. shunt, DSI-5A, DSIi-10A, DSIi-20A | 100 mA (int. shunt), ext. shunt, DSI-5A, DSIi-10A, DSIi-20A | 100 mA (int. shunt), ext. shunt, DSI-5A, DSIi-10A, DSIi-20A | ✓ (Ext. shunt) | ✗ | ✗ | ✗ | ✗ |
| Potentiometer | ✓ | ✓ | ✓ | ✓ | ✗ | ✗ | ✓ | ✗ |
| LVDT | DSI-LVDT | DSI-LVDT | DSI-LVDT | DSI-LVDT | ✗ | ✗ | ✗ | ✗ |
| Charge | DSI-CHG | DSI-CHG | DSI-CHG | DSI-CHG | ✗ | ✗ | ✗ | ✗ |
| TEDS | ✓ | ✓ | ✓ | ✓ | ✗ | ✓ | ✓ | ✗ |
| Isolation voltage | CATII 150V (secondary circuits) | CATII 1000V | 250V | 250 V | 250 V | 250V | Differential | CATII 150V (secondary circuits) |
| Isolation arrangement | Ch-Ch & Ch-GND | Ch-Ch & Ch-GND | Ch-GND | Ch-GND | Ch-GND | Ch-GND | Ch-GND | Ch-Ch & Ch-GND |
| PoE functionality | ✓ | ✓ | ✓ | ✓ | ✓ | ✓ | ✓ | ✓ |
| Advanced functions | Fully isolated, high input range, connect almost any sensor, supports TEDS | High isolation, high input range, connect almost any sensor, supports TEDS | High input range, connect almost any sensor, supports TEDS, high channel density | High input range, powerful sensor supply | High input range, high-channel density | IEPE, supports TEDS, high channel density | Supports all strain types, low power consumption | CAN-FD, CAN 2.0, fully isolated, sensor supply |
| Rev: 1792050960 | ||||||||
SIRIUS-X-16xUNI tekniska specifikationer
| Analog inputs | SIRIUS X 16xUNI | ||||
|---|---|---|---|---|---|
| Input type | Voltage Full/Half/Quarter bridge strain (120 Ω and 350 Ω) IEPE Potentiometer, RTD, Resistance Current (ext. and int. Shunt) | ||||
| Number of channels | 16 | ||||
| ADC Type | 24-bit delta-sigma with anti-aliasing filter (-3 dB @ 160 kHz) Butterworth 2nd order | ||||
| Sampling Rate | Simultaneous 200 kS/sec per channel (software-selectable) | ||||
| Voltage mode | |||||
| Voltage Ranges | ±100 V | ±10 V | ±1 V | ±0.1 V | |
| Input Accuracy | DC | ±0.03 % of reading ±0.02 % of range | |||
| Noise floor, Typ. | 10 kS/s / 100 kS/s / 200 kS/s | 115 dB / 105 dB / 104 dB | 116 dB / 106 dB / 105 dB | 114 dB / 105 dB / 103 dB | 100 dB / 90 dB / 86 dB |
| Dynamic Range, Typ. | 10 kS/s / 100 kS/s / 200 kS/s | -152 dB / -152 dB / -152 dB | -150 dB / -153 dB / -153 dB | -150 dB / -151 dB / -152 dB | -137 dB / -137 dB / -137 dB |
| CMR, Typ. 0 dBFS sine wave @ 50 Hz / 400 Hz / 1 kHz | 200 kS/s | 85 dB / 69 dB /60 dB | 85 dB / 85 dB / 85 dB | 105 dB / 105 dB / 102 dB | 113 dB / 110 dB / 104 dB |
| THD, Typ. -1 dBFS sine wave @ 1 kHz | 200 kS/s | -91 dB | -101 dB | -100 dB | -107 dB |
| Gain Linearity | < 0.02 % | ||||
| Gain Drift | Typ. 10 ppm/K, max. 40 ppm/K | ||||
| Offset Drift | Typ. 0.3 μV/K + 2 ppm of range/K, max. 0.8 μV/K + 10 ppm of range/K | ||||
| Channel Cross talk typ. @ 1 kHz / 10 kHz | -100 dB / -86 dB | -135 dB / -120 dB | -135 dB / -120 dB | -119 dB / -110 dB | |
| Input Coupling | DC, AC 0.15 Hz | ||||
| Input Impedance | 2 MΩ between IN+ and IN- | >1 GΩ between IN+ and IN- | |||
| Max. Common Mode Voltage | ±120 V | ±12 V | |||
| Overvoltage Protection | ±200 V continuously | ±100 V continuously / 200 Vpeak for 10 ms | |||
| Analog input performance | |||||
| Sample rate | SR ≤ 2 kS/s | SR > 2 kS/s | |||
| Bandwidth (-3 dB) | 0.489*fs | 0.433*fs | |||
| Alias-free Bandwidth | DC to 0.44*fs | DC to 0.499*fs | |||
| Alias Rejection | -105 dB (all sample rates) | ||||
| Delay Through ADC | 34 / fs | ||||
| Oversampling | 32 | ||||
| Current mode (internal shunt) | |||||
| Ranges | 20 mA | 2 mA | |||
| Input Accuracy | ±0.03 % of reading, ±0.02 % of range, ±2.1 μA | ||||
| Internal Shunt Resistor | 47 Ω | ||||
| Protection | Typ. trip current: Imax ≥ 31 mA; auto retry on 800 ms (clamp voltage 33 V) | ||||
| Excitation Voltage | |||||
| Excitation Voltage | Free programmable (16-bit DAC) | ||||
| Predefined levels | 24 V DC | ||||
| Accuracy | ±0.05 % ±2 mV | ||||
| Drift | ±10 ppm/K ±100 μV/K | ||||
| Load stability | 0 % to 100 % load <0.01 % | ||||
| Noise @ 10 V / 350 Ω | <150 μVrms @ 10 kS/s | ||||
| Line regulation over 20 Ω of change | <0.005 % @ 120 Ω load | ||||
| Sense Impedance to Exc / to GND | 100 kΩ / >100 MΩ | ||||
| Current limit | 100 mA (max. 800 mW) | ||||
| Protection | Continuous short to ground | ||||
| Excitation Current | |||||
| Excitation Current | Free programmable (16-bit DAC) | ||||
| Predefined levels | 0.1, 1, 2, 5, 10, 20, 60 and 100 mA DC | ||||
| Ranges | 0.1 - 1 mA | >1 - 20 mA | >20 - 100 mA | ||
| Accuracy | ±0.2 % ±0.5 μA | ±0.15 % ±2 μA | ±0.01 % ±40 μA | ||
| Drift Typical | 5 ppm/K | 10 ppm/K | 25 ppm/K | ||
| Compliance voltage | 24 V, max. 600 mW | ||||
| Output Impedance | >100 MΩ | ||||
| Bridge Connection Types | |||||
| Bridge types | Full / Half / Quarter bridge (3-wire or 4-wire) / Quarter Constant Current (2-wire or 4-wire) | ||||
| Ranges | 2 - 1000 mV/V free programmable | ||||
| Internal Bridge Completion | Half and Quarter bridge 120 Ω and 350 Ω | ||||
| Typ. Bridge Completion Accuracy | 0.05 %; TCR: 2 ppm/K (others on request) | ||||
| Internal Shunt Resistor | 100 kΩ, software selectable (others on request) | ||||
| Typ. Shunt Resistor Accuracy | 0.05 %; TCR: 10 ppm/K (others on request) | ||||
| Input Short, Sensor Offset Adjust | Software selectable | ||||
| IEPE mode | |||||
| IEPE Current | Free programmable (16-bit DAC) | ||||
| Predefined levels | 2, 4, 8, 12, 16 or 20 mA | ||||
| Compliance voltage | 25 V | ||||
| Output Impedance | >100 kΩ | ||||
| Sensor detection | Shortcut: <4 V; Open: >19 V | ||||
| Additional Specifications | |||||
| Modular | SIRIUS X 16xUNI | ||||
| Input connector | D-SUB9 | ||||
| Isolation voltage | 250 Vpeak channel to ground (functional isolation) | ||||
| TEDs support | Standard + DSI® adapters | ||||
| Power supply | 9 - 57 V DC, Power over Ethernet: 48 - 57 V DC (PoE++ power delivery) | ||||
| Power consumption | Lemo @Typ. 24 V / MAX. 56 V | Voltage mode: Typ. 15.6 W / Max. 19 W Voltage mode with load: Typ. 45 W / Max. 49 W (Load: 0.8W per channel) | |||
| PoE @Min. 48 V / MAX. 56 V | Voltage mode: Typ. 18.5 W / Max. 19.5 W Voltage mode with load: Typ. 48.2 W / Max. 49.5 W (Load: 0.8W per channel) | ||||
| Weight | 2140 g | ||||
| Slice Dimensions | 266 mm x 149 mm x 55 mm | ||||
| Rack | SIRIUSr X 16xUNI | ||||
| Input connector | D-SUB9 | ||||
| Isolation voltage | 250 Vpeak channel to ground (functional isolation) | ||||
| TEDs support | Standard + DSI® adapters | ||||
| Power supply | 12 - 48 V DC | ||||
| Power consumption | Lemo @Typ. 24 V / Max. 48 V | Voltage mode: Typ. 14.3 W / Max.18 W Voltage mode with load: Typ. 46.5 W / Max. 50 W (Load: 0.8W per channel) | |||
| Weight | 1240 g | ||||
| Slice Dimensions | 266 mm x 143 mm x 45 mm | ||||
| Rev: 1775131200 | |||||
SIRIUSi-X-8xUNI tekniska specifikationer
| Analog inputs | SIRIUSi X 8xUNI | ||||
|---|---|---|---|---|---|
| Input type | Voltage Full / Half / Quarter bridge strain (120 Ω and 350 Ω) Potentiometer, RTD, Resistance Current (int. and ext. Shunt) IEPE | ||||
| Number of channels | 8 | ||||
| ADC Type | 24-bit, delta-sigma with anti-aliasing filter (-3 dB @ 400 kHz) | ||||
| Sampling Rate | Simultaneous 500 kS/sec per channel (software-selectable) | ||||
| Voltage mode | |||||
| Voltage Ranges | ±100 V | ±10 V | ±1 V | ±0.1 V | |
| Input Accuracy | DC | ±0.03 % of reading ±0.02 % of range | |||
| Noise floor, Typ. | 100 S/s / 1 kS/s / 10 kS/s / 100 kS/s / 500 kS/s | -135 dB / -127 dB / -117 dB / -108 dB / -99 dB | -137 dB / -128 dB / -118 dB / -109 dB / -103 dB | -134 dB / -125 dB / -116 dB / -106 dB / -99 dB | -116 dB / -108 dB / -99 dB / -90 dB / -81 dB |
| Dynamic Range, Typ. | 500 kS/s | -151 dB | -152 dB | -148 dB | -131 dB |
| CMR, Typ. -1 dBFS sine wave @ 50 Hz / 400 Hz / 1 kHz | 10 kS/s | 81 dB / 63 dB / 55 dB | 85 dB / 85 dB / 85 dB | 105 dB / 105 dB /102 dB | 99 dB / 99 dB / 98 dB |
| Gain Linearity | ±0.01 % | ||||
| Gain Drift | Typical 10 ppm/K Max. 40 ppm/K | ||||
| Offset Drift | Typical 0.3 μV/K + 2 ppm of range/K Max 0.8 μV/K + 10 ppm of range/K | ||||
| Channel Cross talk typ. @ 1 kHz | -104 dB | -153 dB | -150 dB | -131 dB | |
| Input Coupling | DC, AC 0.15 Hz | ||||
| Input Impedance | 2 MΩ between IN+ and IN- | >1 GΩ between IN+ and IN- | |||
| Max. Common Mode Voltage | ±120 V | ±12 V | |||
| Overvoltage Protection | ±200 V, continuous | ±100 V continuous / ±200 V peak (10 ms) | |||
| Current mode (internal shunt) | |||||
| Ranges | 20 mA | 2 mA | |||
| Input Accuracy | ±0.03 % of reading, ±0.02 % of range, ±2.1 μA | ||||
| Internal Shunt Resistor | 47 Ω | ||||
| Excitation Voltage | |||||
| Excitation Voltage | Free programmable (16-bit DAC) | ||||
| Predefined levels | 1, 2.5, 5, 10, 12, 15, 20, 24 V | ||||
| Accuracy | ±0.05 % ±2 mV | ||||
| Drift | ±10 ppm/K ±100 μV/K | ||||
| Load stability | < 0.01 % | ||||
| Noise @ 10 V / 350 Ω | <150 μVrms @ 10 kS/s | ||||
| Line regulation over 20 Ω of change | <0.005 % @ 120 Ω load | ||||
| Sense Impedance to Exc / to GND | 100 kΩ / >100 MΩ | ||||
| Current limit | 100 mA (max. 800 mW) | ||||
| Protection | Continuous short to ground | ||||
| Excitation Current | |||||
| Excitation Current | Free programmable (16-bit DAC) | ||||
| Predefined levels | 0.1, 1, 2, 5, 10, 20, 60 mA | ||||
| Ranges | 0.1 - 1 mA | >1 - 20 mA | >20 - 100 mA | ||
| Accuracy | ±0.2 % ±0.5 μA | ±0.15 % ±2 μA | ±0.1 % ±10 μA | ||
| Drift Typical | ±5 ppm/K ± 5 nA/K | ±10 ppm/K ± 10 nA/K | ±40ppm/K ± 200 nA/K | ||
| Compliance voltage | 24 V, max. 800 mW | ||||
| Output Impedance | >100 MΩ | ||||
| Bridge Connection Types | |||||
| Bridge types | Full / Half / Quarter bridge (3-wire or 4-wire) / Quarter Constant Current (2-wire or 4-wire) | ||||
| Ranges | 2 - 1000 mV/V free programmable | ||||
| Internal Bridge Completion | ½ bridge and ¼ bridge 120 Ω and 350 Ω | ||||
| Typ. Bridge Completion Accuracy | 0.05 %; TCR: 2 ppm/K | ||||
| Internal Shunt Resistor | 100 kΩ, software selectable | ||||
| Typ. Shunt Resistor Accuracy | 0.05 %; TCR: 10 ppm/K | ||||
| Input Short, Sensor Offset Adjust | Software selectable | ||||
| IEPE mode | |||||
| IEPE Current | Free programmable (16-bit DAC) | ||||
| Predefined levels | 2, 4, 8, 12, 16 and 20 mA | ||||
| Compliance voltage | 24V | ||||
| Output Impedance | >100 kΩ | ||||
| Sensor detection | Shortcut: <4 V; Open: >19 V | ||||
| Additional Specifications | |||||
| Modular | SIRIUSi X 8xUNI | ||||
| Input connector | D-SUB9 | ||||
| Isolation voltage | CH - GND and CH - CH isolated Secondary circuits CAT II 150V | ||||
| TEDs support | Standard + DSI® adapters | ||||
| Power supply | 9 - 57 V DC Power over Ethernet: 48 - 57 V DC (PoE++ power delivery) | ||||
| Power consumption | Lemo @Typ. 24 V / MAX. 56 V | Voltage mode: Typ. 12.1 W / Max. 14.4 W Voltage mode with load: Typ. 27.4 W / Max. 30 W (Load: 0.8W per channel) | |||
| PoE @Min. 48 V / MAX. 56 V | Voltage mode: Typ. 14.7 W / Max. 15.3 W Voltage mode with load: Typ. 34.6 W / Max. 35.3 W (Load: 0.8W per channel) | ||||
| Weight | 1965 g | ||||
| Slice Dimensions | 266 mm x 149 mm x 55 mm | ||||
| Rack | SIRIUSir X 8xUNI | ||||
| Input connector | D-SUB9 | ||||
| Isolation voltage | CH - GND and CH - CH isolated Secondary circuits CAT II 150V | ||||
| TEDs support | Standard + DSI® adapters | ||||
| Power supply | 12 - 48 V DC | ||||
| Power consumption | Lemo @Typ. 24 V / Max. 48 V | Voltage mode: Typ. 10.4 W / Max.11.5 W Voltage mode with load: Typ. 25.9 W / Max. 27.6 W (Load: 0.8W per channel) | |||
| Weight | 1100 g | ||||
| Slice Dimensions | 266 mm x 143 mm x 45 mm | ||||
| Rev: 1775131200 | |||||
SIRIUSf-X-16xACC tekniska specifikationer
| Analog inputs | SIRIUS X 16xACC | ||||
|---|---|---|---|---|---|
| Input type | Voltage, IEPE | ||||
| Number of channels | 16 | ||||
| ADC Type | 24-bit delta-sigma with anti-aliasing filter (-3 dB @ 160 kHz) Butterworth 2nd order | ||||
| Sampling Rate | Simultaneous 200 kS/sec per channel (software-selectable) | ||||
| Voltage mode | |||||
| Voltage Ranges | ±10 V | ±5 V | ±1 V | ±0.2 V | |
| Input Accuracy | DC | ±0.02 % of reading ±0.02 % of range | |||
| Noise floor, Typ. | 10 kS/s / 100 kS/s / 200 kS/s | -117 dB / -107 dB / -106 dB | -117 dB / -107 dB / -106 dB | -115 dB / -105 dB / -103 dB | -105 dB / -96 dB / -93 dB |
| Dynamic Range, Typ. | 10 kS/s / 100 kS/s / 200 kS/s | -160 dB / -154 dB / -154 dB | -160 dB / -155 dB / -155 dB | -157 dB / -152 dB / -150 dB | -150 dB / -142 dB / -140 dB |
| CMR, Typ. 0 dBFS sine wave @ 50 Hz / 400 Hz / 1 kHz | 10 kS/s | 120 dB / 117 dB / 112 dB | 119 dB / 117 dB / 111 dB | 118 dB / 116 dB / 111 dB | 109 dB / 108 dB / 107 dB |
| 100 kS/s | 110 dB / 110 dB / 108 dB | 110 dB / 110 dB / 108 dB | 108 dB / 108 dB / 107 dB | 99 dB / 99 dB / 98 dB | |
| 200 kS/s | 109 dB / 109 dB / 107 dB | 109 dB / 108 dB / 107 dB | 106 dB / 106 dB / 105 dB | 96 dB / 96 dB / 96 dB | |
| THD, Typ. -1 dBFS sine wave @ 1 kHz | 10 kS/s / 100 kS/s / 200 kS/s | -105 dB / -101 dB / -98 dB | -104 dB / -102 dB / -98 dB | -104 dB / -102 dB / -98 dB | -102 dB / -102 dB / -97 dB |
| SFDR, Typ. -1 dBFS sine wave @ 1 kHz | 10 kS/s / 100 kS/s / 200 kS/s | 106 dBc / 104 dBc / 100 dBc | 106 dBc / 104 dBc / 100 dBc | 100 dBc / 100 dBc / 98 dBc | 97 dBc / 95 dBc / 93 dBc |
| SNR, Typ. -1 dBFS sine wave @ 1 kHz | 10 kS/s / 100 kS/s / 200 kS/s | 108 dB / 102 dB / 102 dB | 104 dB / 102 dB / 99 dB | 95 dB / 92 dB / 90 dB | 90 dB / 86 dB / 84 dB |
| Gain Linearity | < 0.02 % | ||||
| Gain Drift | Typical 6 ppm/K, max. 15 ppm/K | ||||
| Offset Drift | Typical 0.5 μV/K + 0.5 ppm of range/K, max 2 μV/K + 2 ppm of range/K | ||||
| Channel Cross talk typ. @ 1 kHz | -130 dB | -130 dB | -130 dB | -125 dB | |
| Input Coupling | DC, AC 0.1 Hz, 1 Hz | ||||
| Input Impedance | 1 MΩ | ||||
| Overvoltage Protection | In+ to In-: 50 V continuous, 200 V peak (10 msec) | ||||
| Analog input performance | |||||
| Sample rate | SR ≤ 2 kS/s | SR > 2 kS/s | |||
| Bandwidth (-3 dB) | 0.489*fs | 0.433*fs | |||
| Alias-free Bandwidth | DC to 0.44*fs | DC to 0.499*fs | |||
| Alias Rejection | -105 dB (all sample rates) | ||||
| Delay Through ADC | 34 / fs | ||||
| Oversampling | 128 | ||||
| IEPE input | |||||
| Excitation | 2 mA, 4 mA, 6 mA | ||||
| Excitation accuracy | 10% | ||||
| Compliance voltage | 24 V | ||||
| Output Impedance | >60 kΩ | ||||
| Sensor detection | Shortcut: <4 V; Open: >19 V | ||||
| Additional Specifications | |||||
| Modular | SIRIUS X 16xACC | ||||
| Input connector | BNC | ||||
| Isolation voltage | 250 Vpeak channel to ground | ||||
| TEDs support | IEPE mode only (class 1) | ||||
| Power supply | 9 - 57 V DC, Power over Ethernet: 48 - 57 V DC (PoE++ power delivery) | ||||
| Power consumption | Lemo @Typ. 24 V / MAX. 56 V | Voltage mode: Typ. 9.1 W / Max. 10.3 W IEPE mode: Typ. 11.7 W @4 mA, 13.2 W @6 mA / Max. 12.9 W @4 mA, 14.3 W @6 mA | |||
| PoE @Min. 48 V / MAX. 56 V | Voltage mode: min. 10.2 W / Max. 10.5 W IEPE mode: min. 12.7 W @4 mA, 14.2 W @6 mA / Max. 13.2 W @4 mA, 14.6 W @6 mA | ||||
| Weight | 1700 g | ||||
| Slice Dimensions | 266 mm x 149 mm x 55 mm | ||||
| Rack | SIRIUSr X 16xACC | ||||
| Input connector | BNC | ||||
| Isolation voltage | 250 Vpeak channel to ground | ||||
| TEDs support | IEPE mode only (class 1) | ||||
| Power supply | 12 - 48 V DC | ||||
| Power consumption | Lemo @Typ. 24 V / Max. 48 V | Voltage mode: Typ. 8 W / Max. 10.3 W IEPE mode: Typ. 11.4 W @4 mA, 12.8 W @6 mA / Max. 13.2 W @4 mA, 14.6 W @6 mA | |||
| Weight | 975 g | ||||
| Slice Dimensions | 266 mm x 143 mm x 45 mm | ||||
| Rev: 1712304000 | |||||
SIRIUSf-X-16xLVe tekniska specifikationer
| Inputs | SIRIUS X 16xLVe | ||
|---|---|---|---|
| Input types | Voltage, Potentiometer (See 1), Current (Ext. shunt) (See 1) | ||
| Number of channels | 16 | ||
| ADC Type | 24-bit delta-sigma with anti-aliasing filter (-3 dB @ 160 kHz) Butterworth 2nd order | ||
| Sampling Rate | Simultaneous up to 200 kS/s | ||
| Voltage Mode | |||
| Ranges | SR | ±100 V | ±5 V |
| Input Accuracy | DC | ±0.03 % of reading ±0.02 % of range | |
| Noise floor, Typ. | 10kS/s/ 100kS/s / 200kS/s | -118 dB / -109 dB / -107 dB | -103 dB / -93 dB / -90 dB |
| Dynamic Range, Typ. | 10kS/s/ 100kS/s / 200kS/s | -141 dB / -144 dB / -141 dB | -132 dB / -131 dB / -127 dB |
| CMR, Typ. 0 dBFS sine wave @ 50 Hz / 400 Hz / 1 kHz | 10kS/s | -115 dB / -103 dB/ -99 dB | -103 dB / -102 dB / -100 dB |
| 100kS/s | -108 dB / -101 dB / -98 dB | -93 dB / -93 dB / -93 dB | |
| 200kS/s | -106 dB / -101 dB / -98 dB | -90 dB / -90 dB / -90 dB | |
| THD, Typ. -1 dBFS sine wave @ 1 kHz | 10kS/s/ 100kS/s / 200kS/s | -95 dB / -95 dB / -96 dB | -110 dB / -106 dB / -99 dB |
| SFDR, Typ. -1 dBFS sine wave @ 1 kHz | 10kS/s/ 100kS/s / 200kS/s | -104 dBc / -98 dBc/ -98 dBc | -109 dBc / -107 dBc/ -102 dBc |
| SNR, Typ. -1 dBFS sine wave @ 1 kHz | 10kS/s/ 100kS/s / 200kS/s | 107 dB / 102 dB / 100 dB | 98 dB / 88 dB / 85 dB |
| Gain Linearity | <0.02 % | <0.01 % | |
| Gain Drift | Typical 5 ppm/K, Max. 20 ppm/K | ||
| Offset Drift | Typical 5 ppm of range/K, Max. 10 ppm of range/K | ||
| Channel CrosstalkChannel Cross talk typ. @ 1 kHz | -131 | ||
| Input Coupling | DC | ||
| Input Impedance | 800 kΩ (between In+ and In-) | ||
| Input Overvoltage Protection | In+ to In-: 200 V continuous, 350 Vpeak (100 ms) | ||
| Max. common mode voltage | ±200 V | ||
| Isolation | 250 V functional isolation (channel to power supply ground) | ||
| Analog input performance | |||
| Sampling rate | SR ≤ 2 kS/s | SR > 2 kS/s | |
| Bandwidth (-3 dB) | 0.489*fs | 0.433*fs | |
| Alias-free Bandwidth | DC to 0.44*fs | DC to 0.499*fs | |
| Alias Rejection | -105 dB (all sample rates) | ||
| Delay Through ADC | 34 / fs | ||
| Oversampling | 32 | ||
| Excitation Voltage | |||
| Excitation voltage | Unipolar 2 ... 24V DC D-SUB9: Fixed bipolar excitation 10V DC | ||
| Current limit | 2.8 W max. power per channel , 22.4 W per slice (MAX: 500mA ) | ||
| Excitation accuracy | 0.05 % ± 2 mV (bipolar 1%) | ||
| Drift | ±20 ppm/K ± 60 μV/K | ||
| Stability 10 % to 90 % load | < 0.06 @ 5 V exc (< 0.01 @ 24 V exc) | ||
| Noise @ 10 V / 350 Ω | <150 μVrms @ 10 kS | ||
| Load regulation over 10 Ω of change | < 0.04 % @ 120 Ω load | ||
| Protection | Continuous short to ground | ||
| Additional Specifications | |||
| Modular | SIRIUS X 16xLVe | ||
| Input connector | Terminal block | DSUB-9 | |
| TEDS support | Yes (See 1) | ||
| Power supply | 9 - 57 V DC, Power over Ethernet: 48 - 57 V DC (PoE++ power delivery) | ||
| Power consumption | Lemo @Typ. 24 V / MAX. 56 V | Analog input only: Typ. 8.7 W / Max. 10.9 W Excitation: Typ 15.5 W 120 Ω load @ 5 V, 74 W 120 Ω load @ 20 V Max: 18.9 W 120 Ω load @ 5 V, 87 W 120 Ω load @ 20 V | |
| PoE @Min. 48 V / MAX. 56 V | Analog input only: min: 10.4 W / MAX. 11.1 W Excitation: Typ. 19 W 120 Ω load @ 5 V, 85.2 W 120 Ω load @ 20 V Max. 19.4 W 120 Ω load @ 5 V, 88.2 W 120 Ω load @ 20 V | ||
| Weight | 1820 g | ||
| Slice Dimensions | 266 mm x 149 mm x 55 mm | ||
| Rack | SIRIUSr X 16xLVe | ||
| Input connector | DSUB-9 | ||
| TEDS support | Yes (See 1) | ||
| Power supply | 12 - 48 V DC | ||
| Power consumption | Lemo @Typ. 24 V / Max. 48 V | Analog input only: Typ. 6.5 W / Max. 8.5 W (100kS/s / 200kS/s) Excitation: Typ. 16 W 120 Ω load @ 5 V, 75 W 120 Ω load @ 20 V Max. 18 W 120 Ω load @ 5 V, 85 W 120 Ω load @ 20 V | |
| Weight | 1130 g | ||
| Slice Dimensions | 266 mm x 143 mm x 45 mm | ||
| Rev: 1712304000 | |||
SIRIUSf-X-32xLV tekniska specifikationer
| Analog input | SIRIUS X 32xLV | ||
|---|---|---|---|
| Input type | Voltage | ||
| Number of channels | 32 | ||
| ADC Type | 24-bit delta-sigma with anti-aliasing filter (-3 dB @ 160 kHz) Butterworth 2nd order | ||
| Sampling Rates | Simultaneous up to 100 kS/s | ||
| Voltage mode | |||
| Voltage ranges | ±200 V | ±10 V | |
| Input Accuracy | DC | ±0.03 % of reading ±40 mV | ±0.03 % of reading ±2 mV |
| Noise floor, Typ. | 10 kS/s / 100 kS/s | -116 dB / -106 dB | -105 dB / -96 dB |
| Dynamic Range | 10 kS/s / 100 kS/s | -158 dB / -152 dB | -147 dB / -140 dB |
| CMR, Typ. -1 dBFS on 10V and -6dB on 200V sine wave @ 1 kHz @ 50 Hz / 400 Hz / 1 kHz | 10 kS/s | -94 dB / -94 dB / -93 dB | -98 dB / -98 dB / -97 dB |
| 100 kS/s | -94 dB / -93 dB / -93 dB | -94 dB / -94 dB / -93 dB | |
| THD, Typ. -1 dBFS on 10V and -6dB on 200V sine wave @ 1 kHz | 10 kS/s / 100 kS/s | -92 dB / -92 dB | -107 dB / -103 dB |
| SFDR, Typ. -1 dBFS on 10V and -6dB on 200V sine wave @ 1 kHz | 100 kS/s | 94 dBc / 94 dBc | 109 dBc / 106 dBc |
| SNR Typ. -1 dBFS on 10V and -6dB on 200V sine wave @ 1 kHz | 10 kS/s / 100 kS/s | 103 dB / 95 dB | 103 dB / 95 dB |
| Gain Linearity | < 0.01 % | ||
| Gain Drift | Typical 10 ppm/K, Max. 40 ppm/K | ||
| Offset Drift | Typical 2 ppm of range/K, Max. 5 ppm of range/K | ||
| Crosstalk (0 dBFS @ 1 kHz) | -115 dB | -130 dB | |
| Input Coupling | DC | ||
| Input Impedance | 800 kΩ | ||
| Max. common mode voltage | ±200 V | ||
| Overvoltage Protection | In+ to In-: 200 V continuous, 350 Vpeak (100 ms) | ||
| Analog input performance | |||
| Bandwidth (-3 dB) | 0.433*fs (0.49*fs @ < 5 kS/s) | ||
| Alias-free Bandwidth | DC to 0.499*fs (DC to 0.44*fs @ < 5 kS/s) | ||
| Alias Rejection | -105 dB (all sample rates) | ||
| Delay Through ADC | 34 / fs | ||
| Oversampling | 32 | ||
| Additional Specifications | |||
| Modular | SIRIUS X 32xLV | ||
| Input Connectors | Terminal block 9 pole OMNIMATE SL 2.50 / BLF 2.50/180 | ||
| Power supply | 9 - 57 V DC, Power over Ethernet: 48 - 57 V DC (PoE++ power delivery) | ||
| Isolation Voltage | 250 Vpeak channel to ground | ||
| Power consumption | Lemo @Typ. 24 V / MAX. 56 V | Voltage mode: Typ. 8.5 W / Max. 10.4 W | |
| PoE @Min. 48 V / MAX. 56 V | Voltage mode: Typ. 10.1 W / Max. 10.7 W | ||
| Weight | 1700 g | ||
| Slice Dimensions | 266 mm x 149 mm x 55 mm | ||
| Rack | SIRIUSr X 32xLV | ||
| Input Connectors | Terminal block 9 pole OMNIMATE SL 2.50 / BLF 2.50/180 | ||
| Power supply | 12 - 48 V DC | ||
| Isolation Voltage | 250 Vpeak channel to ground | ||
| Power consumption | Lemo @Typ. 24 V / Max. 48 V | Voltage mode: Typ. 7.2W / Max. 8.7 W | |
| Weight | 720 g | ||
| Slice Dimensions | 266 mm x 143 mm x 45 mm | ||
| Rev: 1712304000 | |||
SIRIUSf-X-16xSTGS tekniska specifikationer
| Inputs | SIRIUS X 16xSTGS | ||||
|---|---|---|---|---|---|
| Input type | Full / half / quarter bridge (3-wire, 4-wire), Potentiometer | ||||
| Max. number of channels per slice | 16 | ||||
| ADC Type | 24-bit delta-sigma with anti-aliasing filter (-3 dB @ 160 kHz) Butterworth 2nd order | ||||
| Sampling Rate | Simultaneous 50 kS/s per channel (software-selectable) | ||||
| Voltage inputs | |||||
| Ranges | 1 V | 0.1 V | 0.02 V | ||
| Input accuracy | 0.02 % of reading | 0.02 % of reading | 0.02 % of reading | ||
| Offset accuracy | 210 uV | 30 uV | 14 uV | ||
| Offset Accuracy after Balance Amplifier | 21 uV | 3 uV | 1.4 uV | ||
| Noise floor, Typ. | 10 kS/s / 20 kS/s / 50 kS/s | -111 dB / -109 dB / -105 dB | -100 dB / -97 dB / -93 dB | -87 dB / -85 dB / -81 dB | |
| Dynamic Range, Typ. | 10 kS/s / 20 kS/s / 50 kS/s | -144 dB / -142 dB / -141 dB | -135 dB / -134 dB / -134 dB | -122 dB / -122 dB / -122 dB | |
| CMR, Typ. 0 dBFS sine wave @ 50 Hz / 400 Hz / 1 kHz | 10 kS/s | -77 dB / -77 dB / -76 dB | -96 dB / -89 dB / -82 dB | -108 dB / -90 dB / -83 dB | |
| 50 kS/s | -77 dB / -77 dB / -76 dB | -96dB / -89 dB / -82 dB | -107 dB / -90 dB / -83 dB | ||
| Gain Linearity | < 0.01 % | ||||
| Gain Drift | Typical 10 ppm/K (max. 20 ppm/K) | ||||
| Offset Drift | Typical 0.03 μV/K + 2.5 ppm of range/K, max 0.1 μV/K + 9 ppm of range/K | ||||
| Typ. Crosstalk @ 1 kHz | -135 dBc | -125 dBc | -115 dBc | ||
| Input Coupling | DC | ||||
| Overvoltage Protection | In+ to In-: 6 V continuous | ||||
| Analog input performance | |||||
| Sample rate | SR ≤ 2 kS/s | SR > 2 kS/s | |||
| Bandwidth (-3 dB) | 0.49*fs | 0.433*fs | |||
| Alias-free Bandwidth | DC to 0.44*fs | DC to 0.499*fs | |||
| Alias Rejection | -105 dB (all sample rates) | ||||
| Delay Through ADC | 34 / fs | ||||
| Oversampling | 32 | ||||
| Excitation Voltage | |||||
| Excitation Voltage | Unipolar: 1 V, 2 V, 5 V (see 2) | ||||
| Excitation Accuracy | ±3 % of set ±50 mV | ||||
| Sense Accuracy | ±0.02 % of reading ±0.25 mV | ||||
| Sense Drift | Typical 10 μV/K + 10 ppm of reading/K, max 20 μV/K + 20 ppm of reading/K | ||||
| Current Limit | 42 mA | ||||
| Protection | Continuous short to ground | ||||
| Bridge measurement | |||||
| Bridge Connection Types | full bridge strain, ½ bridge strain, ¼ bridge strain (3-wire) | ||||
| Ranges | 2 mV/V ... 1000 mV/V free programmable (software ranges) | ||||
| Accuracy | Range \ Excitation | 1 V | 2 V | 5 V | |
| 200 mV/V | ±0.035 % of reading ±210 μV/V | ±0.028 % of reading ±105 μV/V | ±0.025 % of reading ±42 μV/V | ||
| 20 mV/V | ±0.027 % of reading ±14 μV/V | ±0.027 % of reading ±15 μV/V | ±0.024 % of reading ±6 μV/V | ||
| 2 mV/V | ±0.027 % of reading ±14 μV/V | ±0.024 % of reading ±7 μV/V | ±0.022 % of reading ±2.8 μV/V | ||
| Internal Bridge Completion | ¼ bridge 120 Ω and 350 Ω | ||||
| Bridge Completion Accuracy | ±125 µV/V, TCR: 5 ppm/K (others on request) | ||||
| Internal Shunt Resistor | 100 kΩ software selectable to SNS+ or SNS- and In+ or In- | ||||
| Shunt Resistor Accuracy | 0.05 %; TCR: 10 ppm/K (others on request) | ||||
| Input Short, Sensor Offset Adjust | Software-selectable | ||||
| Load monitoring | |||||
| Bridge configuration | Full / Half | QB | |||
| Current measuring line | EXC+ line | R Line | |||
| Current limit | 50 mAdc | ||||
| Resolution | 0.1 mAdc | ||||
| Additional math calculus | Resistance, Power | ||||
| Sample rate | 1 Hz | ||||
| Additional Specifications | |||||
| Modular | SIRIUS X 16xSTGS | ||||
| Input connectors | D-SUB9 | Terminal Block Weidmueller SL 2.50-04 | |||
| TEDS support | Standard + DSI® adapters | Not available | |||
| Power supply | 9 - 57 V DC, Power over Ethernet: 48 - 57 V DC (PoE++ power delivery) | ||||
| Power consumption | Lemo @Typ. 24 V / MAX. 56 V | Typ. 12.2 W 120 Ω load @ 5 V, Max. 15 W 120 Ω load @ 5 V | |||
| PoE @Min. 48 V / MAX. 56 V | Typ. 14.5 W 120 Ω load @ 5 V, Max. 15.5 W 120 Ω load @ 5 V | ||||
| Weight | 1800 g | ||||
| Slice Dimensions | 266 mm x 149 mm x 55 mm | ||||
| Rack | SIRIUSr X 16xSTGS | ||||
| Input connectors | D-SUB9 | ||||
| TEDS support | Standard + DSI® adapters | ||||
| Power supply | 12 - 48 V DC | ||||
| Power consumption | Lemo @Typ. 24 V / Max. 48 V | Typ. 12.2 W 120 Ω load @ 5 V, Max. 15 W 120 Ω load @ 5 V | |||
| Weight | 1000 g | ||||
| Slice Dimensions | 266 mm x 143 mm x 45 mm | ||||
| Rev: 1712304000 | |||||
SIRIUSif-X-8xCAN-FD tekniska specifikationer
| CAN ports specification | SIRIUSif-X-8xCAN-FD |
|---|---|
| Number of ports | 8 |
| Interface type | CAN-FD, CAN 2.0B |
| Connector | 8x D-SUB9 male (front panel) 1x D-SUB9 male dual CAN-FD (back panel) |
| Special protocols | J1939, XCP, OBDII |
| Isolation | CATII 150V (secondary circuits) |
| Bus pin fault protection | ±36 V |
| Oscillator frequency | 40 MHz |
| Transfer rates | |
| CAN | |
| Max. receive rate | up to 1 Mbit/s per port |
| Max. transmit rate | TBD |
| CAN-FD | |
| Max. receive rate | up to 5 Mbit/s per port |
| Max. transmit rate | TBD |
| Power specification | |
| Power supply | 9 - 57 V DC |
| Maximum sensor power consumption | 5 V / max. 250 mA (1.25 W) 12 V / max. 90 mA (1.1 W) |
| Maximum power consumption | Typ. 6.6 W / Max. 8.5 W (no load) |
| Sensor supply isolation | Ch-Ch & Ch-GND |
| Additional Specifications | |
| Modular | SIRIUSif-X-8xCAN-FD |
| Input connector | DSUB9 |
| Power supply | 9 - 57 V DC |
| Weight | 1500 g |