IOLITE® X 技术规格

  • IOLITE® X 系统技术规格
  • IOLITE® X 放大器概述
  • IOLITE-X-8xUNI 技术规格
  • IOLITEi-X-4xUNI 技术规格
  • IOLITE-X-8xLVe 技术规格
  • IOLITE-X-16xLV 技术规格
  • IOLITE-X-8xACC 技术规格
  • IOLITE-X-8xSTGS 技术规格
  • IOLITEi-X-4xCAN-FD 技术规格

IOLITE® X 系统技术规格

IOLITE X
GLAN
Data interfaceEthernet 1 GbE incl. IEEE1588v2 synchronization (PTP IEEE 802.3) (RJ45)
Data Rate1 Gbps
Sync Accuracy100 ns
Bus connectorsEthernet RJ45, single cable for data, power and sync, daisy chainable
Power over Ethernet48 - 57 V DC (PoE++ power delivery)
Max. cable length between devices100 m
Max. Throughput per Chainup to 90W
Internals
CPU typeARM
Operating systemLinux
SoftwareDewesoft RT
Device protocolsopenDAQ, OPC-UA
Power
Power supply9 - 57 V DC
Power supply connectorSCDN-THR 3.81/04/90G 3.2SN BK BX
Environmental
Operating Temperature-40 to 70 °C
Storage Temperature-40 to 85 °C
Humidity95 %, no condensation
IP ratingIP20
ShockShock: SIST EN 60068-2-27:2009 10 g, 6 ms, half-sine 25x pos./neg in each axis 150x total
VibrationRandom: Profile breakpoints: 10 Hz - 0.75 g²/Hz 150 Hz - 0.015 g²/Hz Profile RMS / Peak = 3.5 g / 10.5 g 1 hrs in each axis, 3 hrs total
Physical
Dimensions137 x 119* x 35 (* depending on the front connectors)
Rev: 1712304000

IOLITE® X 放大器概述

X 4xUNIX 8xUNIX 8xLVeX 16xLVX 8xACCX 8xSTGSX 4xCAN-FD
ConnectorsDSUB-9DSUB-37Terminal block, DSUB-37Terminal blockBNCTerminal block, DSUB-37DSUB-9
Channels per module4x8x8x16x8x8x4x
Data rate per channel500 kS/s200 kS/s200 kS/s100 kS/s200 kS/s50 kS/sup to 5 Mbit/s
Resolution24-bit24-bit24-bit24-bit24-bit24-bitDigital
Bandwidth0.437*fs0.433*fs0.433*fs0.433*fs0.433*fs0.433*fs✗
Voltage mode ranges±100 V, ±10 V, ±1 V, ±100 mV±100 V, ±10 V, ±1 V, ±100 mV±100 V, ±5 V±200 V, ±10 V±10 V, ±5 V, ±1 V, ±200 mV✗✗
Input couplingDC, AC 0.15 HzDC, AC 0.15 HzDCDCDC, AC 1 Hz, AC 0.1 HzDC✗
Sensor excitation0..24 V (unipolar) 0..12 V (bipolar) 0..100 mA (current) Max. 0.8 W/ch 2 mA, 4 mA, 6 mA (IEPE)0..24 V (unipolar) 0..12 V (bipolar) 0..100 mA (current) Max. 0.8 W/ch 2 mA ... 20 mA (IEPE)0...24V (unipolar)✗2 mA, 4 mA, 6 mA1 V, 2 V, 5 V5 V (max. 240 mA), 12 V (max. 100 mA)
Bridge connectionFull, ½, ¼ 350 Ω or ¼ 120 Ω (3-wire or 4-wire), ¼ CC (2-wire or 4-wire) Full, ½, ¼ 350 Ω or ¼ 120 Ω (3-wire or 4-wire), ¼ CC (2-wire or 4-wire) ✗✗✗Full, ½, ¼ 350 Ω, ¼ 120 Ω 3-wire, 4-wire (optional)✗
Programmable shunt100 kΩ100 kΩ✗✗✗100 kΩ✗
IEPE input✓✓✗✗✓✗✗
Resistance✓✓✗✗✗✗✗
Temperature (PTx)✓✓✗✗✗✗✗
ThermocoupleDSI-THDSI-TH✗✗✗✗✗
Current20 mA (int. shunt), ext. shunt, DSI-5A, DSIi-10A, DSIi-20A20 mA (int. shunt), ext. shunt, DSI-5A, DSIi-10A, DSIi-20Aext. shunt✗✗✗✗
Potentiometer✓✓✓ ✗✗✓✗
LVDTDSI-LVDTDSI-LVDT✗✗✗✗✗
ChargeDSI-CHGDSI-CHG✗✗✗✗✗
TEDS✓✓✓✗✓✓ (only TB)✗
Isolation voltage250V250 V250 V250VDifferentialCATII 150 V
Isolation arrangementCh-Ch & Ch-GNDCh-GNDCh-GNDCh-GNDCh-GNDCh-GNDCh-Ch & Ch-GND
PoE functionality✓✓✓✓✓✓✓
Advanced functionsFully isolated, high input range, connect almost any sensor, supports TEDSHigh input range, connect almost any sensor, supports TEDS, high channel densityHigh input range, powerful sensor supplyHigh input range, high-channel densityIEPE, supports TEDS, high channel densitySupports all strain types, low power consumptionCAN-FD, CAN 2.0, fully isolated, sensor supply
Rev: 1779807380

IOLITE-X-8xUNI 技术规格

Analog inputsIOLITE X 8xUNI
Input typeVoltage Full / Half / Quarter bridge strain (120 Ω and 350 Ω) IEPE Potentiometer, RTD, Resistance Current (ext. and int. Shunt)
Number of channels8
ADC Type24-bit delta-sigma with anti-aliasing filter (-3 dB @ 160 kHz) Butterworth 2nd order
Sampling RateSimultaneous 200 kS/s per channel (software-selectable)
Voltage mode
Voltage Ranges±100 V±10 V±1 V±0.1 V
Input AccuracyDC±0.03 % of reading ±0.02 % of range
Noise floor, Typ.10 kS/s / 100 kS/s / 200 kS/s115 dB / 105 dB / 104 dB116 dB / 106 dB / 105 dB114 dB / 105 dB / 103 dB100 dB / 90 dB / 86 dB
Dynamic Range, Typ.10 kS/s / 100 kS/s / 200 kS/s-152 dB / -152 dB / -152 dB-150 dB / -153 dB / -153 dB-150 dB / -151 dB / -152 dB-137 dB / -137 dB / -137 dB
CMR, Typ. 0 dBFS sine wave @ 50 Hz / 400 Hz / 1 kHz200 kS/s85 dB / 69 dB / 60 dB85 dB / 85 dB / 85 dB105 dB / 105 dB / 102 dB113 dB / 110 dB / 104 dB
THD, Typ. -1 dBFS sine wave @ 1 kHz 200 kS/s-91 dB-101 dB-100 dB-107 dB
Gain Linearity< 0.02 %
Gain DriftTyp. 10 ppm/K, max. 40 ppm/K
Offset DriftTyp. 0.3 μV/K + 2 ppm of range/K, max. 0.8 μV/K + 10 ppm of range/K
Channel Cross talk typ. @ 1 kHz / 10 kHz-102 dB / -86 dB-135 dB / -120 dB-135 dB / -120 dB-119 dB / -110 dB
Input CouplingDC, AC 0.15 Hz (1 Hz, 3 Hz, 5 Hz, 10 Hz, 20 Hz - software-selectable)
Input Impedance2 MΩ between IN+ and IN->1 GΩ between IN+ and IN-
Max. Common Mode Voltage±120 V±12 V
Overvoltage Protection±200 V continuously±100 V continuously / 200 Vpeak for 10 ms
Analog input performance
Sample rateSR ≤ 2 kS/sSR > 2 kS/s
Bandwidth (-3 dB)0.489*fs0.433*fs
Alias-free BandwidthDC to 0.44*fsDC to 0.499*fs
Alias Rejection-105 dB (all sample rates)
Delay Through ADC34 / fs
Oversampling128
Current mode (internal shunt)
Ranges20 mA2 mA
Input Accuracy±0.03 % of reading, ±0.02 % of range, ±2.1 μA
Internal Shunt Resistor47 Ω
Protection Typ. trip current: Imax ≥ 31 mA; auto retry on 800 ms (clamp voltage 33 V)
Excitation Voltage
Excitation Voltage Free programmable (16-bit DAC)
Predefined levels24 V DC
Accuracy±0.05 % ±2 mV
Drift±10 ppm/K ±100 μV/K
Load stability0 % to 100 % load <0.01 %
Noise @ 10 V / 350 Ω <150 μVrms @ 10 kS/s
Line regulation over 20 Ω of change<0.005 % @ 120 Ω load
Sense Impedance to Exc / to GND 100 kΩ / >100 MΩ
Current limit100 mA (max. 800 mW)
Protection Continuous short to ground
Excitation Current
Excitation CurrentFree programmable (16-bit DAC)
Predefined levels0.1, 1, 2, 5, 10, 20,60 and 100 mA DC
Ranges 0.1 - 1 mA>1 - 20 mA>20 - 100 mA
Accuracy±0.2 % ±0.5 μA±0.15 % ±2 μA±0.01 % ±40 μA
Drift Typical5 ppm/K10 ppm/K25 ppm/K
Compliance voltage24 V, max. 600 mW
Output Impedance>100 MΩ
Bridge Connection Types
Bridge typesFull / Half / Quarter bridge (3-wire or 4-wire) / Quarter Constant Current (2-wire or 4-wire)
Ranges 2 - 1000 mV/V free programmable
Internal Bridge CompletionHalf and Quarter bridge 120 Ω and 350 Ω
Typ. Bridge Completion Accuracy0.05 %; TCR: 2 ppm/K (others on request)
Internal Shunt Resistor100 kΩ, software selectable (others on request)
Typ. Shunt Resistor Accuracy0.05 %; TCR: 10 ppm/K (others on request)
Input Short, Sensor Offset Adjust Software selectable
IEPE mode
IEPE CurrentFree programmable (16-bit DAC)
Predefined levels2, 4, 8, 12, 16 or 20 mA
Compliance voltage25 V
Output Impedance>100 kΩ
Sensor detectionShortcut: <4 V; Open: >19 V
Additional Specifications
ModularIOLITE X 8xUNI-D37
Input connectorD-SUB37
Isolation voltage250 Vpeak channel to ground (functional isolation)
TEDs supportStandard + DSI® adapters
Power supply9 - 57 V DC, Power over Ethernet: 48 - 57 V DC (PoE++ power delivery)
Power consumptionTB @Typ. 24 V / MAX. 56 VVoltage mode: Typ. 8.7 W / Max. 10.5 W Voltage mode with load: Typ. 25.5 W / Max. 28.2 W (Load: 0.8 W per channel)
PoE @Min. 48 V / MAX. 56 VVoltage mode: Typ. 10.5 W / Max. 11 W Voltage mode with load: Typ. 27.7 W / Max. 28.6 W (Load: 0.8 W per channel)
Weight645 g
Slice Dimensions137 mm x 119 mm x 35 mm
Rev: 1768399200

IOLITEi-X-4xUNI 技术规格

Analog inputsIOLITEi X 4xUNI
Input typeVoltage Full / Half / Quarter bridge strain (120 Ω and 350 Ω) Potentiometer, RTD, Resistance Current (ext. Shunt & int. Shunt) IEPE
Number of channels4
ADC Type24-bit, delta-sigma with anti-aliasing filter (-3 dB @ 400 kHz)
Sampling RateSimultaneous 500 kS/sec per channel (software-selectable)
Voltage mode
Voltage Ranges±100 V±10 V±1 V±0.1 V
Input AccuracyDC±0.03 % of reading ±0.02 % of range
Noise floor, Typ.100 S/s / 1 kS/s / 10 kS/s / 100 kS/s / 500 kS/s-135 dB / -127 dB / -117 dB / -108 dB / -99 dB-137 dB / -128 dB / -118 dB / -109 dB / -103 dB-134 dB / -125 dB / -116 dB / -106 dB / -99 dB-116 dB / -108 dB / -99 dB / -90 dB / -81 dB
Dynamic Range, Typ.-151 dB-152 dB-148 dB-131 dB
CMR, Typ. -1 dBFS sine wave @ 50 Hz / 400 Hz / 1 kHz10 kS/s81 dB / 63 dB / 55 dB85 dB / 85 dB / 85 dB105 dB / 105 dB /102 dB99 dB / 99 dB / 98 dB
Gain Linearity±0.01 %
Gain DriftTypical 10 ppm/K Max. 40 ppm/K
Offset DriftTypical 0.3 μV/K + 2 ppm of range/K Max 0.8 μV/K + 10 ppm of range/K
Channel Cross talk typ. @ 1 kHz-104 dB-153 dB-150 dB-131 dB
Input CouplingDC, AC 0.15 Hz
Input Impedance2 MΩ between IN+ and IN->1 GΩ between IN+ and IN-
Max. Common Mode Voltage±120 V±12 V
Overvoltage Protection±200 V, continuous±100 V continuous / ±200 V peak (10 ms)
Current mode (internal shunt)
Ranges20 mA2 mA
Input Accuracy±0.03 % of reading, ±0.02 % of range, ±2.1 μA
Internal Shunt Resistor47 Ω
Excitation Voltage
Excitation VoltageFree programmable (16-bit DAC)
Predefined levels1, 2.5, 5, 10, 12, 15, 20, 24 V
Accuracy±0.05 % ±2 mV
Drift±10 ppm/K ±100 μV/K
Load stability< 0.01 %
Noise @ 10 V / 350 Ω<150 μVrms @ 10 kS/s
Line regulation over 20 Ω of change<0.005 % @ 120 Ω load
Sense Impedance to Exc / to GND100 kΩ / >100 MΩ
Current limit100 mA (max. 800 mW)
ProtectionContinuous short to ground
Excitation Current
Excitation CurrentFree programmable (16-bit DAC)
Predefined levels0.1, 1, 2, 5, 10, 20, 60 mA
Ranges0.1 - 1 mA>1 - 20 mA>20 - 100 mA
Accuracy±0.2 % ±0.5 μA±0.15 % ±2 μA±0.1 % ±10 μA
Drift Typical±5 ppm/K ± 5 nA/K±10 ppm/K ± 10 nA/K ±40ppm/K ± 200 nA/K
Compliance voltage24 V, max. 800 mW
Output Impedance>100 MΩ
Bridge Connection Types
Bridge typesFull / Half / Quarter bridge (3-wire or 4-wire) / Quarter Constant Current (2-wire or 4-wire)
Ranges2 - 1000 mV/V free programmable
Internal Bridge Completion½ bridge and ¼ bridge 120 Ω and 350 Ω
Typ. Bridge Completion Accuracy0.05 %; TCR: 2 ppm/K
Internal Shunt Resistor100 kΩ, software selectable
Typ. Shunt Resistor Accuracy0.05 %; TCR: 10 ppm/K
Input Short, Sensor Offset AdjustSoftware selectable
IEPE mode
IEPE CurrentFree programmable (16-bit DAC)
Predefined levels2, 4, 8, 12, 16 and 20 mA
Compliance voltage24V
Output Impedance>100 kΩ
Sensor detectionShortcut: <4 V; Open: >19 V
Additional Specifications
ModularIOLITEi X 4xUNI
Input connectorD-SUB9
Isolation voltageCH - GND and CH - CH isolated, CAT II 250V (Secondary circuits)
TEDs supportStandard + DSI® adapters
Power supply9 - 57 V DC Power over Ethernet: 48 - 57 V DC (PoE++ power delivery)
Power consumptionLemo 24 V / 57 VExcitation OFF: Excitation 0.8 W / Ch:7.1 W / 8.7 W 14.8 W / 16.5 W
PoE 48 V / 57 VExcitation OFF: Excitation 0.8 W / Ch:8.4 W / 8.9 W 17.8 W / 18.3 W
Weight640 g
Slice Dimensions136.75 x 115.0 x 35.0 mm
Rev: 1779350400

IOLITE-X-8xLVe 技术规格

InputsIOLITE X 8xLVe
Input typesVoltage, Potentiometer (See 1), Current (Ext. shunt) (See 1)
Number of channels8
ADC Type24-bit delta-sigma with anti-aliasing filter (-3 dB @ 160 kHz) Butterworth 2nd order
Sampling RateSimultaneous up to 200 kS/s
Voltage Mode
RangesSR±100 V±5 V
Accuracy±0.03 % of reading ±0.02 % of range
Noise floor, Typ.10kS/s/ 100kS/s / 200kS/s-118 dB / -109 dB / -107 dB-103 dB / -93 dB / -90 dB
Dynamic Range, Typ.10kS/s/ 100kS/s / 200kS/s-149 dB / -149 dB / -149 dB-133 dB / -133dB / -132 dB
CMR, Typ. 0 dBFS sine wave @ 50 Hz / 400 Hz / 1 kHz10kS/s-113dB /-102dB /-99dB -100dB /-98dB /-97dB
100kS/s-105dB /-100dB /-98dB -90dB /-89dB /-89dB
200kS/s-103dB /-100dB /-97dB -87dB /-86dB /-86dB
THD, Typ. -1 dBFS sine wave @ 1 kHz10kS/s/ 100kS/s / 200kS/s-94 dB / -94 dB / -94 dB-112 dB / -105 dB / -102 dB
SFDR, Typ. -1 dBFS sine wave @ 1 kHz10kS/s/ 100kS/s / 200kS/s-104 dBc / -93 dBc/ -94 dBc-109 dBc / -107 dBc/ -105 dBc
SNR, Typ. -1 dBFS sine wave @ 1 kHz10kS/s/ 100kS/s / 200kS/s107 dB / 102 dB / 100 dB98 dB / 88 dB / 85 dB
Gain Linearity<0.02 %<0.01 %
Gain DriftTypical 5 ppm/K, Max. 20 ppm/K
Offset DriftTypical 5 ppm of range/K, Max. 10 ppm of range/K
Channel Crosstalk typ. @ 1 kHz -131
Input CouplingDC
Input Impedance800 kΩ (between In+ and In-)
Input Overvoltage ProtectionIn+ to In-: 200 V continuous, 350 Vpeak (100 ms)
Max. common mode voltage±200 V
Isolation250 V functional isolation (channel to power supply ground)
Analog input performance
Sample rateSR ≤ 2 kS/sSR > 2 kS/s
Bandwidth (-3 dB)0.489*fs0.433*fs
Alias-free BandwidthDC to 0.44*fsDC to 0.499*fs
Alias Rejection-105 dB (all sample rates)
Delay Through ADC34 / fs
Oversampling32
Excitation Voltage
Excitation voltageUnipolar 2 ... 24V DC
Current limit2.8 W max. power per channel , 22.4 W per slice (MAX: 500mA )
Excitation accuracy0.05 % ± 2 mV
Drift±20 ppm/K ± 60 μV/K
Stability 10 % to 90 % load< 0.06 @ 5 V exc (< 0.01 @ 24 V exc)
Noise @ 10 V / 350 Ω <150 μVrms @ 10 kS/s
Load regulation over 10 Ω of change< 0.04 % @ 120 Ω load
ProtectionContinuous short to ground
Additional Specifications
ModularIOLITE X 8xLVe
Input connectorTerminal blockDSUB-37
TEDS supportYes
Power supply9 - 57 V DC, Power over Ethernet: 48 - 57 V DC (PoE++ power delivery)
Power consumptionTB @Typ. 24 V / MAX. 56 VAnalog input only: Typ. 5.2 W/ MAX. 7.3 W Excitation: Typ 9.4 W 120 Ω load @ 5 V / 40.7 W 120 Ω load @ 20 V MAX: 11.5W 120 Ω load @ 5 V / 45.7 W 120 Ω load @ 20 V
PoE @min 48 V / MAX. 56 VAnalog input only: min: 7.1W / MAX. 7.5 W Excitation min 11.4 W 120 Ω load @ 5 V / 45.2 W 120 Ω load @ 20 V MAX 11.8 W 120 Ω load @ 5 V / 46.4 W 120 Ω load @ 20 V
Weight615 g610 g
Slice Dimensions137 x 115.0 x 35.0 mm
Rev: 1712304000
1) This functionality will be available with a firmware upgrade.

IOLITE-X-16xLV 技术规格

Analog inputIOLITE X 16xLV
Input typeVoltage
Number of channels16
ADC Type24-bit delta-sigma with anti-aliasing filter (-3 dB @ 160 kHz) Butterworth 2nd order
Sampling RatesSimultaneous up to 100 kS/s
Voltage mode
Voltage ranges±200 V±10 V
Input AccuracyDC±0.03 % of reading ±40 mV±0.03 % of reading ±2 mV
Noise floor, Typ.10 kS/s / 100 kS/s-115dB / -106 dB-102 dB / -95 dB
Dynamic Range10 kS/s / 100 kS/s-154 dB / -146 dB-144 dB / -138 dB
CMR, Typ. -1 dBFS on 10V and -6dB on 200V sine wave @ 50 Hz / 400 Hz / 1 kHz10 kS/s-100dB /-100dB /-97dB -100dB /-100dB /-100dB
100 kS/s-101dB /-101dB /-100dB -95dB /-95dB /-94dB
THD, Typ. -1 dBFS on 10V and -6dB on 200V sine wave @ 1 kHz10 kS/s / 100 kS/s-93 dB / -93 dB-106 dB / -102 dB
SFDR, Typ. -1 dBFS on 10V and -6dB on 200V sine wave @ 1 kHz10 kS/s / 100 kS/s95 dBc / 95 dBc107 dBc / 104 dBc
SNR Typ. -1 dBFS on 10V and -6dB on 200V sine wave @ 1 kHz10 kS/s / 100 kS/s94 dB / 94 dB107 dB / 104 dB
Gain Linearity< 0.01 %
Gain DriftTypical 10 ppm/K, Max. 40 ppm/K
Offset DriftTypical 2 ppm of range/K, Max. 5 ppm of range/K
Crosstalk (0 dBFS @ 1 kHz)-115 dB-130 dB
Input CouplingDC
Input Impedance800 kΩ
Max. common mode voltage±200 V
Overvoltage ProtectionIn+ to In-: 200 V continuous, 350 Vpeak (100 ms)
Analog input performance
Sample rateSR ≤ 2 kS/sSR > 2 kS/s
Bandwidth (-3 dB)0.489*fs0.433*fs
Alias-free BandwidthDC to 0.44*fsDC to 0.499*fs
Alias Rejection-105 dB (all sample rates)
Delay Through ADC34 / fs
Oversampling32
Additional Specifications
ModularIOLITE X 16xLV
Input ConnectorsTerminal block 9 pole OMNIMATE SL 2.50 / BLF 2.50/180
Power supply9 - 57 V DC, Power over Ethernet: 48 - 57 V DC (PoE++ power delivery)
Isolation Voltage250 V functional isolation (channel to power supply ground)
Power consumptionTB @Typ. 24 V / MAX. 56 VVoltage mode: Typ. 5.8 W / Max. 7.4 W
PoE @Min. 48 V / MAX. 56 VVoltage mode: Typ. 7.2 W / Max. 7.8W
Weight1250 g
Slice Dimensions137 x 115.0 x 35.0 mm
Rev: 1712304000

IOLITE-X-8xACC 技术规格

Analog inputsIOLITE X 8xACC
Input typeVoltage, IEPE
Number of channels8
ADC Type24-bit delta-sigma with anti-aliasing filter (-3 dB @ 160 kHz) Butterworth 2nd order
Sampling RateSimultaneous 200 kS/s per channel (software-selectable)
Voltage Mode
Voltage Ranges±10 V±5 V±1 V±0.2 V
Input AccuracyDC±0.02 % of reading ±0.02 % of range
Noise floor, Typ.10 kS/s / 100 kS/s / 200 kS/s-116 dB / -107 dB / -105 dB-116 dB / -107 dB / -105 dB-115 dB / -105 dB / -103 dB-105 dB / -95 dB / -93 dB
Dynamic Range, Typ.10 kS/s / 100 kS/s / 200 kS/s-158 dB / -149 dB / -150 dB-156 dB / -150 dB / -150 dB-155 dB / -152 dB / -148 dB-145 dB / -140 dB / -139 dB
CMR, Typ. 0 dBFS sine wave @ 50 Hz / 400 Hz / 1 kHz10 kS/s117 dB / 117 dB / 117 dB117 dB / 115 dB / 116 dB115 dB / 115 dB / 114 dB106 dB / 105 dB / 104 dB
100 kS/s107 dB / 107 dB / 107 dB107 dB / 107 dB / 107 dB105 dB / 105 dB / 104 dB96 dB / 96 dB / 95 dB
200 kS/s106 dB / 106 dB / 106 dB106 dB / 106 dB / 106 dB103 dB / 103 dB / 102 dB93 dB / 93 dB / 93 dB
THD, Typ. -1 dBFS sine wave @ 1 kHz10 kS/s / 100 kS/s / 200 kS/s-105 dB / -103 dB / -99 dB-104 dB / -102 dB / -98 dB-103 dB / -102 dB / -98 dB-101 dB / -101 dB / -99 dB
SFDR, Typ. -1 dBFS sine wave @ 1 kHz10 kS/s / 100 kS/s / 200 kS/s106 dBc / 106 dBc / 101 dBc106 dBc / 105 dBc / 101 dBc106 dBc / 105 dBc / 90 dBc98 dBc / 97 dBc / 90 dBc
SNR, Typ. -1 dBFS sine wave @ 1 kHz10 kS/s / 100 kS/s / 200 kS/s106 dB / 103 dB / 101 dB102 dB / 100 dB / 98 dB104 dB / 99 dB / 88 dB92 dB / 87 dB / 87 dB
Gain Linearity< 0.02 %
Gain DriftTypical 6 ppm/K, max. 15 ppm/K
Offset DriftTypical 0.5 μV/K + 0.5 ppm of range/K, max 2 μV/K + 2 ppm of range/K
Channel Cross talk typ. @ 1 kHz-130 dB-130 dB-130 dB-125 dB
Input CouplingDC, AC 0.1 Hz, 1 Hz
Input Impedance1 MΩ
Overvoltage ProtectionIn+ to In-: 50 V continuous, 200 V peak (10 msec)
Analog input performance
Sample rateSR ≤ 2 kS/sSR > 2 kS/s
Bandwidth (-3 dB)0.489*fs0.433*fs
Alias-free BandwidthDC to 0.44*fsDC to 0.499*fs
Alias Rejection-105 dB (all sample rates)
Delay Through ADC34 / fs
Oversampling128
IEPE input
Excitation2 mA, 4 mA, 6 mA
Excitation accuracy10%
Compliance voltage24 V
Output Impedance>60 kΩ
Sensor detectionShortcut: <4 V; Open: >19 V
Additional Specifications
ModularIOLITE X 8xACC
Input connectorBNC
Isolation voltage250 Vpeak channel to ground
TEDs supportIEPE mode only (class 1)
Power supply9 - 57 V DC, Power over Ethernet: 48 - 57 V DC (PoE++ power delivery)
Power consumptionTB @Typ. 24 V / MAX. 56 VVoltage mode: Typ. 5.88 W / Max. 7.4 W IEPE mode: Typ. 7.3 W @4 mA, 8.6 W @6 mA / Max. 7.9 W@4 mA, 9.3 W @6 mA
PoE @Min. 48 V / MAX. 56 VVoltage mode: min. 7.2 W / Max. 7.6 W IEPE mode: min. 8.5 W @4 mA, 8.9 W @6 mA / Max. 9.2 W@4 mA, 9.5 W @6 mA
Weight600 g
Slice Dimensions137 x 115.0 x 35.0 mm
Rev: 1712304000

IOLITE-X-8xSTGS 技术规格

Analog inputsIOLITE X 8xSTGS
Input typeFull / half / quarter bridge (3-wire, 4-wire), Potentiometer
Number of channels8
ADC Type24-bit delta-sigma with anti-aliasing filter (-3 dB @ 160 kHz) Butterworth 2nd order
Sampling RateSimultaneous 50 kS/s per channel (software-selectable)
Voltage mode (see 1)
Ranges1 V0.1 V0.02 V
Gain accuracy0.02 % of reading0.02 % of reading0.02 % of reading
Offset accuracy 210 uV30 uV14 uV
Offset Accuracy after Balance Amplifier21 uV3 uV1.4 uV
Typical Noise floor @ 100 S/s / 1 kS/s / 10 kS/s / 20 kS/s / 50 kS/s -117 dB /-114 dB /-109 dB /-107 dB /-104 dB-116 dB /-110 dB /-101 dB /-98 dB /-95 dB-108 dB /-99 dB /-89 dB /-86 dB /-83 dB
Typical Dynamic Range @ 100 S/s / 1 kS/s / 10 kS/s / 20 kS/s / 50 kS/s -136 dB /-139 dB /-141 dB /-142 dB /-144 dB-132 dB /-133 dB /-134 dB /-134 dB /-135 dB-122 dB /-121 dB /-122 dB /-122 dB /-122 dB
Typ. Crosstalk(50kS/s) @ 1 kHz-103 dBc-94 dBc-83 dBc
Typ. CMR( 10kS/s) @ 50 Hz / 400 Hz / 1 kHz-89dB /-85dB /-80dB -99dB /-94dB /-89dB -89dB /-88dB /-86dB
Typ. CMR (50kS/s) @ 50 Hz / 400 Hz / 1 kHz-88dB /-85dB /-80dB -94dB /-91dB /-87dB -82dB /-82dB /-81dB
Gain DriftTypical 10 ppm/K (max. 20 ppm/K)
Offset DriftTypical 0.03 μV/K + 2.5 ppm of range/K, max 0.1 μV/K + 9 ppm of range/K
Gain Linearity< 0.01 %
Input CouplingDC
Overvoltage ProtectionIn+ to In-: 6 V continuous
Analog input performance
Sample rateSR ≤ 2 kS/sSR > 2 kS/s
Bandwidth (-3 dB)0.49*fs0.433*fs
Alias-free BandwidthDC to 0.44*fsDC to 0.499*fs
Alias Rejection-105 dB (all sample rates)
Delay Through ADC34 / fs
Oversampling
Excitation Voltage
Excitation VoltageUnipolar: 1 V, 2 V, 5 V (see 2)
Excitation Accuracy±3 % of set ±50 mV
Sense Accuracy±0.02 % of reading ±0.25 mV
Sense DriftTypical 10 μV/K + 10 ppm of reading/K, max 20 μV/K + 20 ppm of reading/K
Current Limit42 mA
ProtectionContinuous short to ground
Bridge measurement
Bridge Connection Typesfull bridge strain, ½ bridge strain, ¼ bridge strain (3-wire/4-wire)
Ranges2 mV/V ... 1000 mV/V free programmable (software ranges)
AccuracyRange \ Excitation1 V2 V5 V
200 mV/V±0.035 % of reading ±210 μV/V±0.028 % of reading ±105 μV/V±0.025 % of reading ±42 μV/V
20 mV/V±0.027 % of reading ±14 μV/V±0.027 % of reading ±15 μV/V±0.024 % of reading ±6 μV/V
2 mV/V±0.027 % of reading ±14 μV/V±0.024 % of reading ±7 μV/V±0.022 % of reading ±2.8 μV/V
Internal Bridge Completion¼ bridge 120 Ω and 350 Ω
Bridge Completion Accuracy±125 µV/V, TCR: 5 ppm/K (others on request)
Internal Shunt Resistor100 kΩ software selectable to SNS+ or SNS- and In+ or In-
Shunt Resistor Accuracy0.05 %; TCR: 10 ppm/K (others on request)
Input Short, Sensor Offset AdjustSoftware-selectable
Internal switch configurationsBridge typeSelectable connectionsInternal SW on:Additonal compensation
Full 6 or 4 Wires4 wires SNS + and EXC + & SNS- and EXC - Lead wire compensation available on 4 wire mode
Half 5 or 3 Wires3 wires SNS + and EXC + & SNS- and EXC - Lead wire compensation available on 3 wire mode
QB 3WSNS + and EXC + & SNS- and EXC - SNS + and EXC + & SNS- and EXC - Lead wire compensation available
QB 4WSNS + and EXC + SNS + and EXC +
Sensor current measurement (sooner)
Bridge configurationFull / HalfQB
Current measuring line EXC+ lineR Line
Current limit 50 mAdc
Resolution0.1 mAdc
Additional math caculusResistance, Power
Sample rate 1 Hz
Additional Specifications
ModularIOLITE-X-8xSTGS
Input connectorsTerminal Block Weidmueller SL 2.50-04DSUB-37
TEDS supportAvailable Not available
Power supply9 - 60 V DC, PoE: 48-60 V DC (802.3bt Type4)
Power ConsumptionTyp. 7.4 W; Max. 9.2W (120 Ω load @ 5 [V] EXC)
Weight g590 g
Slice Dimensions136.75 x 115.0 x 35.0 mm
Rev: 1748858098, 2.6.2025
1) Voltage ranges are describing the input for the bridge mode and are not supported as voltage mode.
2) Ratiometric method

IOLITEi-X-4xCAN-FD 技术规格

CAN ports specificationIOLITEi-X-4xCAN-FD
Number of ports4
Interface typeCAN-FD, CAN 2.0B
Special protocolsJ1939, XCP, OBDII
Galvanic isolationIsolated
Bus pin fault protection±36 V
Oscillator frequency40 MHz
Transfer rates
CAN
Max. receive rateup to 1 Mbit/s per port
Max. transmit rateTBD
CAN-FD
Max. receive rateup to 5 Mbit/s per port
Max. transmit rateTBD
Power specification
Power supply9 - 57 V DC
Maximum sensor power consumption 5 V / max. 250 mA (1.25 W) 12 V / max. 90 mA (1.1 W)
Maximum power consumption Typ. 4.5 W / Max. 5.8 W (no load)
Sensor supply isolationChannel to ground
Additional Specifications
ModularIOLITEi-X-4xCAN-FD
Input connectorDSUB9
Power supply9 - 57 V DC
Weight560 g
Slice Dimensions136.75 x 115.0 x 35.0 mm