SIRIUS® X 技术规格

  • SIRIUS® X 系统技术规格
  • SIRIUS® X 放大器概述
  • SIRIUS-X-16xUNI 技术规格
  • SIRIUSi-X-8xUNI 技术规格
  • SIRIUSf-X-16xACC 技术规格
  • SIRIUSf-X-16xLVe 技术规格
  • SIRIUSf-X-32xLV 技术规格
  • SIRIUSf-X-16xSTGS 技术规格
  • SIRIUSif-X-8xCAN-FD 技术规格

SIRIUS® X 系统技术规格

SIRIUS X
GLAN
Data interfaceEthernet 1 GbE incl. IEEE1588v2 synchronization (PTP IEEE 802.3)
Sync Accuracy100 ns
Bus connectorsEthernet RJ45, single cable for data, power and sync, daisy chainable
Power over Ethernet48-57 V DC (max. 90W)
Max. cable length between devices100 m
CAN bus
Number of ports2
Interface typeCAN 2.0B, up to 1 MBit/s CAN-FD, up to 8 Mbit/s
Special applicationsCCP, OBDII, J1939, CAN RT output
Galvanic isolationIsolated
Bus pin fault protection±36 V
ESD protection8 kV
IRIG Synchronisation
Sync Connectors4-pin LEMO 00 Series female
Sync Accuracy100 ns
Max. Sync-cable length200 m (IRIG)
Level (Input/Output)TTL compatible
Max. Output Current±24 mA (±50 mA for 1 sec)
Other interfaces
USBUSB-C 3.0 Type (See 1)
EtherCATSecondary bus/RT out connection to 3rd party EtherCAT master (See 2)
Internals
CPU typeARM
Operating systemLinux
SoftwareDewesoft RT
Device protocolsopenDAQ, OPC-UA
Power
Power Supply9 - 57 V DC
Power consumption*Configuration dependent
Power supply connector2-pin LEMO 1B Series male
Power OutInput supply on 2-pin LEMO 1B Series female
Environmental
Operating Temperature-10 to 50 °C (optional -40 to 85 °C)
Storage Temperature-40 to 85 °C
Humidity5 to 95 % RH non condensing @ 60 °C
IP ratingIP40
Shock & VibrationVibration sweep sinus (EN 60068-2-6:2008) Vibration random (EN 60721-3-2: 1997 - Class 2M2) Shock (EN 60068-2-27:2009) MIL-STD-810D
Physical
CoolingFan-based, fanless ("f" designation) (See 3)
Dimensions266 mm x 149 mm x 55 mm
Rev: 1712304000
1) Currently, this functionality is available with firmware 2025.2 or newer. 2) EtherCAT functionality is currently not supported. 3) Fanless options of actively cooled units must not be stacked.

SIRIUS® X 放大器概述

8xUNIhi 8xUNI16xUNI16xLVe32xLV16xACC16xSTGS8xCAN-FD
ConnectorsDSUB-9Redel 2P 8-pinDSUB-9D-SUB9Terminal blockBNCD-SUB9DSUB-9
Max. Channels per slice8x8x16x16x32x16x16x4x
Data rate per channel500 kS/s500 kS/s200 kS/s200 kS/s100 kS/s200 kS/s50 kS/sup to XY Mbit/s
Resolution24-bit24-bit24-bit24-bit24-bit24-bit24-bitDigital
Bandwidth0.437*fs0.437*fs0.433*fs0.433*fs0.433*fs0.433*fs0.433*fs✗
Voltage mode ranges±100 V, ±10 V, ±1 V, ±100 mV±100 V, ±10 V, ±1 V, ±100 mV±100 V, ±10 V, ±1 V, ±100 mV±100 V, ±5 V±200 V, ±10 V±10 V, ±5 V, ±1 V, ±200 mV✗✗
Input couplingDC, AC 0.15 HzDC, AC 0.15 HzDC, AC 0.15 HzDCDCDC, AC 1 Hz, AC 0.1 HzDC✗
Sensor excitation0..24 V (unipolar) 0..12 V (bipolar) 0..100 mA (current) Max. 0.8 W/ch 2 mA, 4 mA, 6 mA (IEPE)0..24 V (unipolar) 0..12 V (bipolar) 0..100 mA (current) Max. 0.8 W/ch 2 mA, 4 mA, 6 mA (IEPE)0..24 V (unipolar) 0..12 V (bipolar) 0..100 mA (current) Max. 0.8 W/ch 2 mA ... 20 mA (IEPE)0...24V (unipolar) And *D-SUB9: Fixed 10V DC bipolar excitation✗2 mA, 4 mA, 6 mA1 V, 2 V, 5 V5 V (max. 240 mA), 12 V (max. 100 mA)
Bridge connectionFull, ½, ¼ 350 Ω or ¼ 120 Ω (3-wire or 4-wire), ¼ CC (2-wire or 4-wire) Full, ½, ¼ 350 Ω or ¼ 120 Ω (3-wire or 4-wire), ¼ CC (2-wire or 4-wire) Full, ½, ¼ 350 Ω or ¼ 120 Ω (3-wire or 4-wire), ¼ CC (2-wire or 4-wire) ✗✗✗Full, ½, ¼ 350 Ω, ¼ 120 Ω 3-wire, 4-wire (optional)✗
Programmable shunt100 kΩ100 kΩ100 kΩ✗✗✗100 kΩ✗
IEPE input✓, DSI-ACC✓, DSI-ACC✓, DSI-ACCDSI-ACC✗✓✗✗
Resistance✓✓✓✗✗✗✗✗
Temperature (PTx)✓✓✓DSI-RTD✗✗✗✗
ThermocoupleDSI-THDSI-THDSI-THDSI-TH✗✗✗✗
Current100 mA (int. shunt), ext. shunt, DSI-5A, DSIi-10A, DSIi-20A100 mA (int. shunt), ext. shunt, DSI-5A, DSIi-10A, DSIi-20A100 mA (int. shunt), ext. shunt, DSI-5A, DSIi-10A, DSIi-20A ✓ (Ext. shunt)✗✗✗✗
Potentiometer✓✓✓✓✗✗✓✗
LVDTDSI-LVDTDSI-LVDTDSI-LVDTDSI-LVDT✗✗✗✗
ChargeDSI-CHGDSI-CHGDSI-CHGDSI-CHG✗✗✗✗
TEDS✓✓✓✓✗✓✓✗
Isolation voltageCATII 150V (secondary circuits)CATII 1000V250V250 V250 V250VDifferentialCATII 150V (secondary circuits)
Isolation arrangementCh-Ch & Ch-GNDCh-Ch & Ch-GNDCh-GNDCh-GNDCh-GNDCh-GNDCh-GNDCh-Ch & Ch-GND
PoE functionality✓✓✓✓✓✓✓✓
Advanced functionsFully isolated, high input range, connect almost any sensor, supports TEDSHigh isolation, high input range, connect almost any sensor, supports TEDSHigh input range, connect almost any sensor, supports TEDS, high channel densityHigh input range, powerful sensor supplyHigh input range, high-channel densityIEPE, supports TEDS, high channel densitySupports all strain types, low power consumptionCAN-FD, CAN 2.0, fully isolated, sensor supply
Rev: 1792050960

SIRIUS-X-16xUNI 技术规格

Analog inputsSIRIUS X 16xUNI
Input typeVoltage Full/Half/Quarter bridge strain (120 Ω and 350 Ω) IEPE Potentiometer, RTD, Resistance Current (ext. and int. Shunt)
Number of channels16
ADC Type24-bit delta-sigma with anti-aliasing filter (-3 dB @ 160 kHz) Butterworth 2nd order
Sampling RateSimultaneous 200 kS/sec per channel (software-selectable)
Voltage mode
Voltage Ranges±100 V±10 V±1 V±0.1 V
Input AccuracyDC±0.03 % of reading ±0.02 % of range
Noise floor, Typ.10 kS/s / 100 kS/s / 200 kS/s115 dB / 105 dB / 104 dB116 dB / 106 dB / 105 dB114 dB / 105 dB / 103 dB100 dB / 90 dB / 86 dB
Dynamic Range, Typ.10 kS/s / 100 kS/s / 200 kS/s-152 dB / -152 dB / -152 dB-150 dB / -153 dB / -153 dB-150 dB / -151 dB / -152 dB-137 dB / -137 dB / -137 dB
CMR, Typ. 0 dBFS sine wave @ 50 Hz / 400 Hz / 1 kHz200 kS/s85 dB / 69 dB /60 dB85 dB / 85 dB / 85 dB105 dB / 105 dB / 102 dB113 dB / 110 dB / 104 dB
THD, Typ. -1 dBFS sine wave @ 1 kHz 200 kS/s-91 dB-101 dB-100 dB-107 dB
Gain Linearity< 0.02 %
Gain DriftTyp. 10 ppm/K, max. 40 ppm/K
Offset DriftTyp. 0.3 μV/K + 2 ppm of range/K, max. 0.8 μV/K + 10 ppm of range/K
Channel Cross talk typ. @ 1 kHz / 10 kHz-100 dB / -86 dB-135 dB / -120 dB-135 dB / -120 dB-119 dB / -110 dB
Input CouplingDC, AC 0.15 Hz
Input Impedance2 MΩ between IN+ and IN->1 GΩ between IN+ and IN-
Max. Common Mode Voltage±120 V±12 V
Overvoltage Protection±200 V continuously±100 V continuously / 200 Vpeak for 10 ms
Analog input performance
Sample rateSR ≤ 2 kS/sSR > 2 kS/s
Bandwidth (-3 dB)0.489*fs0.433*fs
Alias-free BandwidthDC to 0.44*fsDC to 0.499*fs
Alias Rejection-105 dB (all sample rates)
Delay Through ADC34 / fs
Oversampling32
Current mode (internal shunt)
Ranges20 mA2 mA
Input Accuracy±0.03 % of reading, ±0.02 % of range, ±2.1 μA
Internal Shunt Resistor47 Ω
Protection Typ. trip current: Imax ≥ 31 mA; auto retry on 800 ms (clamp voltage 33 V)
Excitation Voltage
Excitation Voltage Free programmable (16-bit DAC)
Predefined levels24 V DC
Accuracy±0.05 % ±2 mV
Drift±10 ppm/K ±100 μV/K
Load stability0 % to 100 % load <0.01 %
Noise @ 10 V / 350 Ω <150 μVrms @ 10 kS/s
Line regulation over 20 Ω of change<0.005 % @ 120 Ω load
Sense Impedance to Exc / to GND 100 kΩ / >100 MΩ
Current limit100 mA (max. 800 mW)
Protection Continuous short to ground
Excitation Current
Excitation CurrentFree programmable (16-bit DAC)
Predefined levels0.1, 1, 2, 5, 10, 20, 60 and 100 mA DC
Ranges 0.1 - 1 mA>1 - 20 mA>20 - 100 mA
Accuracy±0.2 % ±0.5 μA±0.15 % ±2 μA±0.01 % ±40 μA
Drift Typical5 ppm/K10 ppm/K25 ppm/K
Compliance voltage24 V, max. 600 mW
Output Impedance>100 MΩ
Bridge Connection Types
Bridge typesFull / Half / Quarter bridge (3-wire or 4-wire) / Quarter Constant Current (2-wire or 4-wire)
Ranges 2 - 1000 mV/V free programmable
Internal Bridge CompletionHalf and Quarter bridge 120 Ω and 350 Ω
Typ. Bridge Completion Accuracy0.05 %; TCR: 2 ppm/K (others on request)
Internal Shunt Resistor100 kΩ, software selectable (others on request)
Typ. Shunt Resistor Accuracy0.05 %; TCR: 10 ppm/K (others on request)
Input Short, Sensor Offset Adjust Software selectable
IEPE mode
IEPE CurrentFree programmable (16-bit DAC)
Predefined levels2, 4, 8, 12, 16 or 20 mA
Compliance voltage25 V
Output Impedance>100 kΩ
Sensor detectionShortcut: <4 V; Open: >19 V
Additional Specifications
ModularSIRIUS X 16xUNI
Input connectorD-SUB9
Isolation voltage250 Vpeak channel to ground (functional isolation)
TEDs supportStandard + DSI® adapters
Power supply9 - 57 V DC, Power over Ethernet: 48 - 57 V DC (PoE++ power delivery)
Power consumptionLemo @Typ. 24 V / MAX. 56 VVoltage mode: Typ. 15.6 W / Max. 19 W Voltage mode with load: Typ. 45 W / Max. 49 W (Load: 0.8W per channel)
PoE @Min. 48 V / MAX. 56 VVoltage mode: Typ. 18.5 W / Max. 19.5 W Voltage mode with load: Typ. 48.2 W / Max. 49.5 W (Load: 0.8W per channel)
Weight2140 g
Slice Dimensions266 mm x 149 mm x 55 mm
RackSIRIUSr X 16xUNI
Input connectorD-SUB9
Isolation voltage250 Vpeak channel to ground (functional isolation)
TEDs supportStandard + DSI® adapters
Power supply12 - 48 V DC
Power consumptionLemo @Typ. 24 V / Max. 48 VVoltage mode: Typ. 14.3 W / Max.18 W Voltage mode with load: Typ. 46.5 W / Max. 50 W (Load: 0.8W per channel)
Weight1240 g
Slice Dimensions266 mm x 143 mm x 45 mm
Rev: 1775131200

SIRIUSi-X-8xUNI 技术规格

Analog inputsSIRIUSi X 8xUNI
Input typeVoltage Full / Half / Quarter bridge strain (120 Ω and 350 Ω) Potentiometer, RTD, Resistance Current (int. and ext. Shunt) IEPE
Number of channels8
ADC Type24-bit, delta-sigma with anti-aliasing filter (-3 dB @ 400 kHz)
Sampling RateSimultaneous 500 kS/sec per channel (software-selectable)
Voltage mode
Voltage Ranges±100 V±10 V±1 V±0.1 V
Input AccuracyDC±0.03 % of reading ±0.02 % of range
Noise floor, Typ.100 S/s / 1 kS/s / 10 kS/s / 100 kS/s / 500 kS/s-135 dB / -127 dB / -117 dB / -108 dB / -99 dB-137 dB / -128 dB / -118 dB / -109 dB / -103 dB-134 dB / -125 dB / -116 dB / -106 dB / -99 dB-116 dB / -108 dB / -99 dB / -90 dB / -81 dB
Dynamic Range, Typ.500 kS/s-151 dB-152 dB-148 dB-131 dB
CMR, Typ. -1 dBFS sine wave @ 50 Hz / 400 Hz / 1 kHz10 kS/s81 dB / 63 dB / 55 dB85 dB / 85 dB / 85 dB105 dB / 105 dB /102 dB99 dB / 99 dB / 98 dB
Gain Linearity±0.01 %
Gain DriftTypical 10 ppm/K Max. 40 ppm/K
Offset DriftTypical 0.3 μV/K + 2 ppm of range/K Max 0.8 μV/K + 10 ppm of range/K
Channel Cross talk typ. @ 1 kHz-104 dB-153 dB-150 dB-131 dB
Input CouplingDC, AC 0.15 Hz
Input Impedance2 MΩ between IN+ and IN->1 GΩ between IN+ and IN-
Max. Common Mode Voltage±120 V±12 V
Overvoltage Protection±200 V, continuous±100 V continuous / ±200 V peak (10 ms)
Current mode (internal shunt)
Ranges20 mA2 mA
Input Accuracy±0.03 % of reading, ±0.02 % of range, ±2.1 μA
Internal Shunt Resistor47 Ω
Excitation Voltage
Excitation Voltage Free programmable (16-bit DAC)
Predefined levels1, 2.5, 5, 10, 12, 15, 20, 24 V
Accuracy±0.05 % ±2 mV
Drift±10 ppm/K ±100 μV/K
Load stability< 0.01 %
Noise @ 10 V / 350 Ω <150 μVrms @ 10 kS/s
Line regulation over 20 Ω of change<0.005 % @ 120 Ω load
Sense Impedance to Exc / to GND 100 kΩ / >100 MΩ
Current limit100 mA (max. 800 mW)
Protection Continuous short to ground
Excitation Current
Excitation CurrentFree programmable (16-bit DAC)
Predefined levels0.1, 1, 2, 5, 10, 20, 60 mA
Ranges 0.1 - 1 mA>1 - 20 mA>20 - 100 mA
Accuracy±0.2 % ±0.5 μA±0.15 % ±2 μA±0.1 % ±10 μA
Drift Typical±5 ppm/K ± 5 nA/K±10 ppm/K ± 10 nA/K ±40ppm/K ± 200 nA/K
Compliance voltage24 V, max. 800 mW
Output Impedance>100 MΩ
Bridge Connection Types
Bridge typesFull / Half / Quarter bridge (3-wire or 4-wire) / Quarter Constant Current (2-wire or 4-wire)
Ranges 2 - 1000 mV/V free programmable
Internal Bridge Completion½ bridge and ¼ bridge 120 Ω and 350 Ω
Typ. Bridge Completion Accuracy0.05 %; TCR: 2 ppm/K
Internal Shunt Resistor100 kΩ, software selectable
Typ. Shunt Resistor Accuracy0.05 %; TCR: 10 ppm/K
Input Short, Sensor Offset Adjust Software selectable
IEPE mode
IEPE CurrentFree programmable (16-bit DAC)
Predefined levels2, 4, 8, 12, 16 and 20 mA
Compliance voltage24V
Output Impedance>100 kΩ
Sensor detectionShortcut: <4 V; Open: >19 V
Additional Specifications
ModularSIRIUSi X 8xUNI
Input connectorD-SUB9
Isolation voltageCH - GND and CH - CH isolated Secondary circuits CAT II 150V
TEDs supportStandard + DSI® adapters
Power supply9 - 57 V DC Power over Ethernet: 48 - 57 V DC (PoE++ power delivery)
Power consumptionLemo @Typ. 24 V / MAX. 56 VVoltage mode: Typ. 12.1 W / Max. 14.4 W Voltage mode with load: Typ. 27.4 W / Max. 30 W (Load: 0.8W per channel)
PoE @Min. 48 V / MAX. 56 VVoltage mode: Typ. 14.7 W / Max. 15.3 W Voltage mode with load: Typ. 34.6 W / Max. 35.3 W (Load: 0.8W per channel)
Weight1965 g
Slice Dimensions266 mm x 149 mm x 55 mm
RackSIRIUSir X 8xUNI
Input connectorD-SUB9
Isolation voltageCH - GND and CH - CH isolated Secondary circuits CAT II 150V
TEDs supportStandard + DSI® adapters
Power supply12 - 48 V DC
Power consumptionLemo @Typ. 24 V / Max. 48 VVoltage mode: Typ. 10.4 W / Max.11.5 W Voltage mode with load: Typ. 25.9 W / Max. 27.6 W (Load: 0.8W per channel)
Weight1100 g
Slice Dimensions266 mm x 143 mm x 45 mm
Rev: 1775131200

SIRIUSf-X-16xACC 技术规格

Analog inputsSIRIUS X 16xACC
Input typeVoltage, IEPE
Number of channels16
ADC Type24-bit delta-sigma with anti-aliasing filter (-3 dB @ 160 kHz) Butterworth 2nd order
Sampling RateSimultaneous 200 kS/sec per channel (software-selectable)
Voltage mode
Voltage Ranges±10 V±5 V±1 V±0.2 V
Input AccuracyDC±0.02 % of reading ±0.02 % of range
Noise floor, Typ.10 kS/s / 100 kS/s / 200 kS/s-117 dB / -107 dB / -106 dB-117 dB / -107 dB / -106 dB-115 dB / -105 dB / -103 dB-105 dB / -96 dB / -93 dB
Dynamic Range, Typ.10 kS/s / 100 kS/s / 200 kS/s-160 dB / -154 dB / -154 dB-160 dB / -155 dB / -155 dB-157 dB / -152 dB / -150 dB-150 dB / -142 dB / -140 dB
CMR, Typ. 0 dBFS sine wave @ 50 Hz / 400 Hz / 1 kHz10 kS/s120 dB / 117 dB / 112 dB119 dB / 117 dB / 111 dB118 dB / 116 dB / 111 dB109 dB / 108 dB / 107 dB
100 kS/s110 dB / 110 dB / 108 dB110 dB / 110 dB / 108 dB108 dB / 108 dB / 107 dB99 dB / 99 dB / 98 dB
200 kS/s109 dB / 109 dB / 107 dB109 dB / 108 dB / 107 dB106 dB / 106 dB / 105 dB96 dB / 96 dB / 96 dB
THD, Typ. -1 dBFS sine wave @ 1 kHz10 kS/s / 100 kS/s / 200 kS/s-105 dB / -101 dB / -98 dB-104 dB / -102 dB / -98 dB-104 dB / -102 dB / -98 dB-102 dB / -102 dB / -97 dB
SFDR, Typ. -1 dBFS sine wave @ 1 kHz10 kS/s / 100 kS/s / 200 kS/s106 dBc / 104 dBc / 100 dBc106 dBc / 104 dBc / 100 dBc100 dBc / 100 dBc / 98 dBc97 dBc / 95 dBc / 93 dBc
SNR, Typ. -1 dBFS sine wave @ 1 kHz10 kS/s / 100 kS/s / 200 kS/s108 dB / 102 dB / 102 dB104 dB / 102 dB / 99 dB95 dB / 92 dB / 90 dB90 dB / 86 dB / 84 dB
Gain Linearity< 0.02 %
Gain DriftTypical 6 ppm/K, max. 15 ppm/K
Offset DriftTypical 0.5 μV/K + 0.5 ppm of range/K, max 2 μV/K + 2 ppm of range/K
Channel Cross talk typ. @ 1 kHz-130 dB-130 dB-130 dB-125 dB
Input CouplingDC, AC 0.1 Hz, 1 Hz
Input Impedance1 MΩ
Overvoltage ProtectionIn+ to In-: 50 V continuous, 200 V peak (10 msec)
Analog input performance
Sample rateSR ≤ 2 kS/sSR > 2 kS/s
Bandwidth (-3 dB)0.489*fs0.433*fs
Alias-free BandwidthDC to 0.44*fsDC to 0.499*fs
Alias Rejection-105 dB (all sample rates)
Delay Through ADC34 / fs
Oversampling128
IEPE input
Excitation2 mA, 4 mA, 6 mA
Excitation accuracy10%
Compliance voltage24 V
Output Impedance>60 kΩ
Sensor detectionShortcut: <4 V; Open: >19 V
Additional Specifications
ModularSIRIUS X 16xACC
Input connectorBNC
Isolation voltage250 Vpeak channel to ground
TEDs supportIEPE mode only (class 1)
Power supply9 - 57 V DC, Power over Ethernet: 48 - 57 V DC (PoE++ power delivery)
Power consumptionLemo @Typ. 24 V / MAX. 56 VVoltage mode: Typ. 9.1 W / Max. 10.3 W IEPE mode: Typ. 11.7 W @4 mA, 13.2 W @6 mA / Max. 12.9 W @4 mA, 14.3 W @6 mA
PoE @Min. 48 V / MAX. 56 VVoltage mode: min. 10.2 W / Max. 10.5 W IEPE mode: min. 12.7 W @4 mA, 14.2 W @6 mA / Max. 13.2 W @4 mA, 14.6 W @6 mA
Weight1700 g
Slice Dimensions266 mm x 149 mm x 55 mm
RackSIRIUSr X 16xACC
Input connectorBNC
Isolation voltage250 Vpeak channel to ground
TEDs supportIEPE mode only (class 1)
Power supply12 - 48 V DC
Power consumptionLemo @Typ. 24 V / Max. 48 VVoltage mode: Typ. 8 W / Max. 10.3 W IEPE mode: Typ. 11.4 W @4 mA, 12.8 W @6 mA / Max. 13.2 W @4 mA, 14.6 W @6 mA
Weight975 g
Slice Dimensions266 mm x 143 mm x 45 mm
Rev: 1712304000

SIRIUSf-X-16xLVe 技术规格

InputsSIRIUS X 16xLVe
Input typesVoltage, Potentiometer (See 1), Current (Ext. shunt) (See 1)
Number of channels16
ADC Type24-bit delta-sigma with anti-aliasing filter (-3 dB @ 160 kHz) Butterworth 2nd order
Sampling RateSimultaneous up to 200 kS/s
Voltage Mode
RangesSR±100 V±5 V
Input AccuracyDC±0.03 % of reading ±0.02 % of range
Noise floor, Typ.10kS/s/ 100kS/s / 200kS/s-118 dB / -109 dB / -107 dB-103 dB / -93 dB / -90 dB
Dynamic Range, Typ.10kS/s/ 100kS/s / 200kS/s-141 dB / -144 dB / -141 dB-132 dB / -131 dB / -127 dB
CMR, Typ. 0 dBFS sine wave @ 50 Hz / 400 Hz / 1 kHz10kS/s-115 dB / -103 dB/ -99 dB-103 dB / -102 dB / -100 dB
100kS/s-108 dB / -101 dB / -98 dB-93 dB / -93 dB / -93 dB
200kS/s-106 dB / -101 dB / -98 dB-90 dB / -90 dB / -90 dB
THD, Typ. -1 dBFS sine wave @ 1 kHz10kS/s/ 100kS/s / 200kS/s-95 dB / -95 dB / -96 dB-110 dB / -106 dB / -99 dB
SFDR, Typ. -1 dBFS sine wave @ 1 kHz10kS/s/ 100kS/s / 200kS/s-104 dBc / -98 dBc/ -98 dBc-109 dBc / -107 dBc/ -102 dBc
SNR, Typ. -1 dBFS sine wave @ 1 kHz10kS/s/ 100kS/s / 200kS/s107 dB / 102 dB / 100 dB98 dB / 88 dB / 85 dB
Gain Linearity<0.02 %<0.01 %
Gain DriftTypical 5 ppm/K, Max. 20 ppm/K
Offset DriftTypical 5 ppm of range/K, Max. 10 ppm of range/K
Channel CrosstalkChannel Cross talk typ. @ 1 kHz -131
Input CouplingDC
Input Impedance800 kΩ (between In+ and In-)
Input Overvoltage ProtectionIn+ to In-: 200 V continuous, 350 Vpeak (100 ms)
Max. common mode voltage±200 V
Isolation250 V functional isolation (channel to power supply ground)
Analog input performance
Sampling rateSR ≤ 2 kS/sSR > 2 kS/s
Bandwidth (-3 dB)0.489*fs0.433*fs
Alias-free BandwidthDC to 0.44*fsDC to 0.499*fs
Alias Rejection-105 dB (all sample rates)
Delay Through ADC34 / fs
Oversampling32
Excitation Voltage
Excitation voltageUnipolar 2 ... 24V DC D-SUB9: Fixed bipolar excitation 10V DC
Current limit2.8 W max. power per channel , 22.4 W per slice (MAX: 500mA )
Excitation accuracy0.05 % ± 2 mV (bipolar 1%)
Drift±20 ppm/K ± 60 μV/K
Stability 10 % to 90 % load< 0.06 @ 5 V exc (< 0.01 @ 24 V exc)
Noise @ 10 V / 350 Ω <150 μVrms @ 10 kS
Load regulation over 10 Ω of change< 0.04 % @ 120 Ω load
ProtectionContinuous short to ground
Additional Specifications
ModularSIRIUS X 16xLVe
Input connectorTerminal blockDSUB-9
TEDS supportYes (See 1)
Power supply9 - 57 V DC, Power over Ethernet: 48 - 57 V DC (PoE++ power delivery)
Power consumptionLemo @Typ. 24 V / MAX. 56 VAnalog input only: Typ. 8.7 W / Max. 10.9 W Excitation: Typ 15.5 W 120 Ω load @ 5 V, 74 W 120 Ω load @ 20 V Max: 18.9 W 120 Ω load @ 5 V, 87 W 120 Ω load @ 20 V
PoE @Min. 48 V / MAX. 56 VAnalog input only: min: 10.4 W / MAX. 11.1 W Excitation: Typ. 19 W 120 Ω load @ 5 V, 85.2 W 120 Ω load @ 20 V Max. 19.4 W 120 Ω load @ 5 V, 88.2 W 120 Ω load @ 20 V
Weight1820 g
Slice Dimensions266 mm x 149 mm x 55 mm
RackSIRIUSr X 16xLVe
Input connectorDSUB-9
TEDS supportYes (See 1)
Power supply12 - 48 V DC
Power consumptionLemo @Typ. 24 V / Max. 48 VAnalog input only: Typ. 6.5 W / Max. 8.5 W (100kS/s / 200kS/s) Excitation: Typ. 16 W 120 Ω load @ 5 V, 75 W 120 Ω load @ 20 V Max. 18 W 120 Ω load @ 5 V, 85 W 120 Ω load @ 20 V
Weight1130 g
Slice Dimensions266 mm x 143 mm x 45 mm
Rev: 1712304000

SIRIUSf-X-32xLV 技术规格

Analog inputSIRIUS X 32xLV
Input typeVoltage
Number of channels32
ADC Type24-bit delta-sigma with anti-aliasing filter (-3 dB @ 160 kHz) Butterworth 2nd order
Sampling RatesSimultaneous up to 100 kS/s
Voltage mode
Voltage ranges±200 V±10 V
Input AccuracyDC±0.03 % of reading ±40 mV±0.03 % of reading ±2 mV
Noise floor, Typ.10 kS/s / 100 kS/s-116 dB / -106 dB-105 dB / -96 dB
Dynamic Range10 kS/s / 100 kS/s-158 dB / -152 dB-147 dB / -140 dB
CMR, Typ. -1 dBFS on 10V and -6dB on 200V sine wave @ 1 kHz @ 50 Hz / 400 Hz / 1 kHz10 kS/s-94 dB / -94 dB / -93 dB-98 dB / -98 dB / -97 dB
100 kS/s-94 dB / -93 dB / -93 dB-94 dB / -94 dB / -93 dB
THD, Typ. -1 dBFS on 10V and -6dB on 200V sine wave @ 1 kHz10 kS/s / 100 kS/s-92 dB / -92 dB-107 dB / -103 dB
SFDR, Typ. -1 dBFS on 10V and -6dB on 200V sine wave @ 1 kHz100 kS/s94 dBc / 94 dBc109 dBc / 106 dBc
SNR Typ. -1 dBFS on 10V and -6dB on 200V sine wave @ 1 kHz10 kS/s / 100 kS/s103 dB / 95 dB103 dB / 95 dB
Gain Linearity< 0.01 %
Gain DriftTypical 10 ppm/K, Max. 40 ppm/K
Offset DriftTypical 2 ppm of range/K, Max. 5 ppm of range/K
Crosstalk (0 dBFS @ 1 kHz)-115 dB-130 dB
Input CouplingDC
Input Impedance800 kΩ
Max. common mode voltage±200 V
Overvoltage ProtectionIn+ to In-: 200 V continuous, 350 Vpeak (100 ms)
Analog input performance
Bandwidth (-3 dB)0.433*fs (0.49*fs @ < 5 kS/s)
Alias-free BandwidthDC to 0.499*fs (DC to 0.44*fs @ < 5 kS/s)
Alias Rejection-105 dB (all sample rates)
Delay Through ADC34 / fs
Oversampling32
Additional Specifications
ModularSIRIUS X 32xLV
Input ConnectorsTerminal block 9 pole OMNIMATE SL 2.50 / BLF 2.50/180
Power supply9 - 57 V DC, Power over Ethernet: 48 - 57 V DC (PoE++ power delivery)
Isolation Voltage250 Vpeak channel to ground
Power consumptionLemo @Typ. 24 V / MAX. 56 VVoltage mode: Typ. 8.5 W / Max. 10.4 W
PoE @Min. 48 V / MAX. 56 VVoltage mode: Typ. 10.1 W / Max. 10.7 W
Weight1700 g
Slice Dimensions266 mm x 149 mm x 55 mm
RackSIRIUSr X 32xLV
Input ConnectorsTerminal block 9 pole OMNIMATE SL 2.50 / BLF 2.50/180
Power supply12 - 48 V DC
Isolation Voltage250 Vpeak channel to ground
Power consumptionLemo @Typ. 24 V / Max. 48 VVoltage mode: Typ. 7.2W / Max. 8.7 W
Weight720 g
Slice Dimensions266 mm x 143 mm x 45 mm
Rev: 1712304000

SIRIUSf-X-16xSTGS 技术规格

InputsSIRIUS X 16xSTGS
Input typeFull / half / quarter bridge (3-wire, 4-wire), Potentiometer
Max. number of channels per slice16
ADC Type24-bit delta-sigma with anti-aliasing filter (-3 dB @ 160 kHz) Butterworth 2nd order
Sampling RateSimultaneous 50 kS/s per channel (software-selectable)
Voltage inputs
Ranges1 V0.1 V0.02 V
Input accuracy0.02 % of reading0.02 % of reading0.02 % of reading
Offset accuracy 210 uV30 uV14 uV
Offset Accuracy after Balance Amplifier21 uV3 uV1.4 uV
Noise floor, Typ.10 kS/s / 20 kS/s / 50 kS/s-111 dB / -109 dB / -105 dB-100 dB / -97 dB / -93 dB-87 dB / -85 dB / -81 dB
Dynamic Range, Typ.10 kS/s / 20 kS/s / 50 kS/s-144 dB / -142 dB / -141 dB-135 dB / -134 dB / -134 dB-122 dB / -122 dB / -122 dB
CMR, Typ. 0 dBFS sine wave @ 50 Hz / 400 Hz / 1 kHz10 kS/s-77 dB / -77 dB / -76 dB -96 dB / -89 dB / -82 dB -108 dB / -90 dB / -83 dB
50 kS/s-77 dB / -77 dB / -76 dB -96dB / -89 dB / -82 dB -107 dB / -90 dB / -83 dB
Gain Linearity< 0.01 %
Gain DriftTypical 10 ppm/K (max. 20 ppm/K)
Offset DriftTypical 0.03 μV/K + 2.5 ppm of range/K, max 0.1 μV/K + 9 ppm of range/K
Typ. Crosstalk @ 1 kHz-135 dBc-125 dBc-115 dBc
Input CouplingDC
Overvoltage ProtectionIn+ to In-: 6 V continuous
Analog input performance
Sample rateSR ≤ 2 kS/sSR > 2 kS/s
Bandwidth (-3 dB)0.49*fs0.433*fs
Alias-free BandwidthDC to 0.44*fsDC to 0.499*fs
Alias Rejection-105 dB (all sample rates)
Delay Through ADC34 / fs
Oversampling32
Excitation Voltage
Excitation VoltageUnipolar: 1 V, 2 V, 5 V (see 2)
Excitation Accuracy±3 % of set ±50 mV
Sense Accuracy±0.02 % of reading ±0.25 mV
Sense DriftTypical 10 μV/K + 10 ppm of reading/K, max 20 μV/K + 20 ppm of reading/K
Current Limit42 mA
ProtectionContinuous short to ground
Bridge measurement
Bridge Connection Typesfull bridge strain, ½ bridge strain, ¼ bridge strain (3-wire)
Ranges2 mV/V ... 1000 mV/V free programmable (software ranges)
AccuracyRange \ Excitation1 V2 V5 V
200 mV/V±0.035 % of reading ±210 μV/V±0.028 % of reading ±105 μV/V±0.025 % of reading ±42 μV/V
20 mV/V±0.027 % of reading ±14 μV/V±0.027 % of reading ±15 μV/V±0.024 % of reading ±6 μV/V
2 mV/V±0.027 % of reading ±14 μV/V±0.024 % of reading ±7 μV/V±0.022 % of reading ±2.8 μV/V
Internal Bridge Completion¼ bridge 120 Ω and 350 Ω
Bridge Completion Accuracy±125 µV/V, TCR: 5 ppm/K (others on request)
Internal Shunt Resistor100 kΩ software selectable to SNS+ or SNS- and In+ or In-
Shunt Resistor Accuracy0.05 %; TCR: 10 ppm/K (others on request)
Input Short, Sensor Offset AdjustSoftware-selectable
Load monitoring
Bridge configurationFull / HalfQB
Current measuring line EXC+ lineR Line
Current limit 50 mAdc
Resolution0.1 mAdc
Additional math calculusResistance, Power
Sample rate 1 Hz
Additional Specifications
ModularSIRIUS X 16xSTGS
Input connectorsD-SUB9Terminal Block Weidmueller SL 2.50-04
TEDS supportStandard + DSI® adapters Not available
Power supply9 - 57 V DC, Power over Ethernet: 48 - 57 V DC (PoE++ power delivery)
Power consumptionLemo @Typ. 24 V / MAX. 56 VTyp. 12.2 W 120 Ω load @ 5 V, Max. 15 W 120 Ω load @ 5 V
PoE @Min. 48 V / MAX. 56 VTyp. 14.5 W 120 Ω load @ 5 V, Max. 15.5 W 120 Ω load @ 5 V
Weight1800 g
Slice Dimensions266 mm x 149 mm x 55 mm
RackSIRIUSr X 16xSTGS
Input connectorsD-SUB9
TEDS supportStandard + DSI® adapters
Power supply12 - 48 V DC
Power consumptionLemo @Typ. 24 V / Max. 48 VTyp. 12.2 W 120 Ω load @ 5 V, Max. 15 W 120 Ω load @ 5 V
Weight1000 g
Slice Dimensions266 mm x 143 mm x 45 mm
Rev: 1712304000

SIRIUSif-X-8xCAN-FD 技术规格

CAN ports specificationSIRIUSif-X-8xCAN-FD
Number of ports8
Interface typeCAN-FD, CAN 2.0B
Connector8x D-SUB9 male (front panel) 1x D-SUB9 male dual CAN-FD (back panel)
Special protocolsJ1939, XCP, OBDII
IsolationCATII 150V (secondary circuits)
Bus pin fault protection±36 V
Oscillator frequency40 MHz
Transfer rates
CAN
Max. receive rateup to 1 Mbit/s per port
Max. transmit rateTBD
CAN-FD
Max. receive rateup to 5 Mbit/s per port
Max. transmit rateTBD
Power specification
Power supply9 - 57 V DC
Maximum sensor power consumption 5 V / max. 250 mA (1.25 W) 12 V / max. 90 mA (1.1 W)
Maximum power consumption Typ. 6.6 W / Max. 8.5 W (no load)
Sensor supply isolationCh-Ch & Ch-GND
Additional Specifications
ModularSIRIUSif-X-8xCAN-FD
Input connectorDSUB9
Power supply9 - 57 V DC
Weight1500 g
Slice Dimensions266 mm x 149 mm x 55 mm